Pedro Camello |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, I´m planning to find a substitute for our old confocal. Since I´m tired of maintenance issues with costly short-lived lasers (around 1000 hours Argon lasers costs 15000 euros in Spain), I was looking for a lamp-based system to avoid wavelength limitations and costs. Am I wrong or is CARV-II the only commercial system of these characteristics at the moment? Otherwise, are recent lasers less expensive and short-lived?. In addition, could any of you share experiences about commercial TIRF systems? I would prefer a widefield system incorporating TIRF as an option (I listened about a ring-shaped filters from Olympus) Thanks |
|
Csúcs Gábor |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Pedro. 1) In our facility most of the laser last considerably longer then 1000 hours, so I wouldn't be afraid to by a new laser based system. Especially because the lamp based systems (of course this depends very much on the application) are not really comparable with the laser based scanning confocals. 2) In addition to the CARV-II, Olympus has also a lamp based spinning disk system. 3) In almost every commercial TIRF system I'm aware of, it is possible to combine wide-field imaging with TIRF imaging. There are pros and contras for each system (especially if you have a limited budget). Nevertheless, I'd say that currently the system offered by Leica is the most elegant (user friendly etc.) solution (but probably right now this is also the most expensive one). Cheers Gabor -- Gabor Csucs Light Microscopy Centre, ETH Zurich Schafmattstrasse 18, HPM F16 CH-8093, Zurich, Switzerland Web: www.lmc.ethz.ch Phone: +41 44 633 6221 Fax: +41 44 632 1298 e-mail: [hidden email] |
|
Holly L. AARON |
|
|
In reply to this post by Pedro Camello
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello, Pedro - In the lamp-based spinning disk confocal arena, you may also consider the Olympus DSU. Re lasers - many companies are now offering diode lasers instead of gas lasers, which last longer but cost more. Well, you didn't expect to get a win-win, did you? Also, 1000 hours does sound a little short for an argon laser. Do you have a lot of on/off? We've found we can extend life by leaving the laser on, even if it won't be used for 2 hours. If you do not have an auto-mated stand-by feature, it is also good to put it into standby when not in use. Cheers, __________________ Holly L. Aaron CRL Molecular Imaging Center http://imaging.berkeley.edu -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Pedro Camello Sent: Monday, May 05, 2008 9:02 AM To: [hidden email] Subject: Please, advice on confocal/deconvol plus TIRF system Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, I´m planning to find a substitute for our old confocal. Since I´m tired of maintenance issues with costly short-lived lasers (around 1000 hours Argon lasers costs 15000 euros in Spain), I was looking for a lamp-based system to avoid wavelength limitations and costs. Am I wrong or is CARV-II the only commercial system of these characteristics at the moment? Otherwise, are recent lasers less expensive and short-lived?. In addition, could any of you share experiences about commercial TIRF systems? I would prefer a widefield system incorporating TIRF as an option (I listened about a ring-shaped filters from Olympus) Thanks |
|
George Peeters-2 |
|
|
In reply to this post by Pedro Camello
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear Pedro: The new CSUX Confocal series has about twice the excitation efficiency as its predecessor the CSU-10. George A. Peeters MD, MS President, Solamere Technology Group Inc 1427 Perry Ave Salt Lake City, UT 84103 801 322-2645 office 801 322-2645 fax 801 232-6911 cell On May 5, 2008, at 10:01 AM, Pedro Camello wrote: Search the CONFOCAL archive at |
|
Michael Weber-4 |
|
|
In reply to this post by Pedro Camello
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Pedro, an Argon laser should really last longer than 1000 hours. But might be that your system needs an older version which doesn't have such a long lifetime. However, new laser based system have a much longer expected one, as already mentioned. On the other hand, costs for service contracts and maintenance will always be higher for laser-based systems. If you are looking for white light solutions: - Spinning Disc Confocal: CARVII, Olympus DSU, Leica SD6000 - TIRF: Nikon, Olympus So if you don't want any lasers, Olympus could be the way to go. I haven't seen a full working solution of White-Light TIRF + Confocal, but if you need speed, you either need to detect TIRF through the scanhead or use two cameras. Andor does offer solutions in this direction - but only with laser-based systems. Leica has the combination of white-light Spinning Disc plus laser-based TIRF. But both solutions are not cheap at all. Michael > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi all, > > I´m planning to find a substitute for our old confocal. Since I´m tired of > maintenance issues with costly short-lived lasers (around 1000 hours Argon > lasers costs 15000 euros in Spain), I was looking for a lamp-based system > to > avoid wavelength limitations and costs. Am I wrong or is CARV-II the only > commercial system of these characteristics at the moment? Otherwise, are > recent lasers less expensive and short-lived?. > > In addition, could any of you share experiences about commercial TIRF > systems? I would prefer a widefield system incorporating TIRF as an option > (I listened about a ring-shaped filters from Olympus) > > Thanks |