http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi! We are trying to run an experiment that requires image acquisition of a very large area (a slice of rat brain; we use mosaic for this purpose), along with optical sectioning. However, the software that we are currently using does not allow us to acquire both mosaic and optical sectioning simultaneously due to the extremely large file size produced (about 2GB), which cannot be processed by the software (software limitation of file size is 200MB). I am wondering if it is possible for us to acquire the mosaic and optical sectioning separately (i.e. mosaic for every plane of interest), export out and compile them with another software. Does any of you have any suggestion of possible software that has the capability of compiling individual planes into a stack? Please advise. Thank you! Cheers, Rachael Lim |
![]() ![]() |
Rietdorf, Jens |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Rachael Lim, Imaris (bitplane AG) has recently implemented functionality to handle huge image files (no commercial interest from my side). If you want to just merge the mosaics into a projection image, this could already be achieved with -for example- ImageJ. Aaron Ponti from our group has recently developed (MatLab based) software for a 3d alignment and shading correction of stacks from mosaics. Dont hesitate to contact me for further detail. Cheers, jens Dr. Jens Rietdorf head of microscopy unit Friedrich-Miescher-Institute, wro-1066.2.32 Maulbeerstr.66, CH-4058 Basel, Switzerland http://www.fmi.ch/ Phone +41 (61) 697-5172 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]]On Behalf Of Rachael Lim Sent: Wednesday, October 17, 2007 9:57 AM To: [hidden email] Subject: Mosaic with optical sectoning Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi! We are trying to run an experiment that requires image acquisition of a very large area (a slice of rat brain; we use mosaic for this purpose), along with optical sectioning. However, the software that we are currently using does not allow us to acquire both mosaic and optical sectioning simultaneously due to the extremely large file size produced (about 2GB), which cannot be processed by the software (software limitation of file size is 200MB). I am wondering if it is possible for us to acquire the mosaic and optical sectioning separately (i.e. mosaic for every plane of interest), export out and compile them with another software. Does any of you have any suggestion of possible software that has the capability of compiling individual planes into a stack? Please advise. Thank you! Cheers, Rachael Lim |
![]() ![]() |
mmodel |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Rachael Lim
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello Rachael, With ImageJ you can form a stack from individual images and then display them as a mosaic. Mike Model -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Rachael Lim Sent: Wednesday, October 17, 2007 3:57 AM To: [hidden email] Subject: Mosaic with optical sectoning Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi! We are trying to run an experiment that requires image acquisition of a very large area (a slice of rat brain; we use mosaic for this purpose), along with optical sectioning. However, the software that we are currently using does not allow us to acquire both mosaic and optical sectioning simultaneously due to the extremely large file size produced (about 2GB), which cannot be processed by the software (software limitation of file size is 200MB). I am wondering if it is possible for us to acquire the mosaic and optical sectioning separately (i.e. mosaic for every plane of interest), export out and compile them with another software. Does any of you have any suggestion of possible software that has the capability of compiling individual planes into a stack? Please advise. Thank you! Cheers, Rachael Lim |
![]() ![]() |
heckman@bgnet.bgsu.edu |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Rachael Lim
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Rachaei- I don't know where you are writing from, but you can do anything with Advanced Visualization Systems software. The learning curve is pretty steep and they may not have it everywhere. > the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hi! We are trying to run an experiment that requires image acquisition of a >very large area (a slice of rat brain; we use mosaic for this purpose), >along with optical sectioning. > >However, the software that we are currently using does not allow us to >acquire both mosaic and optical sectioning simultaneously due to the >extremely large file size produced (about 2GB), which cannot be processed by >the software (software limitation of file size is 200MB). > >I am wondering if it is possible for us to acquire the mosaic and optical >sectioning separately (i.e. mosaic for every plane of interest), export out >and compile them with another software. Does any of you have any suggestion >of possible software that has the capability of compiling individual planes >into a stack? > >Please advise. Thank you! > >Cheers, >Rachael Lim ... [show rest of quote] -- Carol A. Heckman, Ph.D. Director, Center for Microscopy & Microanalysis and Professor of Biological Sciences Bowling Green State University Bowling Green, OH 43403 USA website: http://www.bgsu.edu/departments/biology/facilities/MnM |
![]() ![]() |
George McNamara |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Rachael Lim
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Rachel, you may want to try out the software described in S.L. Wearne et al 2005 Neuroscience 136: 6610-680, http://www.mssm.edu/cnic/tools.html and http://www.mssm.edu/cnic/arborization.html (the latter is a web summary of the paper). The process is done post acquisition. At 03:57 AM 10/17/2007, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hi! We are trying to run an experiment that requires image acquisition of a >very large area (a slice of rat brain; we use mosaic for this purpose), >along with optical sectioning. > >However, the software that we are currently using does not allow us to >acquire both mosaic and optical sectioning simultaneously due to the >extremely large file size produced (about 2GB), which cannot be processed by >the software (software limitation of file size is 200MB). > >I am wondering if it is possible for us to acquire the mosaic and optical >sectioning separately (i.e. mosaic for every plane of interest), export out >and compile them with another software. Does any of you have any suggestion >of possible software that has the capability of compiling individual planes >into a stack? > >Please advise. Thank you! > >Cheers, >Rachael Lim ... [show rest of quote] George McNamara, Ph.D. University of Miami, Miller School of Medicine Image Core Miami, FL 33010 [hidden email] [hidden email] 305-243-8436 office http://home.earthlink.net/~pubspectra/ http://home.earthlink.net/~geomcnamara/ http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility) |
![]() ![]() |
Sven Terclavers |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by heckman@bgnet.bgsu.edu
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Rachael, I don't know if it's possible to install on your equipment, but I get nice results with the Zeiss LSM510 software and their Multi Time 4.0 macro. Large 3D mosaics with multiple tracks (3 to 4 fluorophores) do not give problems, unless the file is larger than 3Gb, but then it's not the software though our Windows-version that blocks the use of more memory (a 64-bit system will allow higher memory allocation, we run the 32-bit version). It might be possible to create lsm-z-stack files from a z-stack saved as single Tiff-files and then run the Mosaic tile-utility on those single lsm-files, but this I've never tried. Sven Terclavers -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Carol Heckman Sent: Thursday, 18 October, 2007 01:36 To: [hidden email] Subject: Re: [CONFOCAL] Mosaic with optical sectoning Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Rachaei- I don't know where you are writing from, but you can do anything with Advanced Visualization Systems software. The learning curve is pretty steep and they may not have it everywhere. > the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hi! We are trying to run an experiment that requires image acquisition of a >very large area (a slice of rat brain; we use mosaic for this purpose), >along with optical sectioning. > >However, the software that we are currently using does not allow us to >acquire both mosaic and optical sectioning simultaneously due to the >extremely large file size produced (about 2GB), which cannot be processed >the software (software limitation of file size is 200MB). > >I am wondering if it is possible for us to acquire the mosaic and optical >sectioning separately (i.e. mosaic for every plane of interest), export out >and compile them with another software. Does any of you have any suggestion >of possible software that has the capability of compiling individual planes >into a stack? > >Please advise. Thank you! > >Cheers, >Rachael Lim -- Carol A. Heckman, Ph.D. Director, Center for Microscopy & Microanalysis and Professor of Biological Sciences Bowling Green State University Bowling Green, OH 43403 USA website: http://www.bgsu.edu/departments/biology/facilities/MnM Disclaimer: http://www.kuleuven.be/cwis/email_disclaimer.htm |
![]() ![]() |
Stanislav Vitha |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Rachael Lim
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Rachael, you do not mention what instrument you are using; For multi-area time lapse Z-stack data from our Olympus FV1000 confocal I have recently been using the ImageJ mosaicking plug-in from the good people at NCMIR, which can import the Olympus confocal files and also understands the XY mosaic log file. download at http://www.ncmir.ucsd.edu/Software/imagej_mosaicking_plugins.htm instructions at http://www.ncmir.ucsd.edu/~skchow/rts2000/mosaicprocess.html With our scope I set up acquisition of overlapping tiles (or for 3D, overlapping z-stacks). The confocal software acquires the z-stacks and also saves a log file with XY coordinates of each z-stack. The mosaicking plugin reads the log file and based on the XY coordinates of each tile assembles them into a mosaic. You end up with assembled mosaic for each z-slice. Their file format is tiled tiff, which can be read by some, but not all software. Stan Vitha Microscopy and Imaging Center Texas A&M University BSBW 119 College Station, TX 77843-2257 On Wed, 17 Oct 2007 15:57:07 +0800, Rachael Lim <[hidden email]> wrote: >Hi! We are trying to run an experiment that requires image acquisition of a >very large area (a slice of rat brain; we use mosaic for this purpose), >along with optical sectioning. > >However, the software that we are currently using does not allow us to >acquire both mosaic and optical sectioning simultaneously due to the >extremely large file size produced (about 2GB), which cannot be processed by >the software (software limitation of file size is 200MB). > >I am wondering if it is possible for us to acquire the mosaic and optical >sectioning separately (i.e. mosaic for every plane of interest), export out >and compile them with another software. Does any of you have any suggestion >of possible software that has the capability of compiling individual planes >into a stack? > >Please advise. Thank you! > >Cheers, >Rachael Lim >========================================================================= |
![]() ![]() |
Kevin Ryan-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Rachael Lim
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Image-Pro, from Media Cybernetics, can assemble mosaics and stacks. Yes, I'm biased; I write this software! There are some limitations - due to Windows OS issues any _single_ frame is limited to ~500MB (leaving room for operating on two images with an output frame, plus the software, inside the ~1.6GB limit of Windows XP). On a 64-bit system that goes up to ~700-900MB. If a single plane of your mosaic is under that size limit Image-Pro can assemble your mosaics and stacks for viewing. If all individual frames are under the size limits Image-Pro will use virtual memory to swap them in and out for operations. I've had >10GB of images open at once, performing tracking on long time lapse series. A mosaic with optical sectioning assembly would be a matter of: (a) Tiling your mosaic images at a plane with good information, features in all tiles of the mosaic, (b) Applying that tiling to all planes, creating a series of mosaics, and (c) Viewing/measuring objects in your series of mosaics as a volume. Note that Image-Pro can read most commonly available file formats. A basic 3D viewer is included in the base Image-Pro; there is an additional module for 3D object segmentation and analysis. Kevin Ryan Media Cybernetics, Inc. > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On > Behalf Of Rachael Lim > Sent: Wednesday, October 17, 2007 3:57 AM > To: [hidden email] > Subject: Mosaic with optical sectoning > > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi! We are trying to run an experiment that requires image acquisition of > a > very large area (a slice of rat brain; we use mosaic for this purpose), > along with optical sectioning. > > However, the software that we are currently using does not allow us to > acquire both mosaic and optical sectioning simultaneously due to the > extremely large file size produced (about 2GB), which cannot be processed > by > the software (software limitation of file size is 200MB). > > I am wondering if it is possible for us to acquire the mosaic and optical > sectioning separately (i.e. mosaic for every plane of interest), export > out > and compile them with another software. Does any of you have any > suggestion > of possible software that has the capability of compiling individual > planes > into a stack? > > Please advise. Thank you! > > Cheers, > Rachael Lim |
![]() ![]() |
Christian Götze |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Rachael Lim
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal COMMERCIAL WARNING! Hi, our company, arivis (http://www.arivis.com) provides the arivis Browser - a software for handling very large image data. Using the Browser there is no problem with the data size anymore. The software has its roots in confocal applications and therefore should be suitable for your problem. We support the acquisition by several devices directly, as Leica confocals using LCS software, different cameras and stages or all TILL Photonics devices. If your device is not supported you can import the single images directly. For getting an idea of the data size we work on usually look at http://www.arivis.com/webview. Don't hestitate to contact me for further questions. Christian Goetze On Wed, 17 Oct 2007 15:57:07 +0800, Rachael Lim <[hidden email]> wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hi! We are trying to run an experiment that requires image acquisition of a >very large area (a slice of rat brain; we use mosaic for this purpose), >along with optical sectioning. > >However, the software that we are currently using does not allow us to >acquire both mosaic and optical sectioning simultaneously due to the >extremely large file size produced (about 2GB), which cannot be processed by >the software (software limitation of file size is 200MB). > >I am wondering if it is possible for us to acquire the mosaic and optical >sectioning separately (i.e. mosaic for every plane of interest), export out >and compile them with another software. Does any of you have any suggestion >of possible software that has the capability of compiling individual planes >into a stack? > >Please advise. Thank you! > >Cheers, >Rachael Lim >========================================================================= ... [show rest of quote]
|