Hi
I have two questions relating to the MT macro which I'm using for screening 96 well plates. Currently I manually set the positions relating to each plate and save them within the macro using the Store/Apply step. However when they are recovered some (~10%) are missing the wells and we loose data within the screens. So my questions are: 1) is it possible to edit the locations within the Store/Apply so that they are all accurate? 2) is it possible to save each co-ordinate as a seperate configuration, with the stage coordinates included, and run the screen as a series of blocks within the macro? I realise this would be initially very labour intensive but worth the effort in the long run. Thanks Gareth |
![]() ![]() |
Danielle Crippen |
![]() |
Dear Gareth,
First of all, I'd recommend calling Zeiss and talking with Zbi, Maya or John about this. You can get advice from them regardless of whether you're on service contract...call 1-800-633-6610. I find them to be endless resources. (BTW--I have no affiliation with Zeiss...just use their equipment a lot). Secondly...I'm wondering...are you zero-ing the stage? Do you have a stage with linear encoders on it for the best accuracy? How are you ensuring you're re-placing your plate in the exact same location as when you originally marked positions. If you're hoping to image the same ROI (with the same cells in it etc), even a 50-100um xy shift can ruin your coordinates. We've had some good success here with 8 well chambered coverglass by zeroing the stage at a particular corner of a particular well. Then when we re-load the chamber, we zero again with the same method...then we can successfully return to previously imaged coordinates for further imaging...even weeks later. I think in answer to your first question, you can of course always manually re-do locations by moving the stage to the well you want to image, selecting that location in the MT list, and clicking the "replace xyz" button. Once you've corrected each location, you can Store/Apply again and overwrite the MT setup file. I can't comment on your second question, b/c I don't utilize the blocks in MT (yet:)...so I'm not sure how or if this would work. Best of luck!! d Danielle Crippen Morphology and Imaging Core Manager Buck Institute for Age Research 8001 Redwood Blvd Novato, CA 94945 415-209-2046 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gareth Howell Sent: Friday, June 19, 2009 4:31 AM To: [hidden email] Subject: Locations in MultiTime macro Hi I have two questions relating to the MT macro which I'm using for screening 96 well plates. Currently I manually set the positions relating to each plate and save them within the macro using the Store/Apply step. However when they are recovered some (~10%) are missing the wells and we loose data within the screens. So my questions are: 1) is it possible to edit the locations within the Store/Apply so that they are all accurate? 2) is it possible to save each co-ordinate as a seperate configuration, with the stage coordinates included, and run the screen as a series of blocks within the macro? I realise this would be initially very labour intensive but worth the effort in the long run. Thanks Gareth |
![]() ![]() |
Armstrong, Brian |
![]() |
Hi Gareth, with regards to question 2, I think you MUST store the
position as a separate configuration in order for it to work properly. This will have your x,y, and z positions stored. If you do this your first question will be unnecassary. Cheers, Brian D Armstrong PhD Light Microscopy Core Manager Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imag ing/Pages/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Danielle Crippen Sent: Friday, June 19, 2009 9:08 AM To: [hidden email] Subject: Re: Locations in MultiTime macro Dear Gareth, First of all, I'd recommend calling Zeiss and talking with Zbi, Maya or John about this. You can get advice from them regardless of whether you're on service contract...call 1-800-633-6610. I find them to be endless resources. (BTW--I have no affiliation with Zeiss...just use their equipment a lot). Secondly...I'm wondering...are you zero-ing the stage? Do you have a stage with linear encoders on it for the best accuracy? How are you ensuring you're re-placing your plate in the exact same location as when you originally marked positions. If you're hoping to image the same ROI (with the same cells in it etc), even a 50-100um xy shift can ruin your coordinates. We've had some good success here with 8 well chambered coverglass by zeroing the stage at a particular corner of a particular well. Then when we re-load the chamber, we zero again with the same method...then we can successfully return to previously imaged coordinates for further imaging...even weeks later. I think in answer to your first question, you can of course always manually re-do locations by moving the stage to the well you want to image, selecting that location in the MT list, and clicking the "replace xyz" button. Once you've corrected each location, you can Store/Apply again and overwrite the MT setup file. I can't comment on your second question, b/c I don't utilize the blocks in MT (yet:)...so I'm not sure how or if this would work. Best of luck!! d Danielle Crippen Morphology and Imaging Core Manager Buck Institute for Age Research 8001 Redwood Blvd Novato, CA 94945 415-209-2046 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gareth Howell Sent: Friday, June 19, 2009 4:31 AM To: [hidden email] Subject: Locations in MultiTime macro Hi I have two questions relating to the MT macro which I'm using for screening 96 well plates. Currently I manually set the positions relating to each plate and save them within the macro using the Store/Apply step. However when they are recovered some (~10%) are missing the wells and we loose data within the screens. So my questions are: 1) is it possible to edit the locations within the Store/Apply so that they are all accurate? 2) is it possible to save each co-ordinate as a seperate configuration, with the stage coordinates included, and run the screen as a series of blocks within the macro? I realise this would be initially very labour intensive but worth the effort in the long run. Thanks Gareth --------------------------------------------------------------------- SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
![]() |
In reply to this post by Danielle Crippen
Regarding the resetting of a zero position: On a different instrument where
we did automated analysis of 96 well plates, we created a fiduciary mark on a blank plate and set that as zero every time we ran the analysis. All the measurement positions were created from that initial point. To make the mark we scratched a cross-hair on the plate's bottom surface using a sharp razor blade. Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Assistant Scientific Investigator Dept. of Cell Biology & Anatomy, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: AHSC 4212 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 http://swehsc.pharmacy.arizona.edu/exppath/ Home of: "Microscopy and Imaging Resources on the WWW" -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Danielle Crippen Sent: Friday, June 19, 2009 9:08 AM To: [hidden email] Subject: Re: Locations in MultiTime macro Dear Gareth, First of all, I'd recommend calling Zeiss and talking with Zbi, Maya or John about this. You can get advice from them regardless of whether you're on service contract...call 1-800-633-6610. I find them to be endless resources. (BTW--I have no affiliation with Zeiss...just use their equipment a lot). Secondly...I'm wondering...are you zero-ing the stage? Do you have a stage with linear encoders on it for the best accuracy? How are you ensuring you're re-placing your plate in the exact same location as when you originally marked positions. If you're hoping to image the same ROI (with the same cells in it etc), even a 50-100um xy shift can ruin your coordinates. We've had some good success here with 8 well chambered coverglass by zeroing the stage at a particular corner of a particular well. Then when we re-load the chamber, we zero again with the same method...then we can successfully return to previously imaged coordinates for further imaging...even weeks later. I think in answer to your first question, you can of course always manually re-do locations by moving the stage to the well you want to image, selecting that location in the MT list, and clicking the "replace xyz" button. Once you've corrected each location, you can Store/Apply again and overwrite the MT setup file. I can't comment on your second question, b/c I don't utilize the blocks in MT (yet:)...so I'm not sure how or if this would work. Best of luck!! d Danielle Crippen Morphology and Imaging Core Manager Buck Institute for Age Research 8001 Redwood Blvd Novato, CA 94945 415-209-2046 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gareth Howell Sent: Friday, June 19, 2009 4:31 AM To: [hidden email] Subject: Locations in MultiTime macro Hi I have two questions relating to the MT macro which I'm using for screening 96 well plates. Currently I manually set the positions relating to each plate and save them within the macro using the Store/Apply step. However when they are recovered some (~10%) are missing the wells and we loose data within the screens. So my questions are: 1) is it possible to edit the locations within the Store/Apply so that they are all accurate? 2) is it possible to save each co-ordinate as a seperate configuration, with the stage coordinates included, and run the screen as a series of blocks within the macro? I realise this would be initially very labour intensive but worth the effort in the long run. Thanks Gareth |
![]() ![]() |
Gert van Cappellen |
![]() |
In reply to this post by Gareth Howell
I don't think option 2 is going to work, because the configurations do
not store the stage position. The Zeiss multi-time macro has however an option to define a grid under the button edit locations. You can define this grid with the dimensions of your 96 well plate. So you focus on the well with the lowest X and Y and than you define the grid. Probably the first time you have to check the dimensions and probably fine tune this a bit. The settings will be remembered so the next time pushing one button will do. Good luck, Gert van Cappellen on 19-6-2009 13:31 Gareth Howell said the following: > Hi > I have two questions relating to the MT macro which I'm using for screening > 96 well plates. Currently I manually set the positions relating to each > plate and save them within the macro using the Store/Apply step. However > when they are recovered some (~10%) are missing the wells and we loose data > within the screens. So my questions are: > 1) is it possible to edit the locations within the Store/Apply so that they > are all accurate? > 2) is it possible to save each co-ordinate as a seperate configuration, with > the stage coordinates included, and run the screen as a series of blocks > within the macro? I realise this would be initially very labour intensive > but worth the effort in the long run. > Thanks > Gareth > -- Gert van Cappellen TEL +31-10-70 43578; FAX +31-10-7044736; Email [hidden email] Optical Imaging Centre (OIC); http://www.erasmusmc.nl/oic/ Dept. of Reproduction and Development; http://www.erasmusmc.nl/rede Room Ee914, Erasmus MC, Dr. Molenwaterplein 50, 3015 GE ROTTERDAM, The Netherlands |