Hi everybody,
I am interested in the overall performances of the Perfect Focus System by Nikon. I would like to know which maximal amplitude and at which maximal speed, the system is able to correct for a change in focus of a specimen. Maybe some of you know and use this system .... Thanks a lot Stephane -- Stéphane Pages PhD Neurobiologie cellulaire Centre de Recherche Université Laval Robert-Giffard (CRULRG) Bureau/office: F-5570 2601, Chemin de la Canardière Beauport (Québec) G1J 2G3 Canada Tél: 418-663-5747 ext. 4712 (bur./office) Fax: 418-663-8756 Courriel/email: [hidden email] |
|
Kevin Braeckmans |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
We are using the PFS system on a Nikon TE2000 E microscope for
about two years now and are quite happy with it. It accurately keeps the focal
position constant over very long times (we have tested up to 24h), so ideal for
long time lapse experiments. I find it also very convenient just for scanning
through the sample: since the sample is never entirely horizontal, when
scanning to different regions the focus position changes (relative to the cover
glass); the PFS nicely corrects for that since it uses the cover slip as a
reference and you can scan your entire sample at a constant focal depth without
having to refocus all the time. On our TE2000 the PFS system is an add-on module. It detects the
cover slip many times per second, so it is really fast in detecting and
correcting for z-drift of the sample. Axial focal range is limited to about 20
micron though, so not suitable for 3-D imaging of very thick samples. Also, due
to a dichroic filter in the unit, fluorescence emission is cut off at around
680 nm. In the latest Nikon microscope stand it is already built in with better
specifications (spectral cut-off at around 750nm and larger z-travel). Best regards, Kevin Van: Confocal Microscopy