Dear All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria


____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University

George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

In reply to this post by muxika
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Maria,

this is going to be very objective lens dependent. Most commercial "through objective" TIRF (o-TIRF) systems do not have UV (such as 355 nm) lasers, so pretty much moot. Three suggestions:

* I suggest you meet with Jennifer Waters, Director of your (Harvard's) Nikon Imaging Center, to discuss your needs,

http://nic.med.harvard.edu/contact

* You might also want to re-think about whether you even need UV ... STORM is much higher spatial precision localization than can be achieved with TIRF optical resolution -- Dr. STORM, aka Prof. Xiaowei Zhuang, is at Harvard, http://zhuang.harvard.edu

* If you really need UV TIRF, I encourage you to discuss this directly with Dr. Alex Asanov, who I am copying, see

http://www.tirf-labs.com/tirfmicroscopy.html

UV-absorbing-blue-light-emitting fluorophores. p-TIRF and lg-TIRF systems can be used with UV excitation, a feature not available in o-TIRF. There is a new class of fluorescent probes with large Stokes shift that are excited by UV light and emit fluorescence in the range 360-450 nm.  These fluorophores, including modified nucleic and amino acids possess great potential for life science and material science applications. p-TIRF and lg-TIRF systems are well suited for TIRFing such fluorophores.

A big advantage of Alex's approach's (prism or light guide) is you do not need a $20K TIRF objective lens. This would save you a lot of money, since UV transmitting (assuming you really need to collect UV emission) is likely a lot less expensive (I suspect TIRF excitation lenses have even more glass than standard high NA objective lenses, every lens is likely 2% or more decrease in transmission ... see also QE curves of conventional sCMOS and back-illuminated sCMOS, EMCCDs, CCDs for UV performance). The money you save by going with either "Alex TIRF" approaches, could also help pay for the UV laser(s) and optionally an LDI laser rack -- see preliminary doc at https://www.cairn-research.co.uk/wp-content/uploads/2016/03/A4-Leaflet-2.pdf   ... http://www.89north.com would be selling the U.S. version (plus would want appropriate emission filters).

best wishes,

George

p.s. if you get this working, please let me know how well BUV395 and BV421 monoclonal antibodies work on it (Brilliant Ultraviolet and Brilliant Violet - see BD Biosciences).




On 12/22/2016 8:36 PM, Maria Mukhina wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria

Maria Mukhina

Postdoctoral Fellow
Kleckner Lab
Department of Molecular and Cellular Biology
Harvard University

-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

0000001ed7f52e4a-dmarc-request 0000001ed7f52e4a-dmarc-request
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

In reply to this post by muxika
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Dear Maria,
Olympus has the 4 laser TIRF module where you can adjust the collimation and TIRF angle individually for each laser line. I think this is unique and it allows fast imaging in multi colours when combined with a camera splitter or even fast filter wheel. They publish the transmission curves of the objectives. I have no experience with Nikon.

Best wishes

Andreas
 

From: [hidden email]
Sent: ‎23/‎12/‎2016 02:49
To: [hidden email]
Subject: Olympus vs Nikon

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Dear All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria


____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University

Dr. K N Ganesh Dr. K N Ganesh
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

In reply to this post by muxika
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
There are couple of high transmission UV optics with Olympus dedicated for Fura imaging. You may not good TIRF because of lower NA.
Regards
Ganesh

On Dec 23, 2016 8:19 AM, "Maria Mukhina" <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria

Maria Mukhina

Postdoctoral Fellow
Kleckner Lab
Department of Molecular and Cellular Biology
Harvard University
Dr. K N Ganesh Dr. K N Ganesh
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

In reply to this post by muxika
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Sorry to repost..
You have couple of cool motorised TIRF from Olympus and a motorised drift compensator called IX3-ZDC2 ... You can check the same in their websites. One of the TIRF is unique in terms providing 4 parallel laser lines (with many power options) helps you to set same depth for all colors with all lasers focused to the BFP of the objective. You can also find High NA (UpTo 1.7) and a high mag (150X) objectives if you are using 16microm pixel EMCCD.
Reg
Ganesh

On Dec 23, 2016 8:19 AM, "Maria Mukhina" <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria

Maria Mukhina

Postdoctoral Fellow
Kleckner Lab
Department of Molecular and Cellular Biology
Harvard University
Alex Asanov Alex Asanov
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon Commercial response

In reply to this post by muxika
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi Maria,
 
We have compared several TIRF objectives using 638, 532, and 465 nm lines. In agreement with the literature we found that the intensitity of stray light was 15% or more at the TIRF surface relative to the intensity of the evanescent wave. Out of 13 objectives tested Nikon lenses produced less stray light. However, it is not a conclusive observation, since the number 13 is small. Make a note that the contamination increases exponentially, since the EW decays exponentially. 
Just a few measurements we performed with 405 nm excitation to realize that the intensity of autofluorescence and scatter were too high.

 

The good news is that prism- and lightguide-based geometries are well suited for TIRFing with UV excitation: http://tirf-labs.com/Select_TIRF_geometry_WP.pdf.

 

Best regards,

Alexander Asanov, Ph.D.
President,  TIRF Labs
Cary, NC 27519
TIRF-Labs.com; TIRFmicroscopy.com 
[hidden email]

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mukhina, Maria
Sent: Thursday, December 22, 2016 10:49 PM
To: [hidden email]
Subject: Olympus vs Nikon

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Dear All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria


____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University

Alex Asanov Alex Asanov
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon Commercial response

In reply to this post by muxika
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
 
Hi Maria,
 
We have compared several TIRF objectives using 638, 532, and 465 nm lines. In agreement with the literature we found that the intensitity of stray light was 15% or more at the TIRF surface relative to the intensity of the evanescent wave. Out of 13 objectives tested Nikon lenses produced less stray light. However, it is not a conclusive observation, since the number 13 is small. Make a note that the contamination increases exponentially, since the EW decays exponentially. 
Just a few measurements we performed with 405 nm excitation to realize that the intensity of autofluorescence and scatter were too high.

 

The good news is that prism- and lightguide-based geometries are well suited for TIRFing with UV excitation: http://tirf-labs.com/Select_TIRF_geometry_WP.pdf.

 

Best regards,

Alexander Asanov, Ph.D.
President,  TIRF Labs
Cary, NC 27519
TIRF-Labs.com; TIRFmicroscopy.com 
[hidden email]

 
 P.S. At the moment I am traveling outside the U.S. It appears that some functions at my email system are blocked. Sorry, if I produce dublicate posting 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mukhina, Maria
Sent: Thursday, December 22, 2016 10:49 PM
To: [hidden email]
Subject: Olympus vs Nikon

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Dear All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria


____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University

muxika muxika
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

In reply to this post by 0000001ed7f52e4a-dmarc-request
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi All,


Thanks a lot for your replies. TIRF is not our priority in this setup. Mostly we are interested in maximal transmission efficiency and high optics threshold. Also we are quite flexible with excitation wavelength in UV. Though the best option is to excite at 300 nm, the excitation wavelength can be shifted to 330-350 nm, and emission wavelength is in the visible range. Since excitation at 300 nm requires very custom optics, we basically excluded this option and are considering using more conventional objectives. It looks like Olympus offers more options for UV imaging, and we are considering UAPON 100XOTIRF objective. In this regard, it would be very helpful to know its power threshold at UV range, especially in TIRF mode.


Another important feature is automatic focus system. We currently use Nikon's Perfect Focus System and are satisfied with results. @Ganesh: Have you tried using Olympus drift compensator with living specimens?


thanks,

Maria


____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University


________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Andreas Bruckbauer <[hidden email]>
Sent: Friday, December 23, 2016 3:15:41 AM
To: [hidden email]
Subject: Re: Olympus vs Nikon

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=CwMFaQ&c=WO-RGvefibhHBZq3fL85hQ&r=lvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas&m=VQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g&s=5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4&e=> Post images on http://www.imgur.com<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=CwMFaQ&c=WO-RGvefibhHBZq3fL85hQ&r=lvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas&m=VQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g&s=s8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg&e=> and include the link in your posting. *****
Dear Maria,
Olympus has the 4 laser TIRF module where you can adjust the collimation and TIRF angle individually for each laser line. I think this is unique and it allows fast imaging in multi colours when combined with a camera splitter or even fast filter wheel. They publish the transmission curves of the objectives. I have no experience with Nikon.

Best wishes

Andreas

________________________________
From: Mukhina, Maria<mailto:[hidden email]>
Sent: ?23/?12/?2016 02:49
To: [hidden email]<mailto:[hidden email]>
Subject: Olympus vs Nikon

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=CwMFaQ&c=WO-RGvefibhHBZq3fL85hQ&r=lvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas&m=VQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g&s=5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4&e=> Post images on http://www.imgur.com<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=CwMFaQ&c=WO-RGvefibhHBZq3fL85hQ&r=lvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas&m=VQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g&s=s8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg&e=> and include the link in your posting. *****

Dear All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria



____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Maria,

We use Olympus UPlanSApo 20/0.75 (or an old 40x oil) for 340 excitation. At this wavelength, some slides can be very autofluorescent, so it's better to use an open coverslipped dish.

Happy holidays!

Mike

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mukhina, Maria
Sent: Tuesday, December 27, 2016 1:58 PM
To: [hidden email]
Subject: Re: Olympus vs Nikon

*****
To join, leave or search the confocal microscopy listserv, go to:
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0
Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0 and include the link in your posting.
*****

Hi All,


Thanks a lot for your replies. TIRF is not our priority in this setup. Mostly we are interested in maximal transmission efficiency and high optics threshold. Also we are quite flexible with excitation wavelength in UV. Though the best option is to excite at 300 nm, the excitation wavelength can be shifted to 330-350 nm, and emission wavelength is in the visible range. Since excitation at 300 nm requires very custom optics, we basically excluded this option and are considering using more conventional objectives. It looks like Olympus offers more options for UV imaging, and we are considering UAPON 100XOTIRF objective. In this regard, it would be very helpful to know its power threshold at UV range, especially in TIRF mode.


Another important feature is automatic focus system. We currently use Nikon's Perfect Focus System and are satisfied with results. @Ganesh: Have you tried using Olympus drift compensator with living specimens?


thanks,

Maria


____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University


________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Andreas Bruckbauer <[hidden email]>
Sent: Friday, December 23, 2016 3:15:41 AM
To: [hidden email]
Subject: Re: Olympus vs Nikon

***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. ***** Dear Maria, Olympus has the 4 laser TIRF module where you can adjust the collimation and TIRF angle individually for each laser line. I think this is unique and it allows fast imaging in multi colours when combined with a camera splitter or even fast filter wheel. They publish the transmission curves of the objectives. I have no experience with Nikon.

Best wishes

Andreas

________________________________
From: Mukhina, Maria<mailto:[hidden email]>
Sent: ?23/?12/?2016 02:49
To: [hidden email]<mailto:[hidden email]>
Subject: Olympus vs Nikon

***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. *****

Dear All,

I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.

thanks in advance,
Maria



____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University
Dr. K N Ganesh Dr. K N Ganesh
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Maria
I have used the n IX3-ZDC which goes with IX83 inverted scope. It works fine both in single shot and continuous mode. I have also seen people using it more than 24 hrs without any issues.
Regards
Ganesh

Sent from my iPhone

> On 28-Dec-2016, at 12:37 AM, MODEL, MICHAEL <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Maria,
>
> We use Olympus UPlanSApo 20/0.75 (or an old 40x oil) for 340 excitation. At this wavelength, some slides can be very autofluorescent, so it's better to use an open coverslipped dish.
>
> Happy holidays!
>
> Mike
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mukhina, Maria
> Sent: Tuesday, December 27, 2016 1:58 PM
> To: [hidden email]
> Subject: Re: Olympus vs Nikon
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0
> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0 and include the link in your posting.
> *****
>
> Hi All,
>
>
> Thanks a lot for your replies. TIRF is not our priority in this setup. Mostly we are interested in maximal transmission efficiency and high optics threshold. Also we are quite flexible with excitation wavelength in UV. Though the best option is to excite at 300 nm, the excitation wavelength can be shifted to 330-350 nm, and emission wavelength is in the visible range. Since excitation at 300 nm requires very custom optics, we basically excluded this option and are considering using more conventional objectives. It looks like Olympus offers more options for UV imaging, and we are considering UAPON 100XOTIRF objective. In this regard, it would be very helpful to know its power threshold at UV range, especially in TIRF mode.
>
>
> Another important feature is automatic focus system. We currently use Nikon's Perfect Focus System and are satisfied with results. @Ganesh: Have you tried using Olympus drift compensator with living specimens?
>
>
> thanks,
>
> Maria
>
>
> ____
>
>
> Maria Mukhina
>
>
> Postdoctoral Fellow
>
> Kleckner Lab
>
> Department of Molecular and Cellular Biology
>
> Harvard University
>
>
> ________________________________
> From: Confocal Microscopy List <[hidden email]> on behalf of Andreas Bruckbauer <[hidden email]>
> Sent: Friday, December 23, 2016 3:15:41 AM
> To: [hidden email]
> Subject: Re: Olympus vs Nikon
>
> ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. ***** Dear Maria, Olympus has the 4 laser TIRF module where you can adjust the collimation and TIRF angle individually for each laser line. I think this is unique and it allows fast imaging in multi colours when combined with a camera splitter or even fast filter wheel. They publish the transmission curves of the objectives. I have no experience with Nikon.
>
> Best wishes
>
> Andreas
>
> ________________________________
> From: Mukhina, Maria<mailto:[hidden email]>
> Sent: ?23/?12/?2016 02:49
> To: [hidden email]<mailto:[hidden email]>
> Subject: Olympus vs Nikon
>
> ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. *****
>
> Dear All,
>
> I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.
>
> thanks in advance,
> Maria
>
>
>
> ____
>
>
> Maria Mukhina
>
>
> Postdoctoral Fellow
>
> Kleckner Lab
>
> Department of Molecular and Cellular Biology
>
> Harvard University
muxika muxika
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi All,


@Mike Could you please tell what source you use at 340 nm?


@Ganesh Thank you.


Happy New Year!


Maria

____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University


________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ganesh Kadasoor <[hidden email]>
Sent: Wednesday, December 28, 2016 1:33:31 AM
To: [hidden email]
Subject: Re: Olympus vs Nikon

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Maria
I have used the n IX3-ZDC which goes with IX83 inverted scope. It works fine both in single shot and continuous mode. I have also seen people using it more than 24 hrs without any issues.
Regards
Ganesh

Sent from my iPhone

> On 28-Dec-2016, at 12:37 AM, MODEL, MICHAEL <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Maria,
>
> We use Olympus UPlanSApo 20/0.75 (or an old 40x oil) for 340 excitation. At this wavelength, some slides can be very autofluorescent, so it's better to use an open coverslipped dish.
>
> Happy holidays!
>
> Mike
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mukhina, Maria
> Sent: Tuesday, December 27, 2016 1:58 PM
> To: [hidden email]
> Subject: Re: Olympus vs Nikon
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0
> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0 and include the link in your posting.
> *****
>
> Hi All,
>
>
> Thanks a lot for your replies. TIRF is not our priority in this setup. Mostly we are interested in maximal transmission efficiency and high optics threshold. Also we are quite flexible with excitation wavelength in UV. Though the best option is to excite at 300 nm, the excitation wavelength can be shifted to 330-350 nm, and emission wavelength is in the visible range. Since excitation at 300 nm requires very custom optics, we basically excluded this option and are considering using more conventional objectives. It looks like Olympus offers more options for UV imaging, and we are considering UAPON 100XOTIRF objective. In this regard, it would be very helpful to know its power threshold at UV range, especially in TIRF mode.
>
>
> Another important feature is automatic focus system. We currently use Nikon's Perfect Focus System and are satisfied with results. @Ganesh: Have you tried using Olympus drift compensator with living specimens?
>
>
> thanks,
>
> Maria
>
>
> ____
>
>
> Maria Mukhina
>
>
> Postdoctoral Fellow
>
> Kleckner Lab
>
> Department of Molecular and Cellular Biology
>
> Harvard University
>
>
> ________________________________
> From: Confocal Microscopy List <[hidden email]> on behalf of Andreas Bruckbauer <[hidden email]>
> Sent: Friday, December 23, 2016 3:15:41 AM
> To: [hidden email]
> Subject: Re: Olympus vs Nikon
>
> ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. ***** Dear Maria, Olympus has the 4 laser TIRF module where you can adjust the collimation and TIRF angle individually for each laser line. I think this is unique and it allows fast imaging in multi colours when combined with a camera splitter or even fast filter wheel. They publish the transmission curves of the objectives. I have no experience with Nikon.
>
> Best wishes
>
> Andreas
>
> ________________________________
> From: Mukhina, Maria<mailto:[hidden email]>
> Sent: ?23/?12/?2016 02:49
> To: [hidden email]<mailto:[hidden email]>
> Subject: Olympus vs Nikon
>
> ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. *****
>
> Dear All,
>
> I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.
>
> thanks in advance,
> Maria
>
>
>
> ____
>
>
> Maria Mukhina
>
>
> Postdoctoral Fellow
>
> Kleckner Lab
>
> Department of Molecular and Cellular Biology
>
> Harvard University
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Olympus vs Nikon

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****


@Mike Could you please tell what source you use at 340 nm?

Mercury lamp


@Ganesh Thank you.


Happy New Year!


Maria

____


Maria Mukhina


Postdoctoral Fellow

Kleckner Lab

Department of Molecular and Cellular Biology

Harvard University


________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ganesh Kadasoor <[hidden email]>
Sent: Wednesday, December 28, 2016 1:33:31 AM
To: [hidden email]
Subject: Re: Olympus vs Nikon

*****
To join, leave or search the confocal microscopy listserv, go to:
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7Caef207ea4b3b4a11252c08d43011a82f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=58x5hGO%2FsgexFf8PSsgqaUV4sOUlue3rlU8YB29c%2FoY%3D&reserved=0
Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7Caef207ea4b3b4a11252c08d43011a82f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=ImpQQwwwrdPppeNdlW5%2FV1cHsoq3J%2BzrN%2Fd80fdHzCQ%3D&reserved=0 and include the link in your posting.
*****

Dear Maria
I have used the n IX3-ZDC which goes with IX83 inverted scope. It works fine both in single shot and continuous mode. I have also seen people using it more than 24 hrs without any issues.
Regards
Ganesh

Sent from my iPhone

> On 28-Dec-2016, at 12:37 AM, MODEL, MICHAEL <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7Caef207ea4b3b4a11252c08d43011a82f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=58x5hGO%2FsgexFf8PSsgqaUV4sOUlue3rlU8YB29c%2FoY%3D&reserved=0
> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7Caef207ea4b3b4a11252c08d43011a82f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=ImpQQwwwrdPppeNdlW5%2FV1cHsoq3J%2BzrN%2Fd80fdHzCQ%3D&reserved=0 and include the link in your posting.
> *****
>
> Hi Maria,
>
> We use Olympus UPlanSApo 20/0.75 (or an old 40x oil) for 340 excitation. At this wavelength, some slides can be very autofluorescent, so it's better to use an open coverslipped dish.
>
> Happy holidays!
>
> Mike
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mukhina, Maria
> Sent: Tuesday, December 27, 2016 1:58 PM
> To: [hidden email]
> Subject: Re: Olympus vs Nikon
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0
> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0 and include the link in your posting.
> *****
>
> Hi All,
>
>
> Thanks a lot for your replies. TIRF is not our priority in this setup. Mostly we are interested in maximal transmission efficiency and high optics threshold. Also we are quite flexible with excitation wavelength in UV. Though the best option is to excite at 300 nm, the excitation wavelength can be shifted to 330-350 nm, and emission wavelength is in the visible range. Since excitation at 300 nm requires very custom optics, we basically excluded this option and are considering using more conventional objectives. It looks like Olympus offers more options for UV imaging, and we are considering UAPON 100XOTIRF objective. In this regard, it would be very helpful to know its power threshold at UV range, especially in TIRF mode.
>
>
> Another important feature is automatic focus system. We currently use Nikon's Perfect Focus System and are satisfied with results. @Ganesh: Have you tried using Olympus drift compensator with living specimens?
>
>
> thanks,
>
> Maria
>
>
> ____
>
>
> Maria Mukhina
>
>
> Postdoctoral Fellow
>
> Kleckner Lab
>
> Department of Molecular and Cellular Biology
>
> Harvard University
>
>
> ________________________________
> From: Confocal Microscopy List <[hidden email]> on behalf of Andreas Bruckbauer <[hidden email]>
> Sent: Friday, December 23, 2016 3:15:41 AM
> To: [hidden email]
> Subject: Re: Olympus vs Nikon
>
> ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. ***** Dear Maria, Olympus has the 4 laser TIRF module where you can adjust the collimation and TIRF angle individually for each laser line. I think this is unique and it allows fast imaging in multi colours when combined with a camera splitter or even fast filter wheel. They publish the transmission curves of the objectives. I have no experience with Nikon.
>
> Best wishes
>
> Andreas
>
> ________________________________
> From: Mukhina, Maria<mailto:[hidden email]>
> Sent: ?23/?12/?2016 02:49
> To: [hidden email]<mailto:[hidden email]>
> Subject: Olympus vs Nikon
>
> ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=zQTnORo1bprZncYw%2BR3kw3LFLQYTiePWzqtO%2FWNMW6E%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3D5u0Szuar1C82RB3RPtJ-M_0nroa8j1MQHVbCEPZ9QZ4%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=YLKO3n7IMVOkyOobeT3DLnnVBOJX6xGdjeNvwQxsYSc%3D&reserved=0=> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=yqQIkHZ2VbKnHDfSzeqc%2FXShoAstahva2OHqenbJW0I%3D&reserved=0<https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DCwMFaQ%26c%3DWO-RGvefibhHBZq3fL85hQ%26r%3Dlvsgf9vctNrfAxRUoAziiQY-eNA2rMSgY-3fOZjtYas%26m%3DVQ87tNDZ4wzO9DfQjPsEPEyrALO87R2NN1VXVptSF3g%26s%3Ds8nje2bpAS463c91VsSOkQnoGn4gGCTZJrnY5YjU9tg%26e&data=01%7C01%7Cmmodel%40KENT.EDU%7C55f9a4e29d4942e5052508d42e8a5179%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata=i09q85%2FthtbPVgOqcWt3hsJo6vhByv04qkqxp53xRkE%3D&reserved=0=> and include the link in your posting. *****
>
> Dear All,
>
> I wonder if anyone can share any experience in regard to relative transmission efficiency/optics threshold (especially in 300-400 nm UV)/software/automatic focus systems/TIRF modules of Olympus and Nikon microscopes. We are considering using microscopes of these two manufacturers as a base for custom setup and would appreciate any comments you might have.
>
> thanks in advance,
> Maria
>
>
>
> ____
>
>
> Maria Mukhina
>
>
> Postdoctoral Fellow
>
> Kleckner Lab
>
> Department of Molecular and Cellular Biology
>
> Harvard University