|
Michelle Aloni |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi list, We have a ZEISS LSM 510 confocal microscope in our busy core. Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and maybe even in the 20X air. I was told that they should be send to Germany for repair. This would be fine, but no one knows how long it will take before we can get them back... ZEISS guys in Los Angeles told me that they don't have lenses to loan... Our schedule for the users is full every day (even on weekends). We can't close the core! What do you guys do when something like this happen? There is no way that we are the first facility facing this problem. After we got in trouble, we learned our lesson the hard way and everybody is now extremely careful, plus we purchased and are using the special collar which protects the lenses of the inverted microscope... One way or another the damage is already done and we need help. Is there someone out there that might be able to loan us a lens or two for AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? If you want to, you can contact me off the list at [hidden email] Your advice or help would be highly appreciated Michelle Aloni MS, HTL (ASCP) Research Specialist USC Keck School of Medicine LA, CA |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Franky, most microscope companies take a very long time to do repairs. In our lab we learned our lesson a while ago. When some grant money came up we purchased a duplicate of our most critical lens and kept it in safe storage. When the lens on the microscope was inevitably broken we were able to swap in the back-up lens while waiting for the damaged one to be repaired. In my experience it takes about 6-8 weeks for a repair to go through. Have you other listers seen better or worse times? Craig On Mon, Oct 17, 2011 at 6:54 PM, Michelle MacVeigh-Aloni <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi list, > > We have a ZEISS LSM 510 confocal microscope in our busy core. > > Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and > maybe even in the 20X air. I was told that they should be send to Germany > for repair. This would be fine, but no one knows how long it will take > before we can get them back... > > ZEISS guys in Los Angeles told me that they don't have lenses to loan... > > Our schedule for the users is full every day (even on weekends). We can't > close the core! > > What do you guys do when something like this happen? There is no way that > we are the first facility facing this problem. > > After we got in trouble, we learned our lesson the hard way and everybody > is now extremely careful, plus we purchased and are using the special collar > which protects the lenses of the inverted microscope... One way or another > the damage is already done and we need help. > > Is there someone out there that might be able to loan us a lens or two for > AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? > > If you want to, you can contact me off the list at [hidden email] > > Your advice or help would be highly appreciated > > Michelle Aloni MS, HTL (ASCP) > Research Specialist > USC Keck School of Medicine LA, CA > |
|
Elena Rebollo Arredondo |
|
|
In reply to this post by Michelle Aloni
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, having oil into the 40x, 63x and 20x at the same time is statistically very unlikely. No joke, I have some really absent minded users and it takes many atemps to really get oil into the lenses. I would recommend having a thorough look under the stereomicroscope plus a good cleaning protocol, sometimes the usual optical paper is not enough. If it is true, then panic. In my experience (with Leica), the usual time is six weeks... Cheers, E On 18 October 2011 02:54, Michelle MacVeigh-Aloni <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi list, > > We have a ZEISS LSM 510 confocal microscope in our busy core. > > Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and > maybe even in the 20X air. I was told that they should be send to Germany > for repair. This would be fine, but no one knows how long it will take > before we can get them back... > > ZEISS guys in Los Angeles told me that they don't have lenses to loan... > > Our schedule for the users is full every day (even on weekends). We can't > close the core! > > What do you guys do when something like this happen? There is no way that > we are the first facility facing this problem. > > After we got in trouble, we learned our lesson the hard way and everybody > is now extremely careful, plus we purchased and are using the special collar > which protects the lenses of the inverted microscope... One way or another > the damage is already done and we need help. > > Is there someone out there that might be able to loan us a lens or two for > AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? > > If you want to, you can contact me off the list at [hidden email] > > Your advice or help would be highly appreciated > > Michelle Aloni MS, HTL (ASCP) > Research Specialist > USC Keck School of Medicine LA, CA > -- Elena Rebollo Ph.D. Confocal Microscopy Unit IBMB-PCB Molecular Biology Institute of Barcelona (IBMB), CSIC c/ Baldiri Reixac,10-12 08028 Barcelona (Spain) Tel (+34) 620938111/ (+34) 934020249 www.ibmb.csic.es<http://www.ibmb.csic.es/index.php?pg=laboratorio&idLaboratorio=27&tab=lab_home> |
|
Julien Cau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, A while ago, we had such a pandemics (few objectives with the front lens leaking in a few weeks time). The likely reason for this was a mistake in the objective cleaning solution we were preparing. Costly error.... You can monitor this under a stereomicroscope, if you see discontinuities in the glue seal sticking the front lens in several fairly new objectives, it's time to change your objective cleaning protocol. Regarding your local representative, I can't believe they won't offer you a temporary solution. Best regards Le 18/10/2011 19:53, Elena Rebollo Arredondo a écrit : > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi, > > having oil into the 40x, 63x and 20x at the same time is statistically very > unlikely. No joke, I have some really absent minded users and it takes many > atemps to really get oil into the lenses. I would recommend having a > thorough look under the stereomicroscope plus a good cleaning protocol, > sometimes the usual optical paper is not enough. > > If it is true, then panic. In my experience (with Leica), the usual time is > six weeks... > > Cheers, > > E > > > On 18 October 2011 02:54, Michelle MacVeigh-Aloni<[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi list, >> >> We have a ZEISS LSM 510 confocal microscope in our busy core. >> >> Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and >> maybe even in the 20X air. I was told that they should be send to Germany >> for repair. This would be fine, but no one knows how long it will take >> before we can get them back... >> >> ZEISS guys in Los Angeles told me that they don't have lenses to loan... >> >> Our schedule for the users is full every day (even on weekends). We can't >> close the core! >> >> What do you guys do when something like this happen? There is no way that >> we are the first facility facing this problem. >> >> After we got in trouble, we learned our lesson the hard way and everybody >> is now extremely careful, plus we purchased and are using the special collar >> which protects the lenses of the inverted microscope... One way or another >> the damage is already done and we need help. >> >> Is there someone out there that might be able to loan us a lens or two for >> AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? >> >> If you want to, you can contact me off the list at [hidden email] >> >> Your advice or help would be highly appreciated >> >> Michelle Aloni MS, HTL (ASCP) >> Research Specialist >> USC Keck School of Medicine LA, CA >> > > -- ____________________________________________ */Julien Cau, PhD./* /Montpellier RIO Imaging Facility manager/Responsable technique MRI/ Montpellier RIO Imaging Montpellier BIOCAMPUS, UMS3426 Arnaud de Villeneuve Campus Imaging Facility Institut de Génétique Humaine-CNRS 141, rue de la Cardonille F-34396 Montpellier(France) e-mail: [hidden email].fr_ <mailto:[hidden email]> phone: +33.4.34.35.99.90 mobile: +33.6.50.19.27.49 fax: +33.4.34.35.99.01 URL: _http://www.mri.cnrs.fr/_ ____________________________________________ |
|
Armstrong, Brian |
|
|
In reply to this post by Michelle Aloni
Michelle, it is possible that I could loan you some Zeiss objectives.
Contact me if you still need help. Brian Armstrong PhD Assistant Research Professor Light Microscopy Core Beckman Research Institute City of Hope 1500 East Duarte Road Duarte, CA 91010 626-256-4673 x62872 Light Microscopy Core Facility -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michelle MacVeigh-Aloni Sent: Monday, October 17, 2011 5:55 PM To: [hidden email] Subject: Practical question (cry for help) about ZEISS LSM 510 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi list, We have a ZEISS LSM 510 confocal microscope in our busy core. Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and maybe even in the 20X air. I was told that they should be send to Germany for repair. This would be fine, but no one knows how long it will take before we can get them back... ZEISS guys in Los Angeles told me that they don't have lenses to loan... Our schedule for the users is full every day (even on weekends). We can't close the core! What do you guys do when something like this happen? There is no way that we are the first facility facing this problem. After we got in trouble, we learned our lesson the hard way and everybody is now extremely careful, plus we purchased and are using the special collar which protects the lenses of the inverted microscope... One way or another the damage is already done and we need help. Is there someone out there that might be able to loan us a lens or two for AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? If you want to, you can contact me off the list at [hidden email] Your advice or help would be highly appreciated Michelle Aloni MS, HTL (ASCP) Research Specialist USC Keck School of Medicine LA, CA --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I agree about repair times. These 'super lenses' are pretty expensive but if you have standard 100x oil and 40x dry with your system (booth typically moderate in price) you will have quite adequate objectives to tide you over during repair times. Most users have neither the skill nor the specimens to benefit from very expensive objectives anyway! Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, 19 October 2011 4:17 AM To: [hidden email] Subject: Re: Practical question (cry for help) about ZEISS LSM 510 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Franky, most microscope companies take a very long time to do repairs. In our lab we learned our lesson a while ago. When some grant money came up we purchased a duplicate of our most critical lens and kept it in safe storage. When the lens on the microscope was inevitably broken we were able to swap in the back-up lens while waiting for the damaged one to be repaired. In my experience it takes about 6-8 weeks for a repair to go through. Have you other listers seen better or worse times? Craig On Mon, Oct 17, 2011 at 6:54 PM, Michelle MacVeigh-Aloni <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi list, > > We have a ZEISS LSM 510 confocal microscope in our busy core. > > Unfortunately, immersion oil got into our 40X oil and 63X oil lenses > maybe even in the 20X air. I was told that they should be send to Germany > for repair. This would be fine, but no one knows how long it will take > before we can get them back... > > ZEISS guys in Los Angeles told me that they don't have lenses to loan... > > Our schedule for the users is full every day (even on weekends). We can't > close the core! > > What do you guys do when something like this happen? There is no way that > we are the first facility facing this problem. > > After we got in trouble, we learned our lesson the hard way and everybody > is now extremely careful, plus we purchased and are using the special collar > which protects the lenses of the inverted microscope... One way or another > the damage is already done and we need help. > > Is there someone out there that might be able to loan us a lens or two for > AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? > > If you want to, you can contact me off the list at [hidden email] > > Your advice or help would be highly appreciated > > Michelle Aloni MS, HTL (ASCP) > Research Specialist > USC Keck School of Medicine LA, CA > ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1411 / Virus Database: 1522/3958 - Release Date: 10/17/11 |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** A one-time insurance investment we made in a previous lab was purchasing a spare favorite objective, in this case an Olympus 60X N.A. 1.4 which worked great on both our Olympus and Zeiss stands. The up front cost may be great, but over a few years, especially in a fee generating facility, it pays off. (Just my $6k.) -Michael C. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Tuesday, October 18, 2011 1:17 PM To: [hidden email] Subject: Re: Practical question (cry for help) about ZEISS LSM 510 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Franky, most microscope companies take a very long time to do repairs. In our lab we learned our lesson a while ago. When some grant money came up we purchased a duplicate of our most critical lens and kept it in safe storage. When the lens on the microscope was inevitably broken we were able to swap in the back-up lens while waiting for the damaged one to be repaired. In my experience it takes about 6-8 weeks for a repair to go through. Have you other listers seen better or worse times? Craig On Mon, Oct 17, 2011 at 6:54 PM, Michelle MacVeigh-Aloni <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi list, > > We have a ZEISS LSM 510 confocal microscope in our busy core. > > Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and > maybe even in the 20X air. I was told that they should be send to Germany > for repair. This would be fine, but no one knows how long it will take > before we can get them back... > > ZEISS guys in Los Angeles told me that they don't have lenses to loan... > > Our schedule for the users is full every day (even on weekends). We can't > close the core! > > What do you guys do when something like this happen? There is no way that > we are the first facility facing this problem. > > After we got in trouble, we learned our lesson the hard way and everybody > is now extremely careful, plus we purchased and are using the special collar > which protects the lenses of the inverted microscope... One way or another > the damage is already done and we need help. > > Is there someone out there that might be able to loan us a lens or two for > AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? > > If you want to, you can contact me off the list at [hidden email] > > Your advice or help would be highly appreciated > > Michelle Aloni MS, HTL (ASCP) > Research Specialist > USC Keck School of Medicine LA, CA > ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
|
Christine Davis |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I agree it can take 2-3 months to get an objective back. I'd look at it under the stereo scope and try cleaning it. Also you can wrap the objective in lens paper, put it in it's case, and turning it upside down (atleast overnight). This might help leach some oil out. If oil has gotten inside on the lens then it wont cure your problem. If oil has loaded itself on/in the shelf below the lens it will help. Maybe you can get a good enough image on one or two lenses to give yourself time to get one properly fixed. Christie Davis On Wed, Oct 19, 2011 at 11:28 AM, Cammer, Michael < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > A one-time insurance investment we made in a previous lab was purchasing a > spare favorite objective, in this case an Olympus 60X N.A. 1.4 which worked > great on both our Olympus and Zeiss stands. The up front cost may be great, > but over a few years, especially in a fee generating facility, it pays off. > (Just my $6k.) > -Michael C. > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Craig Brideau > Sent: Tuesday, October 18, 2011 1:17 PM > To: [hidden email] > Subject: Re: Practical question (cry for help) about ZEISS LSM 510 > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Franky, most microscope companies take a very long time to do repairs. In > our lab we learned our lesson a while ago. When some grant money came up > we > purchased a duplicate of our most critical lens and kept it in safe > storage. When the lens on the microscope was inevitably broken we were > able > to swap in the back-up lens while waiting for the damaged one to be > repaired. In my experience it takes about 6-8 weeks for a repair to go > through. Have you other listers seen better or worse times? > > Craig > > > On Mon, Oct 17, 2011 at 6:54 PM, Michelle MacVeigh-Aloni > <[hidden email]>wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hi list, > > > > We have a ZEISS LSM 510 confocal microscope in our busy core. > > > > Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and > > maybe even in the 20X air. I was told that they should be send to > Germany > > for repair. This would be fine, but no one knows how long it will take > > before we can get them back... > > > > ZEISS guys in Los Angeles told me that they don't have lenses to loan... > > > > Our schedule for the users is full every day (even on weekends). We > can't > > close the core! > > > > What do you guys do when something like this happen? There is no way that > > we are the first facility facing this problem. > > > > After we got in trouble, we learned our lesson the hard way and everybody > > is now extremely careful, plus we purchased and are using the special > collar > > which protects the lenses of the inverted microscope... One way or > another > > the damage is already done and we need help. > > > > Is there someone out there that might be able to loan us a lens or two > for > > AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? > > > > If you want to, you can contact me off the list at [hidden email] > > > > Your advice or help would be highly appreciated > > > > Michelle Aloni MS, HTL (ASCP) > > Research Specialist > > USC Keck School of Medicine LA, CA > > > > ------------------------------------------------------------ > This email message, including any attachments, is for the sole use of the > intended recipient(s) and may contain information that is proprietary, > confidential, and exempt from disclosure under applicable law. Any > unauthorized review, use, disclosure, or distribution is prohibited. If you > have received this email in error please notify the sender by return email > and delete the original message. Please note, the recipient should check > this email and any attachments for the presence of viruses. The organization > accepts no liability for any damage caused by any virus transmitted by this > email. > ================================= > |
|
Jacqueline Ross |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I recommend trying the polystyrene method suggested recently by Guy Cox. It worked brilliantly for one of our dry objective lenses that had somehow got coated in something. You do need to do it under a stereo as suggested by Christine. Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christine Davis Sent: Friday, 21 October 2011 7:33 a.m. To: [hidden email] Subject: Re: Practical question (cry for help) about ZEISS LSM 510 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I agree it can take 2-3 months to get an objective back. I'd look at it under the stereo scope and try cleaning it. Also you can wrap the objective in lens paper, put it in it's case, and turning it upside down (atleast overnight). This might help leach some oil out. If oil has gotten inside on the lens then it wont cure your problem. If oil has loaded itself on/in the shelf below the lens it will help. Maybe you can get a good enough image on one or two lenses to give yourself time to get one properly fixed. Christie Davis On Wed, Oct 19, 2011 at 11:28 AM, Cammer, Michael < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > A one-time insurance investment we made in a previous lab was purchasing a > spare favorite objective, in this case an Olympus 60X N.A. 1.4 which worked > great on both our Olympus and Zeiss stands. The up front cost may be great, > but over a few years, especially in a fee generating facility, it pays off. > (Just my $6k.) > -Michael C. > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Craig Brideau > Sent: Tuesday, October 18, 2011 1:17 PM > To: [hidden email] > Subject: Re: Practical question (cry for help) about ZEISS LSM 510 > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Franky, most microscope companies take a very long time to do repairs. In > our lab we learned our lesson a while ago. When some grant money came up > we > purchased a duplicate of our most critical lens and kept it in safe > storage. When the lens on the microscope was inevitably broken we were > able > to swap in the back-up lens while waiting for the damaged one to be > repaired. In my experience it takes about 6-8 weeks for a repair to go > through. Have you other listers seen better or worse times? > > Craig > > > On Mon, Oct 17, 2011 at 6:54 PM, Michelle MacVeigh-Aloni > <[hidden email]>wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hi list, > > > > We have a ZEISS LSM 510 confocal microscope in our busy core. > > > > Unfortunately, immersion oil got into our 40X oil and 63X oil lenses and > > maybe even in the 20X air. I was told that they should be send to > Germany > > for repair. This would be fine, but no one knows how long it will take > > before we can get them back... > > > > ZEISS guys in Los Angeles told me that they don't have lenses to loan... > > > > Our schedule for the users is full every day (even on weekends). We > can't > > close the core! > > > > What do you guys do when something like this happen? There is no way that > > we are the first facility facing this problem. > > > > After we got in trouble, we learned our lesson the hard way and everybody > > is now extremely careful, plus we purchased and are using the special > collar > > which protects the lenses of the inverted microscope... One way or > another > > the damage is already done and we need help. > > > > Is there someone out there that might be able to loan us a lens or two > for > > AXIO Observer Z1 for the duration of the cleaning/repair of our lenses? > > > > If you want to, you can contact me off the list at [hidden email] > > > > Your advice or help would be highly appreciated > > > > Michelle Aloni MS, HTL (ASCP) > > Research Specialist > > USC Keck School of Medicine LA, CA > > > > ------------------------------------------------------------ > This email message, including any attachments, is for the sole use of the > intended recipient(s) and may contain information that is proprietary, > confidential, and exempt from disclosure under applicable law. Any > unauthorized review, use, disclosure, or distribution is prohibited. If you > have received this email in error please notify the sender by return email > and delete the original message. Please note, the recipient should check > this email and any attachments for the presence of viruses. The organization > accepts no liability for any damage caused by any virus transmitted by this > email. > ================================= > |
|
Ben Abrams |
|
|
This post was updated on .
CONTENTS DELETED
The author has deleted this message.
|
|
lechristophe |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Mr Abrams, would it be possible to turn off that autoreply thing or avoid it next time ? Please be aware that it seems to send a reply to everyone on the Confocal Mailing list, each time there is a message from anyone on the list. Thanks a lot, Christophe Leterrier On Fri, Oct 21, 2011 at 00:29, Ben Abrams <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > This is an automated reply. > > I will be out of the office/ attending a course and will have uncertain > email access Monday October 10 through Friday October 21. > |
| Free forum by Nabble | Edit this page |