To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello List We recently did an experiment where cells were stained with DAPI and Alexa 488. These cells were grown on plastic chamber slides, then fixed and stained. During imaging, we started seeing very weird bleaching patterns and bleedthrough of DAPI into the green viewing filter, even while just viewing the slides with a regular wide-field microscope. We used the same staining protocols as always (mounting media with anti-quenching reagents, etc.). The antibody that we used always worked well in our hands. The only difference that I can think of this time is the plastic slide. We haven't done staining on those before. Does anybody have any similar experience with staining cells on plastic chamber slides? --------------------------------------------------- Yevgeniy Romin Digital Microscopist Memorial Sloan-Kettering Cancer Center Molecular Cytology Core Facility 1275 York Ave. Box 333 New York, NY 10065 Tel.646-888-2186 Fax. 646-422-0640 --------------------------------------------------- ===================================================================== Please note that this e-mail and any files transmitted with it may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Yevgeniy, I have seen this effect with cells grown on glass coverslips as well, usually when the dye is present in very high concentration, so I'm not sure that it is related to the plastic slides. It has also been previously reported to this listserver. The only solution I have found in this case is to minimise viewing of the DAPI by eye with the microscope (using excitation from the Hg or Xe arc lamp) prior to acquiring with the confocal system or camera and image the green fluorophore (Alexa488) first if possible. Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Yevgeniy Romin Sent: Tuesday, 26 April 2011 4:30 a.m. To: [hidden email] Subject: Plastic Chamber Slides ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello List We recently did an experiment where cells were stained with DAPI and Alexa 488. These cells were grown on plastic chamber slides, then fixed and stained. During imaging, we started seeing very weird bleaching patterns and bleedthrough of DAPI into the green viewing filter, even while just viewing the slides with a regular wide-field microscope. We used the same staining protocols as always (mounting media with anti-quenching reagents, etc.). The antibody that we used always worked well in our hands. The only difference that I can think of this time is the plastic slide. We haven't done staining on those before. Does anybody have any similar experience with staining cells on plastic chamber slides? --------------------------------------------------- Yevgeniy Romin Digital Microscopist Memorial Sloan-Kettering Cancer Center Molecular Cytology Core Facility 1275 York Ave. Box 333 New York, NY 10065 Tel.646-888-2186 Fax. 646-422-0640 --------------------------------------------------- ===================================================================== Please note that this e-mail and any files transmitted with it may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
| Free forum by Nabble | Edit this page |