![]() |
I've been getting this error when i shoot larger z stacks and when i
shoot longer time series Serious Error: NO file could be found matching the specification 'C:\Temp\untitled\untitled1_raw*.pic' then i either click ok or close the dialog box and lasersharp 2000 closes/crashes but when i shoot smaller stacks this does NOT happen any help would be greatly appreciated ive tried deleting the C:\Temp folder and this does nothing The reason i need this to finish is so that i can then export it to my desired location on a different server. Thank You --- Matiar Jafari |
![]() |
Try increasing the memory allocated to image acquisition. In Tools
select System Config and set memory to size smaller than Computer RAM. Also make sure your computer hard drive has plenty of space. You cannot collect images/stacks greater than the number set in System Config. good luck, michael On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: > I've been getting this error when i shoot larger z stacks and when i > shoot longer time series > > Serious Error: NO file could be found matching the specification > 'C:\Temp\untitled\untitled1_raw*.pic' > > then i either click ok or close the dialog box and lasersharp 2000 > closes/crashes > > but when i shoot smaller stacks this does NOT happen any help would be > greatly appreciated > > ive tried deleting the C:\Temp folder and this does nothing > > The reason i need this to finish is so that i can then export it to my > desired location on a different server. > > Thank You > --- > Matiar Jafari Michael Bastiani [hidden email] |
![]() |
Hi Matiar
It may be that the drive where your temp directory is located is running short of space thus preventing acquistion. LS2K writes files to c:\temp first, in your case, as they are acquired as untitledN ... and them copy (or move?) to the directory you specified. In general, acquistion does not depend on the amount of RAM installed. As Michael suggested, go to Tools | System configuration, look under the Scan System tab for Collection Limits to make sure that it is set properly. Depending on the verion of LS2K, it could be either 1 GB (pre 5.2, if I remember correctly) or 2 GB. This limit is hardcoded and cannot be increased. If your stack will go over that limit LS2K will not let you. You can also change where the Temporary Directory is located, may be point it to a larger disk. You do need to have system access level in LS2K to change things in System config, I think. You can check that in Tools | User Setup. Good luck. On Tue, 24 Mar 2009, Michael Bastiani wrote: > Try increasing the memory allocated to image acquisition. In Tools select > System Config and set memory to size smaller than Computer RAM. Also make > sure your computer hard drive has plenty of space. > You cannot collect images/stacks greater than the number set in System Config. > good luck, > michael > > > On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: > > > I've been getting this error when i shoot larger z stacks and when i > > shoot longer time series > > > > Serious Error: NO file could be found matching the specification > > 'C:\Temp\untitled\untitled1_raw*.pic' > > > > then i either click ok or close the dialog box and lasersharp 2000 > > closes/crashes > > > > but when i shoot smaller stacks this does NOT happen any help would be > > greatly appreciated > > > > ive tried deleting the C:\Temp folder and this does nothing > > > > The reason i need this to finish is so that i can then export it to my > > desired location on a different server. > > > > Thank You > > --- > > Matiar Jafari -- Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) |
![]() |
In reply to this post by Michael Bastiani
Hi Matiar,
Chang and others have beaten me too it but anyway: I've had this error before on our Radiance 2000 and it was fixed pretty easily and permanently after a chat to the Bio-Rad engineer, but it was sorted out years ago so I am rather fuzzy on the reason for it and the cure. I'm sure though it is a memory error of some sort [I assume you are still running LaserSharp 2000 under NT - our error wasn't on the latest version 6 but earlier, probably v4]. We did add more physical memory - to 768Mb or possibly 1Gb [try eBay for cheap s/h ECC memory, and check if you need matched pairs], plus we played around with the hard drive space, system page file, LaserSharp settings logged in as system, and we put the crucial :\temp on our very much larger D:\ drive [set via Lasersharp system] as the C:\ drive is always compromised to ~4Gb under NT, unless you increased it after SP6 is applied. We had two hard drives fitted, the first split into two partitions C:,E: and the much larger second one D: for user data and the Temp folder So I'm sure all the other postings are on the right track, it's system or hard drive memory and the 1Gb/2Gb limit - see Pangs post [no idea what error message we got though other than the Z-stacks/time-lapses never finished and we lost everything]. I think we still had to split very long time-lapses with z stacks though into two runs [2Gb limit]. Also allocated more virtual memory [page file] from the hard drive to help out the physical RAM [forgotten how to do that under NT, but it's probably under 'system']. I think there used to be a 2Gb limit on collections though even with LaserSharp 6 [1Gb default] even with the tweeks. Zeiss [Microscience] should still support the Radiance 2000 [not the 1024 though] so an email to their service support should clarify the solution. I don't know if upgrading to XP ever helped [you needed a new Firewire card and deep pockets for Bio-Rad to do this for you]. Again look at the old postings below that seem very relevant. Keith --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ ------------------------------------------------------------------- Subject: Re: Radiance 2000 (was Re: [CONFOCAL] Bio-Rad 1024) From: W. Chan Reply-To: Confocal Microscopy List <[hidden email]> Date: Tue, 21 Mar 2006 16:59:39 -0800 Content-Type: TEXT/PLAIN Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal The Control Panel error message is quite misleading. For v5.2, it states that "Collection Requires Too Much Memory" when the predicted dataset size exceeds the collection limit, but in fact it refers to disk space. I agree. Even a 2 GB limit is really not much when you need multi-channels with a merge pane at 1024x1024 with multiple time points. May be a work around is not to have a merge pane which can be generated quite easily after you have the data. A few of our users do not use the merge pane option in their methods. They do not use LaserSharp to analyze their images so they can't read the merge pane .pmf file even if they want to. -- Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) On Sat, 18 Mar 2006, Michael Cammer wrote: > I'll check the upgrade software for writing to disk. But with multiple > probe files our problem was the merge file. For two probes, it needed > 150% more memory than the raw images themselves. Even changing the > configuration didn't provide enough memory. > > >> Are you referring to RAM or harddisk space? As far as I know, LaserSharps >> writes to temporary files rather than holding the experiment in RAM. We >> have a Radiance 2000 on NT (1 GB RAM, 1.5 GB page file) with LaserSharp >> 5.2 build 824. The collection limit is 2 GB. You may want to check >> "System Configuration | Scan system" for the setting of the collection >> limit. Previous versions may have a 1 GB limit. Also check the location >> of the temporary files which needs to have plenty of space (as much as you >> want to acquire each time or more). Good luck. >> >> -- >> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >> >> On Fri, 17 Mar 2006, Michael Cammer wrote: >> >>> We have the Radiance 2000. This is a major limitation of the system. >>> We've had to compromise by restarting long time lapse experiments in the >>> middle, by taking fewer Z planes per timepoint or longer intervals. >>> But when our BioRad confocals die, LaserSharp software should be >>> obsolete. >>> -Michael >>> >>> At 05:51 PM 03/14/06 -0800, you wrote: >>>> >>>> I have a question regarding a BioRad 1024 confocal system. This system >>>> is >>>> working fine. But one trouble is that people cannot run any scanning >>>> program (time lapse experiments) that contains more than 230 mb, >>>> although >>>> the system has plenty of free memory (we just added another 1gb to it). >>>> I >>>> will appreciate if anyone can share with us if you had same experience >>>> and >>>> managed to resolve it. RECENT POSTINGS -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michael Bastiani Sent: 24 March 2009 13:57 To: [hidden email] Subject: Re: LaserSharp 2000 Error Try increasing the memory allocated to image acquisition. In Tools select System Config and set memory to size smaller than Computer RAM. Also make sure your computer hard drive has plenty of space. You cannot collect images/stacks greater than the number set in System Config. good luck, michael On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: > I've been getting this error when i shoot larger z stacks and when i > shoot longer time series > > Serious Error: NO file could be found matching the specification > 'C:\Temp\untitled\untitled1_raw*.pic' > > then i either click ok or close the dialog box and lasersharp 2000 > closes/crashes > > but when i shoot smaller stacks this does NOT happen any help would be > greatly appreciated > > ive tried deleting the C:\Temp folder and this does nothing > > The reason i need this to finish is so that i can then export it to my > desired location on a different server. > > Thank You > --- > Matiar Jafari Michael Bastiani [hidden email] |
![]() |
Dear all,
So according to Chan the limit is hardcoded and cannot be changed.... we have multiple accounts on the computer and im wondering if im not able to change he memory option from 700 mb to the 1gb because its hardcoded or because i need the "admin" login user for lasersharp 2000....help? also when i change the directory for the temp files it doesn't allow me to click ok thus it doesnt change where its located....any help??? On Tue, Mar 24, 2009 at 11:12 AM, Keith Morris <[hidden email]> wrote: > Hi Matiar, > Chang and others have beaten me too it but anyway: > > I've had this error before on our Radiance 2000 and it was fixed pretty > easily and permanently after a chat to the Bio-Rad engineer, but it was > sorted out years ago so I am rather fuzzy on the reason for it and the cure. > I'm sure though it is a memory error of some sort [I assume you are still > running LaserSharp 2000 under NT - our error wasn't on the latest version 6 > but earlier, probably v4]. We did add more physical memory - to 768Mb or > possibly 1Gb [try eBay for cheap s/h ECC memory, and check if you need > matched pairs], plus we played around with the hard drive space, system page > file, LaserSharp settings logged in as system, and we put the crucial :\temp > on our very much larger D:\ drive [set via Lasersharp system] as the C:\ > drive is always compromised to ~4Gb under NT, unless you increased it after > SP6 is applied. We had two hard drives fitted, the first split into two > partitions C:,E: and the much larger second one D: for user data and the > Temp folder > > So I'm sure all the other postings are on the right track, it's system or > hard drive memory and the 1Gb/2Gb limit - see Pangs post [no idea what error > message we got though other than the Z-stacks/time-lapses never finished and > we lost everything]. I think we still had to split very long time-lapses > with z stacks though into two runs [2Gb limit]. Also allocated more virtual > memory [page file] from the hard drive to help out the physical RAM > [forgotten how to do that under NT, but it's probably under 'system']. I > think there used to be a 2Gb limit on collections though even with > LaserSharp 6 [1Gb default] even with the tweeks. > > Zeiss [Microscience] should still support the Radiance 2000 [not the 1024 > though] so an email to their service support should clarify the solution. I > don't know if upgrading to XP ever helped [you needed a new Firewire card > and deep pockets for Bio-Rad to do this for you]. > > Again look at the old postings below that seem very relevant. > > Keith > > --------------------------------------------------------------------------- > Dr Keith J. Morris, > Molecular Cytogenetics and Microscopy Core, > Laboratory 00/069 and 00/070, > The Wellcome Trust Centre for Human Genetics, > Roosevelt Drive, > Oxford OX3 7BN, > United Kingdom. > > Telephone: +44 (0)1865 287568 > Email: [hidden email] > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ > > > > > > > > ------------------------------------------------------------------- > Subject: Re: Radiance 2000 (was Re: [CONFOCAL] Bio-Rad 1024) > From: W. Chan > Reply-To: Confocal Microscopy List <[hidden email]> > Date: Tue, 21 Mar 2006 16:59:39 -0800 > Content-Type: TEXT/PLAIN > > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > The Control Panel error message is quite misleading. For v5.2, it states > that "Collection Requires Too Much Memory" when the predicted dataset size > exceeds the collection limit, but in fact it refers to disk space. > > I agree. Even a 2 GB limit is really not much when you need > multi-channels with a merge pane at 1024x1024 with multiple time points. > May be a work around is not to have a merge pane which can be generated > quite easily after you have the data. A few of our users do not use > the merge pane option in their methods. They do not use LaserSharp to > analyze their images so they can't read the merge pane .pmf file even if > they want to. > > -- > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) > > On Sat, 18 Mar 2006, Michael Cammer wrote: > >> I'll check the upgrade software for writing to disk. But with multiple >> probe files our problem was the merge file. For two probes, it needed >> 150% more memory than the raw images themselves. Even changing the >> configuration didn't provide enough memory. >> >> >>> Are you referring to RAM or harddisk space? As far as I know, > LaserSharps >>> writes to temporary files rather than holding the experiment in RAM. We >>> have a Radiance 2000 on NT (1 GB RAM, 1.5 GB page file) with LaserSharp >>> 5.2 build 824. The collection limit is 2 GB. You may want to check >>> "System Configuration | Scan system" for the setting of the collection >>> limit. Previous versions may have a 1 GB limit. Also check the location >>> of the temporary files which needs to have plenty of space (as much as > you >>> want to acquire each time or more). Good luck. >>> >>> -- >>> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >>> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >>> >>> On Fri, 17 Mar 2006, Michael Cammer wrote: >>> >>>> We have the Radiance 2000. This is a major limitation of the system. >>>> We've had to compromise by restarting long time lapse experiments in the >>>> middle, by taking fewer Z planes per timepoint or longer intervals. >>>> But when our BioRad confocals die, LaserSharp software should be >>>> obsolete. >>>> -Michael >>>> >>>> At 05:51 PM 03/14/06 -0800, you wrote: >>>>> >>>>> I have a question regarding a BioRad 1024 confocal system. This system >>>>> is >>>>> working fine. But one trouble is that people cannot run any scanning >>>>> program (time lapse experiments) that contains more than 230 mb, >>>>> although >>>>> the system has plenty of free memory (we just added another 1gb to it). >>>>> I >>>>> will appreciate if anyone can share with us if you had same experience >>>>> and >>>>> managed to resolve it. > > > > > > > > > RECENT POSTINGS > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On > Behalf Of Michael Bastiani > Sent: 24 March 2009 13:57 > To: [hidden email] > Subject: Re: LaserSharp 2000 Error > > Try increasing the memory allocated to image acquisition. In Tools > select System Config and set memory to size smaller than Computer > RAM. Also make sure your computer hard drive has plenty of space. > You cannot collect images/stacks greater than the number set in System > Config. > good luck, > michael > > > On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: > >> I've been getting this error when i shoot larger z stacks and when i >> shoot longer time series >> >> Serious Error: NO file could be found matching the specification >> 'C:\Temp\untitled\untitled1_raw*.pic' >> >> then i either click ok or close the dialog box and lasersharp 2000 >> closes/crashes >> >> but when i shoot smaller stacks this does NOT happen any help would be >> greatly appreciated >> >> ive tried deleting the C:\Temp folder and this does nothing >> >> The reason i need this to finish is so that i can then export it to my >> desired location on a different server. >> >> Thank You >> --- >> Matiar Jafari > > Michael Bastiani > [hidden email] > -- Matiar Jafari |
![]() |
Hi Matiar
What I meant was e.g., in pre-5.2, you can only give max 1 GB to the application. It is not RAM related. Since you did not mention what version and OS you're running, I just assumed you're still using NT. LaserSharp 6 and XP run really well together; but that's another story. I only have access to v6 now so the terms in the menu may be different than earlier version. System access in lasersharp is completely independent of NT's (or whatever window's) user privilege. Usually, the software is set up with an user called Administrator that have system access, other users for the software may only have read & write access. After you log into lasersharp, go to Tools | User Setup, the second column will tell you what Access level you have. Whoever install the software should have the password to log in as Administrator. There is a way to find that out. Everything for this software are stored in a database e.g., c:\LaserSharpNT\Database\lsMessages.mdb. If you have microsoft access, you can open a COPY of this file and look at the table called LaserSharpUsers and you can see all the user passwords in plain text. Access level 0 is for system. It is probably overly cautious to open a copy because the database will likely be opened as read-only; but why take a chance. Once you log in as an user with system access level, you should be able to change all the configuration parameters. But be very careful, you can mess up things really bad. I've learned that the hard way. Hope this help. On Tue, 24 Mar 2009, Matiar Jafari wrote: > Dear all, > > So according to Chan the limit is hardcoded and cannot be changed.... > > we have multiple accounts on the computer and im wondering if im not > able to change he memory option from 700 mb to the 1gb because its > hardcoded or because i need the "admin" login user for lasersharp > 2000....help? also when i change the directory for the temp files it > doesn't allow me to click ok thus it doesnt change where its > located....any help??? > > On Tue, Mar 24, 2009 at 11:12 AM, Keith Morris <[hidden email]> wrote: > > Hi Matiar, > > Chang and others have beaten me too it but anyway: > > > > I've had this error before on our Radiance 2000 and it was fixed pretty > > easily and permanently after a chat to the Bio-Rad engineer, but it was > > sorted out years ago so I am rather fuzzy on the reason for it and the cure. > > I'm sure though it is a memory error of some sort [I assume you are still > > running LaserSharp 2000 under NT - our error wasn't on the latest version 6 > > but earlier, probably v4]. We did add more physical memory - to 768Mb or > > possibly 1Gb [try eBay for cheap s/h ECC memory, and check if you need > > matched pairs], plus we played around with the hard drive space, system page > > file, LaserSharp settings logged in as system, and we put the crucial :\temp > > on our very much larger D:\ drive [set via Lasersharp system] as the C:\ > > drive is always compromised to ~4Gb under NT, unless you increased it after > > SP6 is applied. We had two hard drives fitted, the first split into two > > partitions C:,E: and the much larger second one D: for user data and the > > Temp folder > > > > So I'm sure all the other postings are on the right track, it's system or > > hard drive memory and the 1Gb/2Gb limit - see Pangs post [no idea what error > > message we got though other than the Z-stacks/time-lapses never finished and > > we lost everything]. I think we still had to split very long time-lapses > > with z stacks though into two runs [2Gb limit]. Also allocated more virtual > > memory [page file] from the hard drive to help out the physical RAM > > [forgotten how to do that under NT, but it's probably under 'system']. I > > think there used to be a 2Gb limit on collections though even with > > LaserSharp 6 [1Gb default] even with the tweeks. > > > > Zeiss [Microscience] should still support the Radiance 2000 [not the 1024 > > though] so an email to their service support should clarify the solution. I > > don't know if upgrading to XP ever helped [you needed a new Firewire card > > and deep pockets for Bio-Rad to do this for you]. > > > > Again look at the old postings below that seem very relevant. > > > > Keith > > > > --------------------------------------------------------------------------- > > Dr Keith J. Morris, > > Molecular Cytogenetics and Microscopy Core, > > Laboratory 00/069 and 00/070, > > The Wellcome Trust Centre for Human Genetics, > > Roosevelt Drive, > > Oxford OX3 7BN, > > United Kingdom. > > > > Telephone: +44 (0)1865 287568 > > Email: [hidden email] > > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ > > > > > > > > > > > > > > > > ------------------------------------------------------------------- > > Subject: Re: Radiance 2000 (was Re: [CONFOCAL] Bio-Rad 1024) > > From: W. Chan > > Reply-To: Confocal Microscopy List <[hidden email]> > > Date: Tue, 21 Mar 2006 16:59:39 -0800 > > Content-Type: TEXT/PLAIN > > > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > > The Control Panel error message is quite misleading. For v5.2, it states > > that "Collection Requires Too Much Memory" when the predicted dataset size > > exceeds the collection limit, but in fact it refers to disk space. > > > > I agree. Even a 2 GB limit is really not much when you need > > multi-channels with a merge pane at 1024x1024 with multiple time points. > > May be a work around is not to have a merge pane which can be generated > > quite easily after you have the data. A few of our users do not use > > the merge pane option in their methods. They do not use LaserSharp to > > analyze their images so they can't read the merge pane .pmf file even if > > they want to. > > > > -- > > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) > > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) > > > > On Sat, 18 Mar 2006, Michael Cammer wrote: > > > >> I'll check the upgrade software for writing to disk. But with multiple > >> probe files our problem was the merge file. For two probes, it needed > >> 150% more memory than the raw images themselves. Even changing the > >> configuration didn't provide enough memory. > >> > >> > >>> Are you referring to RAM or harddisk space? As far as I know, > > LaserSharps > >>> writes to temporary files rather than holding the experiment in RAM. We > >>> have a Radiance 2000 on NT (1 GB RAM, 1.5 GB page file) with LaserSharp > >>> 5.2 build 824. The collection limit is 2 GB. You may want to check > >>> "System Configuration | Scan system" for the setting of the collection > >>> limit. Previous versions may have a 1 GB limit. Also check the location > >>> of the temporary files which needs to have plenty of space (as much as > > you > >>> want to acquire each time or more). Good luck. > >>> > >>> -- > >>> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) > >>> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) > >>> > >>> On Fri, 17 Mar 2006, Michael Cammer wrote: > >>> > >>>> We have the Radiance 2000. This is a major limitation of the system. > >>>> We've had to compromise by restarting long time lapse experiments in the > >>>> middle, by taking fewer Z planes per timepoint or longer intervals. > >>>> But when our BioRad confocals die, LaserSharp software should be > >>>> obsolete. > >>>> -Michael > >>>> > >>>> At 05:51 PM 03/14/06 -0800, you wrote: > >>>>> > >>>>> I have a question regarding a BioRad 1024 confocal system. This system > >>>>> is > >>>>> working fine. But one trouble is that people cannot run any scanning > >>>>> program (time lapse experiments) that contains more than 230 mb, > >>>>> although > >>>>> the system has plenty of free memory (we just added another 1gb to it). > >>>>> I > >>>>> will appreciate if anyone can share with us if you had same experience > >>>>> and > >>>>> managed to resolve it. > > > > RECENT POSTINGS > > -----Original Message----- > > From: Confocal Microscopy List [mailto:[hidden email]] On > > Behalf Of Michael Bastiani > > Sent: 24 March 2009 13:57 > > To: [hidden email] > > Subject: Re: LaserSharp 2000 Error > > > > Try increasing the memory allocated to image acquisition. In Tools > > select System Config and set memory to size smaller than Computer > > RAM. Also make sure your computer hard drive has plenty of space. > > You cannot collect images/stacks greater than the number set in System > > Config. > > good luck, > > michael > > > > > > On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: > > > >> I've been getting this error when i shoot larger z stacks and when i > >> shoot longer time series > >> > >> Serious Error: NO file could be found matching the specification > >> 'C:\Temp\untitled\untitled1_raw*.pic' > >> > >> then i either click ok or close the dialog box and lasersharp 2000 > >> closes/crashes > >> > >> but when i shoot smaller stacks this does NOT happen any help would be > >> greatly appreciated > >> > >> ive tried deleting the C:\Temp folder and this does nothing > >> > >> The reason i need this to finish is so that i can then export it to my > >> desired location on a different server. > >> > >> Thank You > >> --- > >> Matiar Jafari Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) |
![]() |
Dear Matiar Jafari,
Have you actually contacted a Carl Zeiss service representative about this problem?
From my experience and that of others, Zeiss will assist with such problems. I also have it on good authority that although one can no longer buy a service/maintenance contract for an MRC-1024, they will endeavour to service the system on a billable basis for on-site work and, as Bio-Rad did in the past, offer remote assistance at no cost.
Whether you use Zeiss for support or not, knowing a bit more about what type of confocal you actually have and its configuration would be helpful and avoid misdiagnosis.
Good luck,
PH
On Wed, Mar 25, 2009 at 1:24 AM, W. Chan <[hidden email]> wrote: Hi Matiar |
![]() |
New error message after the first one I get. Failed to load PicMat with data fileException in channels, stop capture I changed the ram capacity to 2gb and I tries changing the harddrive temp. I'm shooting then exporting but I can't export as it gives me the error after I shoot anything larger than~80um Thank You --- Matiar Jafari
|
![]() |
In reply to this post by Prof Hindsight
More memory with NT/Lasersharp only helps
if you are really down at the 256Mb level, and mostly it gives stability/speed
provided the modules are matched [all ECC] because you probably still have that
1Gb to 2Gb upper limit built into Lasersharp. Besides the hard drive page-file
can [slowly] provide more memory. Check your PC motherboard can cope with 2Gb
physical RAM [older systems have 1Gb and even 512Mb limits] – crucial.com
often gives this info. Don’t know if you have a MRC1024 or Radiance, but
Zeiss [Microscience] should be able to advise if your error messages are
consistent with Lasersharp v4/v5 operating normally or it’s an error
[still sound like its hitting the Lasersharp image capture memory limit of 1Gb
or 2Gb depending on version – unless of course you used to be able to
capture larger z-stacks]. Once you hit that Lasersharp limit, it’s either
an upgrade or you have to break the stack into a couple of separate ones. From
memory when an error message talks specifically about ‘not enough memory
for capture’ it frequently means hard drive space not system memory. Our Bio-Rad 1024 was no longer fully
supported by Zeiss Micro-Science a few years ago now and so we passed it on to
those less fortunate and bought a Zeiss 510 Meta. Zeiss did say they would continue
to repair the MRC 1024 system while they could still get parts, and that they
would continue to provide technical support advice [which should be free via telephone/email].
Many Bio-Rad staff have obviously left the
company since UK Bio-Rad confocal production was stopped by Zeiss, although
they retained core personal for product support of the newer Radiance 2000
confocals that also run LaserSharp - some of whom will still remember the
elderly 1024 [and perhaps even COMOS, let alone LaserSharp]. Plus all the
knowledgeable Bio-Rad confocal reps generally live on working for other
microscope companies. Keith --------------------------------------------------------------------------- New error message after the first one I get. Failed to load PicMat with
data fileException in channels, stop capture I changed the ram capacity to 2gb and I tries changing the harddrive
temp. I'm shooting then exporting but I can't export as it gives me the error
after I shoot anything larger than~80um --- Matiar Jafari
Dear Matiar Jafari, Have you actually contacted a Carl Zeiss service representative about
this problem? From my experience and that of others, Zeiss will assist with such
problems. I also have it on good authority that although one can no
longer buy a service/maintenance contract for an MRC-1024, they will endeavour
to service the system on a billable basis for on-site work and, as Bio-Rad did
in the past, offer remote assistance at no cost. Whether you use Zeiss for support or not, knowing a bit more about what
type of confocal you actually have and its configuration would be helpful and
avoid misdiagnosis. Good luck, PH On Wed, Mar 25, 2009 at 1:24 AM, W. Chan <[hidden email][hidden email]>
wrote: Hi Matiar
-- Pang (Wai Pang Chan, [hidden email][hidden email],
PAB A087, 206-685-1519) |
![]() |
when i try to acquire >80um stacks its at about either 80-100mb(i
forgot not sure) The system is running on windowns NT version 6 build(i forget i could check later) The error has always been there since ive started working with the confocal and since others have been using it as well. Ive changed the hard drive temp save area from my c drive(has something around 70gb of free space) to the D drive which only has like 10gb of free sapce and ive tried changing it to a server which has 200gb of free space. Ive also played with the RAM changed it from 1000Mb to 2000MB still no dice I called a zeiss representative yesterday and he said that its a known bug and there are two fixes to create a new channel with the laser off or to install a service pack which he was supposed to email me but i never received. system has only 1gb of memory. But still i dont think this matters as i am ONLY capturing a stack that is around 80mb. So far the farthest ive been able to get with largest z stack is 28 um which is about 50 MB On Fri, Mar 27, 2009 at 2:42 AM, Keith Morris <[hidden email]> wrote: > More memory with NT/Lasersharp only helps if you are really down at the > 256Mb level, and mostly it gives stability/speed provided the modules are > matched [all ECC] because you probably still have that 1Gb to 2Gb upper > limit built into Lasersharp. Besides the hard drive page-file can [slowly] > provide more memory. Check your PC motherboard can cope with 2Gb physical > RAM [older systems have 1Gb and even 512Mb limits] – crucial.com often gives > this info. Don’t know if you have a MRC1024 or Radiance, but Zeiss > [Microscience] should be able to advise if your error messages are > consistent with Lasersharp v4/v5 operating normally or it’s an error [still > sound like its hitting the Lasersharp image capture memory limit of 1Gb or > 2Gb depending on version – unless of course you used to be able to capture > larger z-stacks]. Once you hit that Lasersharp limit, it’s either an upgrade > or you have to break the stack into a couple of separate ones. From memory > when an error message talks specifically about ‘not enough memory for > capture’ it frequently means hard drive space not system memory. > > > > Our Bio-Rad 1024 was no longer fully supported by Zeiss Micro-Science a few > years ago now and so we passed it on to those less fortunate and bought a > Zeiss 510 Meta. Zeiss did say they would continue to repair the MRC 1024 > system while they could still get parts, and that they would continue to > provide technical support advice [which should be free via telephone/email]. > > > > Many Bio-Rad staff have obviously left the company since UK Bio-Rad confocal > production was stopped by Zeiss, although they retained core personal for > product support of the newer Radiance 2000 confocals that also run > LaserSharp - some of whom will still remember the elderly 1024 [and perhaps > even COMOS, let alone LaserSharp]. Plus all the knowledgeable Bio-Rad > confocal reps generally live on working for other microscope companies. > > > > Keith > > --------------------------------------------------------------------------- > Dr Keith J. Morris, > Molecular Cytogenetics and Microscopy Core, > Laboratory 00/069 and 00/070, > The Wellcome Trust Centre for Human Genetics, > Roosevelt Drive, > Oxford OX3 7BN, > United Kingdom. > > Telephone: +44 (0)1865 287568 > Email: [hidden email] > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ > > New error message after the first one I get. Failed to load PicMat with data > fileException in channels, stop capture > > > > I changed the ram capacity to 2gb and I tries changing the harddrive temp. > > > > I'm shooting then exporting but I can't export as it gives me the error > after I shoot anything larger than~80um > > Thank You > > --- > > Matiar Jafari > > On Mar 25, 2009, at 2:51 AM, Prof Hindsight <[hidden email]> > wrote: > > Dear Matiar Jafari, > > > > Have you actually contacted a Carl Zeiss service representative about this > problem? > > > > From my experience and that of others, Zeiss will assist with such > problems. I also have it on good authority that although one can no longer > buy a service/maintenance contract for an MRC-1024, they will endeavour to > service the system on a billable basis for on-site work and, as Bio-Rad did > in the past, offer remote assistance at no cost. > > Whether you use Zeiss for support or not, knowing a bit more about what type > of confocal you actually have and its configuration would be helpful and > avoid misdiagnosis. > > > > Good luck, > > > > PH > > On Wed, Mar 25, 2009 at 1:24 AM, W. Chan <[hidden email]> wrote: > > Hi Matiar > > What I meant was e.g., in pre-5.2, you can only give max 1 GB to the > application. It is not RAM related. Since you did not mention what > version and OS you're running, I just assumed you're still using NT. > LaserSharp 6 and XP run really well together; but that's another story. I > only have access to v6 now so the terms in the menu may be different than > earlier version. > > System access in lasersharp is completely independent of NT's (or whatever > window's) user privilege. Usually, the software is set up with an user > called Administrator that have system access, other users for the software > may only have read & write access. After you log into lasersharp, go to > Tools | User Setup, the second column will tell you what Access level you > have. Whoever install the software should have the password to log in as > Administrator. > > There is a way to find that out. Everything for this software are stored > in a database e.g., c:\LaserSharpNT\Database\lsMessages.mdb. If you have > microsoft access, you can open a COPY of this file and look at the table > called LaserSharpUsers and you can see all the user passwords in plain > text. Access level 0 is for system. It is probably overly cautious to > open a copy because the database will likely be opened as read-only; but > why take a chance. > > Once you log in as an user with system access level, you should be able to > change all the configuration parameters. But be very careful, you can > mess up things really bad. I've learned that the hard way. Hope this > help. > > On Tue, 24 Mar 2009, Matiar Jafari wrote: > >> Dear all, >> >> So according to Chan the limit is hardcoded and cannot be changed.... >> >> we have multiple accounts on the computer and im wondering if im not >> able to change he memory option from 700 mb to the 1gb because its >> hardcoded or because i need the "admin" login user for lasersharp >> 2000....help? also when i change the directory for the temp files it >> doesn't allow me to click ok thus it doesnt change where its >> located....any help??? >> >> On Tue, Mar 24, 2009 at 11:12 AM, Keith Morris <[hidden email]> >> wrote: >> > Hi Matiar, >> > Chang and others have beaten me too it but anyway: >> > >> > I've had this error before on our Radiance 2000 and it was fixed pretty >> > easily and permanently after a chat to the Bio-Rad engineer, but it was >> > sorted out years ago so I am rather fuzzy on the reason for it and the >> > cure. >> > I'm sure though it is a memory error of some sort [I assume you are >> > still >> > running LaserSharp 2000 under NT - our error wasn't on the latest >> > version 6 >> > but earlier, probably v4]. We did add more physical memory - to 768Mb or >> > possibly 1Gb [try eBay for cheap s/h ECC memory, and check if you need >> > matched pairs], plus we played around with the hard drive space, system >> > page >> > file, LaserSharp settings logged in as system, and we put the crucial >> > :\temp >> > on our very much larger D:\ drive [set via Lasersharp system] as the C:\ >> > drive is always compromised to ~4Gb under NT, unless you increased it >> > after >> > SP6 is applied. We had two hard drives fitted, the first split into two >> > partitions C:,E: and the much larger second one D: for user data and the >> > Temp folder >> > >> > So I'm sure all the other postings are on the right track, it's system >> > or >> > hard drive memory and the 1Gb/2Gb limit - see Pangs post [no idea what >> > error >> > message we got though other than the Z-stacks/time-lapses never finished >> > and >> > we lost everything]. I think we still had to split very long time-lapses >> > with z stacks though into two runs [2Gb limit]. Also allocated more >> > virtual >> > memory [page file] from the hard drive to help out the physical RAM >> > [forgotten how to do that under NT, but it's probably under 'system']. I >> > think there used to be a 2Gb limit on collections though even with >> > LaserSharp 6 [1Gb default] even with the tweeks. >> > >> > Zeiss [Microscience] should still support the Radiance 2000 [not the >> > 1024 >> > though] so an email to their service support should clarify the >> > solution. I >> > don't know if upgrading to XP ever helped [you needed a new Firewire >> > card >> > and deep pockets for Bio-Rad to do this for you]. >> > >> > Again look at the old postings below that seem very relevant. >> > >> > Keith >> > >> > >> > --------------------------------------------------------------------------- >> > Dr Keith J. Morris, >> > Molecular Cytogenetics and Microscopy Core, >> > Laboratory 00/069 and 00/070, >> > The Wellcome Trust Centre for Human Genetics, >> > Roosevelt Drive, >> > Oxford OX3 7BN, >> > United Kingdom. >> > >> > Telephone: +44 (0)1865 287568 >> > Email: [hidden email] >> > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ >> > >> > >> > >> > >> > >> > >> > >> > ------------------------------------------------------------------- >> > Subject: Re: Radiance 2000 (was Re: [CONFOCAL] Bio-Rad 1024) >> > From: W. Chan >> > Reply-To: Confocal Microscopy List <[hidden email]> >> > Date: Tue, 21 Mar 2006 16:59:39 -0800 >> > Content-Type: TEXT/PLAIN >> > >> > Search the CONFOCAL archive at >> > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> > >> > The Control Panel error message is quite misleading. For v5.2, it >> > states >> > that "Collection Requires Too Much Memory" when the predicted dataset >> > size >> > exceeds the collection limit, but in fact it refers to disk space. >> > >> > I agree. Even a 2 GB limit is really not much when you need >> > multi-channels with a merge pane at 1024x1024 with multiple time points. >> > May be a work around is not to have a merge pane which can be generated >> > quite easily after you have the data. A few of our users do not use >> > the merge pane option in their methods. They do not use LaserSharp to >> > analyze their images so they can't read the merge pane .pmf file even if >> > they want to. >> > >> > -- >> > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >> > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >> > >> > On Sat, 18 Mar 2006, Michael Cammer wrote: >> > >> >> I'll check the upgrade software for writing to disk. But with multiple >> >> probe files our problem was the merge file. For two probes, it needed >> >> 150% more memory than the raw images themselves. Even changing the >> >> configuration didn't provide enough memory. >> >> >> >> >> >>> Are you referring to RAM or harddisk space? As far as I know, >> > LaserSharps >> >>> writes to temporary files rather than holding the experiment in RAM. >> >>> We >> >>> have a Radiance 2000 on NT (1 GB RAM, 1.5 GB page file) with >> >>> LaserSharp >> >>> 5.2 build 824. The collection limit is 2 GB. You may want to check >> >>> "System Configuration | Scan system" for the setting of the collection >> >>> limit. Previous versions may have a 1 GB limit. Also check the >> >>> location >> >>> of the temporary files which needs to have plenty of space (as much as >> > you >> >>> want to acquire each time or more). Good luck. >> >>> >> >>> -- >> >>> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >> >>> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >> >>> >> >>> On Fri, 17 Mar 2006, Michael Cammer wrote: >> >>> >> >>>> We have the Radiance 2000. This is a major limitation of the system. >> >>>> We've had to compromise by restarting long time lapse experiments in >> >>>> the >> >>>> middle, by taking fewer Z planes per timepoint or longer intervals. >> >>>> But when our BioRad confocals die, LaserSharp software should be >> >>>> obsolete. >> >>>> -Michael >> >>>> >> >>>> At 05:51 PM 03/14/06 -0800, you wrote: >> >>>>> >> >>>>> I have a question regarding a BioRad 1024 confocal system. This >> >>>>> system >> >>>>> is >> >>>>> working fine. But one trouble is that people cannot run any >> >>>>> scanning >> >>>>> program (time lapse experiments) that contains more than 230 mb, >> >>>>> although >> >>>>> the system has plenty of free memory (we just added another 1gb to >> >>>>> it). >> >>>>> I >> >>>>> will appreciate if anyone can share with us if you had same >> >>>>> experience >> >>>>> and >> >>>>> managed to resolve it. >> > >> > RECENT POSTINGS >> > -----Original Message----- >> > From: Confocal Microscopy List [mailto:[hidden email]] >> > On >> > Behalf Of Michael Bastiani >> > Sent: 24 March 2009 13:57 >> > To: [hidden email] >> > Subject: Re: LaserSharp 2000 Error >> > >> > Try increasing the memory allocated to image acquisition. In Tools >> > select System Config and set memory to size smaller than Computer >> > RAM. Also make sure your computer hard drive has plenty of space. >> > You cannot collect images/stacks greater than the number set in System >> > Config. >> > good luck, >> > michael >> > >> > >> > On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: >> > >> >> I've been getting this error when i shoot larger z stacks and when i >> >> shoot longer time series >> >> >> >> Serious Error: NO file could be found matching the specification >> >> 'C:\Temp\untitled\untitled1_raw*.pic' >> >> >> >> then i either click ok or close the dialog box and lasersharp 2000 >> >> closes/crashes >> >> >> >> but when i shoot smaller stacks this does NOT happen any help would be >> >> greatly appreciated >> >> >> >> ive tried deleting the C:\Temp folder and this does nothing >> >> >> >> The reason i need this to finish is so that i can then export it to my >> >> desired location on a different server. >> >> >> >> Thank You >> >> --- >> >> Matiar Jafari > > -- > > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) > > > > -- Matiar Jafari |
![]() |
In reply to this post by Keith Morris
update... my apologies i made a mistake im running windows 2000 with
Lasersharp 2000 6.0 build 846 i just shot a stack that was only 19 um and it gave me the same error. I was told the work around was to "....the only solution is to use a method with 2 channels to collect data. This will allow you to collect up to the limits set by LaserSharp ~2GB. Once you have your data you can delete the empty/redundant channel of data." Zeiss Support im shooting with a green method with only one channel any help in how i would do this? thanks in advance Thank You On Fri, Mar 27, 2009 at 2:42 AM, Keith Morris <[hidden email]> wrote: > More memory with NT/Lasersharp only helps if you are really down at the > 256Mb level, and mostly it gives stability/speed provided the modules are > matched [all ECC] because you probably still have that 1Gb to 2Gb upper > limit built into Lasersharp. Besides the hard drive page-file can [slowly] > provide more memory. Check your PC motherboard can cope with 2Gb physical > RAM [older systems have 1Gb and even 512Mb limits] – crucial.com often gives > this info. Don’t know if you have a MRC1024 or Radiance, but Zeiss > [Microscience] should be able to advise if your error messages are > consistent with Lasersharp v4/v5 operating normally or it’s an error [still > sound like its hitting the Lasersharp image capture memory limit of 1Gb or > 2Gb depending on version – unless of course you used to be able to capture > larger z-stacks]. Once you hit that Lasersharp limit, it’s either an upgrade > or you have to break the stack into a couple of separate ones. From memory > when an error message talks specifically about ‘not enough memory for > capture’ it frequently means hard drive space not system memory. > > > > Our Bio-Rad 1024 was no longer fully supported by Zeiss Micro-Science a few > years ago now and so we passed it on to those less fortunate and bought a > Zeiss 510 Meta. Zeiss did say they would continue to repair the MRC 1024 > system while they could still get parts, and that they would continue to > provide technical support advice [which should be free via telephone/email]. > > > > Many Bio-Rad staff have obviously left the company since UK Bio-Rad confocal > production was stopped by Zeiss, although they retained core personal for > product support of the newer Radiance 2000 confocals that also run > LaserSharp - some of whom will still remember the elderly 1024 [and perhaps > even COMOS, let alone LaserSharp]. Plus all the knowledgeable Bio-Rad > confocal reps generally live on working for other microscope companies. > > > > Keith > > --------------------------------------------------------------------------- > Dr Keith J. Morris, > Molecular Cytogenetics and Microscopy Core, > Laboratory 00/069 and 00/070, > The Wellcome Trust Centre for Human Genetics, > Roosevelt Drive, > Oxford OX3 7BN, > United Kingdom. > > Telephone: +44 (0)1865 287568 > Email: [hidden email] > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ > > New error message after the first one I get. Failed to load PicMat with data > fileException in channels, stop capture > > > > I changed the ram capacity to 2gb and I tries changing the harddrive temp. > > > > I'm shooting then exporting but I can't export as it gives me the error > after I shoot anything larger than~80um > > Thank You > > --- > > Matiar Jafari > > On Mar 25, 2009, at 2:51 AM, Prof Hindsight <[hidden email]> > wrote: > > Dear Matiar Jafari, > > > > Have you actually contacted a Carl Zeiss service representative about this > problem? > > > > From my experience and that of others, Zeiss will assist with such > problems. I also have it on good authority that although one can no longer > buy a service/maintenance contract for an MRC-1024, they will endeavour to > service the system on a billable basis for on-site work and, as Bio-Rad did > in the past, offer remote assistance at no cost. > > Whether you use Zeiss for support or not, knowing a bit more about what type > of confocal you actually have and its configuration would be helpful and > avoid misdiagnosis. > > > > Good luck, > > > > PH > > On Wed, Mar 25, 2009 at 1:24 AM, W. Chan <[hidden email]> wrote: > > Hi Matiar > > What I meant was e.g., in pre-5.2, you can only give max 1 GB to the > application. It is not RAM related. Since you did not mention what > version and OS you're running, I just assumed you're still using NT. > LaserSharp 6 and XP run really well together; but that's another story. I > only have access to v6 now so the terms in the menu may be different than > earlier version. > > System access in lasersharp is completely independent of NT's (or whatever > window's) user privilege. Usually, the software is set up with an user > called Administrator that have system access, other users for the software > may only have read & write access. After you log into lasersharp, go to > Tools | User Setup, the second column will tell you what Access level you > have. Whoever install the software should have the password to log in as > Administrator. > > There is a way to find that out. Everything for this software are stored > in a database e.g., c:\LaserSharpNT\Database\lsMessages.mdb. If you have > microsoft access, you can open a COPY of this file and look at the table > called LaserSharpUsers and you can see all the user passwords in plain > text. Access level 0 is for system. It is probably overly cautious to > open a copy because the database will likely be opened as read-only; but > why take a chance. > > Once you log in as an user with system access level, you should be able to > change all the configuration parameters. But be very careful, you can > mess up things really bad. I've learned that the hard way. Hope this > help. > > On Tue, 24 Mar 2009, Matiar Jafari wrote: > >> Dear all, >> >> So according to Chan the limit is hardcoded and cannot be changed.... >> >> we have multiple accounts on the computer and im wondering if im not >> able to change he memory option from 700 mb to the 1gb because its >> hardcoded or because i need the "admin" login user for lasersharp >> 2000....help? also when i change the directory for the temp files it >> doesn't allow me to click ok thus it doesnt change where its >> located....any help??? >> >> On Tue, Mar 24, 2009 at 11:12 AM, Keith Morris <[hidden email]> >> wrote: >> > Hi Matiar, >> > Chang and others have beaten me too it but anyway: >> > >> > I've had this error before on our Radiance 2000 and it was fixed pretty >> > easily and permanently after a chat to the Bio-Rad engineer, but it was >> > sorted out years ago so I am rather fuzzy on the reason for it and the >> > cure. >> > I'm sure though it is a memory error of some sort [I assume you are >> > still >> > running LaserSharp 2000 under NT - our error wasn't on the latest >> > version 6 >> > but earlier, probably v4]. We did add more physical memory - to 768Mb or >> > possibly 1Gb [try eBay for cheap s/h ECC memory, and check if you need >> > matched pairs], plus we played around with the hard drive space, system >> > page >> > file, LaserSharp settings logged in as system, and we put the crucial >> > :\temp >> > on our very much larger D:\ drive [set via Lasersharp system] as the C:\ >> > drive is always compromised to ~4Gb under NT, unless you increased it >> > after >> > SP6 is applied. We had two hard drives fitted, the first split into two >> > partitions C:,E: and the much larger second one D: for user data and the >> > Temp folder >> > >> > So I'm sure all the other postings are on the right track, it's system >> > or >> > hard drive memory and the 1Gb/2Gb limit - see Pangs post [no idea what >> > error >> > message we got though other than the Z-stacks/time-lapses never finished >> > and >> > we lost everything]. I think we still had to split very long time-lapses >> > with z stacks though into two runs [2Gb limit]. Also allocated more >> > virtual >> > memory [page file] from the hard drive to help out the physical RAM >> > [forgotten how to do that under NT, but it's probably under 'system']. I >> > think there used to be a 2Gb limit on collections though even with >> > LaserSharp 6 [1Gb default] even with the tweeks. >> > >> > Zeiss [Microscience] should still support the Radiance 2000 [not the >> > 1024 >> > though] so an email to their service support should clarify the >> > solution. I >> > don't know if upgrading to XP ever helped [you needed a new Firewire >> > card >> > and deep pockets for Bio-Rad to do this for you]. >> > >> > Again look at the old postings below that seem very relevant. >> > >> > Keith >> > >> > >> > --------------------------------------------------------------------------- >> > Dr Keith J. Morris, >> > Molecular Cytogenetics and Microscopy Core, >> > Laboratory 00/069 and 00/070, >> > The Wellcome Trust Centre for Human Genetics, >> > Roosevelt Drive, >> > Oxford OX3 7BN, >> > United Kingdom. >> > >> > Telephone: +44 (0)1865 287568 >> > Email: [hidden email] >> > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ >> > >> > >> > >> > >> > >> > >> > >> > ------------------------------------------------------------------- >> > Subject: Re: Radiance 2000 (was Re: [CONFOCAL] Bio-Rad 1024) >> > From: W. Chan >> > Reply-To: Confocal Microscopy List <[hidden email]> >> > Date: Tue, 21 Mar 2006 16:59:39 -0800 >> > Content-Type: TEXT/PLAIN >> > >> > Search the CONFOCAL archive at >> > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> > >> > The Control Panel error message is quite misleading. For v5.2, it >> > states >> > that "Collection Requires Too Much Memory" when the predicted dataset >> > size >> > exceeds the collection limit, but in fact it refers to disk space. >> > >> > I agree. Even a 2 GB limit is really not much when you need >> > multi-channels with a merge pane at 1024x1024 with multiple time points. >> > May be a work around is not to have a merge pane which can be generated >> > quite easily after you have the data. A few of our users do not use >> > the merge pane option in their methods. They do not use LaserSharp to >> > analyze their images so they can't read the merge pane .pmf file even if >> > they want to. >> > >> > -- >> > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >> > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >> > >> > On Sat, 18 Mar 2006, Michael Cammer wrote: >> > >> >> I'll check the upgrade software for writing to disk. But with multiple >> >> probe files our problem was the merge file. For two probes, it needed >> >> 150% more memory than the raw images themselves. Even changing the >> >> configuration didn't provide enough memory. >> >> >> >> >> >>> Are you referring to RAM or harddisk space? As far as I know, >> > LaserSharps >> >>> writes to temporary files rather than holding the experiment in RAM. >> >>> We >> >>> have a Radiance 2000 on NT (1 GB RAM, 1.5 GB page file) with >> >>> LaserSharp >> >>> 5.2 build 824. The collection limit is 2 GB. You may want to check >> >>> "System Configuration | Scan system" for the setting of the collection >> >>> limit. Previous versions may have a 1 GB limit. Also check the >> >>> location >> >>> of the temporary files which needs to have plenty of space (as much as >> > you >> >>> want to acquire each time or more). Good luck. >> >>> >> >>> -- >> >>> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >> >>> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >> >>> >> >>> On Fri, 17 Mar 2006, Michael Cammer wrote: >> >>> >> >>>> We have the Radiance 2000. This is a major limitation of the system. >> >>>> We've had to compromise by restarting long time lapse experiments in >> >>>> the >> >>>> middle, by taking fewer Z planes per timepoint or longer intervals. >> >>>> But when our BioRad confocals die, LaserSharp software should be >> >>>> obsolete. >> >>>> -Michael >> >>>> >> >>>> At 05:51 PM 03/14/06 -0800, you wrote: >> >>>>> >> >>>>> I have a question regarding a BioRad 1024 confocal system. This >> >>>>> system >> >>>>> is >> >>>>> working fine. But one trouble is that people cannot run any >> >>>>> scanning >> >>>>> program (time lapse experiments) that contains more than 230 mb, >> >>>>> although >> >>>>> the system has plenty of free memory (we just added another 1gb to >> >>>>> it). >> >>>>> I >> >>>>> will appreciate if anyone can share with us if you had same >> >>>>> experience >> >>>>> and >> >>>>> managed to resolve it. >> > >> > RECENT POSTINGS >> > -----Original Message----- >> > From: Confocal Microscopy List [mailto:[hidden email]] >> > On >> > Behalf Of Michael Bastiani >> > Sent: 24 March 2009 13:57 >> > To: [hidden email] >> > Subject: Re: LaserSharp 2000 Error >> > >> > Try increasing the memory allocated to image acquisition. In Tools >> > select System Config and set memory to size smaller than Computer >> > RAM. Also make sure your computer hard drive has plenty of space. >> > You cannot collect images/stacks greater than the number set in System >> > Config. >> > good luck, >> > michael >> > >> > >> > On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: >> > >> >> I've been getting this error when i shoot larger z stacks and when i >> >> shoot longer time series >> >> >> >> Serious Error: NO file could be found matching the specification >> >> 'C:\Temp\untitled\untitled1_raw*.pic' >> >> >> >> then i either click ok or close the dialog box and lasersharp 2000 >> >> closes/crashes >> >> >> >> but when i shoot smaller stacks this does NOT happen any help would be >> >> greatly appreciated >> >> >> >> ive tried deleting the C:\Temp folder and this does nothing >> >> >> >> The reason i need this to finish is so that i can then export it to my >> >> desired location on a different server. >> >> >> >> Thank You >> >> --- >> >> Matiar Jafari > > -- > > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) > > > > -- Matiar Jafari |
![]() |
So i finally got LaserSharp 2000 to work with the help of a Zeiss
Representative and apparently theres a bug in v6 that does this two work arounds one is to open another channel and turn off the laser the other is to get the service pack. Thanks for all the input!!! On Fri, Mar 27, 2009 at 11:41 AM, Matiar Jafari <[hidden email]> wrote: > update... my apologies i made a mistake im running windows 2000 with > Lasersharp 2000 6.0 build 846 i just shot a stack that was only 19 um > and it gave me the same error. > > I was told the work around was to > > "....the only solution is to use a method with 2 channels to collect > data. This will allow you to collect > up to the limits set by LaserSharp ~2GB. Once you have your data you > can delete the empty/redundant channel of data." Zeiss Support > > im shooting with a green method with only one channel any help in how > i would do this? > > thanks in advance > > > Thank You > > On Fri, Mar 27, 2009 at 2:42 AM, Keith Morris <[hidden email]> wrote: >> More memory with NT/Lasersharp only helps if you are really down at the >> 256Mb level, and mostly it gives stability/speed provided the modules are >> matched [all ECC] because you probably still have that 1Gb to 2Gb upper >> limit built into Lasersharp. Besides the hard drive page-file can [slowly] >> provide more memory. Check your PC motherboard can cope with 2Gb physical >> RAM [older systems have 1Gb and even 512Mb limits] – crucial.com often gives >> this info. Don’t know if you have a MRC1024 or Radiance, but Zeiss >> [Microscience] should be able to advise if your error messages are >> consistent with Lasersharp v4/v5 operating normally or it’s an error [still >> sound like its hitting the Lasersharp image capture memory limit of 1Gb or >> 2Gb depending on version – unless of course you used to be able to capture >> larger z-stacks]. Once you hit that Lasersharp limit, it’s either an upgrade >> or you have to break the stack into a couple of separate ones. From memory >> when an error message talks specifically about ‘not enough memory for >> capture’ it frequently means hard drive space not system memory. >> >> >> >> Our Bio-Rad 1024 was no longer fully supported by Zeiss Micro-Science a few >> years ago now and so we passed it on to those less fortunate and bought a >> Zeiss 510 Meta. Zeiss did say they would continue to repair the MRC 1024 >> system while they could still get parts, and that they would continue to >> provide technical support advice [which should be free via telephone/email]. >> >> >> >> Many Bio-Rad staff have obviously left the company since UK Bio-Rad confocal >> production was stopped by Zeiss, although they retained core personal for >> product support of the newer Radiance 2000 confocals that also run >> LaserSharp - some of whom will still remember the elderly 1024 [and perhaps >> even COMOS, let alone LaserSharp]. Plus all the knowledgeable Bio-Rad >> confocal reps generally live on working for other microscope companies. >> >> >> >> Keith >> >> --------------------------------------------------------------------------- >> Dr Keith J. Morris, >> Molecular Cytogenetics and Microscopy Core, >> Laboratory 00/069 and 00/070, >> The Wellcome Trust Centre for Human Genetics, >> Roosevelt Drive, >> Oxford OX3 7BN, >> United Kingdom. >> >> Telephone: +44 (0)1865 287568 >> Email: [hidden email] >> Web-pages: http://www.well.ox.ac.uk/cytogenetics/ >> >> New error message after the first one I get. Failed to load PicMat with data >> fileException in channels, stop capture >> >> >> >> I changed the ram capacity to 2gb and I tries changing the harddrive temp. >> >> >> >> I'm shooting then exporting but I can't export as it gives me the error >> after I shoot anything larger than~80um >> >> Thank You >> >> --- >> >> Matiar Jafari >> >> On Mar 25, 2009, at 2:51 AM, Prof Hindsight <[hidden email]> >> wrote: >> >> Dear Matiar Jafari, >> >> >> >> Have you actually contacted a Carl Zeiss service representative about this >> problem? >> >> >> >> From my experience and that of others, Zeiss will assist with such >> problems. I also have it on good authority that although one can no longer >> buy a service/maintenance contract for an MRC-1024, they will endeavour to >> service the system on a billable basis for on-site work and, as Bio-Rad did >> in the past, offer remote assistance at no cost. >> >> Whether you use Zeiss for support or not, knowing a bit more about what type >> of confocal you actually have and its configuration would be helpful and >> avoid misdiagnosis. >> >> >> >> Good luck, >> >> >> >> PH >> >> On Wed, Mar 25, 2009 at 1:24 AM, W. Chan <[hidden email]> wrote: >> >> Hi Matiar >> >> What I meant was e.g., in pre-5.2, you can only give max 1 GB to the >> application. It is not RAM related. Since you did not mention what >> version and OS you're running, I just assumed you're still using NT. >> LaserSharp 6 and XP run really well together; but that's another story. I >> only have access to v6 now so the terms in the menu may be different than >> earlier version. >> >> System access in lasersharp is completely independent of NT's (or whatever >> window's) user privilege. Usually, the software is set up with an user >> called Administrator that have system access, other users for the software >> may only have read & write access. After you log into lasersharp, go to >> Tools | User Setup, the second column will tell you what Access level you >> have. Whoever install the software should have the password to log in as >> Administrator. >> >> There is a way to find that out. Everything for this software are stored >> in a database e.g., c:\LaserSharpNT\Database\lsMessages.mdb. If you have >> microsoft access, you can open a COPY of this file and look at the table >> called LaserSharpUsers and you can see all the user passwords in plain >> text. Access level 0 is for system. It is probably overly cautious to >> open a copy because the database will likely be opened as read-only; but >> why take a chance. >> >> Once you log in as an user with system access level, you should be able to >> change all the configuration parameters. But be very careful, you can >> mess up things really bad. I've learned that the hard way. Hope this >> help. >> >> On Tue, 24 Mar 2009, Matiar Jafari wrote: >> >>> Dear all, >>> >>> So according to Chan the limit is hardcoded and cannot be changed.... >>> >>> we have multiple accounts on the computer and im wondering if im not >>> able to change he memory option from 700 mb to the 1gb because its >>> hardcoded or because i need the "admin" login user for lasersharp >>> 2000....help? also when i change the directory for the temp files it >>> doesn't allow me to click ok thus it doesnt change where its >>> located....any help??? >>> >>> On Tue, Mar 24, 2009 at 11:12 AM, Keith Morris <[hidden email]> >>> wrote: >>> > Hi Matiar, >>> > Chang and others have beaten me too it but anyway: >>> > >>> > I've had this error before on our Radiance 2000 and it was fixed pretty >>> > easily and permanently after a chat to the Bio-Rad engineer, but it was >>> > sorted out years ago so I am rather fuzzy on the reason for it and the >>> > cure. >>> > I'm sure though it is a memory error of some sort [I assume you are >>> > still >>> > running LaserSharp 2000 under NT - our error wasn't on the latest >>> > version 6 >>> > but earlier, probably v4]. We did add more physical memory - to 768Mb or >>> > possibly 1Gb [try eBay for cheap s/h ECC memory, and check if you need >>> > matched pairs], plus we played around with the hard drive space, system >>> > page >>> > file, LaserSharp settings logged in as system, and we put the crucial >>> > :\temp >>> > on our very much larger D:\ drive [set via Lasersharp system] as the C:\ >>> > drive is always compromised to ~4Gb under NT, unless you increased it >>> > after >>> > SP6 is applied. We had two hard drives fitted, the first split into two >>> > partitions C:,E: and the much larger second one D: for user data and the >>> > Temp folder >>> > >>> > So I'm sure all the other postings are on the right track, it's system >>> > or >>> > hard drive memory and the 1Gb/2Gb limit - see Pangs post [no idea what >>> > error >>> > message we got though other than the Z-stacks/time-lapses never finished >>> > and >>> > we lost everything]. I think we still had to split very long time-lapses >>> > with z stacks though into two runs [2Gb limit]. Also allocated more >>> > virtual >>> > memory [page file] from the hard drive to help out the physical RAM >>> > [forgotten how to do that under NT, but it's probably under 'system']. I >>> > think there used to be a 2Gb limit on collections though even with >>> > LaserSharp 6 [1Gb default] even with the tweeks. >>> > >>> > Zeiss [Microscience] should still support the Radiance 2000 [not the >>> > 1024 >>> > though] so an email to their service support should clarify the >>> > solution. I >>> > don't know if upgrading to XP ever helped [you needed a new Firewire >>> > card >>> > and deep pockets for Bio-Rad to do this for you]. >>> > >>> > Again look at the old postings below that seem very relevant. >>> > >>> > Keith >>> > >>> > >>> > --------------------------------------------------------------------------- >>> > Dr Keith J. Morris, >>> > Molecular Cytogenetics and Microscopy Core, >>> > Laboratory 00/069 and 00/070, >>> > The Wellcome Trust Centre for Human Genetics, >>> > Roosevelt Drive, >>> > Oxford OX3 7BN, >>> > United Kingdom. >>> > >>> > Telephone: +44 (0)1865 287568 >>> > Email: [hidden email] >>> > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ >>> > >>> > >>> > >>> > >>> > >>> > >>> > >>> > ------------------------------------------------------------------- >>> > Subject: Re: Radiance 2000 (was Re: [CONFOCAL] Bio-Rad 1024) >>> > From: W. Chan >>> > Reply-To: Confocal Microscopy List <[hidden email]> >>> > Date: Tue, 21 Mar 2006 16:59:39 -0800 >>> > Content-Type: TEXT/PLAIN >>> > >>> > Search the CONFOCAL archive at >>> > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> > >>> > The Control Panel error message is quite misleading. For v5.2, it >>> > states >>> > that "Collection Requires Too Much Memory" when the predicted dataset >>> > size >>> > exceeds the collection limit, but in fact it refers to disk space. >>> > >>> > I agree. Even a 2 GB limit is really not much when you need >>> > multi-channels with a merge pane at 1024x1024 with multiple time points. >>> > May be a work around is not to have a merge pane which can be generated >>> > quite easily after you have the data. A few of our users do not use >>> > the merge pane option in their methods. They do not use LaserSharp to >>> > analyze their images so they can't read the merge pane .pmf file even if >>> > they want to. >>> > >>> > -- >>> > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >>> > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >>> > >>> > On Sat, 18 Mar 2006, Michael Cammer wrote: >>> > >>> >> I'll check the upgrade software for writing to disk. But with multiple >>> >> probe files our problem was the merge file. For two probes, it needed >>> >> 150% more memory than the raw images themselves. Even changing the >>> >> configuration didn't provide enough memory. >>> >> >>> >> >>> >>> Are you referring to RAM or harddisk space? As far as I know, >>> > LaserSharps >>> >>> writes to temporary files rather than holding the experiment in RAM. >>> >>> We >>> >>> have a Radiance 2000 on NT (1 GB RAM, 1.5 GB page file) with >>> >>> LaserSharp >>> >>> 5.2 build 824. The collection limit is 2 GB. You may want to check >>> >>> "System Configuration | Scan system" for the setting of the collection >>> >>> limit. Previous versions may have a 1 GB limit. Also check the >>> >>> location >>> >>> of the temporary files which needs to have plenty of space (as much as >>> > you >>> >>> want to acquire each time or more). Good luck. >>> >>> >>> >>> -- >>> >>> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >>> >>> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >>> >>> >>> >>> On Fri, 17 Mar 2006, Michael Cammer wrote: >>> >>> >>> >>>> We have the Radiance 2000. This is a major limitation of the system. >>> >>>> We've had to compromise by restarting long time lapse experiments in >>> >>>> the >>> >>>> middle, by taking fewer Z planes per timepoint or longer intervals. >>> >>>> But when our BioRad confocals die, LaserSharp software should be >>> >>>> obsolete. >>> >>>> -Michael >>> >>>> >>> >>>> At 05:51 PM 03/14/06 -0800, you wrote: >>> >>>>> >>> >>>>> I have a question regarding a BioRad 1024 confocal system. This >>> >>>>> system >>> >>>>> is >>> >>>>> working fine. But one trouble is that people cannot run any >>> >>>>> scanning >>> >>>>> program (time lapse experiments) that contains more than 230 mb, >>> >>>>> although >>> >>>>> the system has plenty of free memory (we just added another 1gb to >>> >>>>> it). >>> >>>>> I >>> >>>>> will appreciate if anyone can share with us if you had same >>> >>>>> experience >>> >>>>> and >>> >>>>> managed to resolve it. >>> > >>> > RECENT POSTINGS >>> > -----Original Message----- >>> > From: Confocal Microscopy List [mailto:[hidden email]] >>> > On >>> > Behalf Of Michael Bastiani >>> > Sent: 24 March 2009 13:57 >>> > To: [hidden email] >>> > Subject: Re: LaserSharp 2000 Error >>> > >>> > Try increasing the memory allocated to image acquisition. In Tools >>> > select System Config and set memory to size smaller than Computer >>> > RAM. Also make sure your computer hard drive has plenty of space. >>> > You cannot collect images/stacks greater than the number set in System >>> > Config. >>> > good luck, >>> > michael >>> > >>> > >>> > On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: >>> > >>> >> I've been getting this error when i shoot larger z stacks and when i >>> >> shoot longer time series >>> >> >>> >> Serious Error: NO file could be found matching the specification >>> >> 'C:\Temp\untitled\untitled1_raw*.pic' >>> >> >>> >> then i either click ok or close the dialog box and lasersharp 2000 >>> >> closes/crashes >>> >> >>> >> but when i shoot smaller stacks this does NOT happen any help would be >>> >> greatly appreciated >>> >> >>> >> ive tried deleting the C:\Temp folder and this does nothing >>> >> >>> >> The reason i need this to finish is so that i can then export it to my >>> >> desired location on a different server. >>> >> >>> >> Thank You >>> >> --- >>> >> Matiar Jafari >> >> -- >> >> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >> >> >> >> > > > > -- > Matiar Jafari > -- Matiar Jafari |
![]() |
That was a problem with Bio-Rad's software, bugs. Every Lasersharp version
seemed to start off with one or two major ones - the British software writers tended to get the new version out ASAP and then issue patches ASAP [which naturally you need to apply]. As v6 is the last and best Lasersharp, I expect Zeiss with give you the most up to date version for free. The methodical Germans [e.g. Zeiss] tend to bug the test the software for months or years before releasing it [so neat new features took a longer time to appear]. Given the small user base of the LSM software they all do pretty well though [even Adobe software is pretty bug ridden on first release]. Keith --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Matiar Jafari Sent: 28 March 2009 00:44 To: [hidden email] Subject: Re: LaserSharp 2000 Error So i finally got LaserSharp 2000 to work with the help of a Zeiss Representative and apparently theres a bug in v6 that does this two work arounds one is to open another channel and turn off the laser the other is to get the service pack. Thanks for all the input!!! On Fri, Mar 27, 2009 at 11:41 AM, Matiar Jafari <[hidden email]> wrote: > update... my apologies i made a mistake im running windows 2000 with > Lasersharp 2000 6.0 build 846 i just shot a stack that was only 19 um > and it gave me the same error. > > I was told the work around was to > > "....the only solution is to use a method with 2 channels to collect > data. This will allow you to collect > up to the limits set by LaserSharp ~2GB. Once you have your data you > can delete the empty/redundant channel of data." Zeiss Support > > im shooting with a green method with only one channel any help in how > i would do this? > > thanks in advance > > > Thank You > > On Fri, Mar 27, 2009 at 2:42 AM, Keith Morris <[hidden email]> >> More memory with NT/Lasersharp only helps if you are really down at the >> 256Mb level, and mostly it gives stability/speed provided the modules are >> matched [all ECC] because you probably still have that 1Gb to 2Gb upper >> limit built into Lasersharp. Besides the hard drive page-file can [slowly] >> provide more memory. Check your PC motherboard can cope with 2Gb physical >> RAM [older systems have 1Gb and even 512Mb limits] crucial.com often gives >> this info. Dont know if you have a MRC1024 or Radiance, but Zeiss >> [Microscience] should be able to advise if your error messages are >> consistent with Lasersharp v4/v5 operating normally or its an error [still >> sound like its hitting the Lasersharp image capture memory limit of 1Gb or >> 2Gb depending on version unless of course you used to be able to capture >> larger z-stacks]. Once you hit that Lasersharp limit, its either an upgrade >> or you have to break the stack into a couple of separate ones. From memory >> when an error message talks specifically about not enough memory for >> capture it frequently means hard drive space not system memory. >> >> >> >> Our Bio-Rad 1024 was no longer fully supported by Zeiss Micro-Science a few >> years ago now and so we passed it on to those less fortunate and bought a >> Zeiss 510 Meta. Zeiss did say they would continue to repair the MRC 1024 >> system while they could still get parts, and that they would continue to >> provide technical support advice [which should be free via telephone/email]. >> >> >> >> Many Bio-Rad staff have obviously left the company since UK Bio-Rad confocal >> production was stopped by Zeiss, although they retained core personal for >> product support of the newer Radiance 2000 confocals that also run >> LaserSharp - some of whom will still remember the elderly 1024 [and perhaps >> even COMOS, let alone LaserSharp]. Plus all the knowledgeable Bio-Rad >> confocal reps generally live on working for other microscope companies. >> >> >> >> Keith >> >> --------------------------------------------------------------------------- >> Dr Keith J. Morris, >> Molecular Cytogenetics and Microscopy Core, >> Laboratory 00/069 and 00/070, >> The Wellcome Trust Centre for Human Genetics, >> Roosevelt Drive, >> Oxford OX3 7BN, >> United Kingdom. >> >> Telephone: +44 (0)1865 287568 >> Email: [hidden email] >> Web-pages: http://www.well.ox.ac.uk/cytogenetics/ >> >> New error message after the first one I get. Failed to load PicMat with >> fileException in channels, stop capture >> >> >> >> I changed the ram capacity to 2gb and I tries changing the harddrive temp. >> >> >> >> I'm shooting then exporting but I can't export as it gives me the error >> after I shoot anything larger than~80um >> >> Thank You >> >> --- >> >> Matiar Jafari >> >> On Mar 25, 2009, at 2:51 AM, Prof Hindsight >> wrote: >> >> Dear Matiar Jafari, >> >> >> >> Have you actually contacted a Carl Zeiss service representative about this >> problem? >> >> >> >> From my experience and that of others, Zeiss will assist with such >> problems. I also have it on good authority that although one can no longer >> buy a service/maintenance contract for an MRC-1024, they will endeavour to >> service the system on a billable basis for on-site work and, as Bio-Rad did >> in the past, offer remote assistance at no cost. >> >> Whether you use Zeiss for support or not, knowing a bit more about what type >> of confocal you actually have and its configuration would be helpful and >> avoid misdiagnosis. >> >> >> >> Good luck, >> >> >> >> PH >> >> On Wed, Mar 25, 2009 at 1:24 AM, W. Chan <[hidden email]> wrote: >> >> Hi Matiar >> >> What I meant was e.g., in pre-5.2, you can only give max 1 GB to the >> application. It is not RAM related. Since you did not mention what >> version and OS you're running, I just assumed you're still using NT. >> LaserSharp 6 and XP run really well together; but that's another story. >> only have access to v6 now so the terms in the menu may be different than >> earlier version. >> >> System access in lasersharp is completely independent of NT's (or whatever >> window's) user privilege. Usually, the software is set up with an user >> called Administrator that have system access, other users for the software >> may only have read & write access. After you log into lasersharp, go to >> Tools | User Setup, the second column will tell you what Access level you >> have. Whoever install the software should have the password to log in as >> Administrator. >> >> There is a way to find that out. Everything for this software are stored >> in a database e.g., c:\LaserSharpNT\Database\lsMessages.mdb. If you have >> microsoft access, you can open a COPY of this file and look at the table >> called LaserSharpUsers and you can see all the user passwords in plain >> text. Access level 0 is for system. It is probably overly cautious to >> open a copy because the database will likely be opened as read-only; but >> why take a chance. >> >> Once you log in as an user with system access level, you should be able >> change all the configuration parameters. But be very careful, you can >> mess up things really bad. I've learned that the hard way. Hope this >> help. >> >> On Tue, 24 Mar 2009, Matiar Jafari wrote: >> >>> Dear all, >>> >>> So according to Chan the limit is hardcoded and cannot be changed.... >>> >>> we have multiple accounts on the computer and im wondering if im not >>> able to change he memory option from 700 mb to the 1gb because its >>> hardcoded or because i need the "admin" login user for lasersharp >>> 2000....help? also when i change the directory for the temp files it >>> doesn't allow me to click ok thus it doesnt change where its >>> located....any help??? >>> >>> On Tue, Mar 24, 2009 at 11:12 AM, Keith Morris <[hidden email]> >>> wrote: >>> > Hi Matiar, >>> > Chang and others have beaten me too it but anyway: >>> > >>> > I've had this error before on our Radiance 2000 and it was fixed >>> > easily and permanently after a chat to the Bio-Rad engineer, but it was >>> > sorted out years ago so I am rather fuzzy on the reason for it and the >>> > cure. >>> > I'm sure though it is a memory error of some sort [I assume you are >>> > still >>> > running LaserSharp 2000 under NT - our error wasn't on the latest >>> > version 6 >>> > but earlier, probably v4]. We did add more physical memory - to 768Mb or >>> > possibly 1Gb [try eBay for cheap s/h ECC memory, and check if you need >>> > matched pairs], plus we played around with the hard drive space, system >>> > page >>> > file, LaserSharp settings logged in as system, and we put the crucial >>> > :\temp >>> > on our very much larger D:\ drive [set via Lasersharp system] as the C:\ >>> > drive is always compromised to ~4Gb under NT, unless you increased it >>> > after >>> > SP6 is applied. We had two hard drives fitted, the first split into two >>> > partitions C:,E: and the much larger second one D: for user data and the >>> > Temp folder >>> > >>> > So I'm sure all the other postings are on the right track, it's system >>> > or >>> > hard drive memory and the 1Gb/2Gb limit - see Pangs post [no idea what >>> > error >>> > message we got though other than the Z-stacks/time-lapses never finished >>> > and >>> > we lost everything]. I think we still had to split very long time-lapses >>> > with z stacks though into two runs [2Gb limit]. Also allocated more >>> > virtual >>> > memory [page file] from the hard drive to help out the physical RAM >>> > [forgotten how to do that under NT, but it's probably under 'system']. I >>> > think there used to be a 2Gb limit on collections though even with >>> > LaserSharp 6 [1Gb default] even with the tweeks. >>> > >>> > Zeiss [Microscience] should still support the Radiance 2000 [not the >>> > 1024 >>> > though] so an email to their service support should clarify the >>> > solution. I >>> > don't know if upgrading to XP ever helped [you needed a new Firewire >>> > card >>> > and deep pockets for Bio-Rad to do this for you]. >>> > >>> > Again look at the old postings below that seem very relevant. >>> > >>> > Keith >>> > >>> > >>> > >>> > Dr Keith J. Morris, >>> > Molecular Cytogenetics and Microscopy Core, >>> > Laboratory 00/069 and 00/070, >>> > The Wellcome Trust Centre for Human Genetics, >>> > Roosevelt Drive, >>> > Oxford OX3 7BN, >>> > United Kingdom. >>> > >>> > Telephone: +44 (0)1865 287568 >>> > Email: [hidden email] >>> > Web-pages: http://www.well.ox.ac.uk/cytogenetics/ >>> > >>> > >>> > >>> > >>> > >>> > >>> > >>> > ------------------------------------------------------------------- >>> > Subject: Re: Radiance 2000 (was Re: [CONFOCAL] Bio-Rad 1024) >>> > From: W. Chan >>> > Reply-To: Confocal Microscopy List <[hidden email]> >>> > Date: Tue, 21 Mar 2006 16:59:39 -0800 >>> > Content-Type: TEXT/PLAIN >>> > >>> > Search the CONFOCAL archive at >>> > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> > >>> > The Control Panel error message is quite misleading. For v5.2, it >>> > states >>> > that "Collection Requires Too Much Memory" when the predicted dataset >>> > size >>> > exceeds the collection limit, but in fact it refers to disk space. >>> > >>> > I agree. Even a 2 GB limit is really not much when you need >>> > multi-channels with a merge pane at 1024x1024 with multiple time >>> > May be a work around is not to have a merge pane which can be generated >>> > quite easily after you have the data. A few of our users do not use >>> > the merge pane option in their methods. They do not use LaserSharp to >>> > analyze their images so they can't read the merge pane .pmf file even if >>> > they want to. >>> > >>> > -- >>> > Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >>> > The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >>> > >>> > On Sat, 18 Mar 2006, Michael Cammer wrote: >>> > >>> >> I'll check the upgrade software for writing to disk. But with multiple >>> >> probe files our problem was the merge file. For two probes, it needed >>> >> 150% more memory than the raw images themselves. Even changing the >>> >> configuration didn't provide enough memory. >>> >> >>> >> >>> >>> Are you referring to RAM or harddisk space? As far as I know, >>> > LaserSharps >>> >>> writes to temporary files rather than holding the experiment in RAM. >>> >>> We >>> >>> have a Radiance 2000 on NT (1 GB RAM, 1.5 GB page file) with >>> >>> LaserSharp >>> >>> 5.2 build 824. The collection limit is 2 GB. You may want to check >>> >>> "System Configuration | Scan system" for the setting of the >>> >>> limit. Previous versions may have a 1 GB limit. Also check the >>> >>> location >>> >>> of the temporary files which needs to have plenty of space (as much as >>> > you >>> >>> want to acquire each time or more). Good luck. >>> >>> >>> >>> -- >>> >>> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >>> >>> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >>> >>> >>> >>> On Fri, 17 Mar 2006, Michael Cammer wrote: >>> >>> >>> >>>> We have the Radiance 2000. This is a major limitation of the system. >>> >>>> We've had to compromise by restarting long time lapse experiments in >>> >>>> the >>> >>>> middle, by taking fewer Z planes per timepoint or longer intervals. >>> >>>> But when our BioRad confocals die, LaserSharp software should be >>> >>>> obsolete. >>> >>>> -Michael >>> >>>> >>> >>>> At 05:51 PM 03/14/06 -0800, you wrote: >>> >>>>> >>> >>>>> I have a question regarding a BioRad 1024 confocal system. This >>> >>>>> system >>> >>>>> is >>> >>>>> working fine. But one trouble is that people cannot run any >>> >>>>> scanning >>> >>>>> program (time lapse experiments) that contains more than 230 mb, >>> >>>>> although >>> >>>>> the system has plenty of free memory (we just added another 1gb to >>> >>>>> it). >>> >>>>> I >>> >>>>> will appreciate if anyone can share with us if you had same >>> >>>>> experience >>> >>>>> and >>> >>>>> managed to resolve it. >>> > >>> > RECENT POSTINGS >>> > -----Original Message----- >>> > From: Confocal Microscopy List >>> > On >>> > Behalf Of Michael Bastiani >>> > Sent: 24 March 2009 13:57 >>> > To: [hidden email] >>> > Subject: Re: LaserSharp 2000 Error >>> > >>> > Try increasing the memory allocated to image acquisition. In Tools >>> > select System Config and set memory to size smaller than Computer >>> > RAM. Also make sure your computer hard drive has plenty of space. >>> > You cannot collect images/stacks greater than the number set in System >>> > Config. >>> > good luck, >>> > michael >>> > >>> > >>> > On Mar 24, 2009, at 2:57 AM, Matiar Jafari wrote: >>> > >>> >> I've been getting this error when i shoot larger z stacks and when i >>> >> shoot longer time series >>> >> >>> >> Serious Error: NO file could be found matching the specification >>> >> 'C:\Temp\untitled\untitled1_raw*.pic' >>> >> >>> >> then i either click ok or close the dialog box and lasersharp 2000 >>> >> closes/crashes >>> >> >>> >> but when i shoot smaller stacks this does NOT happen any help would >>> >> greatly appreciated >>> >> >>> >> ive tried deleting the C:\Temp folder and this does nothing >>> >> >>> >> The reason i need this to finish is so that i can then export it to my >>> >> desired location on a different server. >>> >> >>> >> Thank You >>> >> --- >>> >> Matiar Jafari >> >> -- >> >> Pang (Wai Pang Chan, [hidden email], PAB A087, 206-685-1519) >> The Biology Imaging Facility (http://staff.washington.edu/wpchan/if/) >> >> >> >> > > > > -- > Matiar Jafari > -- Matiar Jafari |
Free forum by Nabble | Edit this page |