Dear list During a recent FRET experiment a question came up about
what mounting media should be used on fixed samples during FRET experiments and
whether any one media is better than other for this particular type of
experiment. From what’s published it seems that people use all
kinds of different media. Does anybody here have any idea about any
specific media that would be best/give least interference with FRET results? Thanks a lot to all of you in advance, --------------------------------------------------- Yevgeniy
Romin Digital
Microscopist Memorial
Sloan-Kettering Cancer Center Molecular
Cytology Core Facility 1275
York Ave. Box 333 New
York, NY 10065 Tel.646-888-2186 Fax.
646-422-0640 ---------------------------------------------------
=====================================================================
Please note that this e-mail and any files transmitted with it may be
privileged, confidential, and protected from disclosure under
applicable law. If the reader of this message is not the intended
recipient, or an employee or agent responsible for delivering this
message to the intended recipient, you are hereby notified that any
reading, dissemination, distribution, copying, or other use of this
communication or any of its attachments is strictly prohibited. If
you have received this communication in error, please notify the
sender immediately by replying to this message and deleting this
message, any attachments, and all copies and backups from your
computer.
|
|
John Oreopoulos |
|
|
I have not used fixed specimens for FRET imaging (I have used the sensitized emission method on living cells), but I came across this paper a while ago which might be of interest to you:
Rodighiero, S., et al., Fixation, mounting and sealing with nail polish of cell specimens lead to incorrect FRET measurements using acceptor photobleaching. Cellular Physiology and Biochemistry, 2008. 21(5-6): p. 489-498. John Oreopoulos On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > Dear list > > During a recent FRET experiment a question came up about what mounting media should be used on fixed samples during FRET experiments and whether any one media is better than other for this particular type of experiment. From what’s published it seems that people use all kinds of different media. Does anybody here have any idea about any specific media that would be best/give least interference with FRET results? > > Thanks a lot to all of you in advance, > > --------------------------------------------------- > Yevgeniy Romin > > Digital Microscopist > Memorial Sloan-Kettering Cancer Center > Molecular Cytology Core Facility > 1275 York Ave. Box 333 > New York, NY 10065 > Tel.646-888-2186 > Fax. 646-422-0640 > --------------------------------------------------- > > > > ===================================================================== > > Please note that this e-mail and any files transmitted with it may be > privileged, confidential, and protected from disclosure under > applicable law. If the reader of this message is not the intended > recipient, or an employee or agent responsible for delivering this > message to the intended recipient, you are hereby notified that any > reading, dissemination, distribution, copying, or other use of this > communication or any of its attachments is strictly prohibited. If > you have received this communication in error, please notify the > sender immediately by replying to this message and deleting this > message, any attachments, and all copies and backups from your > computer. > > |
|
Romin, Yevgeniy/Sloan Kettering Institute |
|
|
Thank you very much John, I found the paper you recommended and it is already proving useful. A question for you then - which media were you using for the live cells? What I mean is whether it was something like PBS or some specific media?
--------------------------------------------------- Yevgeniy Romin Digital Microscopist Memorial Sloan-Kettering Cancer Center Molecular Cytology Core Facility 1275 York Ave. Box 333 New York, NY 10065 Tel.646-888-2186 Fax. 646-422-0640 --------------------------------------------------- -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of John Oreopoulos Sent: Friday, April 16, 2010 11:30 AM To: [hidden email] Subject: Re: Mounting media for FRET experiments I have not used fixed specimens for FRET imaging (I have used the sensitized emission method on living cells), but I came across this paper a while ago which might be of interest to you: Rodighiero, S., et al., Fixation, mounting and sealing with nail polish of cell specimens lead to incorrect FRET measurements using acceptor photobleaching. Cellular Physiology and Biochemistry, 2008. 21(5-6): p. 489-498. John Oreopoulos On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > Dear list > > During a recent FRET experiment a question came up about what mounting media should be used on fixed samples during FRET experiments and whether any one media is better than other for this particular type of experiment. From what's published it seems that people use all kinds of different media. Does anybody here have any idea about any specific media that would be best/give least interference with FRET results? > > Thanks a lot to all of you in advance, > > --------------------------------------------------- > Yevgeniy Romin > > Digital Microscopist > Memorial Sloan-Kettering Cancer Center > Molecular Cytology Core Facility > 1275 York Ave. Box 333 > New York, NY 10065 > Tel.646-888-2186 > Fax. 646-422-0640 > --------------------------------------------------- > > > > ===================================================================== > > Please note that this e-mail and any files transmitted with it may be > privileged, confidential, and protected from disclosure under > applicable law. If the reader of this message is not the intended > recipient, or an employee or agent responsible for delivering this > message to the intended recipient, you are hereby notified that any > reading, dissemination, distribution, copying, or other use of this > communication or any of its attachments is strictly prohibited. If > you have received this communication in error, please notify the > sender immediately by replying to this message and deleting this > message, any attachments, and all copies and backups from your > computer. > > |
|
John Oreopoulos |
|
|
Just regular PBS solution. I have never tried a rigorous comparison of different media.
John On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: > Thank you very much John, I found the paper you recommended and it is already proving useful. A question for you then - which media were you using for the live cells? What I mean is whether it was something like PBS or some specific media? > > --------------------------------------------------- > Yevgeniy Romin > > Digital Microscopist > Memorial Sloan-Kettering Cancer Center > Molecular Cytology Core Facility > 1275 York Ave. Box 333 > New York, NY 10065 > Tel.646-888-2186 > Fax. 646-422-0640 > --------------------------------------------------- > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of John Oreopoulos > Sent: Friday, April 16, 2010 11:30 AM > To: [hidden email] > Subject: Re: Mounting media for FRET experiments > > I have not used fixed specimens for FRET imaging (I have used the sensitized emission method on living cells), but I came across this paper a while ago which might be of interest to you: > > Rodighiero, S., et al., Fixation, mounting and sealing with nail polish of cell specimens lead to incorrect FRET measurements using acceptor photobleaching. Cellular Physiology and Biochemistry, 2008. 21(5-6): p. 489-498. > > John Oreopoulos > > > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > >> Dear list >> >> During a recent FRET experiment a question came up about what mounting media should be used on fixed samples during FRET experiments and whether any one media is better than other for this particular type of experiment. From what's published it seems that people use all kinds of different media. Does anybody here have any idea about any specific media that would be best/give least interference with FRET results? >> >> Thanks a lot to all of you in advance, >> >> --------------------------------------------------- >> Yevgeniy Romin >> >> Digital Microscopist >> Memorial Sloan-Kettering Cancer Center >> Molecular Cytology Core Facility >> 1275 York Ave. Box 333 >> New York, NY 10065 >> Tel.646-888-2186 >> Fax. 646-422-0640 >> --------------------------------------------------- >> >> >> >> ===================================================================== >> >> Please note that this e-mail and any files transmitted with it may be >> privileged, confidential, and protected from disclosure under >> applicable law. If the reader of this message is not the intended >> recipient, or an employee or agent responsible for delivering this >> message to the intended recipient, you are hereby notified that any >> reading, dissemination, distribution, copying, or other use of this >> communication or any of its attachments is strictly prohibited. If >> you have received this communication in error, please notify the >> sender immediately by replying to this message and deleting this >> message, any attachments, and all copies and backups from your >> computer. >> >> |
|
Vitaly Boyko |
|
|
Dear John,
Dear Yevgeniy,
Yes, PBS or specialized low serum (1-2%) media is broadly used for live cell imaging under TIR illumination.
I would like to bring your attention to two papers which are relevant to single molecule fluorscence and/or low light live cell imaging - I sent them separately to your e-mail addressess as the LISTSERVER rejected my message due to the attachments. I can also send the papers to anyone by request.
1. The Use of Protocatechuate Dioxygenase for Maintaining Anaerobic Conditions in Biochemical Experiments
Pravin V. Patil and David P. Ballou Department of Biological Chemistry, University of Michigan, Ann Arbor, Michigan 48109-0606 Analytical Biochemistry 286, 187–192 (2000)
2. An Oxygen Scavenging System for Improvement of Dye Stability in Single-Molecule Fluorescence Experiments
Colin Echeverrı´a Aitken,* R. Andrew Marshall, y and Joseph D. Puglisiz§Biophysical Journal Volume 94 March 2008 1826–1835
If you have any questions, please contact me offline or phone.
Also, the accuracy of the FRET eff. measurements could vary or be affected by the mounting media used - live cell FRET seems to be the best approach.
Have a nice weekend,
Vitaly
301-515-7833
From: John Oreopoulos <[hidden email]> To: [hidden email] Sent: Fri, April 16, 2010 12:01:32 PM Subject: Re: Mounting media for FRET experiments Just regular PBS solution. I have never tried a rigorous comparison of different media. John On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: > Thank you very much John, I found the paper you recommended and it is already proving useful. A question for you then - which media were you using for the live cells? What I mean is whether it was something like PBS or some specific media? > > --------------------------------------------------- > Yevgeniy Romin > > Digital Microscopist > Memorial Sloan-Kettering Cancer Center > Molecular Cytology Core Facility > 1275 York Ave. Box 333 > New York, NY 10065 > Tel.646-888-2186 > Fax. 646-422-0640 > --------------------------------------------------- > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of John Oreopoulos > Sent: Friday, April 16, 2010 11:30 AM > To: [hidden email] > Subject: Re: Mounting media for FRET experiments > > I have not used fixed specimens for FRET imaging (I have used the sensitized emission method on living cells), but I came across this paper a while ago which might be of interest to you: > > Rodighiero, S., et al., Fixation, mounting and sealing with nail polish of cell specimens lead to incorrect FRET measurements using acceptor photobleaching. Cellular Physiology and Biochemistry, 2008. 21(5-6): p. 489-498. > > John Oreopoulos > > > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > >> Dear list >> >> During a recent FRET experiment a question came up about what mounting media should be used on fixed samples during FRET experiments and whether any one media is better than other for this particular type of experiment. From what's published it seems that people use all kinds of different media. Does anybody here have any idea about any specific media that would be best/give least interference with FRET results? >> >> Thanks a lot to all of you in advance, >> >> --------------------------------------------------- >> Yevgeniy Romin >> >> Digital Microscopist >> Memorial Sloan-Kettering Cancer Center >> Molecular Cytology Core Facility >> 1275 York Ave. Box 333 >> New York, NY 10065 >> Tel.646-888-2186 >> Fax. 646-422-0640 >> --------------------------------------------------- >> >> >> >> ===================================================================== >> >> Please note that this e-mail and any files transmitted with it may be >> privileged, confidential, and protected from disclosure under >> applicable law. If the reader of this message is not the intended >> recipient, or an employee or agent responsible for delivering this >> message to the intended recipient, you are hereby notified that any >> reading, dissemination, distribution, copying, or other use of this >> communication or any of its attachments is strictly prohibited. If >> you have received this communication in error, please notify the >> sender immediately by replying to this message and deleting this >> message, any attachments, and all copies and backups from your >> computer. >> >> |
|
Daniel Gitler |
|
|
Dear John and Yevgeniy,
My experience when measuring FRET from the same samples before and after fixation is that fixation itself will very much change the FRET efficiency value that you measure (Cerulean and Venus in HEK cells, PBS for both). Efficiency in the fixed samples was significantly lower. This should not come as any big surprise, since the cellular environment around the fluorophores will be very different. The possible reasons for the changes in FRET efficiency are very many... I also do most of my FRET imaging in live samples. Daniel ----- Original Message ----- From: Vitaly Boyko <[hidden email]> Date: Friday, April 16, 2010 19:36 Subject: Re: Mounting media for FRET experiments To: [hidden email] > Dear John, > Dear Yevgeniy, > > Yes, PBS or specialized low serum (1-2%) media is broadly used > for live cell imaging under TIR illumination. > > I would like to bring your attention to two papers which are > relevant to single molecule fluorscence and/or low light live > cell imaging - I sent them separately to your e-mail addressess > as the LISTSERVER rejected my message due to the attachments. I > can also send the papers to anyone by request. > > 1. The Use of Protocatechuate Dioxygenase for Maintaining > Anaerobic Conditions in Biochemical ExperimentsPravin V. Patil > and David P. BallouDepartment of Biological Chemistry, > University of Michigan, Ann Arbor, Michigan 48109-0606 > Analytical Biochemistry 286, 187–192 (2000) > 2. An Oxygen Scavenging System for Improvement of Dye Stability > in Single-Molecule Fluorescence ExperimentsColin Echeverrı´a > Aitken,* R. Andrew Marshall,y and Joseph D. Puglisiz§ > Biophysical Journal Volume 94 March 2008 1826–1835 > > If you have any questions, please contact me offline or phone. > > Also, the accuracy of the FRET eff. measurements could vary or > be affected by the mounting media used - live cell FRET seems to > be the best approach. > > Have a nice weekend, > > Vitaly > 301-515-7833 > > > > > > ________________________________ > From: John Oreopoulos <[hidden email]> > To: [hidden email] > Sent: Fri, April 16, 2010 12:01:32 PM > Subject: Re: Mounting media for FRET experiments > > Just regular PBS solution. I have never tried a rigorous > comparison of different media. > > John > > On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: > > > Thank you very much John, I found the paper you recommended > and it is already proving useful. A question for you then - > which media were you using for the live cells? What I mean is > whether it was something like PBS or some specific media? > > > > --------------------------------------------------- > > Yevgeniy Romin > > > > Digital Microscopist > > Memorial Sloan-Kettering Cancer Center > > Molecular Cytology Core Facility > > 1275 York Ave. Box 333 > > New York, NY 10065 > > Tel.646-888-2186 > > Fax. 646-422-0640 > > --------------------------------------------------- > > > > -----Original Message----- > > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of John Oreopoulos > > Sent: Friday, April 16, 2010 11:30 AM > > To: [hidden email] > > Subject: Re: Mounting media for FRET experiments > > > > I have not used fixed specimens for FRET imaging (I have used > the sensitized emission method on living cells), but I came > across this paper a while ago which might be of interest to you: > > > > Rodighiero, S., et al., Fixation, mounting and sealing with > nail polish of cell specimens lead to incorrect FRET > measurements using acceptor photobleaching. Cellular Physiology > and Biochemistry, 2008. 21(5-6): p. 489-498. > > > > John Oreopoulos > > > > > > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > > > >> Dear list > >> > >> During a recent FRET experiment a question came up about what > mounting media should be used on fixed samples during FRET > experiments and whether any one media is better than other for > this particular type of experiment. From what's published it > seems that people use all kinds of different media. Does > anybody here have any idea about any specific media that would > be best/give least interference with FRET results? > >> > >> Thanks a lot to all of you in advance, > >> > >> --------------------------------------------------- > >> Yevgeniy Romin > >> > >> Digital Microscopist > >> Memorial Sloan-Kettering Cancer Center > >> Molecular Cytology Core Facility > >> 1275 York Ave. Box 333 > >> New York, NY 10065 > >> Tel.646-888-2186 > >> Fax. 646-422-0640 > >> --------------------------------------------------- > >> > >> > >> > >> > =====================================================================>> > >> Please note that this e-mail and any files transmitted > with it may be > >> privileged, confidential, and protected from disclosure under > >> applicable law. If the reader of this message is not the > intended>> recipient, or an employee or agent responsible for > delivering this > >> message to the intended recipient, you are hereby notified > that any > >> reading, dissemination, distribution, copying, or other > use of this > >> communication or any of its attachments is strictly > prohibited. If > >> you have received this communication in error, please > notify the > >> sender immediately by replying to this message and > deleting this > >> message, any attachments, and all copies and backups from your > >> computer. > >> > >> > > > > Daniel Gitler, Ph.D. Department of Physiology and Neurobiology Faculty of Health Sciences Ben Gurion University of the Negev Beer-Sheva 84105 Israel Tel: +972-8-6477345 Cell: +972-54-2110100 Fax: + 972-8-6477628 http://web2.bgu.ac.il/physiology/faculty-members/daniel-gitler/ |
|
Romin, Yevgeniy/Sloan Kettering Institute |
|
|
Thank you very much to everyone replying both on and off the list. Your advice and the papers that were referred to were very helpful. FRET on live samples would be the ideal choice, unfortunately in this case we are doing FRET on white blood cells, which are very round and refuse to attach and sit still, introducing motion artifacts into the FRET analysis. We are currently working on immobilizing them, but there is a fear of matrigel or a more viscous media introducing FRET measurement artifacts as well. Does anyone have experience with doing FRET on cells that do not attach or immobilizing them in a least invasive way?
________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Daniel Gitler [[hidden email]] Sent: Saturday, April 17, 2010 5:22 AM To: [hidden email] Subject: Re: Mounting media for FRET experiments Dear John and Yevgeniy, My experience when measuring FRET from the same samples before and after fixation is that fixation itself will very much change the FRET efficiency value that you measure (Cerulean and Venus in HEK cells, PBS for both). Efficiency in the fixed samples was significantly lower. This should not come as any big surprise, since the cellular environment around the fluorophores will be very different. The possible reasons for the changes in FRET efficiency are very many... I also do most of my FRET imaging in live samples. Daniel ----- Original Message ----- From: Vitaly Boyko <[hidden email]> Date: Friday, April 16, 2010 19:36 Subject: Re: Mounting media for FRET experiments To: [hidden email] > Dear John, > Dear Yevgeniy, > > Yes, PBS or specialized low serum (1-2%) media is broadly used > for live cell imaging under TIR illumination. > > I would like to bring your attention to two papers which are > relevant to single molecule fluorscence and/or low light live > cell imaging - I sent them separately to your e-mail addressess > as the LISTSERVER rejected my message due to the attachments. I > can also send the papers to anyone by request. > > 1. The Use of Protocatechuate Dioxygenase for Maintaining > Anaerobic Conditions in Biochemical ExperimentsPravin V. Patil > and David P. BallouDepartment of Biological Chemistry, > University of Michigan, Ann Arbor, Michigan 48109-0606 > Analytical Biochemistry 286, 187–192 (2000) > 2. An Oxygen Scavenging System for Improvement of Dye Stability > in Single-Molecule Fluorescence ExperimentsColin Echeverrı´a > Aitken,* R. Andrew Marshall,y and Joseph D. Puglisiz§ > Biophysical Journal Volume 94 March 2008 1826–1835 > > If you have any questions, please contact me offline or phone. > > Also, the accuracy of the FRET eff. measurements could vary or > be affected by the mounting media used - live cell FRET seems to > be the best approach. > > Have a nice weekend, > > Vitaly > 301-515-7833 > > > > > > ________________________________ > From: John Oreopoulos <[hidden email]> > To: [hidden email] > Sent: Fri, April 16, 2010 12:01:32 PM > Subject: Re: Mounting media for FRET experiments > > Just regular PBS solution. I have never tried a rigorous > comparison of different media. > > John > > On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: > > > Thank you very much John, I found the paper you recommended > and it is already proving useful. A question for you then - > which media were you using for the live cells? What I mean is > whether it was something like PBS or some specific media? > > > > --------------------------------------------------- > > Yevgeniy Romin > > > > Digital Microscopist > > Memorial Sloan-Kettering Cancer Center > > Molecular Cytology Core Facility > > 1275 York Ave. Box 333 > > New York, NY 10065 > > Tel.646-888-2186 > > Fax. 646-422-0640 > > --------------------------------------------------- > > > > -----Original Message----- > > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of John Oreopoulos > > Sent: Friday, April 16, 2010 11:30 AM > > To: [hidden email] > > Subject: Re: Mounting media for FRET experiments > > > > I have not used fixed specimens for FRET imaging (I have used > the sensitized emission method on living cells), but I came > across this paper a while ago which might be of interest to you: > > > > Rodighiero, S., et al., Fixation, mounting and sealing with > nail polish of cell specimens lead to incorrect FRET > measurements using acceptor photobleaching. Cellular Physiology > and Biochemistry, 2008. 21(5-6): p. 489-498. > > > > John Oreopoulos > > > > > > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > > > >> Dear list > >> > >> During a recent FRET experiment a question came up about what > mounting media should be used on fixed samples during FRET > experiments and whether any one media is better than other for > this particular type of experiment. From what's published it > seems that people use all kinds of different media. Does > anybody here have any idea about any specific media that would > be best/give least interference with FRET results? > >> > >> Thanks a lot to all of you in advance, > >> > >> --------------------------------------------------- > >> Yevgeniy Romin > >> > >> Digital Microscopist > >> Memorial Sloan-Kettering Cancer Center > >> Molecular Cytology Core Facility > >> 1275 York Ave. Box 333 > >> New York, NY 10065 > >> Tel.646-888-2186 > >> Fax. 646-422-0640 > >> --------------------------------------------------- > >> > >> > >> > >> > =====================================================================>> > >> Please note that this e-mail and any files transmitted > with it may be > >> privileged, confidential, and protected from disclosure under > >> applicable law. If the reader of this message is not the > intended>> recipient, or an employee or agent responsible for > delivering this > >> message to the intended recipient, you are hereby notified > that any > >> reading, dissemination, distribution, copying, or other > use of this > >> communication or any of its attachments is strictly > prohibited. If > >> you have received this communication in error, please > notify the > >> sender immediately by replying to this message and > deleting this > >> message, any attachments, and all copies and backups from your > >> computer. > >> > >> > > > > Daniel Gitler, Ph.D. Department of Physiology and Neurobiology Faculty of Health Sciences Ben Gurion University of the Negev Beer-Sheva 84105 Israel Tel: +972-8-6477345 Cell: +972-54-2110100 Fax: + 972-8-6477628 http://web2.bgu.ac.il/physiology/faculty-members/daniel-gitler/ |
|
Cameron Nowell |
|
|
Dear Yevgeniy,
While i haven't had any experience imobilising cells for FRET we have had great success in the past imobolising cells for ICS. We had a cell line that expressed the receptor of interest but was round and didn't attache to anything. We just put them in some 1% low melt temp agarose (made up in PBS) and everything worked really well. Now we were only looking at the bleching of one receptor, but any movement in the system would have been noticed in the final analysis and there was none. I cant see a low percentage of agarose interfereing with you FRET measurments. Cheers Cam Cameron J. Nowell Microscpy Manager Central Resource for Advanced Microscopy Ludwig Insttue for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 AUSTRALIA Office: +61 3 9341 3155 Mobile: +61422882700 Fax: +61 3 9341 3104 http://www.ludwig.edu.au/branch/research/platform/microscopy.htm ________________________________ From: Confocal Microscopy List on behalf of Yevgeniy Romin Sent: Sun 18/04/2010 1:49 AM To: [hidden email] Subject: Re: Mounting media for FRET experiments Thank you very much to everyone replying both on and off the list. Your advice and the papers that were referred to were very helpful. FRET on live samples would be the ideal choice, unfortunately in this case we are doing FRET on white blood cells, which are very round and refuse to attach and sit still, introducing motion artifacts into the FRET analysis. We are currently working on immobilizing them, but there is a fear of matrigel or a more viscous media introducing FRET measurement artifacts as well. Does anyone have experience with doing FRET on cells that do not attach or immobilizing them in a least invasive way? ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Daniel Gitler [[hidden email]] Sent: Saturday, April 17, 2010 5:22 AM To: [hidden email] Subject: Re: Mounting media for FRET experiments Dear John and Yevgeniy, My experience when measuring FRET from the same samples before and after fixation is that fixation itself will very much change the FRET efficiency value that you measure (Cerulean and Venus in HEK cells, PBS for both). Efficiency in the fixed samples was significantly lower. This should not come as any big surprise, since the cellular environment around the fluorophores will be very different. The possible reasons for the changes in FRET efficiency are very many... I also do most of my FRET imaging in live samples. Daniel ----- Original Message ----- From: Vitaly Boyko <[hidden email]> Date: Friday, April 16, 2010 19:36 Subject: Re: Mounting media for FRET experiments To: [hidden email] > Dear John, > Dear Yevgeniy, > > Yes, PBS or specialized low serum (1-2%) media is broadly used > for live cell imaging under TIR illumination. > > I would like to bring your attention to two papers which are > relevant to single molecule fluorscence and/or low light live > cell imaging - I sent them separately to your e-mail addressess > as the LISTSERVER rejected my message due to the attachments. I > can also send the papers to anyone by request. > > 1. The Use of Protocatechuate Dioxygenase for Maintaining > Anaerobic Conditions in Biochemical ExperimentsPravin V. Patil > and David P. BallouDepartment of Biological Chemistry, > University of Michigan, Ann Arbor, Michigan 48109-0606 > Analytical Biochemistry 286, 187-192 (2000) > 2. An Oxygen Scavenging System for Improvement of Dye Stability > in Single-Molecule Fluorescence ExperimentsColin Echeverri´a > Aitken,* R. Andrew Marshall,y and Joseph D. Puglisiz§ > Biophysical Journal Volume 94 March 2008 1826-1835 > > If you have any questions, please contact me offline or phone. > > Also, the accuracy of the FRET eff. measurements could vary or > be affected by the mounting media used - live cell FRET seems to > be the best approach. > > Have a nice weekend, > > Vitaly > 301-515-7833 > > > > > > ________________________________ > From: John Oreopoulos <[hidden email]> > To: [hidden email] > Sent: Fri, April 16, 2010 12:01:32 PM > Subject: Re: Mounting media for FRET experiments > > Just regular PBS solution. I have never tried a rigorous > comparison of different media. > > John > > On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: > > > Thank you very much John, I found the paper you recommended > and it is already proving useful. A question for you then - > which media were you using for the live cells? What I mean is > whether it was something like PBS or some specific media? > > > > --------------------------------------------------- > > Yevgeniy Romin > > > > Digital Microscopist > > Memorial Sloan-Kettering Cancer Center > > Molecular Cytology Core Facility > > 1275 York Ave. Box 333 > > New York, NY 10065 > > Tel.646-888-2186 > > Fax. 646-422-0640 > > --------------------------------------------------- > > > > -----Original Message----- > > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of John Oreopoulos > > Sent: Friday, April 16, 2010 11:30 AM > > To: [hidden email] > > Subject: Re: Mounting media for FRET experiments > > > > I have not used fixed specimens for FRET imaging (I have used > the sensitized emission method on living cells), but I came > across this paper a while ago which might be of interest to you: > > > > Rodighiero, S., et al., Fixation, mounting and sealing with > nail polish of cell specimens lead to incorrect FRET > measurements using acceptor photobleaching. Cellular Physiology > and Biochemistry, 2008. 21(5-6): p. 489-498. > > > > John Oreopoulos > > > > > > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > > > >> Dear list > >> > >> During a recent FRET experiment a question came up about what > mounting media should be used on fixed samples during FRET > experiments and whether any one media is better than other for > this particular type of experiment. From what's published it > seems that people use all kinds of different media. Does > anybody here have any idea about any specific media that would > be best/give least interference with FRET results? > >> > >> Thanks a lot to all of you in advance, > >> > >> --------------------------------------------------- > >> Yevgeniy Romin > >> > >> Digital Microscopist > >> Memorial Sloan-Kettering Cancer Center > >> Molecular Cytology Core Facility > >> 1275 York Ave. Box 333 > >> New York, NY 10065 > >> Tel.646-888-2186 > >> Fax. 646-422-0640 > >> --------------------------------------------------- > >> > >> > >> > >> > =====================================================================>> > >> Please note that this e-mail and any files transmitted > with it may be > >> privileged, confidential, and protected from disclosure under > >> applicable law. If the reader of this message is not the > intended>> recipient, or an employee or agent responsible for > delivering this > >> message to the intended recipient, you are hereby notified > that any > >> reading, dissemination, distribution, copying, or other > use of this > >> communication or any of its attachments is strictly > prohibited. If > >> you have received this communication in error, please > notify the > >> sender immediately by replying to this message and > deleting this > >> message, any attachments, and all copies and backups from your > >> computer. > >> > >> > > > > Daniel Gitler, Ph.D. Department of Physiology and Neurobiology Faculty of Health Sciences Ben Gurion University of the Negev Beer-Sheva 84105 Israel Tel: +972-8-6477345 Cell: +972-54-2110100 Fax: + 972-8-6477628 http://web2.bgu.ac.il/physiology/faculty-members/daniel-gitler/ ? |
|
Cameron Nowell |
|
|
I should also add that we capture a series first in reflection mode to be able to pick the cells that were against the coverslip to remove and potential artifact of imaging through a chunk of agarose.
Cheers Cam ________________________________ From: Confocal Microscopy List on behalf of Cameron Nowell Sent: Sun 18/04/2010 8:47 AM To: [hidden email] Subject: Re: Mounting media for FRET experiments Dear Yevgeniy, While i haven't had any experience imobilising cells for FRET we have had great success in the past imobolising cells for ICS. We had a cell line that expressed the receptor of interest but was round and didn't attache to anything. We just put them in some 1% low melt temp agarose (made up in PBS) and everything worked really well. Now we were only looking at the bleching of one receptor, but any movement in the system would have been noticed in the final analysis and there was none. I cant see a low percentage of agarose interfereing with you FRET measurments. Cheers Cam Cameron J. Nowell Microscpy Manager Central Resource for Advanced Microscopy Ludwig Insttue for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 AUSTRALIA Office: +61 3 9341 3155 Mobile: +61422882700 Fax: +61 3 9341 3104 http://www.ludwig.edu.au/branch/research/platform/microscopy.htm ________________________________ From: Confocal Microscopy List on behalf of Yevgeniy Romin Sent: Sun 18/04/2010 1:49 AM To: [hidden email] Subject: Re: Mounting media for FRET experiments Thank you very much to everyone replying both on and off the list. Your advice and the papers that were referred to were very helpful. FRET on live samples would be the ideal choice, unfortunately in this case we are doing FRET on white blood cells, which are very round and refuse to attach and sit still, introducing motion artifacts into the FRET analysis. We are currently working on immobilizing them, but there is a fear of matrigel or a more viscous media introducing FRET measurement artifacts as well. Does anyone have experience with doing FRET on cells that do not attach or immobilizing them in a least invasive way? ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Daniel Gitler [[hidden email]] Sent: Saturday, April 17, 2010 5:22 AM To: [hidden email] Subject: Re: Mounting media for FRET experiments Dear John and Yevgeniy, My experience when measuring FRET from the same samples before and after fixation is that fixation itself will very much change the FRET efficiency value that you measure (Cerulean and Venus in HEK cells, PBS for both). Efficiency in the fixed samples was significantly lower. This should not come as any big surprise, since the cellular environment around the fluorophores will be very different. The possible reasons for the changes in FRET efficiency are very many... I also do most of my FRET imaging in live samples. Daniel ----- Original Message ----- From: Vitaly Boyko <[hidden email]> Date: Friday, April 16, 2010 19:36 Subject: Re: Mounting media for FRET experiments To: [hidden email] > Dear John, > Dear Yevgeniy, > > Yes, PBS or specialized low serum (1-2%) media is broadly used > for live cell imaging under TIR illumination. > > I would like to bring your attention to two papers which are > relevant to single molecule fluorscence and/or low light live > cell imaging - I sent them separately to your e-mail addressess > as the LISTSERVER rejected my message due to the attachments. I > can also send the papers to anyone by request. > > 1. The Use of Protocatechuate Dioxygenase for Maintaining > Anaerobic Conditions in Biochemical ExperimentsPravin V. Patil > and David P. BallouDepartment of Biological Chemistry, > University of Michigan, Ann Arbor, Michigan 48109-0606 > Analytical Biochemistry 286, 187-192 (2000) > 2. An Oxygen Scavenging System for Improvement of Dye Stability > in Single-Molecule Fluorescence ExperimentsColin Echeverri´a > Aitken,* R. Andrew Marshall,y and Joseph D. Puglisiz§ > Biophysical Journal Volume 94 March 2008 1826-1835 > > If you have any questions, please contact me offline or phone. > > Also, the accuracy of the FRET eff. measurements could vary or > be affected by the mounting media used - live cell FRET seems to > be the best approach. > > Have a nice weekend, > > Vitaly > 301-515-7833 > > > > > > ________________________________ > From: John Oreopoulos <[hidden email]> > To: [hidden email] > Sent: Fri, April 16, 2010 12:01:32 PM > Subject: Re: Mounting media for FRET experiments > > Just regular PBS solution. I have never tried a rigorous > comparison of different media. > > John > > On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: > > > Thank you very much John, I found the paper you recommended > and it is already proving useful. A question for you then - > which media were you using for the live cells? What I mean is > whether it was something like PBS or some specific media? > > > > --------------------------------------------------- > > Yevgeniy Romin > > > > Digital Microscopist > > Memorial Sloan-Kettering Cancer Center > > Molecular Cytology Core Facility > > 1275 York Ave. Box 333 > > New York, NY 10065 > > Tel.646-888-2186 > > Fax. 646-422-0640 > > --------------------------------------------------- > > > > -----Original Message----- > > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of John Oreopoulos > > Sent: Friday, April 16, 2010 11:30 AM > > To: [hidden email] > > Subject: Re: Mounting media for FRET experiments > > > > I have not used fixed specimens for FRET imaging (I have used > the sensitized emission method on living cells), but I came > across this paper a while ago which might be of interest to you: > > > > Rodighiero, S., et al., Fixation, mounting and sealing with > nail polish of cell specimens lead to incorrect FRET > measurements using acceptor photobleaching. Cellular Physiology > and Biochemistry, 2008. 21(5-6): p. 489-498. > > > > John Oreopoulos > > > > > > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: > > > >> Dear list > >> > >> During a recent FRET experiment a question came up about what > mounting media should be used on fixed samples during FRET > experiments and whether any one media is better than other for > this particular type of experiment. From what's published it > seems that people use all kinds of different media. Does > anybody here have any idea about any specific media that would > be best/give least interference with FRET results? > >> > >> Thanks a lot to all of you in advance, > >> > >> --------------------------------------------------- > >> Yevgeniy Romin > >> > >> Digital Microscopist > >> Memorial Sloan-Kettering Cancer Center > >> Molecular Cytology Core Facility > >> 1275 York Ave. Box 333 > >> New York, NY 10065 > >> Tel.646-888-2186 > >> Fax. 646-422-0640 > >> --------------------------------------------------- > >> > >> > >> > >> > =====================================================================>> > >> Please note that this e-mail and any files transmitted > with it may be > >> privileged, confidential, and protected from disclosure under > >> applicable law. If the reader of this message is not the > intended>> recipient, or an employee or agent responsible for > delivering this > >> message to the intended recipient, you are hereby notified > that any > >> reading, dissemination, distribution, copying, or other > use of this > >> communication or any of its attachments is strictly > prohibited. If > >> you have received this communication in error, please > notify the > >> sender immediately by replying to this message and > deleting this > >> message, any attachments, and all copies and backups from your > >> computer. > >> > >> > > > > Daniel Gitler, Ph.D. Department of Physiology and Neurobiology Faculty of Health Sciences Ben Gurion University of the Negev Beer-Sheva 84105 Israel Tel: +972-8-6477345 Cell: +972-54-2110100 Fax: + 972-8-6477628 http://web2.bgu.ac.il/physiology/faculty-members/daniel-gitler/ ? |
|
Martin Spitaler |
|
|
In reply to this post by Romin, Yevgeniy/Sloan Kettering Institute
Dear Yebgeniy,
you could also try CyGel (Biostatus) to immobilise your cells, it's liquid on ice and solid at 37C, so you don't need to heat your cells. I haven't tried it specifically for FRET, but we have used it for various other experiments and it didn't seem to interfere with fluorescence. - No commercial interest - Martin ======================== Martin Spitaler, PhD FILM - Facility for Imaging by Light Microscopy - Facility Manager - Sir Alexander Fleming Building, desk 401 Imperial College London / South Kensington Exhibition Road London SW7 2AZ UK Tel. +44-(0)20-759-42023 E-mail [hidden email] Website: http://imperial.ac.uk/imagingfacility On Sun, 18 Apr 2010 08:47:20 +1000, Cameron Nowell <[hidden email]> wrote: >Dear Yevgeniy, > >While i haven't had any experience imobilising cells for FRET we have had great success in the past imobolising cells for ICS. We had a cell line that expressed the receptor of interest but was round and didn't attache to anything. We just put them in some 1% low melt temp agarose (made up in PBS) and everything worked really well. Now we were only looking at the bleching of one receptor, but any movement in the system would have been noticed in the final analysis and there was none. I cant see a low percentage of agarose interfereing with you FRET measurments. > > >Cheers > > >Cam > > > >Cameron J. Nowell >Microscpy Manager >Central Resource for Advanced Microscopy >Ludwig Insttue for Cancer Research >PO Box 2008 >Royal Melbourne Hospital >Victoria, 3050 >AUSTRALIA > >Office: +61 3 9341 3155 >Mobile: +61422882700 >Fax: +61 3 9341 3104 > >http://www.ludwig.edu.au/branch/research/platform/microscopy.htm > > >________________________________ > >From: Confocal Microscopy List on behalf of Yevgeniy Romin >Sent: Sun 18/04/2010 1:49 AM >To: [hidden email] >Subject: Re: Mounting media for FRET experiments > > > >Thank you very much to everyone replying both on and off the list. Your samples would be the ideal choice, unfortunately in this case we are doing FRET on white blood cells, which are very round and refuse to attach and sit still, introducing motion artifacts into the FRET analysis. We are currently working on immobilizing them, but there is a fear of matrigel or a more viscous media introducing FRET measurement artifacts as well. Does anyone have experience with doing FRET on cells that do not attach or immobilizing them in a least invasive way? >________________________________________ >From: Confocal Microscopy List [[hidden email]] On Behalf Of Daniel Gitler [[hidden email]] >Sent: Saturday, April 17, 2010 5:22 AM >To: [hidden email] >Subject: Re: Mounting media for FRET experiments > >Dear John and Yevgeniy, > >My experience when measuring FRET from the same samples before and after fixation is that fixation itself will very much change the FRET efficiency value that you measure (Cerulean and Venus in HEK cells, PBS for both). Efficiency in the fixed samples was significantly lower. This should not come as any big surprise, since the cellular environment around the fluorophores will be very different. >The possible reasons for the changes in FRET efficiency are very many... >I also do most of my FRET imaging in live samples. > >Daniel > >----- Original Message ----- >From: Vitaly Boyko <[hidden email]> >Date: Friday, April 16, 2010 19:36 >Subject: Re: Mounting media for FRET experiments >To: [hidden email] > >> Dear John, >> Dear Yevgeniy, >> >> Yes, PBS or specialized low serum (1-2%) media is broadly used >> for live cell imaging under TIR illumination. >> >> I would like to bring your attention to two papers which are >> relevant to single molecule fluorscence and/or low light live >> cell imaging - I sent them separately to your e-mail addressess >> as the LISTSERVER rejected my message due to the attachments. I >> can also send the papers to anyone by request. >> >> 1. The Use of Protocatechuate Dioxygenase for Maintaining >> Anaerobic Conditions in Biochemical ExperimentsPravin V. Patil >> and David P. BallouDepartment of Biological Chemistry, >> University of Michigan, Ann Arbor, Michigan 48109-0606 >> Analytical Biochemistry 286, 187-192 (2000) >> 2. An Oxygen Scavenging System for Improvement of Dye Stability >> in Single-Molecule Fluorescence ExperimentsColin Echeverri´a >> Aitken,* R. Andrew Marshall,y and Joseph D. Puglisiz§ >> Biophysical Journal Volume 94 March 2008 1826-1835 >> >> If you have any questions, please contact me offline or phone. >> >> Also, the accuracy of the FRET eff. measurements could vary or >> be affected by the mounting media used - live cell FRET seems to >> be the best approach. >> >> Have a nice weekend, >> >> Vitaly >> 301-515-7833 >> >> >> >> >> >> ________________________________ >> From: John Oreopoulos <[hidden email]> >> To: [hidden email] >> Sent: Fri, April 16, 2010 12:01:32 PM >> Subject: Re: Mounting media for FRET experiments >> >> Just regular PBS solution. I have never tried a rigorous >> comparison of different media. >> >> John >> >> On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: >> >> > Thank you very much John, I found the paper you recommended >> and it is already proving useful. A question for you then - >> which media were you using for the live cells? What I mean is >> whether it was something like PBS or some specific media? >> > >> > --------------------------------------------------- >> > Yevgeniy Romin >> > >> > Digital Microscopist >> > Memorial Sloan-Kettering Cancer Center >> > Molecular Cytology Core Facility >> > 1275 York Ave. Box 333 >> > New York, NY 10065 >> > Tel.646-888-2186 >> > Fax. 646-422-0640 >> > --------------------------------------------------- >> > >> > -----Original Message----- >> > From: Confocal Microscopy List >> [mailto:[hidden email]] On Behalf Of John Oreopoulos >> > Sent: Friday, April 16, 2010 11:30 AM >> > To: [hidden email] >> > Subject: Re: Mounting media for FRET experiments >> > >> > I have not used fixed specimens for FRET imaging (I have used >> the sensitized emission method on living cells), but I came >> across this paper a while ago which might be of interest to you: >> > >> > Rodighiero, S., et al., Fixation, mounting and sealing with >> nail polish of cell specimens lead to incorrect FRET >> measurements using acceptor photobleaching. Cellular Physiology >> and Biochemistry, 2008. 21(5-6): p. 489-498. >> > >> > John Oreopoulos >> > >> > >> > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: >> > >> >> Dear list >> >> >> >> During a recent FRET experiment a question came up about what >> mounting media should be used on fixed samples during FRET >> experiments and whether any one media is better than other for >> this particular type of experiment. From what's published it >> seems that people use all kinds of different media. Does >> anybody here have any idea about any specific media that would >> be best/give least interference with FRET results? >> >> >> >> Thanks a lot to all of you in advance, >> >> >> >> --------------------------------------------------- >> >> Yevgeniy Romin >> >> >> >> Digital Microscopist >> >> Memorial Sloan-Kettering Cancer Center >> >> Molecular Cytology Core Facility >> >> 1275 York Ave. Box 333 >> >> New York, NY 10065 >> >> Tel.646-888-2186 >> >> Fax. 646-422-0640 >> >> --------------------------------------------------- >> >> >> >> >> >> >> >> >> =====================================================================>> >> >> Please note that this e-mail and any files transmitted >> with it may be >> >> privileged, confidential, and protected from disclosure under >> >> applicable law. If the reader of this message is not the >> intended>> recipient, or an employee or agent responsible for >> delivering this >> >> message to the intended recipient, you are hereby notified >> that any >> >> reading, dissemination, distribution, copying, or other >> use of this >> >> communication or any of its attachments is strictly >> prohibited. If >> >> you have received this communication in error, please >> notify the >> >> sender immediately by replying to this message and >> deleting this >> >> message, any attachments, and all copies and backups from your >> >> computer. >> >> >> >> >> >> >> >> > >Daniel Gitler, Ph.D. >Department of Physiology and Neurobiology >Faculty of Health Sciences >Ben Gurion University of the Negev >Beer-Sheva 84105 >Israel > >Tel: +972-8-6477345 >Cell: +972-54-2110100 >Fax: + 972-8-6477628 >http://web2.bgu.ac.il/physiology/faculty-members/daniel-gitler/ > >? |
|
Charu Tanwar |
|
|
In reply to this post by Romin, Yevgeniy/Sloan Kettering Institute
|
|
Janne Hyötylä |
|
|
This article might also provide further ideas for cell adhesion substrates:
Hwang, D.S., Sim, S.B. & Cha, H.J. Cell adhesion biomaterial based on mussel adhesive protein fused with RGD peptide. Biomaterials 28, 4039-46(2007). Cheers Janne On Tue, 20 Apr 2010 10:27:27 +0200, charu tanwar <[hidden email]> wrote: > I have done FRET on Candida cells (live cells, not fixed) which are also > round and small in size and refuse to attach onto the coverslip. I used > Poly Lysine coated coverslip to adhere them on to the surface of the > coverslip so that the movement is less. This you can do in your lab and > these are commercially provided also. > After final washing with PBS, i mounted my cells with antifade reagent > provided commercially ( it has some percentage of glycerol ). This has > worked for me very well, although i can not say about WBC's. > Best > Charu > > CHARU TANWAR > Imaging Specialist > Advanced Instrumentation Research Facility > Jawaharlal Nehru University > New Delhi 110067 > India. > > --- On Sat, 17/4/10, Yevgeniy Romin <[hidden email]> wrote: > > > From: Yevgeniy Romin <[hidden email]> > Subject: Re: Mounting media for FRET experiments > To: [hidden email] > Date: Saturday, 17 April, 2010, 9:19 PM > > > Thank you very much to everyone replying both on and off the list. Your > advice and the papers that were referred to were very helpful. FRET on > live samples would be the ideal choice, unfortunately in this case we > are doing FRET on white blood cells, which are very round and refuse to > attach and sit still, introducing motion artifacts into the FRET > analysis. We are currently working on immobilizing them, but there is a > fear of matrigel or a more viscous media introducing FRET measurement > artifacts as well. Does anyone have experience with doing FRET on cells > that do not attach or immobilizing them in a least invasive way? > ________________________________________ > From: Confocal Microscopy List [[hidden email]] On > Behalf Of Daniel Gitler [[hidden email]] > Sent: Saturday, April 17, 2010 5:22 AM > To: [hidden email] > Subject: Re: Mounting media for FRET experiments > > Dear John and Yevgeniy, > > My experience when measuring FRET from the same samples before and after > fixation is that fixation itself will very much change the FRET > efficiency value that you measure (Cerulean and Venus in HEK cells, PBS > for both). Efficiency in the fixed samples was significantly lower. This > should not come as any big surprise, since the cellular environment > around the fluorophores will be very different. > The possible reasons for the changes in FRET efficiency are very many... > I also do most of my FRET imaging in live samples. > > Daniel > > ----- Original Message ----- > From: Vitaly Boyko <[hidden email]> > Date: Friday, April 16, 2010 19:36 > Subject: Re: Mounting media for FRET experiments > To: [hidden email] > >> Dear John, >> Dear Yevgeniy, >> >> Yes, PBS or specialized low serum (1-2%) media is broadly used >> for live cell imaging under TIR illumination. >> >> I would like to bring your attention to two papers which are >> relevant to single molecule fluorscence and/or low light live >> cell imaging - I sent them separately to your e-mail addressess >> as the LISTSERVER rejected my message due to the attachments. I >> can also send the papers to anyone by request. >> >> 1. The Use of Protocatechuate Dioxygenase for Maintaining >> Anaerobic Conditions in Biochemical ExperimentsPravin V. Patil >> and David P. BallouDepartment of Biological Chemistry, >> University of Michigan, Ann Arbor, Michigan 48109-0606 >> Analytical Biochemistry 286, 187–192 (2000) >> 2. An Oxygen Scavenging System for Improvement of Dye Stability >> in Single-Molecule Fluorescence ExperimentsColin Echeverrı´a >> Aitken,* R. Andrew Marshall,y and Joseph D. Puglisiz§ >> Biophysical Journal Volume 94 March 2008 1826–1835 >> >> If you have any questions, please contact me offline or phone. >> >> Also, the accuracy of the FRET eff. measurements could vary or >> be affected by the mounting media used - live cell FRET seems to >> be the best approach. >> >> Have a nice weekend, >> >> Vitaly >> 301-515-7833 >> >> >> >> >> >> ________________________________ >> From: John Oreopoulos <[hidden email]> >> To: [hidden email] >> Sent: Fri, April 16, 2010 12:01:32 PM >> Subject: Re: Mounting media for FRET experiments >> >> Just regular PBS solution. I have never tried a rigorous >> comparison of different media. >> >> John >> >> On 2010-04-16, at 11:54 AM, Yevgeniy Romin wrote: >> >> > Thank you very much John, I found the paper you recommended >> and it is already proving useful. A question for you then - >> which media were you using for the live cells? What I mean is >> whether it was something like PBS or some specific media? >> > >> > --------------------------------------------------- >> > Yevgeniy Romin >> > >> > Digital Microscopist >> > Memorial Sloan-Kettering Cancer Center >> > Molecular Cytology Core Facility >> > 1275 York Ave. Box 333 >> > New York, NY 10065 >> > Tel.646-888-2186 >> > Fax. 646-422-0640 >> > --------------------------------------------------- >> > >> > -----Original Message----- >> > From: Confocal Microscopy List >> [mailto:[hidden email]] On Behalf Of John Oreopoulos >> > Sent: Friday, April 16, 2010 11:30 AM >> > To: [hidden email] >> > Subject: Re: Mounting media for FRET experiments >> > >> > I have not used fixed specimens for FRET imaging (I have used >> the sensitized emission method on living cells), but I came >> across this paper a while ago which might be of interest to you: >> > >> > Rodighiero, S., et al., Fixation, mounting and sealing with >> nail polish of cell specimens lead to incorrect FRET >> measurements using acceptor photobleaching. Cellular Physiology >> and Biochemistry, 2008. 21(5-6): p. 489-498. >> > >> > John Oreopoulos >> > >> > >> > On 2010-04-16, at 11:20 AM, Yevgeniy Romin wrote: >> > >> >> Dear list >> >> >> >> During a recent FRET experiment a question came up about what >> mounting media should be used on fixed samples during FRET >> experiments and whether any one media is better than other for >> this particular type of experiment. From what's published it >> seems that people use all kinds of different media. Does >> anybody here have any idea about any specific media that would >> be best/give least interference with FRET results? >> >> >> >> Thanks a lot to all of you in advance, >> >> >> >> --------------------------------------------------- >> >> Yevgeniy Romin >> >> >> >> Digital Microscopist >> >> Memorial Sloan-Kettering Cancer Center >> >> Molecular Cytology Core Facility >> >> 1275 York Ave. Box 333 >> >> New York, NY 10065 >> >> Tel.646-888-2186 >> >> Fax. 646-422-0640 >> >> --------------------------------------------------- >> >> >> >> >> >> >> >> >> =====================================================================>> >> >> Please note that this e-mail and any files transmitted >> with it may be >> >> privileged, confidential, and protected from disclosure under >> >> applicable law. If the reader of this message is not the >> intended>> recipient, or an employee or agent responsible for >> delivering this >> >> message to the intended recipient, you are hereby notified >> that any >> >> reading, dissemination, distribution, copying, or other >> use of this >> >> communication or any of its attachments is strictly >> prohibited. If >> >> you have received this communication in error, please >> notify the >> >> sender immediately by replying to this message and >> deleting this >> >> message, any attachments, and all copies and backups from your >> >> computer. >> >> >> >> >> >> >> >> > > Daniel Gitler, Ph.D. > Department of Physiology and Neurobiology > Faculty of Health Sciences > Ben Gurion University of the Negev > Beer-Sheva 84105 > Israel > > Tel: +972-8-6477345 > Cell: +972-54-2110100 > Fax: + 972-8-6477628 > http://web2.bgu.ac.il/physiology/faculty-members/daniel-gitler/ > > > > -- Janne Hyötylä Biozentrum and the Swiss Nanoscience Institute University of Basel Klingelbergstr. 50/70 CH-4056 Basel Switzerland tel: +41 61 267 2082 e-mail: [hidden email] |