Hello Group,
I have acquired a series of time lapse data sets and am beginning some simple analysis. The biology I am looking at is accumulation of an EGFP-tagged protein at the site of UV-laser induced DNA damage. The images were acquired using an EM-CCD and neutral density filter attenuated widefield illumination coming from an EXFO 100W Hg lamp. The EGFP signal accumulates in a well defined spot and I am able analyze the rate of accumulation and fold-increase of florescence intensity at several same-sized ROI's using Image J plugins. Baseline fluorescence intensity subtraction is simple enough. I have used a non-lased cell in the same field of view to monitor EGFP photo-bleaching, calculated the slope for this, but am stuck on how I might use this information to develop a correction factor. I would appreciate any thoughts and help in this regard. Thanks to all. Farid -- Farid Jalali MSc Program Leader- Cellular Imaging Core Applied Molecular Oncology and Radiation Medicine Program Princess Margaret Hospital (University Health Network) Toronto Medical Discovery Tower Toronto, Canada 416-581-7754 STTARR at TMDT 416-581-7791 STTARR Microscopy Suite |
![]() ![]() |
Kevin Braeckmans |
![]() |
Dear Farid, All you need to do is normalisation of the ‘accumulation
signal’ to the intensity of your control cell. This has to be done for
each image in the time series. In case you find that this procedure increases the noise too
much (a division of two measured signals naturally amplifies the noise), you
could first do a best fit to the photobleaching time trace of the control cell
(a single or double exponential should be fine) and then normalize the
accumulation time trace to the fitting result. Best regards, Kevin Prof. Dr. Kevin Braeckmans Lab. General Biochemistry & Physical Pharmacy Ghent University Harelbekestraat 72 9000 Ghent Belgium Tel: +32 (0)9 264.80.78 Fax: +32 (0)9 264.81.89 E-mail: [hidden email] Van: Confocal
Microscopy List [mailto:[hidden email]] Namens Farid
Jalali Hello Group, |
![]() ![]() |
Rietdorf, Jens |
![]() |
In reply to this post by Farid Jalali
Dear Farid, I have put together a small ImageJ macro that does the
normalization to a ROI average much like Kevin suggests. http://www.embl.de/eamnet/html/bleach_correction.html Cheers, jens From: Confocal Microscopy
List [mailto:[hidden email]] On Behalf Of Farid Jalali Hello Group, |
![]() ![]() |
Farid Jalali |
![]() |
Thanks very much Kevin and Jens for your help. I really do appreciate the suggestions.