				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Confocalists, Has anybody experience of spectral imaging on Zeiss Lsm700? Is it possible on all variants of this microscope? I used to work on LSM510meta whose metadector allowed me to use up to 6 different fluorophores (i.e DAPI, Alexa488, Alexa555, Texas Red, Cy5, Alexa680). I can potentially use LSM700 and would like to be sure if it has all capabilities of LSM510meta. Best regards, Michal Dr Michal Gdula Postdoctoral Researcher at Boston University  | 
			
 
	
	 | 
						Armstrong, Brian | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Michal, if I remember correctly the LSM700 configuration does not offer a spectral detector option. You would need to have access to an LSM710 with Quasar and 34 channel detector for spectral capabilities like the 510META you are used to. Contact your Zeiss rep to be sure. Cheers, Brian D Armstrong PhD Assistant Research Professor Light Microscopy Core Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Gdula Sent: Friday, April 29, 2011 2:44 PM To: [hidden email] Subject: Question about Zeiss LSM700 and spectral imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Confocalists, Has anybody experience of spectral imaging on Zeiss Lsm700? Is it possible on all variants of this microscope? I used to work on LSM510meta whose metadector allowed me to use up to 6 different fluorophores (i.e DAPI, Alexa488, Alexa555, Texas Red, Cy5, Alexa680). I can potentially use LSM700 and would like to be sure if it has all capabilities of LSM510meta. Best regards, Michal Dr Michal Gdula Postdoctoral Researcher at Boston University --------------------------------------------------------------------- SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. ---------------------------------------------------------------------  | 
			
 
	
	 | 
						G. Esteban Fernandez | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Indeed. My 700 has no spectral capabilties but my 710 does. -Esteban On Fri, Apr 29, 2011 at 3:13 PM, Armstrong, Brian <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Michal, if I remember correctly the LSM700 configuration does not offer a spectral detector option. > You would need to have access to an LSM710 with Quasar and 34 channel detector for spectral capabilities like the 510META you are used to. > > Contact your Zeiss rep to be sure. > > Cheers, > > Brian D Armstrong PhD > Assistant Research Professor > Light Microscopy Core > Beckman Research Institute > City of Hope > Dept of Neuroscience > 1450 E Duarte Rd > Duarte, CA 91010 > 626-256-4673 x62872 > > http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Gdula > Sent: Friday, April 29, 2011 2:44 PM > To: [hidden email] > Subject: Question about Zeiss LSM700 and spectral imaging > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Confocalists, > > Has anybody experience of spectral imaging on Zeiss Lsm700? Is it possible > on all variants of this microscope? > > I used to work on LSM510meta whose metadector allowed me to use up to 6 > different fluorophores (i.e DAPI, Alexa488, Alexa555, Texas Red, Cy5, > Alexa680). I can potentially use LSM700 and would like to be sure if it has > all capabilities of LSM510meta. > > Best regards, > > Michal > > Dr Michal Gdula > > Postdoctoral Researcher at Boston University > > > --------------------------------------------------------------------- > SECURITY/CONFIDENTIALITY WARNING: > This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. > > --------------------------------------------------------------------- >  | 
			
 
	
	 | 
						Michal Gdula-2 | 
| 
					
		 | 
				
					
	 
				In reply to this post by Michal Gdula-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Thank you for your fast response! This was also what the owner of the microscope claimed, however in the specification from zeiss site it is wrtten: "Spectral detection Simultaneously in two confocal reflection channels, with high- sensitivity, low-noise PMTs, adjustable (increment 1 nm)" http://www.zeiss.com/C12567BE00472A5C/EmbedTitelIntern/LSM700_Specificatio n/$File/LSM700_Specification.pdf and it is also repeated in the product brochure - 2 mistakes in Zeiss documentation? additionaly in manual are information on scanning in lambda mode... I will try to call Zeiss, but still hopes to get in touch with somebody who used LSM700 for spectral imaging... Regards, Michal  | 
			
 
	
	 | 
						leoncio vergara | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Michal, I was intrigued by your insistence despite the opinions against so I could not resist the temptation to look at the LSM700 brochure and indeed it is advertised as Lambda Scan capable by means of the VSD (variable secondary dichroic beamsplitter)... so I think you might be in the right track... hope your Zeiss rep gives you a positive response, good luck Leoncio A. Vergara MD Assistant Director Center for Biomedical Engineering Assistant Professor Microbiology and Immunology University of Texas Medical Branch 409-750-2153 (cell) -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Gdula Sent: Friday, April 29, 2011 7:01 PM To: [hidden email] Subject: Re: Question about Zeiss LSM700 and spectral imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Thank you for your fast response! This was also what the owner of the microscope claimed, however in the specification from zeiss site it is wrtten: "Spectral detection Simultaneously in two confocal reflection channels, with high- sensitivity, low-noise PMTs, adjustable (increment 1 nm)" http://www.zeiss.com/C12567BE00472A5C/EmbedTitelIntern/LSM700_Specificatio n/$File/LSM700_Specification.pdf and it is also repeated in the product brochure - 2 mistakes in Zeiss documentation? additionaly in manual are information on scanning in lambda mode... I will try to call Zeiss, but still hopes to get in touch with somebody who used LSM700 for spectral imaging... Regards, Michal  | 
			
 
	
	 | 
						Axel Kurt Preuss | 
| 
					
		 | 
				
					
	
	 
		Sure it does spectral imaging, but be careful with the upper end of long wavelenghts. Due to VSD config it cannot lambda scan  as far into the red as one would wish for
 
				Thanks Cheers Axel Sent from +6592715622 ————— Axel K Preuss, PhD, Central Imaging, IMCB, A*Star, 61 Biopolis Dr, 6-19B, Singapore 138673 On Apr 30, 2011, at 8:50 AM, "Vergara, Leoncio A." <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Michal, I was intrigued by your insistence despite the opinions against so I could not resist the temptation to look at the LSM700 brochure and indeed it is advertised as Lambda Scan capable by means of the VSD (variable secondary dichroic beamsplitter)... so I think you might be in the right track... hope your Zeiss rep gives you a positive response, good luck > > > Leoncio A. Vergara MD > Assistant Director > Center for Biomedical Engineering > Assistant Professor > Microbiology and Immunology > University of Texas Medical Branch > 409-750-2153 (cell) > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Gdula > Sent: Friday, April 29, 2011 7:01 PM > To: [hidden email] > Subject: Re: Question about Zeiss LSM700 and spectral imaging > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Thank you for your fast response! > > This was also what the owner of the microscope claimed, however in the > specification from zeiss site it is wrtten: > "Spectral detection Simultaneously in two confocal reflection channels, with high- > sensitivity, low-noise PMTs, > adjustable (increment 1 nm)" > http://www.zeiss.com/C12567BE00472A5C/EmbedTitelIntern/LSM700_Specificatio > n/$File/LSM700_Specification.pdf > > and it is also repeated in the product brochure - 2 mistakes in Zeiss > documentation? additionaly in manual are information on scanning in lambda > mode... I will try to call Zeiss, but still hopes to get in touch with somebody who > used LSM700 for spectral imaging... > > Regards, > > Michal Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The internal detectors on our 710 system only practically collect to 740 nm. From our lab web pages: Three internal detectors have adjustable gates for wavelength ranges. They collect from below the shortest wavelength laser on the system at 405 nm to (practically) 740 nm. According to the software controls, the internal detectors collect up through 758 nm. However, testing with reflection mode with the longest wavelength physical gate, collection is practical only to 740 nm. After 740 nm collection intensity falls rapidly and is not detectable above 750 nm. _________________________________________ Michael Cammer, Assistant Research Scientist Skirball Institute of Biomolecular Medicine Lab: (212) 263-3208 Cell: (914) 309-3270 ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Axel Kurt Preuss [[hidden email]] Sent: Friday, April 29, 2011 10:10 PM To: [hidden email] Subject: Re: Question about Zeiss LSM700 and spectral imaging Sure it does spectral imaging, but be careful with the upper end of long wavelenghts. Due to VSD config it cannot lambda scan as far into the red as one would wish for Thanks Cheers Axel Sent from +6592715622 ————— Axel K Preuss, PhD, Central Imaging, IMCB, A*Star, 61 Biopolis Dr, 6-19B, Singapore 138673 On Apr 30, 2011, at 8:50 AM, "Vergara, Leoncio A." <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Michal, I was intrigued by your insistence despite the opinions against so I could not resist the temptation to look at the LSM700 brochure and indeed it is advertised as Lambda Scan capable by means of the VSD (variable secondary dichroic beamsplitter)... so I think you might be in the right track... hope your Zeiss rep gives you a positive response, good luck > > > Leoncio A. Vergara MD > Assistant Director > Center for Biomedical Engineering > Assistant Professor > Microbiology and Immunology > University of Texas Medical Branch > 409-750-2153 (cell) > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Gdula > Sent: Friday, April 29, 2011 7:01 PM > To: [hidden email] > Subject: Re: Question about Zeiss LSM700 and spectral imaging > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Thank you for your fast response! > > This was also what the owner of the microscope claimed, however in the > specification from zeiss site it is wrtten: > "Spectral detection Simultaneously in two confocal reflection channels, with high- > sensitivity, low-noise PMTs, > adjustable (increment 1 nm)" > http://www.zeiss.com/C12567BE00472A5C/EmbedTitelIntern/LSM700_Specificatio > n/$File/LSM700_Specification.pdf > > and it is also repeated in the product brochure - 2 mistakes in Zeiss > documentation? additionaly in manual are information on scanning in lambda > mode... I will try to call Zeiss, but still hopes to get in touch with somebody who > used LSM700 for spectral imaging... > > Regards, > > Michal Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you. ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. =================================  | 
			
 
	
	 | 
						G. Esteban Fernandez | 
| 
					
		 | 
				
					
	 
				In reply to this post by leoncio vergara
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The 700 does have a continuously variable dichroic so it is posible to do emission wavelength fingerprinting in series but not traditional spectral analysis of multiple wavelengths in parallel. To do spectral imaging on a routine basis the 710 is the way to go. -Esteban On Friday, April 29, 2011, Vergara, Leoncio A. <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Michal, I was intrigued by your insistence despite the opinions against so I could not resist the temptation to look at the LSM700 brochure and indeed it is advertised as Lambda Scan capable by means of the VSD (variable secondary dichroic beamsplitter)... so I think you might be in the right track... hope your Zeiss rep gives you a positive response, good luck > > > Leoncio A. Vergara MD > Assistant Director > Center for Biomedical Engineering > Assistant Professor > Microbiology and Immunology > University of Texas Medical Branch > 409-750-2153 (cell) > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Gdula > Sent: Friday, April 29, 2011 7:01 PM > To: [hidden email] > Subject: Re: Question about Zeiss LSM700 and spectral imaging > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Thank you for your fast response! > > This was also what the owner of the microscope claimed, however in the > specification from zeiss site it is wrtten: > "Spectral detection Simultaneously in two confocal reflection channels, with high- > sensitivity, low-noise PMTs, > adjustable (increment 1 nm)" > http://www.zeiss.com/C12567BE00472A5C/EmbedTitelIntern/LSM700_Specificatio > n/$File/LSM700_Specification.pdf > > and it is also repeated in the product brochure - 2 mistakes in Zeiss > documentation? additionaly in manual are information on scanning in lambda > mode... I will try to call Zeiss, but still hopes to get in touch with somebody who > used LSM700 for spectral imaging... > > Regards, > > Michal >  | 
			
 
	
	 | 
						Axel Kurt Preuss | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Let s just stick to the definition of lambda scanning, and that is -scanning lambda ( :) ) This means using a certain excitation wavelength X and scanning the emission spectrum of the signal we produce. This can be done with the 700, very nicely indeed, for example by using ROI based lambda scanning. Of course one scan the whole image too. If one feels it makes sense... But one has to realize that the 700 has several limiting factors 1) the variable dichroic, how far it can move "up" to long wavelength 2) any optical filters which are or one chooses to have or have not in addition in the beam path to the PMT. 3) it s fixed gate (primary dichroic) (yes, there IS a main beamsplitter!) Important for the discussion here is the capability to lambda scan into the red, and, as I pointed out in a post before, it aint easy to scan far into the red. Yes, lambda scanning is possible, and very nicely, but we have to know the limitations into the red. To be more precise, as Seichow and Goetze from Zeiss write, "the new "Variable Secondary Dichroic" (VSD) splits the emission between two detectors (in the 2 channel configuration only). The VSD is a substrate coated with a special dichroic gradient layer. Embedded in a new detection beam path design the VSD is mounted on a motorized slider, and depending on its position, it can be a short pass/long reflectance dichroic at any wavelength between 420 and 630 nm. " Please note the "630nm". It is hard to imagine to lambda scan above the 630nm, as it will not be possible to differentiate -in a lambda scan- 650 from 700nm, as BOTH are, as we all know, or should know, above 630nm, and both get reflected to the same PMT. For anybody who is NOT interested in this color range, lambda scanning with the 700 should be sufficient. Makes sense? Axel Preuss, Central Imaging at IMCB, Singapore -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of G. Esteban Fernandez Sent: Monday, 2 May, 2011 5:37 AM To: [hidden email] Subject: Question about Zeiss LSM700 and spectral imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The 700 does have a continuously variable dichroic so it is posible to do emission wavelength fingerprinting in series but not traditional spectral analysis of multiple wavelengths in parallel. To do spectral imaging on a routine basis the 710 is the way to go. -Esteban On Friday, April 29, 2011, Vergara, Leoncio A. <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Michal, I was intrigued by your insistence despite the opinions > against so I could not resist the temptation to look at the LSM700 > brochure and indeed it is advertised as Lambda Scan capable by means > of the VSD (variable secondary dichroic beamsplitter)... so I think > you might be in the right track... hope your Zeiss rep gives you a > positive response, good luck > > > Leoncio A. Vergara MD > Assistant Director > Center for Biomedical Engineering > Assistant Professor > Microbiology and Immunology > University of Texas Medical Branch > 409-750-2153 (cell) > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Michal Gdula > Sent: Friday, April 29, 2011 7:01 PM > To: [hidden email] > Subject: Re: Question about Zeiss LSM700 and spectral imaging > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Thank you for your fast response! > > This was also what the owner of the microscope claimed, however in the > specification from zeiss site it is wrtten: > "Spectral detection Simultaneously in two confocal reflection > channels, with high- sensitivity, low-noise PMTs, adjustable > (increment 1 nm)" > http://www.zeiss.com/C12567BE00472A5C/EmbedTitelIntern/LSM700_Specific > atio > n/$File/LSM700_Specification.pdf > > and it is also repeated in the product brochure - 2 mistakes in Zeiss > documentation? additionaly in manual are information on scanning in > lambda mode... I will try to call Zeiss, but still hopes to get in > touch with somebody who used LSM700 for spectral imaging... > > Regards, > > Michal > Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.  | 
			
 
	
	 | 
						Aryeh Weiss | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** An important point about spectral scanning with the LSM700 is that you are not measuring emission within a small spectral bandwidth centered on some wavelength. Instead, you are measuring (ideally) with a high pass in one detector and low pass in the other. This means that your various lambda points are not independent, and you must be careful when interpreting such measurements. In principle you can derive the spectrum from what you measure (eg by taking differences), but you will have difficulty seeing weak spectral features which are near much stronger features. --aryeh On 5/3/11 9:02 AM, Axel Kurt Preuss wrote: > ***** To join, leave or search the confocal microscopy listserv, go > to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** > > Let s just stick to the definition of lambda scanning, and that is > -scanning lambda ( :) ) > > This means using a certain excitation wavelength X and scanning the > emission spectrum of the signal we produce. > > This can be done with the 700, very nicely indeed, for example by > using ROI based lambda scanning. Of course one scan the whole image > too. If one feels it makes sense... > > But one has to realize that the 700 has several limiting factors 1) > the variable dichroic, how far it can move "up" to long wavelength 2) > any optical filters which are or one chooses to have or have not in > addition in the beam path to the PMT. 3) it s fixed gate (primary > dichroic) (yes, there IS a main beamsplitter!) > > > Important for the discussion here is the capability to lambda scan > into the red, and, as I pointed out in a post before, it aint easy > to scan far into the red. Yes, lambda scanning is possible, and very > nicely, but we have to know the limitations into the red. > > To be more precise, as Seichow and Goetze from Zeiss write, "the new > "Variable Secondary Dichroic" (VSD) splits the emission between two > detectors (in the 2 channel configuration only). The VSD is a > substrate coated with a special dichroic gradient layer. Embedded in > a new detection beam path design the VSD is mounted on a motorized > slider, and depending on its position, it can be a short pass/long > reflectance dichroic at any wavelength between 420 and 630 nm. " > > Please note the "630nm". It is hard to imagine to lambda scan above > the 630nm, as it will not be possible to differentiate -in a lambda > scan- 650 from 700nm, as BOTH are, as we all know, or should know, > above 630nm, and both get reflected to the same PMT. > > For anybody who is NOT interested in this color range, lambda > scanning with the 700 should be sufficient. > > Makes sense? > > Axel Preuss, Central Imaging at IMCB, Singapore > > -----Original Message----- From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of G. Esteban > Fernandez Sent: Monday, 2 May, 2011 5:37 AM To: > [hidden email] Subject: Question about Zeiss LSM700 > and spectral imaging > > ***** To join, leave or search the confocal microscopy listserv, go > to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** > > The 700 does have a continuously variable dichroic so it is posible > to do emission wavelength fingerprinting in series but not > traditional spectral analysis of multiple wavelengths in parallel. To > do spectral imaging on a routine basis the 710 is the way to go. > > -Esteban > > > On Friday, April 29, 2011, Vergara, Leoncio A.<[hidden email]> > wrote: >> ***** To join, leave or search the confocal microscopy listserv, go >> to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** >> >> Michal, I was intrigued by your insistence despite the opinions >> against so I could not resist the temptation to look at the LSM700 >> brochure and indeed it is advertised as Lambda Scan capable by >> means of the VSD (variable secondary dichroic beamsplitter)... so I >> think you might be in the right track... hope your Zeiss rep gives >> you a positive response, good luck >> >> >> Leoncio A. Vergara MD Assistant Director Center for Biomedical >> Engineering Assistant Professor Microbiology and Immunology >> University of Texas Medical Branch 409-750-2153 (cell) >> >> -----Original Message----- From: Confocal Microscopy List >> [mailto:[hidden email]] On Behalf Of Michal >> Gdula Sent: Friday, April 29, 2011 7:01 PM To: >> [hidden email] Subject: Re: Question about Zeiss >> LSM700 and spectral imaging >> >> ***** To join, leave or search the confocal microscopy listserv, go >> to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** >> >> Thank you for your fast response! >> >> This was also what the owner of the microscope claimed, however in >> the specification from zeiss site it is wrtten: "Spectral detection >> Simultaneously in two confocal reflection channels, with high- >> sensitivity, low-noise PMTs, adjustable (increment 1 nm)" >> http://www.zeiss.com/C12567BE00472A5C/EmbedTitelIntern/LSM700_Specific >> >> >> n/$File/LSM700_Specification.pdf >> >> and it is also repeated in the product brochure - 2 mistakes in >> Zeiss documentation? additionaly in manual are information on >> scanning in lambda mode... I will try to call Zeiss, but still >> hopes to get in touch with somebody who used LSM700 for spectral >> imaging... >> >> Regards, >> >> Michal >> -- Aryeh Weiss School of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051  | 
			
 
	
	 | 
						Jeremy Adler-4 | 
| 
					
		 | 
				
					
	 
				In reply to this post by leoncio vergara
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Invitation to everyone On May 25th there is a half day meeting at Uppsala U in Sweden on live imaging details below http://www.scilifelab.uu.se/Technology_Platforms/BioVis/BioVissymposium/ Jeremy Adler Genetics & Pathology Rudbeckslaboratoriet Daghammersköljdsväg 20 751 85 Uppsala Sweden 0046 (0)18 471 4607  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, all, We are decommisioning a good working order Philips 300 TEM. Please contact me if you are interested in taking it for free. However, you will need to be responisble for the cost of dismantling and shipping of the scope. We need to move it before the end of the month. Thanks Andre Andre Wong, Senior Manager, Imaging Facility, Faculty of Dentistry, University of BC 2199, Wesbrook Mall, Vancouver, BC Canada, V6T 1Z3. Tel. 604-822-2873 Fax. 604-822-3562  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** WOW! A Philips TEM 300. Donate it to a museum. Loved that instrument. Best to all, -Rich Richard F. Konz, Jr. Assistant Professor Department of Medicine Director, Flow Cytometry Core Facility UMASS Medical School 55 Lake Avenue North Room S5-322 Worcester, MA 01655 Office: 508-856-1598 Lab: 508-856-3276 Mobil: 508-868-6021 Lab Web Link: http://www.umassmed.edu/facslab/index.aspx My Faculty Page: http://www.umassmed.edu/ivp/faculty/Konz.cfm ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of andre wong [[hidden email]] Sent: Thursday, May 12, 2011 3:06 PM To: [hidden email] Subject: Re: Philips 300 TEM for free ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, all, We are decommisioning a good working order Philips 300 TEM. Please contact me if you are interested in taking it for free. However, you will need to be responisble for the cost of dismantling and shipping of the scope. We need to move it before the end of the month. Thanks Andre Andre Wong, Senior Manager, Imaging Facility, Faculty of Dentistry, University of BC 2199, Wesbrook Mall, Vancouver, BC Canada, V6T 1Z3. Tel. 604-822-2873 Fax. 604-822-3562  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Michal Gdula-2
			 
	
		Perhaps a little late, but you can indeed do some spectral scanning/unmixing on the LSM700. It has a continuously variable secondary dichroic (ie. spectral grating) in front of the pmt's. Using the Zen software you can access the controls from the Smart Setup panel. One limitation you will find though is that it only allows you to set up 4 channels at a time so if your specimen is multiplexed to a greater degree you may have to do a 2-step acquisition or something.
 