![]() ![]() |
Christian Liebig |
![]() |
Hi all,
I wonder whether anybody has tried/ succeeded to control a Leica SP2 confocal with Labview or a software other than the Leica software? Is it possible at all? Thanks for any helpful hints, kind regards, Christian -- Christian Liebig, PhD Hertie-Institut für klinische Hirnforschung Abteilung Zellbiologie Neurologischer Erkrankungen Otfried-Müller-Strasse 27 72076 Tübingen Tel.: 07071/29-87607 Fax: 07071/29-4521 [hidden email] |
![]() ![]() |
Mario Faretta |
![]() |
Leica provides a macro package (sold as a separate option to be activated in
your hardware key) containing the commands to control the confocal with visual basic and allowing access to most of the elementary operations. I found it very useful to produce ad-hoc acquisition semiautomated protocols. Hope it helps Mario Christian Liebig ([hidden email]) wrote: > > Hi all, > > I wonder whether anybody has tried/ succeeded to control a Leica SP2 > confocal with Labview or a software other than the Leica software? Is it > possible at all? > > Thanks for any helpful hints, kind regards, > > Christian > > > -- > Christian Liebig, PhD > > Hertie-Institut für klinische Hirnforschung > Abteilung Zellbiologie Neurologischer Erkrankungen > Otfried-Müller-Strasse 27 > 72076 Tübingen > > Tel.: 07071/29-87607 > Fax: 07071/29-4521 > > [hidden email] > -- Mario Faretta Department of Experimental Oncology European Institute of Oncology c/o IFOM-IEO Campus for Oncogenomics via Adamello 16 20139 Milan Italy Phone: ++39-02574303054 email: [hidden email] http://www.ifom-ieo-campus.it [- Il Tuo 5 per Mille a favore della Ricerca dell'Istituto Europeo di Oncologia Tutti coloro che presentano il modello Unico, il modello 730 o più semplicemente che ricevono dal proprio datore di lavoro il modello CUD, hanno la facoltà di scegliere la destinazione del proprio 5 per mille. Nella casella riservata al Finanziamento agli Enti della Ricerca Sanitaria inserisci il codice fiscale dello IEO (08 69 14 40 153) ed apponi la Tua firma. Il Tuo 5 per Mille verrà destinato alla ricerca contro il cancro dell'Istituto Europeo di Oncologia NON COSTA NULLA E NON COMPORTA ALCUN AUMENTO DELLE IMPOSTE DA VERSARE NON È UN'ALTERNATIVA ALL'8 PER MILLE E' UN GESTO CONCRETO E DI GRANDE VALORE Per saperne di più vai al sito dello IEO www.ieo.it < http://www.ieo.it > oppure scrivi a [hidden email] < mailto:[hidden email]> Segnala ad un amico questa opportunità -] |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by Christian Liebig
Labview is relatively slow so you will need a good computer to be able
to keep up with things on the data display side. The DAQ hardware has no problems streaming data in, but the labview user interface is inefficient enough that it often can't update the display fast enough. I managed to get a microscope running at a frame per second. The microscope was mostly built from parts though so I didn't have to try to decipher any proprietary interfaces. We had another microscope that we tried to run at video rate, but the programmer was only able to get it up to about 7 fps after considerable labview code optimization. Craig On Fri, Feb 5, 2010 at 5:34 AM, Christian Liebig <[hidden email]> wrote: > Hi all, > > I wonder whether anybody has tried/ succeeded to control a Leica SP2 > confocal with Labview or a software other than the Leica software? Is it > possible at all? > > Thanks for any helpful hints, kind regards, > > Christian > > > -- > Christian Liebig, PhD > > Hertie-Institut für klinische Hirnforschung > Abteilung Zellbiologie Neurologischer Erkrankungen > Otfried-Müller-Strasse 27 > 72076 Tübingen > > Tel.: 07071/29-87607 > Fax: 07071/29-4521 > > [hidden email] > |
![]() |
Craig,
Can you give some of the specifications of the computer you were using? Multiprocessor? Bus speed (probably fairly critical)? System software? Were all the routines done in LabView/Realtime or was the LabView interface used to call external subroutines (often better). When I was at the Advanced Light Source in Berkeley, a very large percentage of the beamlines used LabView programming to simultaneously run steering mirrors, pinholes, cameras, photomultiplier signal collection, piezo scanning stages, etc. Mostly this was done on Unix platforms but some Windows, as well. External subroutines were used liberally in addition. In any event, I have had it on my to do list to assemble a microscope system using LabView since, in principle, it should allow for a lot of flexibility. I am aware that subVI routines can introduce a lot of overhead and that to make LabView run well you are better off using National Instrument boards, which may or may not be a bit self-serving. Can you comment a bit further? Thanks, Mario >Labview is relatively slow so you will need a good computer to be able >to keep up with things on the data display side. The DAQ hardware has >no problems streaming data in, but the labview user interface is >inefficient enough that it often can't update the display fast enough. > I managed to get a microscope running at a frame per second. The >microscope was mostly built from parts though so I didn't have to try >to decipher any proprietary interfaces. We had another microscope >that we tried to run at video rate, but the programmer was only able >to get it up to about 7 fps after considerable labview code >optimization. > >Craig > > >On Fri, Feb 5, 2010 at 5:34 AM, Christian Liebig ><[hidden email]> wrote: >> Hi all, >> >> I wonder whether anybody has tried/ succeeded to control a Leica SP2 >> confocal with Labview or a software other than the Leica software? Is it >> possible at all? >> >> Thanks for any helpful hints, kind regards, >> >> Christian >> >> >> -- >> Christian Liebig, PhD >> >> Hertie-Institut für klinische Hirnforschung >> Abteilung Zellbiologie Neurologischer Erkrankungen >> Otfried-Müller-Strasse 27 >> 72076 Tübingen >> >> Tel.: 07071/29-87607 >> Fax: 07071/29-4521 >> >> [hidden email] >> -- ________________________________________________________________________________ Mario M. Moronne, Ph.D. [hidden email] [hidden email] |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by Craig Brideau
The slow scan system was constructed using a regular PC with 2GB of
RAM and a P4 processor with all NI boards for DAQ. The scan controller was a pre-programmed SC2000 from GSI or Cambridge (I forget which) generating a triggered raster pattern. Basically a DIO line from one of the NI boards would trigger the scan controller line-by-line and the DAQ acquired data driven by a pixel clock generated by a NI 6608 timer-counter board. We did photon counting so one of the other counters on the 6608 actually collected the data, gated by the first pixel clock timer. This was all displayed in Labview using one of the Chart display widgets. The main limit seemed to be getting the data from the 6608's buffer and displaying the data in the graph onscreen. The laser would sweep out a line, the data would stream into the 6608's buffer, then labview would display the information for that line. Then the next line would trigger and the process would repeat. The fast scan system we recently tried to assemble was video-rate with analog detection. The computer was a mid-high-end Dell workstation with 4GB RAM and recent processor (don't remember specs off hand but was fairly new). The DAQ was an Alzar two-channel board. The scanner was a Thorlabs 'Confocal Camera' which is a resonant-galvo-based system. A D-A board from NI ran an analog waveform to drive the slow galvo, while the resonant galvo free-ran at 8kHz. A triggering pulse from the resonant scanner system synchronized the Alzar data acquisition and the slow-galvo sweep. Again the speed limit seemed to be the interface. The scanner was doing about 20 fps but we could only display 7 fps to the user. The data WAS saved at 20 fps so we weren't losing frames from the record, it's just that the user got a slower real-time frame rate. We will be trying to improve this one so any input would be appreciated. Craig On Fri, Feb 5, 2010 at 12:52 PM, Mario <[hidden email]> wrote: > Craig, > > Can you give some of the specifications of the computer you were using? > Multiprocessor? Bus speed (probably fairly critical)? System software? > > Were all the routines done in LabView/Realtime or was the LabView interface > used to call external subroutines (often better). > > When I was at the Advanced Light Source in Berkeley, a very large percentage > of the beamlines used LabView programming to simultaneously run steering > mirrors, pinholes, cameras, photomultiplier signal collection, piezo > scanning stages, etc. Mostly this was done on Unix platforms but some > Windows, as well. External subroutines were used liberally in addition. > > In any event, I have had it on my to do list to assemble a microscope system > using LabView since, in principle, it should allow for a lot of flexibility. > I am aware that subVI routines can introduce a lot of overhead and that to > make LabView run well you are better off using National Instrument boards, > which may or may not be a bit self-serving. Can you comment a bit further? > > Thanks, > > Mario > >> Labview is relatively slow so you will need a good computer to be able >> to keep up with things on the data display side. The DAQ hardware has >> no problems streaming data in, but the labview user interface is >> inefficient enough that it often can't update the display fast enough. >> I managed to get a microscope running at a frame per second. The >> microscope was mostly built from parts though so I didn't have to try >> to decipher any proprietary interfaces. We had another microscope >> that we tried to run at video rate, but the programmer was only able >> to get it up to about 7 fps after considerable labview code >> optimization. >> >> Craig >> >> >> On Fri, Feb 5, 2010 at 5:34 AM, Christian Liebig >> <[hidden email]> wrote: >>> >>> Hi all, >>> >>> I wonder whether anybody has tried/ succeeded to control a Leica SP2 >>> confocal with Labview or a software other than the Leica software? Is it >>> possible at all? >>> >>> Thanks for any helpful hints, kind regards, >>> >>> Christian >>> >>> >>> -- >>> Christian Liebig, PhD >>> >>> Hertie-Institut für klinische Hirnforschung >>> Abteilung Zellbiologie Neurologischer Erkrankungen >>> Otfried-Müller-Strasse 27 >>> 72076 Tübingen >>> >>> Tel.: 07071/29-87607 >>> Fax: 07071/29-4521 >>> >>> [hidden email] >>> > > > -- > ________________________________________________________________________________ > Mario M. Moronne, Ph.D. > > [hidden email] > [hidden email] > |
![]() ![]() |
Emmanuel Gustin |
![]() |
In reply to this post by Craig Brideau
In my own experience -- with a now ancient version of LabView -- we solved the
image throughput problem by using a C++ DLL, which transferred images from the camera, put them on screen in a dialog window, and stored them. To avoid delay in user interface interactions, as much as possible the work was delegated to a background thread in the DLL, and some image processing was done by the processor of the framegrabber. LabView code integrated all the part of the system, and provided the user interface. It still works well. Debugging a setup like that may require a bit of ingenuity! I have also used the VBA macro language of a Leica SP2 -- specifically, to set up automated 4D scanning on eight-chamber slides. I have to say that it wasn't as easy as I had hoped, because the SP2 macro language was a bit underdeveloped, and that was never solved because Leica concentrated its efforts on the new SP5 software. For example, after a lot of trouble it was suggested to us that if we wanted to run a macro that changes the scan mode, problems could be avoided by using the Format button in the menu -- just change the image size back and forth -- before starting the macro. (Don't ask me why, I don't know.) Not all SP2 methods are implemented in VBA, and some that do work, appear to leak memory. However, it is an useful tool. (For the SP5, Leica provides its own automated scanning solution.) From the VBA environment, you can also call a COM object. I used this to analyze XZ slices as an improvised autofocus mechanism, and it appeared to be trouble-free. That is at least one way to connect an SP2 to code written in another language, and you could use it to synchronize the actions of the SP2 with other systems. Best Regards, Emmanuel -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: vrijdag 5 februari 2010 18:15 To: [hidden email] Subject: Re: Leica SP2 control via Labview Labview is relatively slow so you will need a good computer to be able to keep up with things on the data display side. The DAQ hardware has no problems streaming data in, but the labview user interface is inefficient enough that it often can't update the display fast enough. I managed to get a microscope running at a frame per second. The microscope was mostly built from parts though so I didn't have to try to decipher any proprietary interfaces. We had another microscope that we tried to run at video rate, but the programmer was only able to get it up to about 7 fps after considerable labview code optimization. Craig On Fri, Feb 5, 2010 at 5:34 AM, Christian Liebig <[hidden email]> wrote: > Hi all, > > I wonder whether anybody has tried/ succeeded to control a Leica SP2 > confocal with Labview or a software other than the Leica software? Is it > possible at all? > > Thanks for any helpful hints, kind regards, > > Christian > > > -- > Christian Liebig, PhD > > Hertie-Institut für klinische Hirnforschung > Abteilung Zellbiologie Neurologischer Erkrankungen > Otfried-Müller-Strasse 27 > 72076 Tübingen > > Tel.: 07071/29-87607 > Fax: 07071/29-4521 > > [hidden email] > |
Free forum by Nabble | Edit this page |