EricMarino |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Has anybody measured, noticed or have any information regarding Laser bleed through of the Yokogawa CSU-22 or X1. I've placed notch and stop filters in the system and it didn't seem to work. Since I'm using an EM camera I feel I might be enhancing acceptable bleed through. Any opinions or thoughts would be appreciated. Eric Marino Senior Imaging Specialist Immune Disease Institute Harvard Medical School 200 Longwood Ave WAB 133D Boston, MA 02115 Lab: 617 713-8885 Cell: 617 913-9647 [hidden email]
Eric Marino
Senior Imaging Specialist Program in Cellular and Molecular Medicine Boston Children's Hospital |
|
Cameron, Lisa |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Eric, Are you using single band pass or multi-band pass emission filters? For instance, we have an emission filter wheel with 4 single band pass emission filters and then a combination of dual and triple band pass. For the CSU-22, it depends on what filters are installed in the head. - Lisa --------------------------------------- Lisa Cameron, Ph.D. Director of Confocal and Light Microscopy Dana Farber Cancer Institute Boston, MA 02215 [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Eric Marino Sent: Tuesday, July 26, 2011 12:33 PM To: [hidden email] Subject: [CONFOCALMICROSCOPY] Laser Bleed-through on a Spinning Disc (CSU-22 & X1) ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Has anybody measured, noticed or have any information regarding Laser bleed through of the Yokogawa CSU-22 or X1. I've placed notch and stop filters in the system and it didn't seem to work. Since I'm using an EM camera I feel I might be enhancing acceptable bleed through. Any opinions or thoughts would be appreciated. Eric Marino Senior Imaging Specialist Immune Disease Institute Harvard Medical School 200 Longwood Ave WAB 133D Boston, MA 02115 Lab: 617 713-8885 Cell: 617 913-9647 [hidden email] The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
| Free forum by Nabble | Edit this page |