To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, Slightly off-topic but I'm interested in how people store their objectives. In particular if anyone knows where to buy empty containers/cases/boxes to hold loose objectives? Some of our objectives came in plastic containers with threads in the lids to screw the objectives into... but some did not and I'm not sure what is the best way to handle storing them. Regards, Adrian Smith Centenary Institute, Sydney, Australia |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ask your local reps for more containers. That should be no problem. Since many of the mounts are standard threads, it doesn't even have to be the same company. For water immersion objectives used in experiments in live tissue, storing them with a bit of water in the bottom of the container is a good way to keep them from growing stubborn salt crystals. As shown here: http://labrigger.com/blog/2010/07/15/cleaning-objectives-and-other-optics/ |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Leaving objective lenses in a humid environment is a good way to encourage growth of mold, trading one problem for another. I believe it wiser to remove the water immersion objective lens from the microscope after use and clean it carefully with a gentle water based cleaning solution like 'Sparkle'. Make sure that the lens is dry and store it in its original capsule. Be sure to put the nosepiece plug back in the nosepiece. Many high numerical aperture dipping lenses do not have the same thread sizes as typical lenses from their manufacturers. Your rep should provide you with a replacement capsule identical to the one your objective lens was delivered in if you've lost yours. You'll also want to ask for a replacement nosepiece plug, or maybe a few of them. Jeffrey M Larson Confocal Systems Product Manager Nikon Instruments Inc. 1300 Walt Whitman Road Melville NY 11747-3064 Office: 631-547-8540 Fax: 631-547-4033 Mobile: 516-617-2228 [hidden email] www.nikoninstruments.com -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of SL Smith Sent: Friday, September 24, 2010 5:07 AM To: [hidden email] Subject: Re: Objective storage? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ask your local reps for more containers. That should be no problem. Since many of the mounts are standard threads, it doesn't even have to be the same company. For water immersion objectives used in experiments in live tissue, storing them with a bit of water in the bottom of the container is a good way to keep them from growing stubborn salt crystals. As shown here: http://labrigger.com/blog/2010/07/15/cleaning-objectives-and-other-optic s/ CONFIDENTIAL: This e-mail including any attachments is intended only for the party or parties to whom it is addressed and may contain information which is privileged and/or confidential. If you are not the intended recipient, you are hereby notified that any use, disclosure, dissemination, distribution, copying, or printing of any information contained in or attached to this e-mail is STRICTLY PROHIBITED and may constitute a breach of confidentiality and/or privilege. If you have received this e-mail in error, please notify immediately the sender by reply e-mail and then delete this e-mail and any attachments in their entirety from your system. Thank you. This e-mail message including any attachments is believed to be free of any viruses; however, it is the sole responsibility of the recipient to ensure that it is virus free, and Nikon does not accept any responsibility for any loss, disruption or damage to your data or computer system which may occur in connection with this e-mail including any attachments. |
![]() |
In reply to this post by Adrian Smith-6
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Adrian, if it's about zeiss lenses just let me know the Order numbers printed on the lenses and we will take care Best regards Bruene Venus Product Manager with Carl Zeiss MicroImaging in Germany Von meinem iPhone gesendet Am 24.09.2010 um 06:17 schrieb Adrian Smith <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > Slightly off-topic but I'm interested in how people store their objectives. > > In particular if anyone knows where to buy empty containers/cases/boxes to hold loose objectives? > > Some of our objectives came in plastic containers with threads in the lids to screw the objectives into... but some did not and I'm not sure what is the best way to handle storing them. > > Regards, > > Adrian Smith > Centenary Institute, Sydney, Australia |
![]() |
In reply to this post by Adrian Smith-6
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The trouble with storing objectives in the manufacturers' containers is that it is then natural to stand them on the lid (the widest part). But immersion objectives should always be stored pointing down so that nothing can run into the lens. We make up racks to hold the containers so that they have to be put in that way up. We've also had some lenses supplied in cardboard boxes with no proper container, but mostly these were cheap 'lab' or 'student' objectives which weren't so important. In such cases you have to decide whether the objective is worth the cost of a case. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Adrian Smith Sent: Friday, 24 September 2010 2:17 PM To: [hidden email] Subject: Objective storage? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, Slightly off-topic but I'm interested in how people store their objectives. In particular if anyone knows where to buy empty containers/cases/boxes to hold loose objectives? Some of our objectives came in plastic containers with threads in the lids to screw the objectives into... but some did not and I'm not sure what is the best way to handle storing them. Regards, Adrian Smith Centenary Institute, Sydney, Australia |
![]() |
In reply to this post by Adrian Smith-6
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We've designed an objective case that can be 3D-printed - you can download a model of it here: http://derisilab.ucsf.edu/index.php?3D=94 We've had some problems with storing the objectives in the manufacturer's cases as the plastic erodes into the objective threads and causes the objectives to stick in the microscope turret. Kurt On 9/23/2010 9:17 PM, Adrian Smith wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > Slightly off-topic but I'm interested in how people store their objectives. > > In particular if anyone knows where to buy empty containers/cases/boxes to hold loose objectives? > > Some of our objectives came in plastic containers with threads in the lids to screw the objectives into... but some did not and I'm not sure what is the best way to handle storing them. > > Regards, > > Adrian Smith > Centenary Institute, Sydney, Australia > |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Kurt, Your vase stores the objectives upside down. like Guy suggested it is best to store oil objectives pointing down to stop oil from seeping down inside the objective over the years. Are the objectives held in place somehow so you could invert the case if required? Cheers Cam Cameron J. Nowell Microscpy Manager Central Resource for Advanced Microscopy Ludwig Insttue for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 AUSTRALIA Office: +61 3 9341 3155 Mobile: +61422882700 Fax: +61 3 9341 3104 http://www.ludwig.edu.au/branch/research/platform/microscopy.htm ________________________________ From: Confocal Microscopy List on behalf of Kurt Thorn Sent: Tue 28/09/2010 3:00 AM To: [hidden email] Subject: Re: Objective storage? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We've designed an objective case that can be 3D-printed - you can download a model of it here: http://derisilab.ucsf.edu/index.php?3D=94 We've had some problems with storing the objectives in the manufacturer's cases as the plastic erodes into the objective threads and causes the objectives to stick in the microscope turret. Kurt On 9/23/2010 9:17 PM, Adrian Smith wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > Slightly off-topic but I'm interested in how people store their objectives. > > In particular if anyone knows where to buy empty containers/cases/boxes to hold loose objectives? > > Some of our objectives came in plastic containers with threads in the lids to screw the objectives into... but some did not and I'm not sure what is the best way to handle storing them. > > Regards, > > Adrian Smith > Centenary Institute, Sydney, Australia > |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** No, they just sit in there. We've had more trouble with objective threads getting gummed up with plastic particles than with oil running into the objectives (fingers crossed!) so we explicitly didn't want the objectives to thread into anything. For other projects we've tried to tap the ABS our 3D printer prints without much success. Kurt On 9/27/2010 3:18 PM, Cameron Nowell wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Kurt, > > Your vase stores the objectives upside down. like Guy suggested it is best to store oil objectives pointing down to stop oil from seeping down inside the objective over the years. Are the objectives held in place somehow so you could invert the case if required? > > > Cheers > > > Cam > > > Cameron J. Nowell > Microscpy Manager > Central Resource for Advanced Microscopy > Ludwig Insttue for Cancer Research > PO Box 2008 > Royal Melbourne Hospital > Victoria, 3050 > AUSTRALIA > > Office: +61 3 9341 3155 > Mobile: +61422882700 > Fax: +61 3 9341 3104 > > http://www.ludwig.edu.au/branch/research/platform/microscopy.htm > > > ________________________________ > > From: Confocal Microscopy List on behalf of Kurt Thorn > Sent: Tue 28/09/2010 3:00 AM > To: [hidden email] > Subject: Re: Objective storage? > > > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > We've designed an objective case that can be 3D-printed - you can > download a model of it here: http://derisilab.ucsf.edu/index.php?3D=94 > > We've had some problems with storing the objectives in the > manufacturer's cases as the plastic erodes into the objective threads > and causes the objectives to stick in the microscope turret. > > Kurt > > On 9/23/2010 9:17 PM, Adrian Smith wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi all, >> >> Slightly off-topic but I'm interested in how people store their objectives. >> >> In particular if anyone knows where to buy empty containers/cases/boxes to hold loose objectives? >> >> Some of our objectives came in plastic containers with threads in the lids to screw the objectives into... but some did not and I'm not sure what is the best way to handle storing them. >> >> Regards, >> >> Adrian Smith >> Centenary Institute, Sydney, Australia >> |