Hello,
We have a Zeiss axioObserver and we are trying to figure out whether we can
attach a Piezo nanopositioner that is currently installed on a Nikon
TE2000-s scope. The piezo is old and may not have the quicklock adaptor.  

I was wondering if someone knew whether a Nikon to Zeiss adaptor is readily
available at someplace for this purpose?


Thanks.

Sripad
Chris Tully Chris Tully
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nikon to Zeiss thread adaptor

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Sripad,

Just for the purposes of full disclosure, I have previously (1998-2001)
worked for a Nikon Microscopes dealer.  I now work for myself as an Image
Analysis and Microscopy Consultant.

The Nikon TE2000-s microscope is part of the Eclipse series and uses the
most recent revision of Nikon objectives.  Nikon Eclipse microscopes use
the "new" CFI-60 objectives which have M25 threads, while Zeiss still uses
the RMS (Royal Microscopy Society) standard. The M25 designation refers to
the diameter of the thread in mm.  The RMS specification calls of a 0.8
inch thread diameter (
http://onlinelibrary.wiley.com/doi/10.1111/j.1365-2818.1896.tb00905.x/abstract),
which is significantly smaller.

While it is possible to buy (I think) or make an adapter that will fit an
RMS threaded objective on to a scope that uses M25 threads, I doubt that
the reverse is true.  An adapter that mounts a CFI-60 objective (or
compatible mount) on an RMS scope would necessarily move the object being
adapted at least 3-4 mm further away from the nose piece of the microscope,
resulting in:
1. Unknown magnification (although this can be easily measured).
2. Parfocality issues.
3. Danger of banging the adapted objective into the slide after using a
standard objective.

Also keep in mind the the CFI-60 objectives are not just bigger around than
the RMS objectives they tend to be longer as well.

Chris Tully
Microscopy and Image Analysis Consultant
[hidden email]
www.ImageIncyte.com
[hidden email]
240-475-9753 (c)

[image: View my profile on LinkedIn]<http://www.linkedin.com/in/christully/>


On Mon, Jul 8, 2013 at 2:22 PM, Sripad Ram <[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello,
> We have a Zeiss axioObserver and we are trying to figure out whether we can
> attach a Piezo nanopositioner that is currently installed on a Nikon
> TE2000-s scope. The piezo is old and may not have the quicklock adaptor.
>
> I was wondering if someone knew whether a Nikon to Zeiss adaptor is readily
> available at someplace for this purpose?
>
>
> Thanks.
>
> Sripad
>
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nikon to Zeiss thread adaptor

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Zeiss uses the M27 thread in their new microscope stands. This is significantly larger than the old Zeiss mount, and a bit larger than Nikon. Zeiss makes adapters from the old to the new M27 mount, so it should be possible to get similar adapters from for Nikon objectives on the new Zeiss scopes. In fact, I think Zeiss even has adapters to mount the new larger objectives on the older scopes. However, all caveats mentioned by Chris remain valid. In particular, Nikon objectives are huge compared to Zeiss (Nikon 60x/1.4 Plan Apo is about 2 longer than the equivalent Zeiss.  

Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109

http://www.fhcrc.org/en.html

==



On Jul 8, 2013, at 11:47 AM, Chris Tully wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Sripad,
>
> Just for the purposes of full disclosure, I have previously (1998-2001)
> worked for a Nikon Microscopes dealer.  I now work for myself as an Image
> Analysis and Microscopy Consultant.
>
> The Nikon TE2000-s microscope is part of the Eclipse series and uses the
> most recent revision of Nikon objectives.  Nikon Eclipse microscopes use
> the "new" CFI-60 objectives which have M25 threads, while Zeiss still uses
> the RMS (Royal Microscopy Society) standard. The M25 designation refers to
> the diameter of the thread in mm.  The RMS specification calls of a 0.8
> inch thread diameter (
> http://onlinelibrary.wiley.com/doi/10.1111/j.1365-2818.1896.tb00905.x/abstract),
> which is significantly smaller.
>
> While it is possible to buy (I think) or make an adapter that will fit an
> RMS threaded objective on to a scope that uses M25 threads, I doubt that
> the reverse is true.  An adapter that mounts a CFI-60 objective (or
> compatible mount) on an RMS scope would necessarily move the object being
> adapted at least 3-4 mm further away from the nose piece of the microscope,
> resulting in:
> 1. Unknown magnification (although this can be easily measured).
> 2. Parfocality issues.
> 3. Danger of banging the adapted objective into the slide after using a
> standard objective.
>
> Also keep in mind the the CFI-60 objectives are not just bigger around than
> the RMS objectives they tend to be longer as well.
>
> Chris Tully
> Microscopy and Image Analysis Consultant
> [hidden email]
> www.ImageIncyte.com
> [hidden email]
> 240-475-9753 (c)
>
> [image: View my profile on LinkedIn]<http://www.linkedin.com/in/christully/>
>
>
> On Mon, Jul 8, 2013 at 2:22 PM, Sripad Ram <[hidden email]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello,
>> We have a Zeiss axioObserver and we are trying to figure out whether we can
>> attach a Piezo nanopositioner that is currently installed on a Nikon
>> TE2000-s scope. The piezo is old and may not have the quicklock adaptor.
>>
>> I was wondering if someone knew whether a Nikon to Zeiss adaptor is readily
>> available at someplace for this purpose?
>>
>>
>> Thanks.
>>
>> Sripad
>>
Jeff Spector Jeff Spector
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nikon to Zeiss thread adaptor

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

If you are just looking for a threaded adapter check out Thorlabs (no
commercial interest)

http://www.thorlabs.com/newgrouppage9.cfm?objectgroup_id=1524

but all the previous caveats still remain valid...

Jeff Spector
NIH-NINDS
Bethesda,MD


On Mon, Jul 8, 2013 at 4:35 PM, Julio Vazquez <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Zeiss uses the M27 thread in their new microscope stands. This is
> significantly larger than the old Zeiss mount, and a bit larger than Nikon.
> Zeiss makes adapters from the old to the new M27 mount, so it should be
> possible to get similar adapters from for Nikon objectives on the new Zeiss
> scopes. In fact, I think Zeiss even has adapters to mount the new larger
> objectives on the older scopes. However, all caveats mentioned by Chris
> remain valid. In particular, Nikon objectives are huge compared to Zeiss
> (Nikon 60x/1.4 Plan Apo is about 2 longer than the equivalent Zeiss.
>
> Julio Vazquez
> Fred Hutchinson Cancer Research Center
> Seattle, WA 98109
>
> http://www.fhcrc.org/en.html
>
> ==
>
>
>
> On Jul 8, 2013, at 11:47 AM, Chris Tully wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Sripad,
> >
> > Just for the purposes of full disclosure, I have previously (1998-2001)
> > worked for a Nikon Microscopes dealer.  I now work for myself as an Image
> > Analysis and Microscopy Consultant.
> >
> > The Nikon TE2000-s microscope is part of the Eclipse series and uses the
> > most recent revision of Nikon objectives.  Nikon Eclipse microscopes use
> > the "new" CFI-60 objectives which have M25 threads, while Zeiss still
> uses
> > the RMS (Royal Microscopy Society) standard. The M25 designation refers
> to
> > the diameter of the thread in mm.  The RMS specification calls of a 0.8
> > inch thread diameter (
> >
> http://onlinelibrary.wiley.com/doi/10.1111/j.1365-2818.1896.tb00905.x/abstract
> ),
> > which is significantly smaller.
> >
> > While it is possible to buy (I think) or make an adapter that will fit an
> > RMS threaded objective on to a scope that uses M25 threads, I doubt that
> > the reverse is true.  An adapter that mounts a CFI-60 objective (or
> > compatible mount) on an RMS scope would necessarily move the object being
> > adapted at least 3-4 mm further away from the nose piece of the
> microscope,
> > resulting in:
> > 1. Unknown magnification (although this can be easily measured).
> > 2. Parfocality issues.
> > 3. Danger of banging the adapted objective into the slide after using a
> > standard objective.
> >
> > Also keep in mind the the CFI-60 objectives are not just bigger around
> than
> > the RMS objectives they tend to be longer as well.
> >
> > Chris Tully
> > Microscopy and Image Analysis Consultant
> > [hidden email]
> > www.ImageIncyte.com
> > [hidden email]
> > 240-475-9753 (c)
> >
> > [image: View my profile on LinkedIn]<
> http://www.linkedin.com/in/christully/>
> >
> >
> > On Mon, Jul 8, 2013 at 2:22 PM, Sripad Ram <
> [hidden email]>wrote:
> >
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> *****
> >>
> >> Hello,
> >> We have a Zeiss axioObserver and we are trying to figure out whether we
> can
> >> attach a Piezo nanopositioner that is currently installed on a Nikon
> >> TE2000-s scope. The piezo is old and may not have the quicklock adaptor.
> >>
> >> I was wondering if someone knew whether a Nikon to Zeiss adaptor is
> readily
> >> available at someplace for this purpose?
> >>
> >>
> >> Thanks.
> >>
> >> Sripad
> >>
>
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nikon to Zeiss thread adaptor

In reply to this post by Julio Vazquez
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Sorry, I read the post a bit too quickly... Are you planning to mount the Nikon piezo with a Nikon objective on a Zeiss scope? because of the size of Nikon objectives, maybe an older Zeiss mount would work better...i.e. find adapter to mount Nikon piezo on Zeiss scope, and then adapter to mount old Zeiss objective on Nikon piezo... these should both be possible.

Thor Labs may have such adapters...

http://www.thorlabs.us/newgrouppage9.cfm?objectgroup_id=1524

Julio


On Jul 8, 2013, at 1:35 PM, Julio Vazquez wrote:

> Zeiss uses the M27 thread in their new microscope stands. This is significantly larger than the old Zeiss mount, and a bit larger than Nikon. Zeiss makes adapters from the old to the new M27 mount, so it should be possible to get similar adapters from for Nikon objectives on the new Zeiss scopes. In fact, I think Zeiss even has adapters to mount the new larger objectives on the older scopes. However, all caveats mentioned by Chris remain valid. In particular, Nikon objectives are huge compared to Zeiss (Nikon 60x/1.4 Plan Apo is about 2 longer than the equivalent Zeiss.  
>
> Julio Vazquez
> Fred Hutchinson Cancer Research Center
> Seattle, WA 98109
>
> http://www.fhcrc.org/en.html
Sripad Ram Sripad Ram
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nikon to Zeiss thread adaptor

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello Julio,
Yes I am trying to mount a Nikon piezo on a Zeiss scope and then put an old
Zeiss objective on the Nikon piezo.

Sripad


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Julio Vazquez
Sent: Monday, July 08, 2013 3:43 PM
To: [hidden email]
Subject: Re: Nikon to Zeiss thread adaptor

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Sorry, I read the post a bit too quickly... Are you planning to mount the
Nikon piezo with a Nikon objective on a Zeiss scope? because of the size of
Nikon objectives, maybe an older Zeiss mount would work better...i.e. find
adapter to mount Nikon piezo on Zeiss scope, and then adapter to mount old
Zeiss objective on Nikon piezo... these should both be possible.

Thor Labs may have such adapters...

http://www.thorlabs.us/newgrouppage9.cfm?objectgroup_id=1524

Julio


On Jul 8, 2013, at 1:35 PM, Julio Vazquez wrote:

> Zeiss uses the M27 thread in their new microscope stands. This is
significantly larger than the old Zeiss mount, and a bit larger than Nikon.
Zeiss makes adapters from the old to the new M27 mount, so it should be
possible to get similar adapters from for Nikon objectives on the new Zeiss
scopes. In fact, I think Zeiss even has adapters to mount the new larger
objectives on the older scopes. However, all caveats mentioned by Chris
remain valid. In particular, Nikon objectives are huge compared to Zeiss
(Nikon 60x/1.4 Plan Apo is about 2 longer than the equivalent Zeiss.  
>
> Julio Vazquez
> Fred Hutchinson Cancer Research Center Seattle, WA 98109
>
> http://www.fhcrc.org/en.html
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nikon to Zeiss thread adaptor

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

OK. In that case, you should be able to get both adapters from ThorLabs (step down from Zeiss M27 to Nikon, and step down from Nikon to Zeiss 0.8). These are quite cheap, so it's probably worth a try.

Julio

On Jul 8, 2013, at 2:04 PM, Sripad Ram wrote:

> Hello Julio,
> Yes I am trying to mount a Nikon piezo on a Zeiss scope and then put an old
> Zeiss objective on the Nikon piezo.
>
> Sripad