Seb Stephens |
Hi,
100% laser for 1 ms OR 1% laser for 100ms? This is a subject that has been bothering me and I have never got quite the answer for it. After speaking to someone who uses 100% laser power (sorry I don’t know what the actual laser output is) I noted their goal was to expose for the least amount of time possible (which was only a few millesecs). This enticed me to check myself how my protein structures (podosomes) could handle 100% laser power as opposed to 30% which I usually use. As expected the exposure times went down significantly (from 320 ms to 30ms) and in fact I was able to still get decent information from 10ms exposures taken every sec for 3 mins. Surprisingly, I saw no detectable bleach and it did not seem any different to long exposure times with less laser power. Im only new to microscopy and surely others must have experimented with: high laser+ low exp compared to low laser+high exp. I really want to know, which one should I do (high or low power)? It seems that there could be mathematically and practically equivalents of high laser+low exp to low laser+high exp (ie as above 100ms 1% power pairs with 1ms 100% laser power). So, is there a relationship between power and exposure time exists that can sort of be mathematically related? Cheers Seb |
Shalin Mehta |
Hi Sebastian,
You are nearly correct in saying that the product of excitation power and exposure time (i.e. total energy put in the sample) is invariant. This linear relationship breaks down when you increase power too much, so that fluorophore gets 'saturated' from too many photons trying to excite it in short time. The energy (E) of the photon and wavelength (L) are related by Planck's constant (h) and speed of light (c): E=hc/L. The basic model of fluorescence is that the fluorophore absorbs a photon (with varying probabilities) across a range of wavelengths and after some time (called lifetime ) typically of pico to nano second values emits a lower energy photon. Ideally, you get one emission photon per excitation photon. When you put in certain power, you are imparting certain energy per second, i.e., putting in a given number of photons per second. As long as the time-rate of photons hitting the fluorophore is low you should get an emission photon per excitation photon. But at very high power ( high pohton-rates) this relationship becomes non-linear, and you get less number of emission photons per excitation photon. You may find introductory chapters of "Principles of fluorescence spectroscopy" by Joseph Lakowicz illuminating in this regard. best shalin On Mon, Jul 13, 2009 at 2:18 AM, Sebastien Stephens<[hidden email]> wrote: > Hi, > > 100% laser for 1 ms OR 1% laser for 100ms? > > This is a subject that has been bothering me and I have never got quite the > answer for it. > > After speaking to someone who uses 100% laser power (sorry I don’t know > what the actual laser output is) I noted their goal was to expose for the least > amount of time possible (which was only a few millesecs). This enticed me to > check myself how my protein structures (podosomes) could handle 100% laser > power as opposed to 30% which I usually use. As expected the exposure > times went down significantly (from 320 ms to 30ms) and in fact I was able to > still get decent information from 10ms exposures taken every sec for 3 mins. > Surprisingly, I saw no detectable bleach and it did not seem any different to > long exposure times with less laser power. > > Im only new to microscopy and surely others must have experimented with: > high laser+ low exp compared to low laser+high exp. > > I really want to know, which one should I do (high or low power)? > > It seems that there could be mathematically and practically equivalents of > high laser+low exp to low laser+high exp (ie as above 100ms 1% power pairs > with 1ms 100% laser power). So, is there a relationship between power and > exposure time exists that can sort of be mathematically related? > > Cheers > > Seb > |
John Oreopoulos |
It's a good question and I think Shalin and Claire have given good
answers already. Perhaps there is a way to determine on the fly during an experiment if you are indeed saturating the fluorescence signal in any parts of the cell, but without that knowledge you could be doing damage to the cells. Generally, it has been my experience, and also it was emphasized many times at Jim Pawley's live-cell microscopy course that I attended a few years ago, that the products of photobleaching are toxic to the cell, and so using as little light as possible is preferred so long as you can compromise with a longer exposure time or increased detector gain and still get an acceptable SNR for the phenomenon you're trying to observe. That being said, there may be situations where you are trying to observe a fast dynamic process in the cell by time-lapse imaging which necessitates short exposure times and using a larger laser power to get a good SNR again. As I understand it, some of the super-resolution imaging techniques even depend on the non-linear effects of fluorescence saturation to achieve imaging resolution below the diffraction limit, so there are a few times when a larger laser power might be called for. So, I would say there is no one definite recipe for all imaging situations, but keeping the cells alive on the microscope stage is a priority. Jennifer Waters has a very nice recent review on "Accuracy and Precision in Light Microscopy" in the Journal of Cell Biology where she rightly points out that it's up to the experimenter to test and figure out the optimum settings of acquisition and perform the proper control experiments to back up the interpretation results of an imaging experiment. John Oreopoulos On 13-Jul-09, at 4:01 AM, Shalin Mehta wrote: > Hi Sebastian, > > You are nearly correct in saying that the product of excitation power > and exposure time (i.e. total energy put in the sample) is invariant. > This linear relationship breaks down when you increase power too much, > so that fluorophore gets 'saturated' from too many photons trying to > excite it in short time. > > The energy (E) of the photon and wavelength (L) are related by > Planck's constant (h) and speed of light (c): E=hc/L. The basic model > of fluorescence is that the fluorophore absorbs a photon (with varying > probabilities) across a range of wavelengths and after some time > (called lifetime ) typically of pico to nano second values emits a > lower energy photon. > > Ideally, you get one emission photon per excitation photon. When you > put in certain power, you are imparting certain energy per second, > i.e., putting in a given number of photons per second. As long as the > time-rate of photons hitting the fluorophore is low you should get an > emission photon per excitation photon. But at very high power ( high > pohton-rates) this relationship becomes non-linear, and you get less > number of emission photons per excitation photon. > > You may find introductory chapters of "Principles of fluorescence > spectroscopy" by Joseph Lakowicz illuminating in this regard. > > best > shalin > > On Mon, Jul 13, 2009 at 2:18 AM, Sebastien > Stephens<[hidden email]> wrote: >> Hi, >> >> 100% laser for 1 ms OR 1% laser for 100ms? >> >> This is a subject that has been bothering me and I have never got >> quite the >> answer for it. >> >> After speaking to someone who uses 100% laser power (sorry I don’t >> know >> what the actual laser output is) I noted their goal was to expose >> for the least >> amount of time possible (which was only a few millesecs). This >> enticed me to >> check myself how my protein structures (podosomes) could handle >> 100% laser >> power as opposed to 30% which I usually use. As expected the exposure >> times went down significantly (from 320 ms to 30ms) and in fact I >> was able to >> still get decent information from 10ms exposures taken every sec >> for 3 mins. >> Surprisingly, I saw no detectable bleach and it did not seem any >> different to >> long exposure times with less laser power. >> >> Im only new to microscopy and surely others must have experimented >> with: >> high laser+ low exp compared to low laser+high exp. >> >> I really want to know, which one should I do (high or low power)? >> >> It seems that there could be mathematically and practically >> equivalents of >> high laser+low exp to low laser+high exp (ie as above 100ms 1% >> power pairs >> with 1ms 100% laser power). So, is there a relationship between >> power and >> exposure time exists that can sort of be mathematically related? >> >> Cheers >> >> Seb >> |
Seb Stephens |
Dear Tom, John, Claire, Esteban and Shalin,
Thank you all very much for your input. I am now very clear on how I will approach my image sampling. My best regards Sébastien Stephens
Harvard University School of Dental Medicine 188 Longwood Ave REB 313 Boston, MA 02115 Ph: 617-4327328 Fax: 617-4321897 [hidden email] Téléchargez le nouveau Windows Live Messenger ! Téléchargez Messenger, c'est gratuit ! |
In reply to this post by John Oreopoulos
Hello all;
> > I have a colleague who is unhappy with the images of td-Tomato > AI9 mice on our Fluoview 500. They first view their brain sections > on our AX70, with mercury lamp illumination, a TRITC 535/50 EX filter, > 610/75 EM filter, and like what they see. On our FV500 with 543 laser > illumination (HeNe-Green 50% default AOTF increased to 90%), we have > switched to a 560LP filter from the 560-600EM filter. This improved > their results, but they still are disappointed in signal intensity. I > recently had our system checked over by Olympus service, and we > measured the laser output as 0.45mW out of the fiber (non-wavelength > specific power meter), which is just off the best it's ever been. I > have suggested opening the pinhole and sacrificing Z resolution for > brightness. > > Would you have any other suggestions for increasing the intensity of > their signal? I have explained filters and Z-thickness versus > brightness. I don't know what to tell them, and they are faulting the > system. From all spectra I can find, our system should be great as seeing this construct. Kathy Spencer |
Are they using an index-matched mounting medium?
Do they understand that a wide field "like what they see" is quite different to a confocal section? Cheers >> I have a colleague who is unhappy with the images of td-Tomato >> AI9 mice on our Fluoview 500. They first view their brain sections >> on our AX70, with mercury lamp illumination, a TRITC 535/50 EX filter, >> 610/75 EM filter, and like what they see. On our FV500 with 543 laser >> illumination (HeNe-Green 50% default AOTF increased to 90%), we have >> switched to a 560LP filter from the 560-600EM filter. This improved >> their results, but they still are disappointed in signal intensity. I >> recently had our system checked over by Olympus service, and we >> measured the laser output as 0.45mW out of the fiber (non-wavelength >> specific power meter), which is just off the best it's ever been. I >> have suggested opening the pinhole and sacrificing Z resolution for >> brightness. >> >> Would you have any other suggestions for increasing the intensity of >> their signal? I have explained filters and Z-thickness versus >> brightness. I don't know what to tell them, and they are faulting the >> system. From all spectra I can find, our system should be great as seeing this construct. >> > > Kathy Spencer > |
In reply to this post by kspencer007
Dear Kathy,
your mercury arc has a nice bright peak at that ex and the HeNe 543 is the weakest laser on the list. Likewise the PMT has a QE below 10% in that range of the em spectrum, your camera may have up to 90%. What happens if you use the arc & a camera on the FV500? Do they use lenses with similar NA on both systems? Is the pixel size x,y similar? Is the pinhole aligned? Is the PSF of the confocal ok? Just a few points to check. Good luck, jens -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kathryn Spencer Sent: Wednesday, November 04, 2009 6:57 PM To: [hidden email] Subject: tdTomato and LSM confocal Hello all; > > I have a colleague who is unhappy with the images of td-Tomato > AI9 mice on our Fluoview 500. They first view their brain sections > on our AX70, with mercury lamp illumination, a TRITC 535/50 EX filter, > 610/75 EM filter, and like what they see. On our FV500 with 543 laser > illumination (HeNe-Green 50% default AOTF increased to 90%), we have > switched to a 560LP filter from the 560-600EM filter. This improved > their results, but they still are disappointed in signal intensity. I > recently had our system checked over by Olympus service, and we > measured the laser output as 0.45mW out of the fiber (non-wavelength > specific power meter), which is just off the best it's ever been. I > have suggested opening the pinhole and sacrificing Z resolution for > brightness. > > Would you have any other suggestions for increasing the intensity of > their signal? I have explained filters and Z-thickness versus > brightness. I don't know what to tell them, and they are faulting the > system. From all spectra I can find, our system should be great as Kathy Spencer |
Free forum by Nabble | Edit this page |