PSF of beads far way from glass surface

classic Classic list List threaded Threaded
♦
♦
12 messages Options
Volodymyr Nechyporuk-Zloy Volodymyr Nechyporuk-Zloy
Reply | Threaded
Open this post in threaded view
♦
♦
|

PSF of beads far way from glass surface

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Colleagues,

I would like to acquire PSF of beads far away from glass surface, e.g. 50-100 um, in ibidi chambers. What transparent hard gel or any other media should I use to reduce [eliminate] Brownian motion?


Best regards,
Volodymyr

Dr. Volodymyr Nechyporuk-Zloy (Imaging Facility Manager)

My office: 668.30.06D

The Kennedy Institute of Rheumatology
University of Oxford,
Roosevelt Drive, Headington,
Oxford, OX3 7FY,
United Kingdom


Office: (+44) 1865612665
E-mails: [hidden email] [hidden email]
Skype: vzloy1work
Professional Web: http://focus-on-single-molecule.info
Department Web: http://www.kennedy.ox.ac.uk/
Eric Marino Eric Marino
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Good Morning Volodymyr,

Try suspending them in 1% agarose.

Eric Marino
[hidden email]

On 9/16/16, 7:03 AM, "Confocal Microscopy List on behalf of Volodymyr Nechyporuk-Zloy" <[hidden email] on behalf of [hidden email]> wrote:

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Dear Colleagues,
   
    I would like to acquire PSF of beads far away from glass surface, e.g. 50-100 um, in ibidi chambers. What transparent hard gel or any other media should I use to reduce [eliminate] Brownian motion?
   
   
    Best regards,
    Volodymyr
   
    Dr. Volodymyr Nechyporuk-Zloy (Imaging Facility Manager)
   
    My office: 668.30.06D
   
    The Kennedy Institute of Rheumatology
    University of Oxford,
    Roosevelt Drive, Headington,
    Oxford, OX3 7FY,
    United Kingdom
   
   
    Office: (+44) 1865612665
    E-mails: [hidden email] [hidden email]
    Skype: vzloy1work
    Professional Web: http://focus-on-single-molecule.info
    Department Web: http://www.kennedy.ox.ac.uk/
   
Carina Monico Carina Monico
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

In reply to this post by Volodymyr Nechyporuk-Zloy
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Volodymyr,

you could try with agarose. You can find a protocol in this paper: www.nature.com/articles/srep16088  

Carina
________________________________
________________________________
Dr. Carina Monico
Postdoctoral Research Associate
Chemistry Research Laboratory
12 Mansfield Road
University of Oxford
Oxford
OX1 3TA

T: 01865 275486
E: [hidden email]

________________________________________
From: Confocal Microscopy List [[hidden email]] on behalf of Volodymyr Nechyporuk-Zloy [[hidden email]]
Sent: Friday, September 16, 2016 12:03 PM
To: [hidden email]
Subject: PSF of beads far way from glass surface

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Colleagues,

I would like to acquire PSF of beads far away from glass surface, e.g. 50-100 um, in ibidi chambers. What transparent hard gel or any other media should I use to reduce [eliminate] Brownian motion?


Best regards,
Volodymyr

Dr. Volodymyr Nechyporuk-Zloy (Imaging Facility Manager)

My office: 668.30.06D

The Kennedy Institute of Rheumatology
University of Oxford,
Roosevelt Drive, Headington,
Oxford, OX3 7FY,
United Kingdom


Office: (+44) 1865612665
E-mails: [hidden email] [hidden email]
Skype: vzloy1work
Professional Web: http://focus-on-single-molecule.info
Department Web: http://www.kennedy.ox.ac.uk/
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

In reply to this post by Volodymyr Nechyporuk-Zloy
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Depends on what you want to have in between. Beads can be attached to glass on the top surface (I am not sure if the geometry of your chambers allow that) from 1% polyvinyl acetate in methanol and then you can add water, oil or nothing (Journal ofMicroscopy, Vol. 241, Pt 1 2011, pp. 94–100)





From: Confocal Microscopy List <[hidden email]> on behalf of Volodymyr Nechyporuk-Zloy <[hidden email]>
Sent: Friday, September 16, 2016 7:03 AM
To: [hidden email]
Subject: PSF of beads far way from glass surface
 
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

lists.umn.edu
[hidden email]: listserv archives. confocalmicroscopy



Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Colleagues,

I would like to acquire PSF of beads far away from glass surface, e.g. 50-100 um, in ibidi chambers. What transparent hard gel or any other media should I use to reduce [eliminate] Brownian motion?


Best regards,
Volodymyr

Dr. Volodymyr Nechyporuk-Zloy (Imaging Facility Manager)

My office: 668.30.06D

The Kennedy Institute of Rheumatology
University of Oxford,
Roosevelt Drive, Headington,
Oxford, OX3 7FY,
United Kingdom


Office: (+44) 1865612665
E-mails: [hidden email] [hidden email]
Skype: vzloy1work
Professional Web: http://focus-on-single-molecule.info
Department Web: http://www.kennedy.ox.ac.uk/
Cvic Innocent Cvic Innocent
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hey Volodymyr

For a full matrix of beads at various depths I suggest using CyGel. We have a bit over at Micron so if you'd care to testdrive then come over.

Cordially,
cvic

Cassandravictoria Innocent, PhD
Micron ABU // Schermelleh Group
Dept of Biochemistry
Oxford University

On Fri, Sep 16, 2016 at 1:13 PM, MODEL, MICHAEL <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Depends on what you want to have in between. Beads can be attached to glass on the top surface (I am not sure if the geometry of your chambers allow that) from 1% polyvinyl acetate in methanol and then you can add water, oil or nothing (Journal ofMicroscopy, Vol. 241, Pt 1 2011, pp. 94–100)





From: Confocal Microscopy List <[hidden email]> on behalf of Volodymyr Nechyporuk-Zloy <[hidden email]>
Sent: Friday, September 16, 2016 7:03 AM
To: [hidden email]
Subject: PSF of beads far way from glass surface
 
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy



Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Colleagues,

I would like to acquire PSF of beads far away from glass surface, e.g. 50-100 um, in ibidi chambers. What transparent hard gel or any other media should I use to reduce [eliminate] Brownian motion?


Best regards,
Volodymyr

Dr. Volodymyr Nechyporuk-Zloy (Imaging Facility Manager)

My office: 668.30.06D

The Kennedy Institute of Rheumatology
University of Oxford,
Roosevelt Drive, Headington,
Oxford, OX3 7FY,
United Kingdom


Office: <a href="tel:%28%2B44%29%201865612665" value="+441865612665" target="_blank">(+44) 1865612665
E-mails: [hidden email] [hidden email]
Skype: vzloy1work
Professional Web: http://focus-on-single-molecule.info
Department Web: http://www.kennedy.ox.ac.uk/

Volodymyr Nechyporuk-Zloy Volodymyr Nechyporuk-Zloy
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

In reply to this post by Volodymyr Nechyporuk-Zloy
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I have tried CyGel. It works to immobilise cells, but for beads it does not work - at least, under room temperature. The beads are moving.
Michael Giacomelli Michael Giacomelli
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
I've used both PDMS and optical adhesive to immobilize beads.  The resulting phantoms are then stable indefinitely.

Mike 

On Fri, Sep 16, 2016 at 9:11 AM, Volodymyr Nechyporuk-Zloy <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I have tried CyGel. It works to immobilise cells, but for beads it does not work - at least, under room temperature. The beads are moving.

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Sylgard completely solidifies and is fairly clear. Not sure what its exact index is but it seems to work well.

Craig

On Fri, Sep 16, 2016 at 11:05 AM, Michael Giacomelli <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
I've used both PDMS and optical adhesive to immobilize beads.  The resulting phantoms are then stable indefinitely.

Mike 

On Fri, Sep 16, 2016 at 9:11 AM, Volodymyr Nechyporuk-Zloy <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I have tried CyGel. It works to immobilise cells, but for beads it does not work - at least, under room temperature. The beads are moving.


Stanislav Vitha-2 Stanislav Vitha-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

In reply to this post by Volodymyr Nechyporuk-Zloy
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I normally suspend the beads in  2% agarose, but sometimes the beads still have some limited motion.
If you need a higher refractive index than just water, you could suspend the beads in gelatin or acrylamide  and when solidified in imaging chamber, infiltrate with whatever mounting media you need.


Stan

Stanislav Vitha
Microscopy and Imaging Center
Texas A&M University

 
On Fri, 16 Sep 2016 06:03:15 -0500, Volodymyr Nechyporuk-Zloy <[hidden email]> wrote:

>
>Dear Colleagues,
>
>I would like to acquire PSF of beads far away from glass surface, e.g. 50-100 um, in ibidi chambers. What transparent hard gel or any other media should I use to reduce [eliminate] Brownian motion?
>
>
>Best regards,
>Volodymyr
>
>
Zdenek Svindrych-2 Zdenek Svindrych-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Agarose is quite forgiving, it will solidify in a range of media, including quite high concentrations (70% v/v) of glycerol or TDE. Cooling below room temperature may be needed for better jelling. I haven't tried something like 95% TDE, but once I've left a piece of TDE-agarose dry out on the desk, most of the water evaporated, and I got nice piece of jelly with n=1.51...
Best, zdenek

Zdenek Svindrych, Ph.D.
W.M. Keck Center for Cellular Imaging (PLSB 003)
University of Virginia, Charlottesville, VA 
http://www.kcci.virginia.edu/
tel: 434-982-4869
Annual FRET Workshop: http://kcci.virginia.edu/workshop-2017


---------- Původní zpráva ----------
Od: Stanislav Vitha <[hidden email]>
Komu: [hidden email]
Datum: 19. 9. 2016 11:04:32
Předmět: Re: PSF of beads far way from glass surface


*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I normally suspend the beads in 2% agarose, but sometimes the beads still have some limited motion.
If you need a higher refractive index than just water, you could suspend the beads in gelatin or acrylamide and when solidified in imaging chamber, infiltrate with whatever mounting media you need.


Stan

Stanislav Vitha
Microscopy and Imaging Center
Texas A&M University


On Fri, 16 Sep 2016 06:03:15 -0500, Volodymyr Nechyporuk-Zloy <[hidden email]> wrote:

>
>Dear Colleagues,
>
>I would like to acquire PSF of beads far away from glass surface, e.g. 50-100 um, in ibidi chambers. What transparent hard gel or any other media should I use to reduce [eliminate] Brownian motion?
>
>
>Best regards,
>Volodymyr
>
>
Volodymyr Nechyporuk-Zloy Volodymyr Nechyporuk-Zloy
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

In reply to this post by Volodymyr Nechyporuk-Zloy
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Michael,

What adhesive have you used?

Best regards,
Volodymyr
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF of beads far way from glass surface

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Volodymyr,


Polyvinyl acetate is a polymer soluble in methanol but not in water. You can mix beads into 1% PVA in methanol, then put a drop of it on a glass slide/coverslip and let it dry. Then the beads remain firmly trapped in a thin film of polymer stuck to the slide and will stay there after you add a liquid (any liquid that will not dissolve PVA).


Best regards


Mike





From: Confocal Microscopy List <[hidden email]> on behalf of Volodymyr Nechyporuk-Zloy <[hidden email]>
Sent: Tuesday, September 20, 2016 6:18 AM
To: [hidden email]
Subject: Re: PSF of beads far way from glass surface
 
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
lists.umn.edu
[hidden email]: listserv archives. confocalmicroscopy


Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Michael,

What adhesive have you used?

Best regards,
Volodymyr