Smith, Benjamin E. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hey listers, I was wondering if anyone has used hydrid detectors in the NDD position on a multi-photon microscope, and if so, what they thought of them. Right now we have PMT NDDs installed on our Leica SP8, but we are considering upgrading to HyD detectors. We have a couple of HyDs in the descan position, and they are far more sensitive than PMTs, but they also can shut-off when the fluorescent signal is too bright. Normally this is not an issue, but for large tiling runs it can be prohibitively time consuming to check the gain on the detector for the entire imaging volume, and yet it is equally frustrating to have a detector quit 3/4 of the way into an hour long tiling run due to an unexpectedly bright cell, or worse yet to have a detector quit during live imaging because a cell became increasingly bright. We would love the increased sensitivity of HyD NDDs, especially for live imaging and autofluorescence detection, but I'm not sure if the tendency of the HyD detectors to shut-off makes HyD NDDs more of a hassle then they're worth. Thanks for any feedback, Ben Smith -- Benjamin E. Smith, Ph.D. Samuel Roberts Noble Microscopy Laboratory Research Scientist II University of Oklahoma Norman, OK 73019 E-mail: [hidden email] Voice 405-325-4391 FAX 405-325-7619 http://www.microscopy.ou.edu/ |
|
Steffen Dietzel |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ben, sorry I can't give you any advice, but you did make an excellent argument why companies should implement "Controled light exposure microscopy" as suggested by Erik Manders. If the Laser would automatically be attenuated when the image gets very bright, this could solve the problem. Question: Would the Hybrid detector stay off for the rest of the scan? On our LaVisionBiotech multi-Photon, the GaAsps also shut automatically off, but are turned on again after a moment. So you don't loose the rest of the stack/time series but end up only with a few black image lines instead. Steffen On 24.06.2013 21:12, Ben Smith wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hey listers, > > I was wondering if anyone has used hydrid detectors in the NDD position > on a multi-photon microscope, and if so, what they thought of them. > > Right now we have PMT NDDs installed on our Leica SP8, but we are > considering upgrading to HyD detectors. We have a couple of HyDs in the > descan position, and they are far more sensitive than PMTs, but they > also can shut-off when the fluorescent signal is too bright. Normally > this is not an issue, but for large tiling runs it can be prohibitively > time consuming to check the gain on the detector for the entire imaging > volume, and yet it is equally frustrating to have a detector quit 3/4 of > the way into an hour long tiling run due to an unexpectedly bright cell, > or worse yet to have a detector quit during live imaging because a cell > became increasingly bright. > > We would love the increased sensitivity of HyD NDDs, especially for live > imaging and autofluorescence detection, but I'm not sure if the tendency > of the HyD detectors to shut-off makes HyD NDDs more of a hassle then > they're worth. > > Thanks for any feedback, > Ben Smith > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |
|
Smith, Benjamin E. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hey Steffen, Unfortunately, our hybrid detectors shut-off for the rest of the series if a signal is too bright. However, a representative from Leica informed me that there are several ways around this including line/frame acqusition (which would be fine for live imaging, but time prohibitive for tiling) and using the detectors in photon-counting mode with 12 or 16-bit intensity resolution (which also can be a bit more time consuming). They also mentioned that hydrid detectors are effectively always in photon counting mode, so it makes sense why they shut-down. If the moment a SPAD refreshes, it is then hit by another photon, it would burn out pretty quick from all the current. What would be really nice would be if the detector could sense the frequency of photon detection events, and if the frequency got too high then the pixel would be flagged as saturated, and then the detector could stay below the bias voltage until the laser moved on to the next pixel. This would make hybrid detectors comparable to PMT detectors in high intensity situations. -Ben Smith On 6/25/2013 3:10 AM, Steffen Dietzel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Ben, > > sorry I can't give you any advice, but you did make an excellent > argument why companies should implement "Controled light exposure > microscopy" as suggested by Erik Manders. If the Laser would > automatically be attenuated when the image gets very bright, this > could solve the problem. > > Question: Would the Hybrid detector stay off for the rest of the scan? > On our LaVisionBiotech multi-Photon, the GaAsps also shut > automatically off, but are turned on again after a moment. So you > don't loose the rest of the stack/time series but end up only with a > few black image lines instead. > > Steffen > > On 24.06.2013 21:12, Ben Smith wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hey listers, >> >> I was wondering if anyone has used hydrid detectors in the NDD position >> on a multi-photon microscope, and if so, what they thought of them. >> >> Right now we have PMT NDDs installed on our Leica SP8, but we are >> considering upgrading to HyD detectors. We have a couple of HyDs in the >> descan position, and they are far more sensitive than PMTs, but they >> also can shut-off when the fluorescent signal is too bright. Normally >> this is not an issue, but for large tiling runs it can be prohibitively >> time consuming to check the gain on the detector for the entire imaging >> volume, and yet it is equally frustrating to have a detector quit 3/4 of >> the way into an hour long tiling run due to an unexpectedly bright cell, >> or worse yet to have a detector quit during live imaging because a cell >> became increasingly bright. >> >> We would love the increased sensitivity of HyD NDDs, especially for live >> imaging and autofluorescence detection, but I'm not sure if the tendency >> of the HyD detectors to shut-off makes HyD NDDs more of a hassle then >> they're worth. >> >> Thanks for any feedback, >> Ben Smith >> > > |
|
Zdenek Svindrych |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ben, you can use the fastest possible scanning (preferably resonant) and average (or sum) several dim images. But I remember that the FOV shrinks appreciably when the normal scanner (accidentally being referred to as a "large FOV scanner") is pushed to the limits... Your suggested per-pixel HyD voltage is not practical (nor possible perhaps) as one cannot change the high voltage (6kV) quickly enough (up to about 10 MHz), and very accurately and predictably at the same time. On the other hand, it is not necessary to turn off the detector for the rest of the experiment... A few black lines is more like the kind of behavior I'd expect. Leica definitively should improve upon this. Good luck with your acquisition! Zdenek Svindrych ---------- Původní zpráva ---------- Od: Ben Smith <[hidden email]> Datum: 25. 6. 2013 Předmět: Re: LM Hybrid vs PMT non-descan detectors (NDDs) for multi-photon microscopy "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hey Steffen,Odeslat zprávu(https://email.seznam.cz/#) Unfortunately, our hybrid detectors shut-off for the rest of the series if a signal is too bright. However, a representative from Leica informed me that there are several ways around this including line/frame acqusition (which would be fine for live imaging, but time prohibitive for tiling) and using the detectors in photon-counting mode with 12 or 16-bit intensity resolution (which also can be a bit more time consuming). They also mentioned that hydrid detectors are effectively always in photon counting mode, so it makes sense why they shut-down. If the moment a SPAD refreshes, it is then hit by another photon, it would burn out pretty quick from all the current. What would be really nice would be if the detector could sense the frequency of photon detection events, and if the frequency got too high then the pixel would be flagged as saturated, and then the detector could stay below the bias voltage until the laser moved on to the next pixel. This would make hybrid detectors comparable to PMT detectors in high intensity situations. -Ben Smith On 6/25/2013 3:10 AM, Steffen Dietzel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Ben, > > sorry I can't give you any advice, but you did make an excellent > argument why companies should implement "Controled light exposure > microscopy" as suggested by Erik Manders. If the Laser would > automatically be attenuated when the image gets very bright, this > could solve the problem. > > Question: Would the Hybrid detector stay off for the rest of the scan? > On our LaVisionBiotech multi-Photon, the GaAsps also shut > automatically off, but are turned on again after a moment. So you > don't loose the rest of the stack/time series but end up only with a > few black image lines instead. > > Steffen > > On 24.06.2013 21:12, Ben Smith wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hey listers, >> >> I was wondering if anyone has used hydrid detectors in the NDD position >> on a multi-photon microscope, and if so, what they thought of them. >> >> Right now we have PMT NDDs installed on our Leica SP8, but we are >> considering upgrading to HyD detectors. We have a couple of HyDs in the >> descan position, and they are far more sensitive than PMTs, but they >> also can shut-off when the fluorescent signal is too bright. Normally >> this is not an issue, but for large tiling runs it can be prohibitively >> time consuming to check the gain on the detector for the entire imaging >> volume, and yet it is equally frustrating to have a detector quit 3/4 of >> the way into an hour long tiling run due to an unexpectedly bright cell, >> or worse yet to have a detector quit during live imaging because a cell >> became increasingly bright. >> >> We would love the increased sensitivity of HyD NDDs, especially for live >> imaging and autofluorescence detection, but I'm not sure if the tendency >> of the HyD detectors to shut-off makes HyD NDDs more of a hassle then >> they're worth. >> >> Thanks for any feedback, >> Ben Smith >> > >" |
| Free forum by Nabble | Edit this page |