Hi list.
I want to change grease in condenser and  binocular tubes of Leica DM IRE2
microscope. The old grease solidified and made those parts stuck. Can
anyone recommend a type and/or supplier of a grease for such purpose?
Thank you.
Petro.
Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Petro,

I use white lithium grease for this. Available in hardware and auto parts stores.

Phil
-------------
Philip Oshel    
Imaging Facility Director
Biology Department
1304 Biosciences
1455 Calumet Ct.
Central Michigan University
Mt. Pleasant, MI 48859
989 774-3576 office
989 774-7567 lab

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> on behalf of Piter_ <[hidden email]>
Reply-To: Confocal Microscopy List <[hidden email]>
Date: Sunday,  03February, 2019 at 07:06
To: "[hidden email]" <[hidden email]>
Subject: Oil/grease for microscope?

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Hi list.
    I want to change grease in condenser and  binocular tubes of Leica DM IRE2
    microscope. The old grease solidified and made those parts stuck. Can
    anyone recommend a type and/or supplier of a grease for such purpose?
    Thank you.
    Petro.
   

Piter_ Piter_
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Phil.
Lithium grease was the first thing came into my mind. I used to have a big
can of Litol for my bike. But then I have  read some scary info that
microscope grease has to be very special  or  it will evaporate, leak and
destroy your equipment.... So that kind of stopped me from using it. Do you
use it for long time? My guess is you had no problems with it :) My
intension was to put some dumping grease there. Like Nye A975, which is
recommended for focusing mechanism. Maybe it is an overkill.

Thanks.
Petro.

On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <[hidden email] wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Petro,
>
> I use white lithium grease for this. Available in hardware and auto parts
> stores.
>
> Phil
> -------------
> Philip Oshel
> Imaging Facility Director
> Biology Department
> 1304 Biosciences
> 1455 Calumet Ct.
> Central Michigan University
> Mt. Pleasant, MI 48859
> 989 774-3576 office
> 989 774-7567 lab
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> on
> behalf of Piter_ <[hidden email]>
> Reply-To: Confocal Microscopy List <[hidden email]>
> Date: Sunday,  03February, 2019 at 07:06
> To: "[hidden email]" <[hidden email]>
> Subject: Oil/grease for microscope?
>
>     *****
>     To join, leave or search the confocal microscopy listserv, go to:
>     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>     Post images on http://www.imgur.com and include the link in your
> posting.
>     *****
>
>     Hi list.
>     I want to change grease in condenser and  binocular tubes of Leica DM
> IRE2
>     microscope. The old grease solidified and made those parts stuck. Can
>     anyone recommend a type and/or supplier of a grease for such purpose?
>     Thank you.
>     Petro.
>
>
>
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

There is grease available for vacuum applications that is very low
off-gassing. If you use that stuff you will probably be fine. It is
typically used for moving parts inside vacuum chambers or other
low-pressure environments.

Craig

On Mon, Feb 4, 2019 at 12:18 PM Piter_ <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear Phil.
> Lithium grease was the first thing came into my mind. I used to have a big
> can of Litol for my bike. But then I have  read some scary info that
> microscope grease has to be very special  or  it will evaporate, leak and
> destroy your equipment.... So that kind of stopped me from using it. Do you
> use it for long time? My guess is you had no problems with it :) My
> intension was to put some dumping grease there. Like Nye A975, which is
> recommended for focusing mechanism. Maybe it is an overkill.
>
> Thanks.
> Petro.
>
> On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <[hidden email] wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Petro,
> >
> > I use white lithium grease for this. Available in hardware and auto parts
> > stores.
> >
> > Phil
> > -------------
> > Philip Oshel
> > Imaging Facility Director
> > Biology Department
> > 1304 Biosciences
> > 1455 Calumet Ct.
> > Central Michigan University
> > Mt. Pleasant, MI 48859
> > 989 774-3576 office
> > 989 774-7567 lab
> >
> > -----Original Message-----
> > From: Confocal Microscopy List <[hidden email]> on
> > behalf of Piter_ <[hidden email]>
> > Reply-To: Confocal Microscopy List <[hidden email]>
> > Date: Sunday,  03February, 2019 at 07:06
> > To: "[hidden email]" <[hidden email]
> >
> > Subject: Oil/grease for microscope?
> >
> >     *****
> >     To join, leave or search the confocal microscopy listserv, go to:
> >     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >     Post images on http://www.imgur.com and include the link in your
> > posting.
> >     *****
> >
> >     Hi list.
> >     I want to change grease in condenser and  binocular tubes of Leica DM
> > IRE2
> >     microscope. The old grease solidified and made those parts stuck. Can
> >     anyone recommend a type and/or supplier of a grease for such purpose?
> >     Thank you.
> >     Petro.
> >
> >
> >
>
Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Note that for-real vacuum grease is NOT silicone vacuum grease. The silicone grease is only suitable for sealing desiccators.
Fomblin is probably the best choice of vacuum grease for moving parts. Be aware that any high vacuum grease is high viscosity and can be "sticky". Use a very small amount and apply as a *thin* layer.
It's also expensive, but all true high-vacuum greases are.
Also: if you're applying the grease to any polymer or rubber parts, first check it on a bit that doesn't matter. The grease may goo up the polymer over time, especially if the part is warmed or heated.

Phil
-------------
Philip Oshel    
Imaging Facility Director
Biology Department
1304 Biosciences
1455 Calumet Ct.
Central Michigan University
Mt. Pleasant, MI 48859
989 774-3576 office
989 774-7567 lab

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> on behalf of Craig Brideau <[hidden email]>
Reply-To: Confocal Microscopy List <[hidden email]>
Date: Monday,  04February, 2019 at 17:44
To: "[hidden email]" <[hidden email]>
Subject: Re: Oil/grease for microscope?

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    There is grease available for vacuum applications that is very low
    off-gassing. If you use that stuff you will probably be fine. It is
    typically used for moving parts inside vacuum chambers or other
    low-pressure environments.
   
    Craig
   
    On Mon, Feb 4, 2019 at 12:18 PM Piter_ <[hidden email]> wrote:
   
    > *****
    > To join, leave or search the confocal microscopy listserv, go to:
    > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    > Post images on http://www.imgur.com and include the link in your posting.
    > *****
    >
    > Dear Phil.
    > Lithium grease was the first thing came into my mind. I used to have a big
    > can of Litol for my bike. But then I have  read some scary info that
    > microscope grease has to be very special  or  it will evaporate, leak and
    > destroy your equipment.... So that kind of stopped me from using it. Do you
    > use it for long time? My guess is you had no problems with it :) My
    > intension was to put some dumping grease there. Like Nye A975, which is
    > recommended for focusing mechanism. Maybe it is an overkill.
    >
    > Thanks.
    > Petro.
    >
    > On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <[hidden email] wrote:
    >
    > > *****
    > > To join, leave or search the confocal microscopy listserv, go to:
    > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    > > Post images on http://www.imgur.com and include the link in your
    > posting.
    > > *****
    > >
    > > Petro,
    > >
    > > I use white lithium grease for this. Available in hardware and auto parts
    > > stores.
    > >
    > > Phil
    > > -------------
    > > Philip Oshel
    > > Imaging Facility Director
    > > Biology Department
    > > 1304 Biosciences
    > > 1455 Calumet Ct.
    > > Central Michigan University
    > > Mt. Pleasant, MI 48859
    > > 989 774-3576 office
    > > 989 774-7567 lab
    > >
    > > -----Original Message-----
    > > From: Confocal Microscopy List <[hidden email]> on
    > > behalf of Piter_ <[hidden email]>
    > > Reply-To: Confocal Microscopy List <[hidden email]>
    > > Date: Sunday,  03February, 2019 at 07:06
    > > To: "[hidden email]" <[hidden email]
    > >
    > > Subject: Oil/grease for microscope?
    > >
    > >     *****
    > >     To join, leave or search the confocal microscopy listserv, go to:
    > >     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    > >     Post images on http://www.imgur.com and include the link in your
    > > posting.
    > >     *****
    > >
    > >     Hi list.
    > >     I want to change grease in condenser and  binocular tubes of Leica DM
    > > IRE2
    > >     microscope. The old grease solidified and made those parts stuck. Can
    > >     anyone recommend a type and/or supplier of a grease for such purpose?
    > >     Thank you.
    > >     Petro.
    > >
    > >
    > >
    >
   

Mark Cannell-2 Mark Cannell-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Seems to me that Dow silicone High vacuum grease is a good choice -they recommend it for optical components and being silicone protects o rings well. It's no so sticky. No need to worry about super high vacuum in this application ... ?

HTH

Mark B. Cannell.
Department of Physiology, Pharmacology & Neuroscience
School of Medical Sciences
University Walk
Bristol BS8 1TD
 
[hidden email]
 
 

On 5/02/19, 1:23 PM, "Confocal Microscopy List on behalf of Oshel, Philip Eugene" <[hidden email] on behalf of [hidden email]> wrote:

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Note that for-real vacuum grease is NOT silicone vacuum grease. The silicone grease is only suitable for sealing desiccators.
    Fomblin is probably the best choice of vacuum grease for moving parts. Be aware that any high vacuum grease is high viscosity and can be "sticky". Use a very small amount and apply as a *thin* layer.
    It's also expensive, but all true high-vacuum greases are.
    Also: if you're applying the grease to any polymer or rubber parts, first check it on a bit that doesn't matter. The grease may goo up the polymer over time, especially if the part is warmed or heated.
   
    Phil
    -------------
    Philip Oshel    
    Imaging Facility Director
    Biology Department
    1304 Biosciences
    1455 Calumet Ct.
    Central Michigan University
    Mt. Pleasant, MI 48859
    989 774-3576 office
    989 774-7567 lab
   
    -----Original Message-----
    From: Confocal Microscopy List <[hidden email]> on behalf of Craig Brideau <[hidden email]>
    Reply-To: Confocal Microscopy List <[hidden email]>
    Date: Monday,  04February, 2019 at 17:44
    To: "[hidden email]" <[hidden email]>
    Subject: Re: Oil/grease for microscope?
   
        *****
        To join, leave or search the confocal microscopy listserv, go to:
        http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
        Post images on http://www.imgur.com and include the link in your posting.
        *****
       
        There is grease available for vacuum applications that is very low
        off-gassing. If you use that stuff you will probably be fine. It is
        typically used for moving parts inside vacuum chambers or other
        low-pressure environments.
       
        Craig
       
        On Mon, Feb 4, 2019 at 12:18 PM Piter_ <[hidden email]> wrote:
       
        > *****
        > To join, leave or search the confocal microscopy listserv, go to:
        > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
        > Post images on http://www.imgur.com and include the link in your posting.
        > *****
        >
        > Dear Phil.
        > Lithium grease was the first thing came into my mind. I used to have a big
        > can of Litol for my bike. But then I have  read some scary info that
        > microscope grease has to be very special  or  it will evaporate, leak and
        > destroy your equipment.... So that kind of stopped me from using it. Do you
        > use it for long time? My guess is you had no problems with it :) My
        > intension was to put some dumping grease there. Like Nye A975, which is
        > recommended for focusing mechanism. Maybe it is an overkill.
        >
        > Thanks.
        > Petro.
        >
        > On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <[hidden email] wrote:
        >
        > > *****
        > > To join, leave or search the confocal microscopy listserv, go to:
        > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
        > > Post images on http://www.imgur.com and include the link in your
        > posting.
        > > *****
        > >
        > > Petro,
        > >
        > > I use white lithium grease for this. Available in hardware and auto parts
        > > stores.
        > >
        > > Phil
        > > -------------
        > > Philip Oshel
        > > Imaging Facility Director
        > > Biology Department
        > > 1304 Biosciences
        > > 1455 Calumet Ct.
        > > Central Michigan University
        > > Mt. Pleasant, MI 48859
        > > 989 774-3576 office
        > > 989 774-7567 lab
        > >
        > > -----Original Message-----
        > > From: Confocal Microscopy List <[hidden email]> on
        > > behalf of Piter_ <[hidden email]>
        > > Reply-To: Confocal Microscopy List <[hidden email]>
        > > Date: Sunday,  03February, 2019 at 07:06
        > > To: "[hidden email]" <[hidden email]
        > >
        > > Subject: Oil/grease for microscope?
        > >
        > >     *****
        > >     To join, leave or search the confocal microscopy listserv, go to:
        > >     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
        > >     Post images on http://www.imgur.com and include the link in your
        > > posting.
        > >     *****
        > >
        > >     Hi list.
        > >     I want to change grease in condenser and  binocular tubes of Leica DM
        > > IRE2
        > >     microscope. The old grease solidified and made those parts stuck. Can
        > >     anyone recommend a type and/or supplier of a grease for such purpose?
        > >     Thank you.
        > >     Petro.
        > >
        > >
        > >
        >
       
   
   

Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Correct, the vacuum is not the issue, but Fomblin and like are still good greases, just pricey. They don't outgas and so won't get gunk on the optics or otherwise contaminate the optics.

I don't think the silicone high vacuum greases are good for lubricating. First, they give off acetic acid vapor - I don't see why Dow would recommend it for optical components. Second, the silicone greases are sticky. Not good for moving parts. It does protect o-rings, but watch the acetic acid and o-ring composition.
Silicone grease seals coverslips well, though.

Phil
-------------
Philip Oshel    
Imaging Facility Director
Biology Department
1304 Biosciences
1455 Calumet Ct.
Central Michigan University
Mt. Pleasant, MI 48859
989 774-3576 office
989 774-7567 lab

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> on behalf of Mark Cannell <[hidden email]>
Reply-To: Confocal Microscopy List <[hidden email]>
Date: Tuesday,  05February, 2019 at 11:02
To: "[hidden email]" <[hidden email]>
Subject: Re: Oil/grease for microscope?

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Seems to me that Dow silicone High vacuum grease is a good choice -they recommend it for optical components and being silicone protects o rings well. It's no so sticky. No need to worry about super high vacuum in this application ... ?
   
    HTH
   
    Mark B. Cannell.
    Department of Physiology, Pharmacology & Neuroscience
    School of Medical Sciences
    University Walk
    Bristol BS8 1TD
     
    [hidden email]
     
     
   
    On 5/02/19, 1:23 PM, "Confocal Microscopy List on behalf of Oshel, Philip Eugene" <[hidden email] on behalf of [hidden email]> wrote:
   
        *****
        To join, leave or search the confocal microscopy listserv, go to:
        http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
        Post images on http://www.imgur.com and include the link in your posting.
        *****
       
        Note that for-real vacuum grease is NOT silicone vacuum grease. The silicone grease is only suitable for sealing desiccators.
        Fomblin is probably the best choice of vacuum grease for moving parts. Be aware that any high vacuum grease is high viscosity and can be "sticky". Use a very small amount and apply as a *thin* layer.
        It's also expensive, but all true high-vacuum greases are.
        Also: if you're applying the grease to any polymer or rubber parts, first check it on a bit that doesn't matter. The grease may goo up the polymer over time, especially if the part is warmed or heated.
       
        Phil
        -------------
        Philip Oshel    
        Imaging Facility Director
        Biology Department
        1304 Biosciences
        1455 Calumet Ct.
        Central Michigan University
        Mt. Pleasant, MI 48859
        989 774-3576 office
        989 774-7567 lab
       
        -----Original Message-----
        From: Confocal Microscopy List <[hidden email]> on behalf of Craig Brideau <[hidden email]>
        Reply-To: Confocal Microscopy List <[hidden email]>
        Date: Monday,  04February, 2019 at 17:44
        To: "[hidden email]" <[hidden email]>
        Subject: Re: Oil/grease for microscope?
       
            *****
            To join, leave or search the confocal microscopy listserv, go to:
            http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
            Post images on http://www.imgur.com and include the link in your posting.
            *****
           
            There is grease available for vacuum applications that is very low
            off-gassing. If you use that stuff you will probably be fine. It is
            typically used for moving parts inside vacuum chambers or other
            low-pressure environments.
           
            Craig
           
            On Mon, Feb 4, 2019 at 12:18 PM Piter_ <[hidden email]> wrote:
           
            > *****
            > To join, leave or search the confocal microscopy listserv, go to:
            > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
            > Post images on http://www.imgur.com and include the link in your posting.
            > *****
            >
            > Dear Phil.
            > Lithium grease was the first thing came into my mind. I used to have a big
            > can of Litol for my bike. But then I have  read some scary info that
            > microscope grease has to be very special  or  it will evaporate, leak and
            > destroy your equipment.... So that kind of stopped me from using it. Do you
            > use it for long time? My guess is you had no problems with it :) My
            > intension was to put some dumping grease there. Like Nye A975, which is
            > recommended for focusing mechanism. Maybe it is an overkill.
            >
            > Thanks.
            > Petro.
            >
            > On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <[hidden email] wrote:
            >
            > > *****
            > > To join, leave or search the confocal microscopy listserv, go to:
            > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
            > > Post images on http://www.imgur.com and include the link in your
            > posting.
            > > *****
            > >
            > > Petro,
            > >
            > > I use white lithium grease for this. Available in hardware and auto parts
            > > stores.
            > >
            > > Phil
            > > -------------
            > > Philip Oshel
            > > Imaging Facility Director
            > > Biology Department
            > > 1304 Biosciences
            > > 1455 Calumet Ct.
            > > Central Michigan University
            > > Mt. Pleasant, MI 48859
            > > 989 774-3576 office
            > > 989 774-7567 lab
            > >
            > > -----Original Message-----
            > > From: Confocal Microscopy List <[hidden email]> on
            > > behalf of Piter_ <[hidden email]>
            > > Reply-To: Confocal Microscopy List <[hidden email]>
            > > Date: Sunday,  03February, 2019 at 07:06
            > > To: "[hidden email]" <[hidden email]
            > >
            > > Subject: Oil/grease for microscope?
            > >
            > >     *****
            > >     To join, leave or search the confocal microscopy listserv, go to:
            > >     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
            > >     Post images on http://www.imgur.com and include the link in your
            > > posting.
            > >     *****
            > >
            > >     Hi list.
            > >     I want to change grease in condenser and  binocular tubes of Leica DM
            > > IRE2
            > >     microscope. The old grease solidified and made those parts stuck. Can
            > >     anyone recommend a type and/or supplier of a grease for such purpose?
            > >     Thank you.
            > >     Petro.
            > >
            > >
            > >
            >
           
       
       
   
   

Mark Cannell-2 Mark Cannell-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

No acetic acid from the Dow corning stuff I've used -it's a specialty silicon polymer plus thickener -not polymerized "bathroom caulk" as far as I know -more like low MW sylgard I think. While it's exact composition may be a trade secret, it's vapour pressure is very low at 100oC so I doubt it could be outgassing much, if any, acetic acid. I've not seen it corroding brass parts.

Mark B. Cannell.
Department of Physiology, Pharmacology & Neuroscience
School of Medical Sciences
University Walk
Bristol BS8 1TD
 
[hidden email]
 
 

On 5/02/19, 4:06 PM, "Confocal Microscopy List on behalf of Oshel, Philip Eugene" <[hidden email] on behalf of [hidden email]> wrote:

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Correct, the vacuum is not the issue, but Fomblin and like are still good greases, just pricey. They don't outgas and so won't get gunk on the optics or otherwise contaminate the optics.
   
    I don't think the silicone high vacuum greases are good for lubricating. First, they give off acetic acid vapor - I don't see why Dow would recommend it for optical components. Second, the silicone greases are sticky. Not good for moving parts. It does protect o-rings, but watch the acetic acid and o-ring composition.
    Silicone grease seals coverslips well, though.
   
    Phil
    -------------
    Philip Oshel    
    Imaging Facility Director
    Biology Department
    1304 Biosciences
    1455 Calumet Ct.
    Central Michigan University
    Mt. Pleasant, MI 48859
    989 774-3576 office
    989 774-7567 lab
   
    -----Original Message-----
    From: Confocal Microscopy List <[hidden email]> on behalf of Mark Cannell <[hidden email]>
    Reply-To: Confocal Microscopy List <[hidden email]>
    Date: Tuesday,  05February, 2019 at 11:02
    To: "[hidden email]" <[hidden email]>
    Subject: Re: Oil/grease for microscope?
   
        *****
        To join, leave or search the confocal microscopy listserv, go to:
        http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
        Post images on http://www.imgur.com and include the link in your posting.
        *****
       
        Seems to me that Dow silicone High vacuum grease is a good choice -they recommend it for optical components and being silicone protects o rings well. It's no so sticky. No need to worry about super high vacuum in this application ... ?
       
        HTH
       
        Mark B. Cannell.
        Department of Physiology, Pharmacology & Neuroscience
        School of Medical Sciences
        University Walk
        Bristol BS8 1TD
         
        [hidden email]
         
         
       
        On 5/02/19, 1:23 PM, "Confocal Microscopy List on behalf of Oshel, Philip Eugene" <[hidden email] on behalf of [hidden email]> wrote:
       
            *****
            To join, leave or search the confocal microscopy listserv, go to:
            http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
            Post images on http://www.imgur.com and include the link in your posting.
            *****
           
            Note that for-real vacuum grease is NOT silicone vacuum grease. The silicone grease is only suitable for sealing desiccators.
            Fomblin is probably the best choice of vacuum grease for moving parts. Be aware that any high vacuum grease is high viscosity and can be "sticky". Use a very small amount and apply as a *thin* layer.
            It's also expensive, but all true high-vacuum greases are.
            Also: if you're applying the grease to any polymer or rubber parts, first check it on a bit that doesn't matter. The grease may goo up the polymer over time, especially if the part is warmed or heated.
           
            Phil
            -------------
            Philip Oshel    
            Imaging Facility Director
            Biology Department
            1304 Biosciences
            1455 Calumet Ct.
            Central Michigan University
            Mt. Pleasant, MI 48859
            989 774-3576 office
            989 774-7567 lab
           
            -----Original Message-----
            From: Confocal Microscopy List <[hidden email]> on behalf of Craig Brideau <[hidden email]>
            Reply-To: Confocal Microscopy List <[hidden email]>
            Date: Monday,  04February, 2019 at 17:44
            To: "[hidden email]" <[hidden email]>
            Subject: Re: Oil/grease for microscope?
           
                *****
                To join, leave or search the confocal microscopy listserv, go to:
                http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
                Post images on http://www.imgur.com and include the link in your posting.
                *****
               
                There is grease available for vacuum applications that is very low
                off-gassing. If you use that stuff you will probably be fine. It is
                typically used for moving parts inside vacuum chambers or other
                low-pressure environments.
               
                Craig
               
                On Mon, Feb 4, 2019 at 12:18 PM Piter_ <[hidden email]> wrote:
               
                > *****
                > To join, leave or search the confocal microscopy listserv, go to:
                > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
                > Post images on http://www.imgur.com and include the link in your posting.
                > *****
                >
                > Dear Phil.
                > Lithium grease was the first thing came into my mind. I used to have a big
                > can of Litol for my bike. But then I have  read some scary info that
                > microscope grease has to be very special  or  it will evaporate, leak and
                > destroy your equipment.... So that kind of stopped me from using it. Do you
                > use it for long time? My guess is you had no problems with it :) My
                > intension was to put some dumping grease there. Like Nye A975, which is
                > recommended for focusing mechanism. Maybe it is an overkill.
                >
                > Thanks.
                > Petro.
                >
                > On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <[hidden email] wrote:
                >
                > > *****
                > > To join, leave or search the confocal microscopy listserv, go to:
                > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
                > > Post images on http://www.imgur.com and include the link in your
                > posting.
                > > *****
                > >
                > > Petro,
                > >
                > > I use white lithium grease for this. Available in hardware and auto parts
                > > stores.
                > >
                > > Phil
                > > -------------
                > > Philip Oshel
                > > Imaging Facility Director
                > > Biology Department
                > > 1304 Biosciences
                > > 1455 Calumet Ct.
                > > Central Michigan University
                > > Mt. Pleasant, MI 48859
                > > 989 774-3576 office
                > > 989 774-7567 lab
                > >
                > > -----Original Message-----
                > > From: Confocal Microscopy List <[hidden email]> on
                > > behalf of Piter_ <[hidden email]>
                > > Reply-To: Confocal Microscopy List <[hidden email]>
                > > Date: Sunday,  03February, 2019 at 07:06
                > > To: "[hidden email]" <[hidden email]
                > >
                > > Subject: Oil/grease for microscope?
                > >
                > >     *****
                > >     To join, leave or search the confocal microscopy listserv, go to:
                > >     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
                > >     Post images on http://www.imgur.com and include the link in your
                > > posting.
                > >     *****
                > >
                > >     Hi list.
                > >     I want to change grease in condenser and  binocular tubes of Leica DM
                > > IRE2
                > >     microscope. The old grease solidified and made those parts stuck. Can
                > >     anyone recommend a type and/or supplier of a grease for such purpose?
                > >     Thank you.
                > >     Petro.
                > >
                > >
                > >
                >
               
           
           
       
       
   
   

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I'm guessing you DON'T mean Dow Corning 732, as that is the one that has
acetic acid. As Phil clarified, there is 'vacuum grease', and then there is
'grease for use in a vacuum'. They are not quite the same thing. Vacuum
application servo lubricant grease is safe (grease for use in a vacuum),
whereas the Dow Corning 732 is actually mean for creating a seal in a
vacuum chamber etc. and can off-gas.

Craig

On Tue, Feb 5, 2019 at 9:18 AM Mark Cannell <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> No acetic acid from the Dow corning stuff I've used -it's a specialty
> silicon polymer plus thickener -not polymerized "bathroom caulk" as far as
> I know -more like low MW sylgard I think. While it's exact composition may
> be a trade secret, it's vapour pressure is very low at 100oC so I doubt it
> could be outgassing much, if any, acetic acid. I've not seen it corroding
> brass parts.
>
> Mark B. Cannell.
> Department of Physiology, Pharmacology & Neuroscience
> School of Medical Sciences
> University Walk
> Bristol BS8 1TD
>
> [hidden email]
>
>
>
> On 5/02/19, 4:06 PM, "Confocal Microscopy List on behalf of Oshel, Philip
> Eugene" <[hidden email] on behalf of [hidden email]>
> wrote:
>
>     *****
>     To join, leave or search the confocal microscopy listserv, go to:
>     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>     Post images on http://www.imgur.com and include the link in your
> posting.
>     *****
>
>     Correct, the vacuum is not the issue, but Fomblin and like are still
> good greases, just pricey. They don't outgas and so won't get gunk on the
> optics or otherwise contaminate the optics.
>
>     I don't think the silicone high vacuum greases are good for
> lubricating. First, they give off acetic acid vapor - I don't see why Dow
> would recommend it for optical components. Second, the silicone greases are
> sticky. Not good for moving parts. It does protect o-rings, but watch the
> acetic acid and o-ring composition.
>     Silicone grease seals coverslips well, though.
>
>     Phil
>     -------------
>     Philip Oshel
>     Imaging Facility Director
>     Biology Department
>     1304 Biosciences
>     1455 Calumet Ct.
>     Central Michigan University
>     Mt. Pleasant, MI 48859
>     989 774-3576 office
>     989 774-7567 lab
>
>     -----Original Message-----
>     From: Confocal Microscopy List <[hidden email]> on
> behalf of Mark Cannell <[hidden email]>
>     Reply-To: Confocal Microscopy List <[hidden email]>
>     Date: Tuesday,  05February, 2019 at 11:02
>     To: "[hidden email]" <
> [hidden email]>
>     Subject: Re: Oil/grease for microscope?
>
>         *****
>         To join, leave or search the confocal microscopy listserv, go to:
>         http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>         Post images on http://www.imgur.com and include the link in your
> posting.
>         *****
>
>         Seems to me that Dow silicone High vacuum grease is a good choice
> -they recommend it for optical components and being silicone protects o
> rings well. It's no so sticky. No need to worry about super high vacuum in
> this application ... ?
>
>         HTH
>
>         Mark B. Cannell.
>         Department of Physiology, Pharmacology & Neuroscience
>         School of Medical Sciences
>         University Walk
>         Bristol BS8 1TD
>
>         [hidden email]
>
>
>
>         On 5/02/19, 1:23 PM, "Confocal Microscopy List on behalf of Oshel,
> Philip Eugene" <[hidden email] on behalf of
> [hidden email]> wrote:
>
>             *****
>             To join, leave or search the confocal microscopy listserv, go
> to:
>             http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>             Post images on http://www.imgur.com and include the link in
> your posting.
>             *****
>
>             Note that for-real vacuum grease is NOT silicone vacuum
> grease. The silicone grease is only suitable for sealing desiccators.
>             Fomblin is probably the best choice of vacuum grease for
> moving parts. Be aware that any high vacuum grease is high viscosity and
> can be "sticky". Use a very small amount and apply as a *thin* layer.
>             It's also expensive, but all true high-vacuum greases are.
>             Also: if you're applying the grease to any polymer or rubber
> parts, first check it on a bit that doesn't matter. The grease may goo up
> the polymer over time, especially if the part is warmed or heated.
>
>             Phil
>             -------------
>             Philip Oshel
>             Imaging Facility Director
>             Biology Department
>             1304 Biosciences
>             1455 Calumet Ct.
>             Central Michigan University
>             Mt. Pleasant, MI 48859
>             989 774-3576 office
>             989 774-7567 lab
>
>             -----Original Message-----
>             From: Confocal Microscopy List <
> [hidden email]> on behalf of Craig Brideau <
> [hidden email]>
>             Reply-To: Confocal Microscopy List <
> [hidden email]>
>             Date: Monday,  04February, 2019 at 17:44
>             To: "[hidden email]" <
> [hidden email]>
>             Subject: Re: Oil/grease for microscope?
>
>                 *****
>                 To join, leave or search the confocal microscopy listserv,
> go to:
>                 http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>                 Post images on http://www.imgur.com and include the link
> in your posting.
>                 *****
>
>                 There is grease available for vacuum applications that is
> very low
>                 off-gassing. If you use that stuff you will probably be
> fine. It is
>                 typically used for moving parts inside vacuum chambers or
> other
>                 low-pressure environments.
>
>                 Craig
>
>                 On Mon, Feb 4, 2019 at 12:18 PM Piter_ <[hidden email]>
> wrote:
>
>                 > *****
>                 > To join, leave or search the confocal microscopy
> listserv, go to:
>                 > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>                 > Post images on http://www.imgur.com and include the
> link in your posting.
>                 > *****
>                 >
>                 > Dear Phil.
>                 > Lithium grease was the first thing came into my mind. I
> used to have a big
>                 > can of Litol for my bike. But then I have  read some
> scary info that
>                 > microscope grease has to be very special  or  it will
> evaporate, leak and
>                 > destroy your equipment.... So that kind of stopped me
> from using it. Do you
>                 > use it for long time? My guess is you had no problems
> with it :) My
>                 > intension was to put some dumping grease there. Like Nye
> A975, which is
>                 > recommended for focusing mechanism. Maybe it is an
> overkill.
>                 >
>                 > Thanks.
>                 > Petro.
>                 >
>                 > On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <
> [hidden email] wrote:
>                 >
>                 > > *****
>                 > > To join, leave or search the confocal microscopy
> listserv, go to:
>                 > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>                 > > Post images on http://www.imgur.com and include the
> link in your
>                 > posting.
>                 > > *****
>                 > >
>                 > > Petro,
>                 > >
>                 > > I use white lithium grease for this. Available in
> hardware and auto parts
>                 > > stores.
>                 > >
>                 > > Phil
>                 > > -------------
>                 > > Philip Oshel
>                 > > Imaging Facility Director
>                 > > Biology Department
>                 > > 1304 Biosciences
>                 > > 1455 Calumet Ct.
>                 > > Central Michigan University
>                 > > Mt. Pleasant, MI 48859
>                 > > 989 774-3576 office
>                 > > 989 774-7567 lab
>                 > >
>                 > > -----Original Message-----
>                 > > From: Confocal Microscopy List <
> [hidden email]> on
>                 > > behalf of Piter_ <[hidden email]>
>                 > > Reply-To: Confocal Microscopy List <
> [hidden email]>
>                 > > Date: Sunday,  03February, 2019 at 07:06
>                 > > To: "[hidden email]" <
> [hidden email]
>                 > >
>                 > > Subject: Oil/grease for microscope?
>                 > >
>                 > >     *****
>                 > >     To join, leave or search the confocal microscopy
> listserv, go to:
>                 > >
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>                 > >     Post images on http://www.imgur.com and include
> the link in your
>                 > > posting.
>                 > >     *****
>                 > >
>                 > >     Hi list.
>                 > >     I want to change grease in condenser and
> binocular tubes of Leica DM
>                 > > IRE2
>                 > >     microscope. The old grease solidified and made
> those parts stuck. Can
>                 > >     anyone recommend a type and/or supplier of a
> grease for such purpose?
>                 > >     Thank you.
>                 > >     Petro.
>                 > >
>                 > >
>                 > >
>                 >
>
>
>
>
>
>
>
>
>
Mark Cannell-2 Mark Cannell-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

732 is certainly not high vacuum grease! Here is a spec sheet for Dow Corning high vacuum grease : https://www.emsdiasum.com/microscopy/technical/datasheet/60705.aspx

HTH


Mark B. Cannell
Department of Physiology, Pharmacology & Neuroscience
School of Medical Sciences
University Walk
Bristol BS8 1TD
 
[hidden email]
 
 

On 5/02/19, 4:58 PM, "Confocal Microscopy List on behalf of Craig Brideau" <[hidden email] on behalf of [hidden email]> wrote:

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    I'm guessing you DON'T mean Dow Corning 732, as that is the one that has
    acetic acid. As Phil clarified, there is 'vacuum grease', and then there is
    'grease for use in a vacuum'. They are not quite the same thing. Vacuum
    application servo lubricant grease is safe (grease for use in a vacuum),
    whereas the Dow Corning 732 is actually mean for creating a seal in a
    vacuum chamber etc. and can off-gas.
   
    Craig
   
    On Tue, Feb 5, 2019 at 9:18 AM Mark Cannell <[hidden email]>
    wrote:
   
    > *****
    > To join, leave or search the confocal microscopy listserv, go to:
    > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    > Post images on http://www.imgur.com and include the link in your posting.
    > *****
    >
    > No acetic acid from the Dow corning stuff I've used -it's a specialty
    > silicon polymer plus thickener -not polymerized "bathroom caulk" as far as
    > I know -more like low MW sylgard I think. While it's exact composition may
    > be a trade secret, it's vapour pressure is very low at 100oC so I doubt it
    > could be outgassing much, if any, acetic acid. I've not seen it corroding
    > brass parts.
    >
    > Mark B. Cannell.
    > Department of Physiology, Pharmacology & Neuroscience
    > School of Medical Sciences
    > University Walk
    > Bristol BS8 1TD
    >
    > [hidden email]
    >
    >
    >
    > On 5/02/19, 4:06 PM, "Confocal Microscopy List on behalf of Oshel, Philip
    > Eugene" <[hidden email] on behalf of [hidden email]>
    > wrote:
    >
    >     *****
    >     To join, leave or search the confocal microscopy listserv, go to:
    >     http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    >     Post images on http://www.imgur.com and include the link in your
    > posting.
    >     *****
    >
    >     Correct, the vacuum is not the issue, but Fomblin and like are still
    > good greases, just pricey. They don't outgas and so won't get gunk on the
    > optics or otherwise contaminate the optics.
    >
    >     I don't think the silicone high vacuum greases are good for
    > lubricating. First, they give off acetic acid vapor - I don't see why Dow
    > would recommend it for optical components. Second, the silicone greases are
    > sticky. Not good for moving parts. It does protect o-rings, but watch the
    > acetic acid and o-ring composition.
    >     Silicone grease seals coverslips well, though.
    >
    >     Phil
    >     -------------
    >     Philip Oshel
    >     Imaging Facility Director
    >     Biology Department
    >     1304 Biosciences
    >     1455 Calumet Ct.
    >     Central Michigan University
    >     Mt. Pleasant, MI 48859
    >     989 774-3576 office
    >     989 774-7567 lab
    >
    >     -----Original Message-----
    >     From: Confocal Microscopy List <[hidden email]> on
    > behalf of Mark Cannell <[hidden email]>
    >     Reply-To: Confocal Microscopy List <[hidden email]>
    >     Date: Tuesday,  05February, 2019 at 11:02
    >     To: "[hidden email]" <
    > [hidden email]>
    >     Subject: Re: Oil/grease for microscope?
    >
    >         *****
    >         To join, leave or search the confocal microscopy listserv, go to:
    >         http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    >         Post images on http://www.imgur.com and include the link in your
    > posting.
    >         *****
    >
    >         Seems to me that Dow silicone High vacuum grease is a good choice
    > -they recommend it for optical components and being silicone protects o
    > rings well. It's no so sticky. No need to worry about super high vacuum in
    > this application ... ?
    >
    >         HTH
    >
    >         Mark B. Cannell.
    >         Department of Physiology, Pharmacology & Neuroscience
    >         School of Medical Sciences
    >         University Walk
    >         Bristol BS8 1TD
    >
    >         [hidden email]
    >
    >
    >
    >         On 5/02/19, 1:23 PM, "Confocal Microscopy List on behalf of Oshel,
    > Philip Eugene" <[hidden email] on behalf of
    > [hidden email]> wrote:
    >
    >             *****
    >             To join, leave or search the confocal microscopy listserv, go
    > to:
    >             http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    >             Post images on http://www.imgur.com and include the link in
    > your posting.
    >             *****
    >
    >             Note that for-real vacuum grease is NOT silicone vacuum
    > grease. The silicone grease is only suitable for sealing desiccators.
    >             Fomblin is probably the best choice of vacuum grease for
    > moving parts. Be aware that any high vacuum grease is high viscosity and
    > can be "sticky". Use a very small amount and apply as a *thin* layer.
    >             It's also expensive, but all true high-vacuum greases are.
    >             Also: if you're applying the grease to any polymer or rubber
    > parts, first check it on a bit that doesn't matter. The grease may goo up
    > the polymer over time, especially if the part is warmed or heated.
    >
    >             Phil
    >             -------------
    >             Philip Oshel
    >             Imaging Facility Director
    >             Biology Department
    >             1304 Biosciences
    >             1455 Calumet Ct.
    >             Central Michigan University
    >             Mt. Pleasant, MI 48859
    >             989 774-3576 office
    >             989 774-7567 lab
    >
    >             -----Original Message-----
    >             From: Confocal Microscopy List <
    > [hidden email]> on behalf of Craig Brideau <
    > [hidden email]>
    >             Reply-To: Confocal Microscopy List <
    > [hidden email]>
    >             Date: Monday,  04February, 2019 at 17:44
    >             To: "[hidden email]" <
    > [hidden email]>
    >             Subject: Re: Oil/grease for microscope?
    >
    >                 *****
    >                 To join, leave or search the confocal microscopy listserv,
    > go to:
    >                 http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    >                 Post images on http://www.imgur.com and include the link
    > in your posting.
    >                 *****
    >
    >                 There is grease available for vacuum applications that is
    > very low
    >                 off-gassing. If you use that stuff you will probably be
    > fine. It is
    >                 typically used for moving parts inside vacuum chambers or
    > other
    >                 low-pressure environments.
    >
    >                 Craig
    >
    >                 On Mon, Feb 4, 2019 at 12:18 PM Piter_ <[hidden email]>
    > wrote:
    >
    >                 > *****
    >                 > To join, leave or search the confocal microscopy
    > listserv, go to:
    >                 > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    >                 > Post images on http://www.imgur.com and include the
    > link in your posting.
    >                 > *****
    >                 >
    >                 > Dear Phil.
    >                 > Lithium grease was the first thing came into my mind. I
    > used to have a big
    >                 > can of Litol for my bike. But then I have  read some
    > scary info that
    >                 > microscope grease has to be very special  or  it will
    > evaporate, leak and
    >                 > destroy your equipment.... So that kind of stopped me
    > from using it. Do you
    >                 > use it for long time? My guess is you had no problems
    > with it :) My
    >                 > intension was to put some dumping grease there. Like Nye
    > A975, which is
    >                 > recommended for focusing mechanism. Maybe it is an
    > overkill.
    >                 >
    >                 > Thanks.
    >                 > Petro.
    >                 >
    >                 > On Mon, Feb 4, 2019, 15:23 Oshel, Philip Eugene <
    > [hidden email] wrote:
    >                 >
    >                 > > *****
    >                 > > To join, leave or search the confocal microscopy
    > listserv, go to:
    >                 > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    >                 > > Post images on http://www.imgur.com and include the
    > link in your
    >                 > posting.
    >                 > > *****
    >                 > >
    >                 > > Petro,
    >                 > >
    >                 > > I use white lithium grease for this. Available in
    > hardware and auto parts
    >                 > > stores.
    >                 > >
    >                 > > Phil
    >                 > > -------------
    >                 > > Philip Oshel
    >                 > > Imaging Facility Director
    >                 > > Biology Department
    >                 > > 1304 Biosciences
    >                 > > 1455 Calumet Ct.
    >                 > > Central Michigan University
    >                 > > Mt. Pleasant, MI 48859
    >                 > > 989 774-3576 office
    >                 > > 989 774-7567 lab
    >                 > >
    >                 > > -----Original Message-----
    >                 > > From: Confocal Microscopy List <
    > [hidden email]> on
    >                 > > behalf of Piter_ <[hidden email]>
    >                 > > Reply-To: Confocal Microscopy List <
    > [hidden email]>
    >                 > > Date: Sunday,  03February, 2019 at 07:06
    >                 > > To: "[hidden email]" <
    > [hidden email]
    >                 > >
    >                 > > Subject: Oil/grease for microscope?
    >                 > >
    >                 > >     *****
    >                 > >     To join, leave or search the confocal microscopy
    > listserv, go to:
    >                 > >
    > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    >                 > >     Post images on http://www.imgur.com and include
    > the link in your
    >                 > > posting.
    >                 > >     *****
    >                 > >
    >                 > >     Hi list.
    >                 > >     I want to change grease in condenser and
    > binocular tubes of Leica DM
    >                 > > IRE2
    >                 > >     microscope. The old grease solidified and made
    > those parts stuck. Can
    >                 > >     anyone recommend a type and/or supplier of a
    > grease for such purpose?
    >                 > >     Thank you.
    >                 > >     Petro.
    >                 > >
    >                 > >
    >                 > >
    >                 >
    >
    >
    >
    >
    >
    >
    >
    >
    >
   

Wai Chan Wai Chan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

In reply to this post by Piter_
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Petro

You can try Losimol (http://www.losimol.de/en/products/#microscopes). Just as Dan Focht pointed out, there are many options to choose from. Cheers.
Piter_ Piter_
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Oil/grease for microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi list.
One more trick with solidified grease.
If things do not move sometimes heating the mic part with hair dryer helps.
Best.
Petro.



On Thu, Feb 7, 2019, 07:48 Wai Chan <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Petro
>
> You can try Losimol (http://www.losimol.de/en/products/#microscopes).
> Just as Dan Focht pointed out, there are many options to choose from.
> Cheers.
>