Laser Protection

classic Classic list List threaded Threaded
♦
♦
5 messages Options
Katherina Garcia Garcia Katherina Garcia Garcia
Reply | Threaded
Open this post in threaded view
♦
♦
|

Laser Protection

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

-- Hello:

I am the person in charge of a Microscopy Facility. The Occupational Risks Department in our University came to check the hazards of the confocal microscopes (lasers). They told us to use some kind of glasses (for me when I have to check lasers) and to put a window around the confocal for users protection against lasers.

I contacted one company (Lasing), but it is difficult to find something suitable for all the laser lines (range between 405 and 635). There are several kind of glasses and you can buy a few to cover the range. Regarding windows, I only found one that protects between 315 and 540.

Can you help me with this? Do you have references I can contact?

Thanks in advance for your help.
Best regards,
Katherina




---------------------------------------------------------------
Katherina García García
Técnico Especialista en Microscopía y Citometría
Tlf. 954977432
[hidden email]
Centro Andaluz de Biología del Desarrollo
Consejo Superior de Investigaciones Científicas
Universidad Pablo Olavide
Ctra. Utrera km1
41013 Sevilla
España
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Laser Protection

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Katherina,

 

                 All the confocal microscope manufacturers who sell microscopes in the EEC ensure (as they are bound to) that their microscopes comply with EEC laser regulations.  You should simply put your local Occupational Risks department in contact with the suppliers of your systems.  It is unlikely that you will have to change anything.

 

                                                                                          Guy

 

Sponsor my next half-marathon on May 15th

There's a special reason - find it out at

http://www.everydayhero.com.au/Guy_Cox_4846 <http://www.everydayhero.com.au/Guy_Cox_4846>

______________________________________________

Associate Professor Guy Cox, MA, DPhil(Oxon)

Australian Centre for Microscopy & Microanalysis,

Madsen Building F09, University of Sydney, NSW 2006

 

Phone +61 2 9351 3176     Fax +61 2 9351 7682

             Mobile 0413 281 861

______________________________________________

      http://www.guycox.net <http://www.guycox.net/>

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Katherina Garcia Garcia
Sent: Monday, 16 May 2011 6:46 PM
To: [hidden email]
Subject: Laser Protection

 

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

-- Hello:

I am the person in charge of a Microscopy Facility. The Occupational Risks Department in our University came to check the hazards of the confocal microscopes (lasers). They told us to use some kind of glasses (for me when I have to check lasers) and to put a window around the confocal for users protection against lasers.

I contacted one company (Lasing), but it is difficult to find something suitable for all the laser lines (range between 405 and 635). There are several kind of glasses and you can buy a few to cover the range. Regarding windows, I only found one that protects between 315 and 540.

Can you help me with this? Do you have references I can contact?

Thanks in advance for your help.
Best regards,
Katherina




---------------------------------------------------------------
Katherina García García
Técnico Especialista en Microscopía y Citometría
Tlf. 954977432
[hidden email]
Centro Andaluz de Biología del Desarrollo
Consejo Superior de Investigaciones Científicas
Universidad Pablo Olavide
Ctra. Utrera km1
41013 Sevilla
España

________________________________

No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1209 / Virus Database: 1500/3640 - Release Date: 05/15/11
Emmanuel Gustin Emmanuel Gustin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Laser Protection

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You need to make a distinction here between end users and maintainers.

From the end user's perspective, you expect that a standard microscope checks out as a Class 1 laser system, i.e. intrinsically safe, with no need to provide additional safety devices. The manual should specify this. Putting it in a black box -- black shields being the only windows that will stop light at all visible wavelengths -- should be superfluous, at least for safety purposes.

For people who open the instrument to check or align lasers it is different, as this involves disabling or circumventing safety systems. Then you are working with Class 3 or even Class 4 lasers, and in that case I can understand that OH&S would insist on safety glasses. (AFAIK they should insist on a yearly visit to an ophthalmologist to inspect your retina for damage as well.) It is true that with relatively low power visible lasers (as opposed to the far more dangerous IR lasers) few of the maintenance engineers bother to wear safety glasses. But presumably they are trained in safe working practices; if you are not then safety glasses are a wise precaution.

You always need specific glasses that matches the wavelengths (and power!) of the laser you are working with; there is no general solution.

Best Regards,

Emmanuel


--
 Emmanuel Gustin,    Tel. (+32) 14 64 1586,    e-mail: [hidden email]  

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox
Sent: maandag 16 mei 2011 10:58
To: [hidden email]
Subject: Re: Laser Protection

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Katherina,

 

                 All the confocal microscope manufacturers who sell microscopes in the EEC ensure (as they are bound to) that their microscopes comply with EEC laser regulations.  You should simply put your local Occupational Risks department in contact with the suppliers of your systems.  It is unlikely that you will have to change anything.

 

                                                                                          Guy

 

Sponsor my next half-marathon on May 15th

There's a special reason - find it out at

http://www.everydayhero.com.au/Guy_Cox_4846 <http://www.everydayhero.com.au/Guy_Cox_4846>

______________________________________________

Associate Professor Guy Cox, MA, DPhil(Oxon)

Australian Centre for Microscopy & Microanalysis,

Madsen Building F09, University of Sydney, NSW 2006

 

Phone +61 2 9351 3176     Fax +61 2 9351 7682

             Mobile 0413 281 861

______________________________________________

      http://www.guycox.net <http://www.guycox.net/>

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Katherina Garcia Garcia
Sent: Monday, 16 May 2011 6:46 PM
To: [hidden email]
Subject: Laser Protection

 

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

-- Hello:

I am the person in charge of a Microscopy Facility. The Occupational Risks Department in our University came to check the hazards of the confocal microscopes (lasers). They told us to use some kind of glasses (for me when I have to check lasers) and to put a window around the confocal for users protection against lasers.

I contacted one company (Lasing), but it is difficult to find something suitable for all the laser lines (range between 405 and 635). There are several kind of glasses and you can buy a few to cover the range. Regarding windows, I only found one that protects between 315 and 540.

Can you help me with this? Do you have references I can contact?

Thanks in advance for your help.
Best regards,
Katherina




---------------------------------------------------------------
Katherina García García
Técnico Especialista en Microscopía y Citometría
Tlf. 954977432
[hidden email]
Centro Andaluz de Biología del Desarrollo
Consejo Superior de Investigaciones Científicas
Universidad Pablo Olavide
Ctra. Utrera km1
41013 Sevilla
España

________________________________

No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1209 / Virus Database: 1500/3640 - Release Date: 05/15/11
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Laser Protection

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


 Glasses only protect yourself, make sure nobody else enters the room during laser maintenance, e.g. putting a sign at the door. This is also a good idea when the service engineer comes to do the alignment and forgets to put up a sign.

 
best wishes

Andreas


 

 

-----Original Message-----
From: Gustin, Emmanuel [TIBBE] <[hidden email]>
To: [hidden email]
Sent: Mon, 16 May 2011 10:32
Subject: Re: Laser Protection


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



You need to make a distinction here between end users and maintainers.



From the end user's perspective, you expect that a standard microscope checks

out as a Class 1 laser system, i.e. intrinsically safe, with no need to provide

additional safety devices. The manual should specify this. Putting it in a black

box -- black shields being the only windows that will stop light at all visible

wavelengths -- should be superfluous, at least for safety purposes.



For people who open the instrument to check or align lasers it is different, as

this involves disabling or circumventing safety systems. Then you are working

with Class 3 or even Class 4 lasers, and in that case I can understand that OH&S

would insist on safety glasses. (AFAIK they should insist on a yearly visit to

an ophthalmologist to inspect your retina for damage as well.) It is true that

with relatively low power visible lasers (as opposed to the far more dangerous

IR lasers) few of the maintenance engineers bother to wear safety glasses. But

presumably they are trained in safe working practices; if you are not then

safety glasses are a wise precaution.



You always need specific glasses that matches the wavelengths (and power!) of

the laser you are working with; there is no general solution.



Best Regards,



Emmanuel





--

 Emmanuel Gustin,    Tel. (+32) 14 64 1586,    e-mail: [hidden email]  



-----Original Message-----

From: Confocal Microscopy List [mailto:[hidden email]] On

Behalf Of Guy Cox

Sent: maandag 16 mei 2011 10:58

To: [hidden email]

Subject: Re: Laser Protection



*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



Katherina,



 



                 All the confocal microscope manufacturers who sell microscopes

in the EEC ensure (as they are bound to) that their microscopes comply with EEC

laser regulations.  You should simply put your local Occupational Risks

department in contact with the suppliers of your systems.  It is unlikely that

you will have to change anything.



 



                                                                                         

Guy



 



Sponsor my next half-marathon on May 15th



There's a special reason - find it out at



http://www.everydayhero.com.au/Guy_Cox_4846 <http://www.everydayhero.com.au/Guy_Cox_4846>





______________________________________________



Associate Professor Guy Cox, MA, DPhil(Oxon)



Australian Centre for Microscopy & Microanalysis,



Madsen Building F09, University of Sydney, NSW 2006



 



Phone +61 2 9351 3176     Fax +61 2 9351 7682



             Mobile 0413 281 861



______________________________________________



      http://www.guycox.net <http://www.guycox.net/>



 



From: Confocal Microscopy List [mailto:[hidden email]] On

Behalf Of Katherina Garcia Garcia

Sent: Monday, 16 May 2011 6:46 PM

To: [hidden email]

Subject: Laser Protection



 



*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



-- Hello:



I am the person in charge of a Microscopy Facility. The Occupational Risks

Department in our University came to check the hazards of the confocal

microscopes (lasers). They told us to use some kind of glasses (for me when I

have to check lasers) and to put a window around the confocal for users

protection against lasers.



I contacted one company (Lasing), but it is difficult to find something suitable

for all the laser lines (range between 405 and 635). There are several kind of

glasses and you can buy a few to cover the range. Regarding windows, I only

found one that protects between 315 and 540.



Can you help me with this? Do you have references I can contact?



Thanks in advance for your help.

Best regards,

Katherina









---------------------------------------------------------------

Katherina García García

Técnico Especialista en Microscopía y Citometría

Tlf. 954977432

[hidden email]

Centro Andaluz de Biología del Desarrollo

Consejo Superior de Investigaciones Científicas

Universidad Pablo Olavide

Ctra. Utrera km1

41013 Sevilla

España



________________________________



No virus found in this message.

Checked by AVG - www.avg.com

Version: 10.0.1209 / Virus Database: 1500/3640 - Release Date: 05/15/11


 
R. Eric King R. Eric King
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Laser Protection

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear List,

Some of our customers have a micro-switch on the door of the lab; which is part of the interlock loop of the laser system. Most, if not all laser systems have a connector for this. But as mentioned before, this is typically for the service / maintenance of the laser system, or used on custom built "open table" set-ups.  

Yes, safety glasses are wavelength range specific:

http://en.wikipedia.org/wiki/Laser_safety#Protective_eyewear


Best Regards,




R. Eric King
OEM & International Sales Manager

[hidden email]

www.LaserInnovations.com



LASER

     INNOVATIONS

1150 East Main Street
Santa Paula, CA 93060
(805) 933-0015
(805) 933-0042 fax



www.CoherentLaser.com

www.244nm.com

www.337nm.com

www.355nm.com

www.488nm.com

www.532nm.com

www.Solidimaging.com

 




----- Original Message -----
From: "Andreas Bruckbauer" <[hidden email]>
To: <[hidden email]>
Sent: Monday, May 16, 2011 4:30 AM
Subject: Re: Laser Protection


*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


 Glasses only protect yourself, make sure nobody else enters the room during laser maintenance, e.g. putting a sign at the door. This is also a good idea when the service engineer comes to do the alignment and forgets to put up a sign.

 
best wishes

Andreas


 

 

-----Original Message-----
From: Gustin, Emmanuel [TIBBE] <[hidden email]>
To: [hidden email]
Sent: Mon, 16 May 2011 10:32
Subject: Re: Laser Protection


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



You need to make a distinction here between end users and maintainers.



From the end user's perspective, you expect that a standard microscope checks

out as a Class 1 laser system, i.e. intrinsically safe, with no need to provide

additional safety devices. The manual should specify this. Putting it in a black

box -- black shields being the only windows that will stop light at all visible

wavelengths -- should be superfluous, at least for safety purposes.



For people who open the instrument to check or align lasers it is different, as

this involves disabling or circumventing safety systems. Then you are working

with Class 3 or even Class 4 lasers, and in that case I can understand that OH&S

would insist on safety glasses. (AFAIK they should insist on a yearly visit to

an ophthalmologist to inspect your retina for damage as well.) It is true that

with relatively low power visible lasers (as opposed to the far more dangerous

IR lasers) few of the maintenance engineers bother to wear safety glasses. But

presumably they are trained in safe working practices; if you are not then

safety glasses are a wise precaution.



You always need specific glasses that matches the wavelengths (and power!) of

the laser you are working with; there is no general solution.



Best Regards,



Emmanuel





--

 Emmanuel Gustin,    Tel. (+32) 14 64 1586,    e-mail: [hidden email]  



-----Original Message-----

From: Confocal Microscopy List [mailto:[hidden email]] On

Behalf Of Guy Cox

Sent: maandag 16 mei 2011 10:58

To: [hidden email]

Subject: Re: Laser Protection



*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



Katherina,



 



                 All the confocal microscope manufacturers who sell microscopes

in the EEC ensure (as they are bound to) that their microscopes comply with EEC

laser regulations.  You should simply put your local Occupational Risks

department in contact with the suppliers of your systems.  It is unlikely that

you will have to change anything.



 



                                                                                         

Guy



 



Sponsor my next half-marathon on May 15th



There's a special reason - find it out at



http://www.everydayhero.com.au/Guy_Cox_4846 <http://www.everydayhero.com.au/Guy_Cox_4846>





______________________________________________



Associate Professor Guy Cox, MA, DPhil(Oxon)



Australian Centre for Microscopy & Microanalysis,



Madsen Building F09, University of Sydney, NSW 2006



 



Phone +61 2 9351 3176     Fax +61 2 9351 7682



             Mobile 0413 281 861



______________________________________________



      http://www.guycox.net <http://www.guycox.net/>



 



From: Confocal Microscopy List [mailto:[hidden email]] On

Behalf Of Katherina Garcia Garcia

Sent: Monday, 16 May 2011 6:46 PM

To: [hidden email]

Subject: Laser Protection



 



*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



-- Hello:



I am the person in charge of a Microscopy Facility. The Occupational Risks

Department in our University came to check the hazards of the confocal

microscopes (lasers). They told us to use some kind of glasses (for me when I

have to check lasers) and to put a window around the confocal for users

protection against lasers.



I contacted one company (Lasing), but it is difficult to find something suitable

for all the laser lines (range between 405 and 635). There are several kind of

glasses and you can buy a few to cover the range. Regarding windows, I only

found one that protects between 315 and 540.



Can you help me with this? Do you have references I can contact?



Thanks in advance for your help.

Best regards,

Katherina









---------------------------------------------------------------

Katherina García García

Técnico Especialista en Microscopía y Citometría

Tlf. 954977432

[hidden email]

Centro Andaluz de Biología del Desarrollo

Consejo Superior de Investigaciones Científicas

Universidad Pablo Olavide

Ctra. Utrera km1

41013 Sevilla

España



________________________________



No virus found in this message.

Checked by AVG - www.avg.com

Version: 10.0.1209 / Virus Database: 1500/3640 - Release Date: 05/15/11