Faizan Malik |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Everyone, I am currently working on optimizing a protocol for live cell imaging of embryonic brains (mice) with a inverted confocal microscope. Does anyone have a specific set up or protocol they use? Or a certain type of perfusion chamber design? I have found a lot of literature regarding live cell imaging with normal confocal microscopes, but there does not seem to be much there on live imaging with inverted confocal microscopes. Is there any specific way that you attach your samples to the chamber in a way that prevents damage to morphology as well as prevent reduced image quality due to the using substances like glue or collagen to hold the specimen in place (a problem not seen with non-inverted confocal microscopes). We use a Zeiss LSM 700 in our lab. Your help is appreciated! Cheers, Faizan |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Inverted systems work best for live when you can pull the organ out of the body partially and lay it next to the animal. Kidney etc. works well for this. For brain I think the embryo would be tricky to position although perfusion might be easier since the part you want to image would be deeper in the chamber. You might want to look into 3d printing for solutions if you need to build a custom chamber. I've done a fair bit of this for spinal cord holders/supports and sciatic nerve and being able to create my own shapes was very helpful! Craig On Wed, May 27, 2015 at 4:11 PM, Faizan Malik <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > I am currently working on optimizing a protocol for live cell imaging of > embryonic brains (mice) with a inverted confocal microscope. > > Does anyone have a specific set up or protocol they use? Or a certain type > of perfusion chamber design? I have found a lot of literature regarding > live > cell imaging with normal confocal microscopes, but there does not seem to > be > much there on live imaging with inverted confocal microscopes. Is there any > specific way that you attach your samples to the chamber in a way that > prevents damage to morphology as well as prevent reduced image quality due > to the using substances like glue or collagen to hold the specimen in place > (a problem not seen with non-inverted confocal microscopes). We use a Zeiss > LSM 700 in our lab. > > Your help is appreciated! > > Cheers, > Faizan > |
|
Luis Farias |
|
|
In reply to this post by Faizan Malik
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello Faizan, Just to clarify my doubt, you don't want to use collagen with your cells? Cause I'd succeeded in live imaging using an LSM 5 Pascal (inverted) using Collagen I protocols. But there is other options without using Collagen, you just need the suitable dish plates with cover slips on the base. hope it helps, regards Luis Henrique Seabra de Farias*, Msc.* E-mail: [hidden email] <[hidden email]> Universidade Federal do Pará Instituto de Ciências Biológicas Laboratório de Parasitologia/Laboratório de Biologia Estrutural Telefone: 913201-7927/913201-8232 2015-05-27 19:11 GMT-03:00 Faizan Malik <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > I am currently working on optimizing a protocol for live cell imaging of > embryonic brains (mice) with a inverted confocal microscope. > > Does anyone have a specific set up or protocol they use? Or a certain type > of perfusion chamber design? I have found a lot of literature regarding > live > cell imaging with normal confocal microscopes, but there does not seem to > be > much there on live imaging with inverted confocal microscopes. Is there any > specific way that you attach your samples to the chamber in a way that > prevents damage to morphology as well as prevent reduced image quality due > to the using substances like glue or collagen to hold the specimen in place > (a problem not seen with non-inverted confocal microscopes). We use a Zeiss > LSM 700 in our lab. > > Your help is appreciated! > > Cheers, > Faizan > |
|
Richard Mort |
|
|
In reply to this post by Faizan Malik
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Faizan, Its a challenge, especially if you require an air-liquid interface. We solved this problem using lummox gas permeable membranes (from Greiner - I have no association) and this works well for embryonic skin culture. We also tried on embryonic brain slices recently and it looked promising. The membranes can be cut out of the dishes to use for custom apparatus. The most recent of the following papers may be the most useful but if you don't have a workshop you may be able to adapt a 50mm lummox dish if you can find a way to immobilize your sample. http://www.ncbi.nlm.nih.gov/pubmed/24894489 http://www.ncbi.nlm.nih.gov/pubmed/20067551 We found that if handled carefully the membranes were compatible with Nikon PFS which was a great help. Best wishes Richard On 27/05/15 23:11, Faizan Malik wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > I am currently working on optimizing a protocol for live cell imaging of > embryonic brains (mice) with a inverted confocal microscope. > > Does anyone have a specific set up or protocol they use? Or a certain type > of perfusion chamber design? I have found a lot of literature regarding live > cell imaging with normal confocal microscopes, but there does not seem to be > much there on live imaging with inverted confocal microscopes. Is there any > specific way that you attach your samples to the chamber in a way that > prevents damage to morphology as well as prevent reduced image quality due > to the using substances like glue or collagen to hold the specimen in place > (a problem not seen with non-inverted confocal microscopes). We use a Zeiss > LSM 700 in our lab. > > Your help is appreciated! > > Cheers, > Faizan -- Dr Richard Mort Research Fellow MRC Human Genetics Unit Institute of Genetics and Molecular Medicine University of Edinburgh Western General Hospital Crewe Road Edinburgh EH4 2XU Tel: 0131 651 8648 The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336. |
|
George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Richard, Thanks for the tip on the Greiner Bio-One products - product line has only one m -- Lumox. Product pdf is http://greinerbioone.com/UserFiles/File/IVSSbrochure.pdf Hi Fazian, You may want to invest in an Objective Inverter from LSM Tech http://www.lsmtech.com/product_objective_inverter.html enables your specimen to be viewed at any angle. Depending on your microscope layout, you might also want one of their XYZ Side Stages, http://www.lsmtech.com/xyz_sidestage.html though maybe a lab jack next to the microscope would be sufficient. Enjoy, George On 5/28/2015 4:17 AM, Richard Mort wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Faizan, > > Its a challenge, especially if you require an air-liquid interface. We > solved this problem using lummox gas permeable membranes (from Greiner > - I have no association) and this works well for embryonic skin > culture. We also tried on embryonic brain slices recently and it > looked promising. The membranes can be cut out of the dishes to use > for custom apparatus. > > The most recent of the following papers may be the most useful but if > you don't have a workshop you may be able to adapt a 50mm lummox dish > if you can find a way to immobilize your sample. > > http://www.ncbi.nlm.nih.gov/pubmed/24894489 > http://www.ncbi.nlm.nih.gov/pubmed/20067551 > > We found that if handled carefully the membranes were compatible with > Nikon PFS which was a great help. > Best wishes > Richard > > > > On 27/05/15 23:11, Faizan Malik wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >> posting. >> ***** >> >> Hi Everyone, >> >> I am currently working on optimizing a protocol for live cell imaging of >> embryonic brains (mice) with a inverted confocal microscope. >> >> Does anyone have a specific set up or protocol they use? Or a certain >> type >> of perfusion chamber design? I have found a lot of literature >> regarding live >> cell imaging with normal confocal microscopes, but there does not >> seem to be >> much there on live imaging with inverted confocal microscopes. Is >> there any >> specific way that you attach your samples to the chamber in a way that >> prevents damage to morphology as well as prevent reduced image >> quality due >> to the using substances like glue or collagen to hold the specimen in >> place >> (a problem not seen with non-inverted confocal microscopes). We use a >> Zeiss >> LSM 700 in our lab. >> >> Your help is appreciated! >> >> Cheers, >> Faizan > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/42 |
|
|
In reply to this post by Faizan Malik
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Faizan I just want to add to George’s suggestion of the LSM Tech inverter. If you use one of these devices you may find it helpful to place your specimen on top of a Corning™ Snapwell membrane in a Bioptechs™ intrinsically temperature controlled Delta T dish. You can then perfuse media under the membrane to nourish the specimen resting on top of the membrane and image directly from above. We have a perfusion system that will maintain this configuration without overflowing making it a convenient methodology. Another approach is to use our Brain slice adapter in a Delta T dish micro-environmental system. Link: http://www.bioptechs.com/Products/Delta_T/Options/options.html#Brain%20A Dan On May 27, 2015, at 6:11 PM, Faizan Malik <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Everyone, I am currently working on optimizing a protocol for live cell imaging of embryonic brains (mice) with a inverted confocal microscope. Does anyone have a specific set up or protocol they use? Or a certain type of perfusion chamber design? I have found a lot of literature regarding live cell imaging with normal confocal microscopes, but there does not seem to be much there on live imaging with inverted confocal microscopes. Is there any specific way that you attach your samples to the chamber in a way that prevents damage to morphology as well as prevent reduced image quality due to the using substances like glue or collagen to hold the specimen in place (a problem not seen with non-inverted confocal microscopes). We use a Zeiss LSM 700 in our lab. Your help is appreciated! Cheers, Faizan Dan Focht Bioptechs, Inc. 3560 Beck Rd. Butler, PA 16002 724-282-7145 Toll Free 877-LIVE-CELL (548-3235) |