|
John Oreopoulos |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I have a question for the listserver pertaining to the application of Invitrogen TetraSpeck fluorescent beads (no commercial interest): http://www.lifetechnologies.com/order/catalog/product/T7282 I often use fluorescent beads like these for aligning two cameras (ie: simultaneous detection) on a widefield microscope or a spinning disk confocal microscope system. The spectral channels are usually green (488 nm excitation, 525/50 nm emission filter) and red (561 nm excitation, 593/40 nm emission filter) or far red (642 nm excitation, 700/75 nm emission filter). A high-quality 565 nm cut-on wavelength or dichroic mirror or a 620 nm dichroic mirror splits the multi-color emission between the two cameras. My issue is that as of late, I've noticed that with just 488 nm excitation alone, I can get quite a bit of fluorescence in my red or far red channel. It's possible that this signal is due to cross-talk or bleedthrough (I don't know the exact spectra of the beads), but I'm also wondering if there is a FRET effect going on here since the beads contain multiple dyes that are likely in close proximity within the bead. I find that I have to bias (increase) the EM gain on the green channel camera to get the green fluorescence signal comparable/balanced to the much brighter red or far-red fluorescence signal (which requires no EM gain and very little 561 nm or 642 nm excitation power). I can usually see from the image histogram that a good portion of the signal in the red or far red channel is due to this undesired signal induced by 488 nm excitation (which is undesirable for two camera alignment procedures). Does anyone out there have a similar experience with these beads and can someone suggest some ways avoid this undesired signal (that don't involve changing the filters). What I really want is multi-color beads that are spectrally distinct in these channels and only respond to the corresponding laser lines. Much thanks, John Oreopoulos Staff Scientist Spectral Applied Research Inc. A Division of Andor Technology Richmond Hill, Ontario Canada www.spectral.ca |
|
Iain Johnson |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** John: I'm sure the TetraSpeck beads exhibit FRET. Even the single-color Molecular Probes beads (FluoSpheres) exhibit FRET (homo-FRET in this case), deduced from the fact that they have very low fluorescence polarization, contrary to the predictions of hydrodynamics. Whether the red signal you are seeing with 488 nm is actually FRET or just direct excitation of the red and far red dyes at 488 nm is another question. The latter effect is inevitably present to some extent, as the dyes have a non-zero absorption cross-sections at all wavelengths below the 0-0 transition (as does every other fluorescent dye ever invented). The spectra of the dyes used to make the tetraspeck beads are available in the Molecular Probes Spectraviewer utility. I guess I'm wondering why you need to use two different lasers. Is there a significant difference in the illumination distribution delivered by the two lasers that invalidates aligning the cameras using green and red signals derived from 488 nm excitation only? Iain On Tue, Sep 23, 2014 at 12:19 PM, John Oreopoulos < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I have a question for the listserver pertaining to the application of > Invitrogen TetraSpeck fluorescent beads (no commercial interest): > > http://www.lifetechnologies.com/order/catalog/product/T7282 > > I often use fluorescent beads like these for aligning two cameras (ie: > simultaneous detection) on a widefield microscope or a spinning disk > confocal microscope system. The spectral channels are usually green (488 nm > excitation, 525/50 nm emission filter) and red (561 nm excitation, 593/40 > nm emission filter) or far red (642 nm excitation, 700/75 nm emission > filter). A high-quality 565 nm cut-on wavelength or dichroic mirror or a > 620 nm dichroic mirror splits the multi-color emission between the two > cameras. > > My issue is that as of late, I've noticed that with just 488 nm excitation > alone, I can get quite a bit of fluorescence in my red or far red channel. > It's possible that this signal is due to cross-talk or bleedthrough (I > don't know the exact spectra of the beads), but I'm also wondering if there > is a FRET effect going on here since the beads contain multiple dyes that > are likely in close proximity within the bead. I find that I have to bias > (increase) the EM gain on the green channel camera to get the green > fluorescence signal comparable/balanced to the much brighter red or far-red > fluorescence signal (which requires no EM gain and very little 561 nm or > 642 nm excitation power). I can usually see from the image histogram that a > good portion of the signal in the red or far red channel is due to this > undesired signal induced by 488 nm excitation (which is undesirable for two > camera alignment procedures). > > Does anyone out there have a similar experience with these beads and can > someone suggest some ways avoid this undesired signal (that don't involve > changing the filters). What I really want is multi-color beads that are > spectrally distinct in these channels and only respond to the corresponding > laser lines. > > Much thanks, > > > > John Oreopoulos > Staff Scientist > Spectral Applied Research Inc. > A Division of Andor Technology > Richmond Hill, Ontario > Canada > www.spectral.ca |
|
John Oreopoulos |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Iain, I guess the occurrence of FRET concerns me to some extent because there might be a (wavelength-dependent) lateral magnification difference of the bead images onto each camera. The camera lenses I use have an adjustable focal length and it is possible to achieve pixel-to-pixel registration by making appropriate adjustments of these lenses, but I want to be sure that the magnification difference I'm seeing is for a certain excitation/emission combination. A similar argument for wanting to remove the FRET effect might be made when trying to match the axial focus of both cameras. But maybe my concern is trivial? Perhaps the magnification difference between 488 nm excitation / 700 nm emission compared to 640 nm excitation / 700 emission is too small to detect (ie: sub-pixel difference). I'm not sure, but I seem to recall one instance where I thought I could see this difference. John Oreopoulos Staff Scientist Spectral Applied Research Inc. A Division of Andor Technology Richmond Hill, Ontario Canada www.spectral.ca On 2014-09-24, at 10:20 PM, Iain Johnson wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > John: > > I'm sure the TetraSpeck beads exhibit FRET. Even the single-color > Molecular Probes beads (FluoSpheres) exhibit FRET (homo-FRET in this > case), deduced from the fact that they have very low fluorescence > polarization, contrary to the predictions of hydrodynamics. Whether the > red signal you are seeing with 488 nm is actually FRET or just direct > excitation of the red and far red dyes at 488 nm is another question. The > latter effect is inevitably present to some extent, as the dyes have a > non-zero absorption cross-sections at all wavelengths below the 0-0 > transition (as does every other fluorescent dye ever invented). The > spectra of the dyes used to make the tetraspeck beads are available in the > Molecular Probes Spectraviewer utility. > > I guess I'm wondering why you need to use two different lasers. Is there a > significant difference in the illumination distribution delivered by the > two lasers that invalidates aligning the cameras using green and red > signals derived from 488 nm excitation only? > > Iain > > On Tue, Sep 23, 2014 at 12:19 PM, John Oreopoulos < > [hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> I have a question for the listserver pertaining to the application of >> Invitrogen TetraSpeck fluorescent beads (no commercial interest): >> >> http://www.lifetechnologies.com/order/catalog/product/T7282 >> >> I often use fluorescent beads like these for aligning two cameras (ie: >> simultaneous detection) on a widefield microscope or a spinning disk >> confocal microscope system. The spectral channels are usually green (488 nm >> excitation, 525/50 nm emission filter) and red (561 nm excitation, 593/40 >> nm emission filter) or far red (642 nm excitation, 700/75 nm emission >> filter). A high-quality 565 nm cut-on wavelength or dichroic mirror or a >> 620 nm dichroic mirror splits the multi-color emission between the two >> cameras. >> >> My issue is that as of late, I've noticed that with just 488 nm excitation >> alone, I can get quite a bit of fluorescence in my red or far red channel. >> It's possible that this signal is due to cross-talk or bleedthrough (I >> don't know the exact spectra of the beads), but I'm also wondering if there >> is a FRET effect going on here since the beads contain multiple dyes that >> are likely in close proximity within the bead. I find that I have to bias >> (increase) the EM gain on the green channel camera to get the green >> fluorescence signal comparable/balanced to the much brighter red or far-red >> fluorescence signal (which requires no EM gain and very little 561 nm or >> 642 nm excitation power). I can usually see from the image histogram that a >> good portion of the signal in the red or far red channel is due to this >> undesired signal induced by 488 nm excitation (which is undesirable for two >> camera alignment procedures). >> >> Does anyone out there have a similar experience with these beads and can >> someone suggest some ways avoid this undesired signal (that don't involve >> changing the filters). What I really want is multi-color beads that are >> spectrally distinct in these channels and only respond to the corresponding >> laser lines. >> >> Much thanks, >> >> >> >> John Oreopoulos >> Staff Scientist >> Spectral Applied Research Inc. >> A Division of Andor Technology >> Richmond Hill, Ontario >> Canada >> www.spectral.ca |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I guess the best way to test would be to use a single-labelled bead with the same fluorophore as the putative donor. If FRET is occurring in the tetraspeck beads you'll see a reduction in donor lifetime. If you don't have a FLIM system you can probably persuade someone else to run the test for you. I don't know how these beads are made. If they really are four distinct specks it seems unlikely that significant FRET would occur, buy of course the centroids won't align which could be a problem for your registration tests. If that's not the case, and the beads are just coated with all 4 fluorophores then FRET is a distinct possibility. Guy Guy Cox, Honorary Associate Professor School of Medical Sciences Australian Centre for Microscopy and Microanalysis, Madsen, F09, University of Sydney, NSW 2006 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of John Oreopoulos Sent: Saturday, 27 September 2014 2:52 PM To: [hidden email] Subject: Re: Occurrence of FRET in Invitrogen TetraSpeck beads? (commercial response) ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Iain, I guess the occurrence of FRET concerns me to some extent because there might be a (wavelength-dependent) lateral magnification difference of the bead images onto each camera. The camera lenses I use have an adjustable focal length and it is possible to achieve pixel-to-pixel registration by making appropriate adjustments of these lenses, but I want to be sure that the magnification difference I'm seeing is for a certain excitation/emission combination. A similar argument for wanting to remove the FRET effect might be made when trying to match the axial focus of both cameras. But maybe my concern is trivial? Perhaps the magnification difference between 488 nm excitation / 700 nm emission compared to 640 nm excitation / 700 emission is too small to detect (ie: sub-pixel difference). I'm not sure, but I seem to recall one instance where I thought I could see this difference. John Oreopoulos Staff Scientist Spectral Applied Research Inc. A Division of Andor Technology Richmond Hill, Ontario Canada www.spectral.ca On 2014-09-24, at 10:20 PM, Iain Johnson wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > John: > > I'm sure the TetraSpeck beads exhibit FRET. Even the single-color > Molecular Probes beads (FluoSpheres) exhibit FRET (homo-FRET in this > case), deduced from the fact that they have very low fluorescence > polarization, contrary to the predictions of hydrodynamics. Whether > the red signal you are seeing with 488 nm is actually FRET or just > direct excitation of the red and far red dyes at 488 nm is another > question. The latter effect is inevitably present to some extent, as > the dyes have a non-zero absorption cross-sections at all wavelengths below the 0-0 > transition (as does every other fluorescent dye ever invented). The > spectra of the dyes used to make the tetraspeck beads are available in > the Molecular Probes Spectraviewer utility. > > I guess I'm wondering why you need to use two different lasers. Is > there a significant difference in the illumination distribution > delivered by the two lasers that invalidates aligning the cameras > using green and red signals derived from 488 nm excitation only? > > Iain > > On Tue, Sep 23, 2014 at 12:19 PM, John Oreopoulos < > [hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> I have a question for the listserver pertaining to the application of >> Invitrogen TetraSpeck fluorescent beads (no commercial interest): >> >> http://www.lifetechnologies.com/order/catalog/product/T7282 >> >> I often use fluorescent beads like these for aligning two cameras (ie: >> simultaneous detection) on a widefield microscope or a spinning disk >> confocal microscope system. The spectral channels are usually green >> (488 nm excitation, 525/50 nm emission filter) and red (561 nm >> excitation, 593/40 nm emission filter) or far red (642 nm excitation, >> 700/75 nm emission filter). A high-quality 565 nm cut-on wavelength >> or dichroic mirror or a >> 620 nm dichroic mirror splits the multi-color emission between the >> two cameras. >> >> My issue is that as of late, I've noticed that with just 488 nm >> excitation alone, I can get quite a bit of fluorescence in my red or far red channel. >> It's possible that this signal is due to cross-talk or bleedthrough >> (I don't know the exact spectra of the beads), but I'm also wondering >> if there is a FRET effect going on here since the beads contain >> multiple dyes that are likely in close proximity within the bead. I >> find that I have to bias >> (increase) the EM gain on the green channel camera to get the green >> fluorescence signal comparable/balanced to the much brighter red or >> far-red fluorescence signal (which requires no EM gain and very >> little 561 nm or >> 642 nm excitation power). I can usually see from the image histogram >> that a good portion of the signal in the red or far red channel is >> due to this undesired signal induced by 488 nm excitation (which is >> undesirable for two camera alignment procedures). >> >> Does anyone out there have a similar experience with these beads and >> can someone suggest some ways avoid this undesired signal (that don't >> involve changing the filters). What I really want is multi-color >> beads that are spectrally distinct in these channels and only respond >> to the corresponding laser lines. >> >> Much thanks, >> >> >> >> John Oreopoulos >> Staff Scientist >> Spectral Applied Research Inc. >> A Division of Andor Technology >> Richmond Hill, Ontario >> Canada >> www.spectral.ca |
|
Mark Cannell-2 |
|
|
In reply to this post by John Oreopoulos
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi John Wavelength dependent magnification differences would be seen as blurring toward the edge of the field in transmitted white light images for a monochrome camera and colour fringing in a colour camera. HTH Mark On 27/09/2014, at 5:51 am, John Oreopoulos <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Iain, > > I guess the occurrence of FRET concerns me to some extent because there might be a (wavelength-dependent) lateral magnification difference of the bead images onto each camera. The camera lenses I use have an adjustable focal length and it is possible to achieve pixel-to-pixel registration by making appropriate adjustments of these lenses, but I want to be sure that the magnification difference I'm seeing is for a certain excitation/emission combination. > > A similar argument for wanting to remove the FRET effect might be made when trying to match the axial focus of both cameras. > > But maybe my concern is trivial? Perhaps the magnification difference between 488 nm excitation / 700 nm emission compared to 640 nm excitation / 700 emission is too small to detect (ie: sub-pixel difference). I'm not sure, but I seem to recall one instance where I thought I could see this difference. > > John Oreopoulos > Staff Scientist > Spectral Applied Research Inc. > A Division of Andor Technology > Richmond Hill, Ontario > Canada > www.spectral.ca > > > > On 2014-09-24, at 10:20 PM, Iain Johnson wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> John: >> >> I'm sure the TetraSpeck beads exhibit FRET. Even the single-color >> Molecular Probes beads (FluoSpheres) exhibit FRET (homo-FRET in this >> case), deduced from the fact that they have very low fluorescence >> polarization, contrary to the predictions of hydrodynamics. Whether the >> red signal you are seeing with 488 nm is actually FRET or just direct >> excitation of the red and far red dyes at 488 nm is another question. The >> latter effect is inevitably present to some extent, as the dyes have a >> non-zero absorption cross-sections at all wavelengths below the 0-0 >> transition (as does every other fluorescent dye ever invented). The >> spectra of the dyes used to make the tetraspeck beads are available in the >> Molecular Probes Spectraviewer utility. >> >> I guess I'm wondering why you need to use two different lasers. Is there a >> significant difference in the illumination distribution delivered by the >> two lasers that invalidates aligning the cameras using green and red >> signals derived from 488 nm excitation only? >> >> Iain >> >> On Tue, Sep 23, 2014 at 12:19 PM, John Oreopoulos < >> [hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. >>> ***** >>> >>> I have a question for the listserver pertaining to the application of >>> Invitrogen TetraSpeck fluorescent beads (no commercial interest): >>> >>> http://www.lifetechnologies.com/order/catalog/product/T7282 >>> >>> I often use fluorescent beads like these for aligning two cameras (ie: >>> simultaneous detection) on a widefield microscope or a spinning disk >>> confocal microscope system. The spectral channels are usually green (488 nm >>> excitation, 525/50 nm emission filter) and red (561 nm excitation, 593/40 >>> nm emission filter) or far red (642 nm excitation, 700/75 nm emission >>> filter). A high-quality 565 nm cut-on wavelength or dichroic mirror or a >>> 620 nm dichroic mirror splits the multi-color emission between the two >>> cameras. >>> >>> My issue is that as of late, I've noticed that with just 488 nm excitation >>> alone, I can get quite a bit of fluorescence in my red or far red channel. >>> It's possible that this signal is due to cross-talk or bleedthrough (I >>> don't know the exact spectra of the beads), but I'm also wondering if there >>> is a FRET effect going on here since the beads contain multiple dyes that >>> are likely in close proximity within the bead. I find that I have to bias >>> (increase) the EM gain on the green channel camera to get the green >>> fluorescence signal comparable/balanced to the much brighter red or far-red >>> fluorescence signal (which requires no EM gain and very little 561 nm or >>> 642 nm excitation power). I can usually see from the image histogram that a >>> good portion of the signal in the red or far red channel is due to this >>> undesired signal induced by 488 nm excitation (which is undesirable for two >>> camera alignment procedures). >>> >>> Does anyone out there have a similar experience with these beads and can >>> someone suggest some ways avoid this undesired signal (that don't involve >>> changing the filters). What I really want is multi-color beads that are >>> spectrally distinct in these channels and only respond to the corresponding >>> laser lines. >>> >>> Much thanks, >>> >>> >>> >>> John Oreopoulos >>> Staff Scientist >>> Spectral Applied Research Inc. >>> A Division of Andor Technology >>> Richmond Hill, Ontario >>> Canada >>> www.spectral.ca Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK |
|
Reece, Jeff (NIH/NIDDK) [E] |
|
|
In reply to this post by John Oreopoulos
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi John, Seems to me, since we are talking about camera-based WF and spinning disk confocal, that where the photons land in the image plane is determined only by the emission path from the sample, and the emission wavelengths, not where the excitation came from or how it was applied. If comparing to a FRET signal, the image location would be the same, but bleedthrough could theoretically show a slight difference because the emission profile within the bandpass filter would be different. However, I would be surprised if you can see the difference with normal bandpass filters, using regular WF and spinning disk imaging and assuming good quality of optics. Cheers, -Jeff -----Original Message----- From: John Oreopoulos [mailto:[hidden email]] Sent: Saturday, September 27, 2014 12:52 AM To: [hidden email] Subject: Re: Occurrence of FRET in Invitrogen TetraSpeck beads? (commercial response) ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Iain, I guess the occurrence of FRET concerns me to some extent because there might be a (wavelength-dependent) lateral magnification difference of the bead images onto each camera. The camera lenses I use have an adjustable focal length and it is possible to achieve pixel-to-pixel registration by making appropriate adjustments of these lenses, but I want to be sure that the magnification difference I'm seeing is for a certain excitation/emission combination. A similar argument for wanting to remove the FRET effect might be made when trying to match the axial focus of both cameras. But maybe my concern is trivial? Perhaps the magnification difference between 488 nm excitation / 700 nm emission compared to 640 nm excitation / 700 emission is too small to detect (ie: sub-pixel difference). I'm not sure, but I seem to recall one instance where I thought I could see this difference. John Oreopoulos Staff Scientist Spectral Applied Research Inc. A Division of Andor Technology Richmond Hill, Ontario Canada www.spectral.ca On 2014-09-24, at 10:20 PM, Iain Johnson wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > John: > > I'm sure the TetraSpeck beads exhibit FRET. Even the single-color > Molecular Probes beads (FluoSpheres) exhibit FRET (homo-FRET in this > case), deduced from the fact that they have very low fluorescence > polarization, contrary to the predictions of hydrodynamics. Whether > the red signal you are seeing with 488 nm is actually FRET or just > direct excitation of the red and far red dyes at 488 nm is another > question. The latter effect is inevitably present to some extent, as > the dyes have a non-zero absorption cross-sections at all wavelengths below the 0-0 > transition (as does every other fluorescent dye ever invented). The > spectra of the dyes used to make the tetraspeck beads are available in > the Molecular Probes Spectraviewer utility. > > I guess I'm wondering why you need to use two different lasers. Is > there a significant difference in the illumination distribution > delivered by the two lasers that invalidates aligning the cameras > using green and red signals derived from 488 nm excitation only? > > Iain > > On Tue, Sep 23, 2014 at 12:19 PM, John Oreopoulos < > [hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> I have a question for the listserver pertaining to the application of >> Invitrogen TetraSpeck fluorescent beads (no commercial interest): >> >> http://www.lifetechnologies.com/order/catalog/product/T7282 >> >> I often use fluorescent beads like these for aligning two cameras (ie: >> simultaneous detection) on a widefield microscope or a spinning disk >> confocal microscope system. The spectral channels are usually green >> (488 nm excitation, 525/50 nm emission filter) and red (561 nm >> excitation, 593/40 nm emission filter) or far red (642 nm excitation, >> 700/75 nm emission filter). A high-quality 565 nm cut-on wavelength >> or dichroic mirror or a >> 620 nm dichroic mirror splits the multi-color emission between the >> two cameras. >> >> My issue is that as of late, I've noticed that with just 488 nm >> excitation alone, I can get quite a bit of fluorescence in my red or far red channel. >> It's possible that this signal is due to cross-talk or bleedthrough >> (I don't know the exact spectra of the beads), but I'm also wondering >> if there is a FRET effect going on here since the beads contain >> multiple dyes that are likely in close proximity within the bead. I >> find that I have to bias >> (increase) the EM gain on the green channel camera to get the green >> fluorescence signal comparable/balanced to the much brighter red or >> far-red fluorescence signal (which requires no EM gain and very >> little 561 nm or >> 642 nm excitation power). I can usually see from the image histogram >> that a good portion of the signal in the red or far red channel is >> due to this undesired signal induced by 488 nm excitation (which is >> undesirable for two camera alignment procedures). >> >> Does anyone out there have a similar experience with these beads and >> can someone suggest some ways avoid this undesired signal (that don't >> involve changing the filters). What I really want is multi-color >> beads that are spectrally distinct in these channels and only respond >> to the corresponding laser lines. >> >> Much thanks, >> >> >> >> John Oreopoulos >> Staff Scientist >> Spectral Applied Research Inc. >> A Division of Andor Technology >> Richmond Hill, Ontario >> Canada >> www.spectral.ca |
| Free forum by Nabble | Edit this page |