NIS elements software

classic Classic list List threaded Threaded
♦
♦
10 messages Options
Valeria Berno Valeria Berno
Reply | Threaded
Open this post in threaded view
♦
♦
|

NIS elements software

Hi

did any of you have experience with NIS element software from Nikon?

Any complain?

quick check in order to take a decision for a widefield microscope....

thanks in advance

Valeria

Berno Valeria, PhD

EMBL- Mouse Biology Unit
Campus A.  Buzzati-Traverso
Via Ramarini,  32
00015, Monterotondo Scalo  (RM)
Italy

Tel:  +39 06 90091243
Fax: +39 06 90091406
email: [hidden email]
www.embl.it
Marco Dal Maschio Marco Dal Maschio
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

Dear Valeria, 
we currently work with NIS controlling a motorized widefield TiE microscope and a hamamatsu camera 
for acquisition of Ca signal but also for multicolor cell immunostaining that you can perform in full automatic mode
with a macro. We tested the same software with the A1 confocal (same microscope with XY sample holder traslator): very reliable.
Still missing a relaible integration with UNIBLITZ shutter.

No complain at all!

If you need more details do not esitate to contact me
My best
Marco




On Wed, Nov 18, 2009 at 3:36 PM, Valeria Berno <[hidden email]> wrote:
Hi

did any of you have experience with NIS element software from Nikon?

Any complain?

quick check in order to take a decision for a widefield microscope....

thanks in advance

Valeria

Berno Valeria, PhD

EMBL- Mouse Biology Unit
Campus A.  Buzzati-Traverso
Via Ramarini,  32
00015, Monterotondo Scalo  (RM)
Italy

Tel:  +39 06 90091243
Fax: +39 06 90091406
email: [hidden email]
www.embl.it



--
Marco Dal Maschio      [hidden email]
Home: via Garzetta 23
         30015 Chioggia VE
          mob:  0039-3488720013

       


Evangelos Gatzogiannis Evangelos Gatzogiannis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

In reply to this post by Valeria Berno
   NIS Elements is very straightforward and reliable software from my
experience, what's nice is that Nikon has the basic version ready for
all their microscopes and depending on what kind of widefield or
confocal microscope you have, more features are just added on depending
on whether you have a motorized focus, or a raster-scanning stage, or
types cameras. I love the handy little scrolling "magnifying glass" on
NIS elements - it's kind of cute.   I'm a big fan of Metamorph and of
NIS Elements, even though I only have Elements on two very basic
microscopes, they have given me and my 75 graduate student/post-doc
users no problems at all - to be clear, I have no commercial interest at
all.

Best,
Evangelos Gatzogiannis
Advanced Biological Imaging Scientist
Harvard University Center for Nanoscale Systems
11 Oxford Street
Cambridge, MA 02138



Valeria Berno wrote:

> Hi
>
> did any of you have experience with NIS element software from Nikon?
>
> Any complain?
>
> quick check in order to take a decision for a widefield microscope....
>
> thanks in advance
>
> Valeria
>
> Berno Valeria, PhD
>
> EMBL- Mouse Biology Unit
> Campus A.  Buzzati-Traverso
> Via Ramarini,  32
> 00015, Monterotondo Scalo  (RM)
> Italy
>
> Tel:  +39 06 90091243
> Fax: +39 06 90091406
> email: [hidden email]
> www.embl.it
Tim Feinstein-2 Tim Feinstein-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

In reply to this post by Valeria Berno
Dear Valeria,

My lab has used Elements to control a Ti-E for widefield and confocal (A1) imaging for about a year, and so far our reaction is mixed.  The software feels well-designed and intuitive, and modularity made it easy to find a package that suits our needs.  However, bug-checking has been an issue.  The edition of Elements-AR that we received in January, one of the first written for Windows Vista, had glitches that sometimes forced us to find work-arounds to get ordinary tasks done.  One cause may be that Vista handles peripherals particuarly badly; we have an unusual number of scope-associated peripherals installed.  

With a very few exceptions, Nikon has fixed our issues with updates.  The latest edition for Vista should therefore work fine.  The XP edition that came with our demo scope also worked flawlessly.  I cannot comment on Elements in Windows 7, but based on experience we plan to let a few rounds of bug fixing proceed before we upgrade.  

None of these software issues change our impression that the hardware, especially the perfect focus system, is top of the class.  

All the best,


Tim

Tim Feinstein, PhD
Postdoctoral Associate
University of Pittsburgh Dept. of Pharmacology

On Nov 18, 2009, at 9:36 AM, Valeria Berno wrote:

> Hi
>
> did any of you have experience with NIS element software from Nikon?
>
> Any complain?
>
> quick check in order to take a decision for a widefield microscope....
>
> thanks in advance
>
> Valeria
>
> Berno Valeria, PhD
>
> EMBL- Mouse Biology Unit
> Campus A.  Buzzati-Traverso
> Via Ramarini,  32
> 00015, Monterotondo Scalo  (RM)
> Italy
>
> Tel:  +39 06 90091243
> Fax: +39 06 90091406
> email: [hidden email]
> www.embl.it
Evangelos Gatzogiannis Evangelos Gatzogiannis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

  I agree with Tim, the Nikon Perfect Focus is by far the best in class
and in my opinion a must if you are doing long - time (several days)
live cell observation or TIRF microscopy - and it works relatively
seamlessly with Elements. Windows Vista, however, I'll reserve comments
on that in private and try to run everything in lab off of XP as that
was the best version of Windows yet.

Best,
Evangelos
Neeraj Gohad-2 Neeraj Gohad-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

In reply to this post by Marco Dal Maschio

Hi Valeria,

 

We too have NIS elements on our motorized Nikon TiE microscope as well as on the Nikon AZ-100 and UDM microscope. We also have  an offline post-processing workstation. As Evangelos suggested you can add modules and features as per your need. On the microscope control side, setting up multidimensional experiments is really easy, I routinely  conduct 3-4 day time lapse experiments on taking images every couple of minutes on different stage points in different channels. Setting up different optical configuration such as objectives and filter cubes and assigning microscope and camera setting to theses configurations is also straightforward and intuitive.  For analysis, you can trim down datasets and easily extract specific time points, focal planes, channels etc. There is wide variety of annotation and measurement tools. One of the features which comes handy  is merging of different channels, you  just have to  drag and drop to get a multichannel image.  In all I think it’s a good software for both  control and analysis, but there some room for improvement with the free image viewer, it should do lot more than what it does.

 

As always, no commercial interests.

 

Best,

 

Neeraj.

 

 

Neeraj V. Gohad, Ph.D.

Postdoctoral Fellow

Okeanos Research Group

Department of Biological Sciences

132 Long Hall

Clemson University

Clemson,SC-29634

Phone: 864-656-3597

Fax: 864-656-0435

 

Please note my new email address: [hidden email]

 

 

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Marco Dal Maschio
Sent: Wednesday, November 18, 2009 10:16 AM
To: [hidden email]
Subject: Re: NIS elements software

 

Dear Valeria, 

we currently work with NIS controlling a motorized widefield TiE microscope and a hamamatsu camera 

for acquisition of Ca signal but also for multicolor cell immunostaining that you can perform in full automatic mode

with a macro. We tested the same software with the A1 confocal (same microscope with XY sample holder traslator): very reliable.

Still missing a relaible integration with UNIBLITZ shutter.

 

No complain at all!

 

If you need more details do not esitate to contact me

My best

Marco

 

 

 

On Wed, Nov 18, 2009 at 3:36 PM, Valeria Berno <[hidden email]> wrote:

Hi

did any of you have experience with NIS element software from Nikon?

Any complain?

quick check in order to take a decision for a widefield microscope....

thanks in advance

Valeria

Berno Valeria, PhD

EMBL- Mouse Biology Unit
Campus A.  Buzzati-Traverso
Via Ramarini,  32
00015, Monterotondo Scalo  (RM)
Italy

Tel:  +39 06 90091243
Fax: +39 06 90091406
email: [hidden email]
www.embl.it




--
Marco Dal Maschio      [hidden email]
Home: via Garzetta 23
         30015 Chioggia VE
          mob:  0039-3488720013

       

LANCE RODENKIRCH LANCE RODENKIRCH
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

In reply to this post by Tim Feinstein-2
Thanks Tim--I have had similar experiences with an A1R.

An item I hope to see in the Windows 7 release of Elements, would be individual logins for each laboratory.  A key component if you have a large userbase.  best,  Lance


Lance Rodenkirch
W.M Keck Laboratory for Biological Imaging
University of Wisconsin School of Medicine
1300 University Ave Rm 172 MSC
Madison, WI  53706
http://www.keck.bioimaging.wisc.edu/
Phone: 608.265.5651
Fax: 608.265.5536

----- Original Message -----
From: Tim Feinstein <[hidden email]>
Date: Wednesday, November 18, 2009 11:06 am
Subject: Re: NIS elements software
To: [hidden email]


> Dear Valeria,
>  
>  My lab has used Elements to control a Ti-E for widefield and confocal
> (A1) imaging for about a year, and so far our reaction is mixed.  The
> software feels well-designed and intuitive, and modularity made it
> easy to find a package that suits our needs.  However, bug-checking
> has been an issue.  The edition of Elements-AR that we received in
> January, one of the first written for Windows Vista, had glitches that
> sometimes forced us to find work-arounds to get ordinary tasks done.  
> One cause may be that Vista handles peripherals particuarly badly; we
> have an unusual number of scope-associated peripherals installed.  
>  
>  With a very few exceptions, Nikon has fixed our issues with updates.  
> The latest edition for Vista should therefore work fine.  The XP
> edition that came with our demo scope also worked flawlessly.  I
> cannot comment on Elements in Windows 7, but based on experience we
> plan to let a few rounds of bug fixing proceed before we upgrade.  
>  
>  None of these software issues change our impression that the
> hardware, especially the perfect focus system, is top of the class.  
>  
>  All the best,
>  
>  
>  Tim
>  
>  Tim Feinstein, PhD
>  Postdoctoral Associate
>  University of Pittsburgh Dept. of Pharmacology
>  
>  On Nov 18, 2009, at 9:36 AM, Valeria Berno wrote:
>  
>  > Hi
>  >
>  > did any of you have experience with NIS element software from Nikon?
>  >
>  > Any complain?
>  >
>  > quick check in order to take a decision for a widefield microscope....
>  >
>  > thanks in advance
>  >
>  > Valeria
>  >
>  > Berno Valeria, PhD
>  >
>  > EMBL- Mouse Biology Unit
>  > Campus A.  Buzzati-Traverso
>  > Via Ramarini,  32
>  > 00015, Monterotondo Scalo  (RM)
>  > Italy
>  >
>  > Tel:  +39 06 90091243
>  > Fax: +39 06 90091406
>  > email: [hidden email]
>  > www.embl.it
>  
Neeraj Gohad-2 Neeraj Gohad-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

In reply to this post by Neeraj Gohad-2

I agree, Nikons Perfect Focus works really well and I think it’s must for any long term time lapse experiments. I also forgot to mention that all are systems are Windows XP, only the offline post processing workstation is Windows Vista.

 

Neeraj.

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Neeraj Gohad
Sent: Wednesday, November 18, 2009 1:13 PM
To: [hidden email]
Subject: Re: NIS elements software

 

Hi Valeria,

 

We too have NIS elements on our motorized Nikon TiE microscope as well as on the Nikon AZ-100 and UDM microscope. We also have  an offline post-processing workstation. As Evangelos suggested you can add modules and features as per your need. On the microscope control side, setting up multidimensional experiments is really easy, I routinely  conduct 3-4 day time lapse experiments on taking images every couple of minutes on different stage points in different channels. Setting up different optical configuration such as objectives and filter cubes and assigning microscope and camera setting to theses configurations is also straightforward and intuitive.  For analysis, you can trim down datasets and easily extract specific time points, focal planes, channels etc. There is wide variety of annotation and measurement tools. One of the features which comes handy  is merging of different channels, you  just have to  drag and drop to get a multichannel image.  In all I think it’s a good software for both  control and analysis, but there some room for improvement with the free image viewer, it should do lot more than what it does.

 

As always, no commercial interests.

 

Best,

 

Neeraj.

 

 

Neeraj V. Gohad, Ph.D.

Postdoctoral Fellow

Okeanos Research Group

Department of Biological Sciences

132 Long Hall

Clemson University

Clemson,SC-29634

Phone: 864-656-3597

Fax: 864-656-0435

 

Please note my new email address: [hidden email]

 

 

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Marco Dal Maschio
Sent: Wednesday, November 18, 2009 10:16 AM
To: [hidden email]
Subject: Re: NIS elements software

 

Dear Valeria, 

we currently work with NIS controlling a motorized widefield TiE microscope and a hamamatsu camera 

for acquisition of Ca signal but also for multicolor cell immunostaining that you can perform in full automatic mode

with a macro. We tested the same software with the A1 confocal (same microscope with XY sample holder traslator): very reliable.

Still missing a relaible integration with UNIBLITZ shutter.

 

No complain at all!

 

If you need more details do not esitate to contact me

My best

Marco

 

 

 

On Wed, Nov 18, 2009 at 3:36 PM, Valeria Berno <[hidden email]> wrote:

Hi

did any of you have experience with NIS element software from Nikon?

Any complain?

quick check in order to take a decision for a widefield microscope....

thanks in advance

Valeria

Berno Valeria, PhD

EMBL- Mouse Biology Unit
Campus A.  Buzzati-Traverso
Via Ramarini,  32
00015, Monterotondo Scalo  (RM)
Italy

Tel:  +39 06 90091243
Fax: +39 06 90091406
email: [hidden email]
www.embl.it




--
Marco Dal Maschio      [hidden email]
Home: via Garzetta 23
         30015 Chioggia VE
          mob:  0039-3488720013

       

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

We have NIS elements for Windows XP and it works fine.  I would be
suspicious of the version for Vista; maybe the Windows 7 version will
be more solid.  Our university has avoided vista entirely and still
only offers windows XP as its license.  I expect we will skip Vista
entirely and eventually switch to Windows 7.
We don't have Perfect Focus yet but we are thinking of upgrading based
on comments from various users (both on this list and others).  It
sounds like a solid performer.

Craig


On Wed, Nov 18, 2009 at 12:25 PM, Neeraj Gohad <[hidden email]> wrote:

> I agree, Nikons Perfect Focus works really well and I think it’s must for
> any long term time lapse experiments. I also forgot to mention that all are
> systems are Windows XP, only the offline post processing workstation is
> Windows Vista.
>
>
>
> Neeraj.
>
>
>
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Neeraj Gohad
> Sent: Wednesday, November 18, 2009 1:13 PM
> To: [hidden email]
> Subject: Re: NIS elements software
>
>
>
> Hi Valeria,
>
>
>
> We too have NIS elements on our motorized Nikon TiE microscope as well as on
> the Nikon AZ-100 and UDM microscope. We also have  an offline
> post-processing workstation. As Evangelos suggested you can add modules and
> features as per your need. On the microscope control side, setting up
> multidimensional experiments is really easy, I routinely  conduct 3-4 day
> time lapse experiments on taking images every couple of minutes on different
> stage points in different channels. Setting up different optical
> configuration such as objectives and filter cubes and assigning microscope
> and camera setting to theses configurations is also straightforward and
> intuitive.  For analysis, you can trim down datasets and easily extract
> specific time points, focal planes, channels etc. There is wide variety of
> annotation and measurement tools. One of the features which comes handy  is
> merging of different channels, you  just have to  drag and drop to get a
> multichannel image.  In all I think it’s a good software for both  control
> and analysis, but there some room for improvement with the free image
> viewer, it should do lot more than what it does.
>
>
>
> As always, no commercial interests.
>
>
>
> Best,
>
>
>
> Neeraj.
>
>
>
>
>
> Neeraj V. Gohad, Ph.D.
>
> Postdoctoral Fellow
>
> Okeanos Research Group
>
> Department of Biological Sciences
>
> 132 Long Hall
>
> Clemson University
>
> Clemson,SC-29634
>
> Phone: 864-656-3597
>
> Fax: 864-656-0435
>
>
>
> Please note my new email address: [hidden email]
>
>
>
>
>
>
>
>
>
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Marco Dal Maschio
> Sent: Wednesday, November 18, 2009 10:16 AM
> To: [hidden email]
> Subject: Re: NIS elements software
>
>
>
> Dear Valeria,
>
> we currently work with NIS controlling a motorized widefield TiE microscope
> and a hamamatsu camera
>
> for acquisition of Ca signal but also for multicolor cell immunostaining
> that you can perform in full automatic mode
>
> with a macro. We tested the same software with the A1 confocal (same
> microscope with XY sample holder traslator): very reliable.
>
> Still missing a relaible integration with UNIBLITZ shutter.
>
>
>
> No complain at all!
>
>
>
> If you need more details do not esitate to contact me
>
> My best
>
> Marco
>
>
>
>
>
>
>
> On Wed, Nov 18, 2009 at 3:36 PM, Valeria Berno <[hidden email]>
> wrote:
>
> Hi
>
> did any of you have experience with NIS element software from Nikon?
>
> Any complain?
>
> quick check in order to take a decision for a widefield microscope....
>
> thanks in advance
>
> Valeria
>
> Berno Valeria, PhD
>
> EMBL- Mouse Biology Unit
> Campus A.  Buzzati-Traverso
> Via Ramarini,  32
> 00015, Monterotondo Scalo  (RM)
> Italy
>
> Tel:  +39 06 90091243
> Fax: +39 06 90091406
> email: [hidden email]
> www.embl.it
>
>
> --
> Marco Dal Maschio      [hidden email]
> Home: via Garzetta 23
>          30015 Chioggia VE
>           mob:  0039-3488720013
>
>
Keith Morris Keith Morris
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIS elements software

In reply to this post by Tim Feinstein-2
Hi all,

We use NISElements Basic Research with the optional EDF [extended depth of
focus] module. It's naturally pretty ideal for a Nikon microscope, although
we use it with a manual inverted Nikon Eclipse TE2000U so it only controls
the two B&W/Colour cameras, not the objectives nosepiece/filter wheel/focus.
The NISElements free version is very very limited [our Basic Research
version is actually the expensive £2k+ fully featured one]. Software modules
can be added, e.g. one that allows split screens between stored and the live
images. The basic research package has a lot of image analysis features and
so is fairly complicated [but you can ignore all that]. Capturing
fluorescence via a B&W camera is also a bit convoluted with our old v2.3
[c2008] if you want colour LUTs applied [so many stick with IPLabs that’s on
the same PC].

NISElements is great for colour bright-field and works with many cameras
[well our Nikon and Hamamatsu CCD cameras anyway]. We have NISElements v2.3
which is a 1.5 year old version. Nikons support seems similar to Zeiss and
Olympus around here [prompt replies to emails etc..]. We don't have a
maintenance contract. The £600 EDF function works well but it's only for
very thick samples, e.g. things like embryos, tissue slices, ladybirds etc..
where's there's a lot of out of focus info - the focused bits of a widefield
Z stack get all compiled into a single 2D image [no 3D], fun but rarely used
with flat cells and thin tissue sections that don't really have much out of
focus info. I don't use NISElements image analysis stuff much as we have
MetaMorph v7.5  [2D quantification] and Volocity v4.2 [3D deconvolution and
4D time lapse]. Our NISElements does crash on very rare occasions, but it
doesn't affect users much, and we are investigating the Windows XP install
rather than the software. It's dongle controlled [unless the limited
freebie] so it's one install per licence. Generally very happy with
NISElements Basic Research, and its image handling [capturing, live zoom,
image database, saving and loading] seems pretty good.

Regards

Keith

http://www.well.ox.ac.uk/nikon-microscope

---------------------------------------------------------------------------
Dr Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory 00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt Drive,
Oxford  OX3 7BN,
United Kingdom.

Telephone:  +44 (0)1865 287568
Email:  [hidden email]
Web-pages: http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Tim Feinstein
Sent: 18 November 2009 17:06
To: [hidden email]
Subject: Re: NIS elements software

Dear Valeria,

My lab has used Elements to control a Ti-E for widefield and confocal (A1)
imaging for about a year, and so far our reaction is mixed.  The software
feels well-designed and intuitive, and modularity made it easy to find a
package that suits our needs.  However, bug-checking has been an issue.  The
edition of Elements-AR that we received in January, one of the first written
for Windows Vista, had glitches that sometimes forced us to find
work-arounds to get ordinary tasks done.  One cause may be that Vista
handles peripherals particuarly badly; we have an unusual number of
scope-associated peripherals installed.  

With a very few exceptions, Nikon has fixed our issues with updates.  The
latest edition for Vista should therefore work fine.  The XP edition that
came with our demo scope also worked flawlessly.  I cannot comment on
Elements in Windows 7, but based on experience we plan to let a few rounds
of bug fixing proceed before we upgrade.  

None of these software issues change our impression that the hardware,
especially the perfect focus system, is top of the class.  

All the best,


Tim

Tim Feinstein, PhD
Postdoctoral Associate
University of Pittsburgh Dept. of Pharmacology

On Nov 18, 2009, at 9:36 AM, Valeria Berno wrote:

> Hi
>
> did any of you have experience with NIS element software from Nikon?
>
> Any complain?
>
> quick check in order to take a decision for a widefield microscope....
>