Question about a super-resolution technique

Dear all,

I jsut heard of a technique to obtain a super-resolution image from a series of low-resolution shifted images :
http://www.und.edu/instruct/young/toast/Super-Resolution.html
See also here : http://scien.stanford.edu/class/psych221/projects/99/stever/report.html
Basically, you extract details from sub-pixels shift in the image.

My understanding is that it could be useful for microscopy, but only for undersampled images where the limiting factor is the pixel size, and not the diffraction. Is that right that you can't break the diffusion barrier with this method ?

Christophe
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Question about a super-resolution technique

Christophe,
 
We have a Zeiss LSM 5 LIVE with a detector of fixed 512 pixels.  To achieve true 1024x1024 or 1536x1536 it uses the shift technique you describe, shifting by 1/2 or 1/3 pixel size to get 1024 or 1536, respectively.  It is true that this increases resolution when pixels are the limiting factor, it does not break the diffraction limit of resolution.
 
-Esteban

On Wed, Jul 29, 2009 at 4:34 AM, Christophe Leterrier <[hidden email]> wrote:
Dear all,

I jsut heard of a technique to obtain a super-resolution image from a series of low-resolution shifted images :
http://www.und.edu/instruct/young/toast/Super-Resolution.html
See also here : http://scien.stanford.edu/class/psych221/projects/99/stever/report.html
Basically, you extract details from sub-pixels shift in the image.

My understanding is that it could be useful for microscopy, but only for undersampled images where the limiting factor is the pixel size, and not the diffraction. Is that right that you can't break the diffusion barrier with this method ?

Christophe



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9676 fax
Armstrong, Brian Armstrong, Brian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Question about a super-resolution technique

Hello, I believe that camera manufacturers have used “pixel shifting” to obtain super high resolution from CCDs for years. For example the Leica DFC500 was a pixel shift camera that could utilize up to 36 shots to compose a single image (final resolution around 4k x 3k).

 

Brian D Armstrong PhD

Light Microscopy Core Manager

Beckman Research Institute

City of Hope

Dept of Neuroscience

1450 E Duarte Rd

Duarte, CA 91010

626-256-4673 x62872

http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of G. Esteban Fernandez
Sent: Wednesday, July 29, 2009 8:21 AM
To: [hidden email]
Subject: Re: Question about a super-resolution technique

 

Christophe,

 

We have a Zeiss LSM 5 LIVE with a detector of fixed 512 pixels.  To achieve true 1024x1024 or 1536x1536 it uses the shift technique you describe, shifting by 1/2 or 1/3 pixel size to get 1024 or 1536, respectively.  It is true that this increases resolution when pixels are the limiting factor, it does not break the diffraction limit of resolution.

 

-Esteban

On Wed, Jul 29, 2009 at 4:34 AM, Christophe Leterrier <[hidden email]> wrote:

Dear all,

I jsut heard of a technique to obtain a super-resolution image from a series of low-resolution shifted images :
http://www.und.edu/instruct/young/toast/Super-Resolution.html
See also here : http://scien.stanford.edu/class/psych221/projects/99/stever/report.html
Basically, you extract details from sub-pixels shift in the image.

My understanding is that it could be useful for microscopy, but only for undersampled images where the limiting factor is the pixel size, and not the diffraction. Is that right that you can't break the diffusion barrier with this method ?

Christophe




--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9676 fax


---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING:
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.
---------------------------------------------------------------------

Leoncio A. Vergara Leoncio A. Vergara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Question about a super-resolution technique

In reply to this post by lechristophe
I think the result of the pixel shifting cameras is equivalent to a smoothed
version of the image that would have been acquired with a CCD with the
higher number of pixels... in the example that would be like applying a 6x6
kernel... I am right?
 
Leoncio Vergara MD
Technical Director
Integrated Microscopy Core (IMC)
 
Wright, Colin Wright, Colin
Reply | Threaded
Open this post in threaded view
♦
♦
|

unsubscribe

In reply to this post by G. Esteban Fernandez
 
 
Best Regards
 
Colin
 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of G. Esteban Fernandez
Sent: 29 July 2009 16:21
To: [hidden email]
Subject: Re: Question about a super-resolution technique

Christophe,
 
We have a Zeiss LSM 5 LIVE with a detector of fixed 512 pixels.  To achieve true 1024x1024 or 1536x1536 it uses the shift technique you describe, shifting by 1/2 or 1/3 pixel size to get 1024 or 1536, respectively.  It is true that this increases resolution when pixels are the limiting factor, it does not break the diffraction limit of resolution.
 
-Esteban

On Wed, Jul 29, 2009 at 4:34 AM, Christophe Leterrier <[hidden email]> wrote:
Dear all,

I jsut heard of a technique to obtain a super-resolution image from a series of low-resolution shifted images :
http://www.und.edu/instruct/young/toast/Super-Resolution.html
See also here : http://scien.stanford.edu/class/psych221/projects/99/stever/report.html
Basically, you extract details from sub-pixels shift in the image.

My understanding is that it could be useful for microscopy, but only for undersampled images where the limiting factor is the pixel size, and not the diffraction. Is that right that you can't break the diffusion barrier with this method ?

Christophe



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9676 fax


This E-mail is strictly confidential and intended solely for persons or organisations to whom it is addressed.If you are not the intended recipient,you must not copy, distribute or take any action in reliance upon it. If you receive this E-mail in error, please notify the IT department of Nikon U.K. Limited (02084816861) as soon as possible, delete this E-mail from your computer and destroy any printed copies.

Nikon U.K. Limited cannot accept any responsibility for the accuracy or completeness or security of this message as it has been transmitted over the Internet and has been relayed by a number of servers. If you suspect that the message has been altered, please inform us.

At Nikon we operate by best practice and exercise an 'opt-out' policy. If you do not wish to be e-mailed in future by Nikon U.K. Limited please indicate this fact in your response.

Nikon is the trading name of Nikon U.K. Limited which is a company registered in England and Wales with company number 1448987 and VAT registration number GB 243-0236-06 whose registered office is at 380 Richmond Road,Kingston,Surrey,KT2 5PR,England.
Adrienne Rollie Adrienne Rollie
Reply | Threaded
Open this post in threaded view
♦
♦
|

unsubscribe

Thanks,
Adrienne


Date: Thu, 30 Jul 2009 13:31:44 +0200
From: [hidden email]
Subject: unsubscribe
To: [hidden email]

 
 
Best Regards
 
Colin
 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of G. Esteban Fernandez
Sent: 29 July 2009 16:21
To: [hidden email]
Subject: Re: Question about a super-resolution technique

Christophe,
 
We have a Zeiss LSM 5 LIVE with a detector of fixed 512 pixels.  To achieve true 1024x1024 or 1536x1536 it uses the shift technique you describe, shifting by 1/2 or 1/3 pixel size to get 1024 or 1536, respectively.  It is true that this increases resolution when pixels are the limiting factor, it does not break the diffraction limit of resolution.
 
-Esteban

On Wed, Jul 29, 2009 at 4:34 AM, Christophe Leterrier <[hidden email]> wrote:
Dear all,

I jsut heard of a technique to obtain a super-resolution image from a series of low-resolution shifted images :
http://www.und.edu/instruct/young/toast/Super-Resolution.html
See also here : http://scien.stanford.edu/class/psych221/projects/99/stever/report.html
Basically, you extract details from sub-pixels shift in the image.

My understanding is that it could be useful for microscopy, but only for undersampled images where the limiting factor is the pixel size, and not the diffraction. Is that right that you can't break the diffusion barrier with this method ?

Christophe



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9676 fax


This E-mail is strictly confidential and intended solely for persons or organisations to whom it is addressed.If you are not the intended recipient,you must not copy, distribute or take any action in reliance upon it. If you receive this E-mail in error, please notify the IT department of Nikon U.K. Limited (02084816861) as soon as possible, delete this E-mail from your computer and destroy any printed copies.

Nikon U.K. Limited cannot accept any responsibility for the accuracy or completeness or security of this message as it has been transmitted over the Internet and has been relayed by a number of servers. If you suspect that the message has been altered, please inform us.

At Nikon we operate by best practice and exercise an 'opt-out' policy. If you do not wish to be e-mailed in future by Nikon U.K. Limited please indicate this fact in your response.

Nikon is the trading name of Nikon U.K. Limited which is a company registered in England and Wales with company number 1448987 and VAT registration number GB 243-0236-06 whose registered office is at 380 Richmond Road,Kingston,Surrey,KT2 5PR,England.


Windows Live™ Hotmail®: Celebrate the moment with your favorite sports pics. Check it out.
eric.f.eaton eric.f.eaton
Reply | Threaded
Open this post in threaded view
♦
♦
|

unsubscribe


From: Adrienne Rollie
Date: Thu, 30 Jul 2009 10:12:36 -0700
To: <[hidden email]>
Subject: unsubscribe

Thanks,
Adrienne


Date: Thu, 30 Jul 2009 13:31:44 +0200
From: [hidden email]
Subject: unsubscribe
To: [hidden email]

 
 
Best Regards
 
Colin
 


From: Confocal Microscopy List [mailto:[hidden email]]On Behalf Of G. Esteban Fernandez
Sent: 29 July 2009 16:21
To: [hidden email]
Subject: Re: Question about a super-resolution technique

Christophe,
 
We have a Zeiss LSM 5 LIVE with a detector of fixed 512 pixels.  To achieve true 1024x1024 or 1536x1536 it uses the shift technique you describe, shifting by 1/2 or 1/3 pixel size to get 1024 or 1536, respectively.  It is true that this increases resolution when pixels are the limiting factor, it does not break the diffraction limit of resolution.
 
-Esteban

On Wed, Jul 29, 2009 at 4:34 AM, Christophe Leterrier <[hidden email]> wrote:
Dear all,

I jsut heard of a technique to obtain a super-resolution image from a series of low-resolution shifted images :
http://www.und.edu/instruct/young/toast/Super-Resolution.html
See also here : http://scien.stanford.edu/class/psych221/projects/99/stever/report.html
Basically, you extract details from sub-pixels shift in the image.

My understanding is that it could be useful for microscopy, but only for undersampled images where the limiting factor is the pixel size, and not the diffraction. Is that right that you can't break the diffusion barrier with this method ?

Christophe



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9676 fax


This E-mail is strictly confidential and intended solely for persons or organisations to whom it is addressed.If you are not the intended recipient,you must not copy, distribute or take any action in reliance upon it. If you receive this E-mail in error, please notify the IT department of Nikon U.K. Limited (02084816861) as soon as possible, delete this E-mail from your computer and destroy any printed copies.

Nikon U.K. Limited cannot accept any responsibility for the accuracy or completeness or security of this message as it has been transmitted over the Internet and has been relayed by a number of servers. If you suspect that the message has been altered, please inform us.

At Nikon we operate by best practice and exercise an 'opt-out' policy. If you do not wish to be e-mailed in future by Nikon U.K. Limited please indicate this fact in your response.

Nikon is the trading name of Nikon U.K. Limited which is a company registered in England and Wales with company number 1448987 and VAT registration number GB 243-0236-06 whose registered office is at 380 Richmond Road,Kingston,Surrey,KT2 5PR,England.


Windows Live™ Hotmail®: Celebrate the moment with your favorite sports pics. Check it out.
Arrate Mallabiabarrena Ormaechea Arrate Mallabiabarrena Ormaechea
Reply | Threaded
Open this post in threaded view
♦
♦
|

unsubscribe

In reply to this post by Wright, Colin

 

 

Arrate Mallabiabarrena

 

 

Unitat Microscòpia Òptica Avançada/ Advanced Light Microscopy Unit

 

CRG- Centre de Regulació Genòmica

C/ Dr. Aiguader, 88

08003 Barcelona

 

Tel. +34 93 316 01 95

Fax + 34 93 316 99 83

 

Ian Dobbie Ian Dobbie
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: unsubscribe

In reply to this post by eric.f.eaton
Hi everyone,

Mailing list control addresses are listed in the headers of every
message.

The relevant parts are:
--------------------
List-Help: <http://lists.umn.edu/cgi-bin/wa?LIST=CONFOCALMICROSCOPY>,
           <mailto:[hidden email]?body=INFO%20CONFOCALMICROSCOPY>
List-Unsubscribe: <mailto:[hidden email]>
List-Subscribe: <mailto:[hidden email]>
List-Owner: <mailto:[hidden email]>
List-Archive: <http://lists.umn.edu/cgi-bin/wa?LIST=CONFOCALMICROSCOPY>
--------------------

Ian
Peter Gabriel Pitrone Peter Gabriel Pitrone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: unsubscribe

What's up with all the defectors? ;-)

On Jul 31, 2009, at 10:42 AM, Ian Dobbie wrote:

Hi everyone,

Mailing list control addresses are listed in the headers of every
message.

The relevant parts are:
--------------------
List-Help: <http://lists.umn.edu/cgi-bin/wa?LIST=CONFOCALMICROSCOPY>,
            <mailto:[hidden email]?body=INFO%
20CONFOCALMICROSCOPY>
List-Unsubscribe: <mailto:CONFOCALMICROSCOPY-unsubscribe-
[hidden email]>
List-Subscribe: <mailto:CONFOCALMICROSCOPY-subscribe-
[hidden email]>
List-Owner: <mailto:[hidden email]>
List-Archive: <http://lists.umn.edu/cgi-bin/wa?LIST=CONFOCALMICROSCOPY>
--------------------

Ian
Eric Scarfone Eric Scarfone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: unsubscribe

> What's up with all the defectors? ;-)

looks like the Olympus crewmen are leaving the ship!!   

    : ^ )

 

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: Peter Pitrone <[hidden email]>
Date: Friday, July 31, 2009 11:27 am
Subject: Re: unsubscribe
To: [hidden email]


> What's up with all the defectors? ;-)
>
> On Jul 31, 2009, at 10:42 AM, Ian Dobbie wrote:
>
> Hi everyone,
>
> Mailing list control addresses are listed in the headers of every
> message.
>
> The relevant parts are:
> --------------------
> List-Help: <http://lists.umn.edu/cgi-bin/wa?LIST=CONFOCALMICROSCOPY>,

> <mailto:[hidden email]?body=INFO%
> 20CONFOCALMICROSCOPY>
> List-Unsubscribe: <mailto:CONFOCALMICROSCOPY-unsubscribe-
> [hidden email]>
> List-Subscribe: <mailto:CONFOCALMICROSCOPY-subscribe-
> [hidden email]>
> List-Owner: <mailto:[hidden email]>
> List-Archive: <http://lists.umn.edu/cgi-
> bin/wa?LIST=CONFOCALMICROSCOPY>--------------------
>
> Ian
>

Darren Korber Darren Korber
Reply | Threaded
Open this post in threaded view
♦
♦
|

Unsubscribe

In reply to this post by Wright, Colin
Unsubscribe
Gregg Sobocinski Gregg Sobocinski
Reply | Threaded
Open this post in threaded view
♦
♦
|

Reply regarding unsubscribing

In reply to this post by Eric Scarfone

I suspect the defections are a response for those who are going to be “out of office”. The policy for some listservers is that those who will be “out of the office” should unsubscribe, then resubscribe when they return. The address for this is different from list postings, so it can be performed without human intervention. I don’t remember what the official policy is for this list, but as Ian Dobbie pointed out, the address for subscribing and unsubscribing IS DIFFERENT from that for posting.

 

Below is a copy of the precise instructions I received at one point.

 

Disclaimer: I’m not the administrator, just passing along the info that I have.

~Gregg

 

Gregg Sobocinski

Microscope Imaging Specialist

University of Michigan, MCDB Dept.

Ann Arbor, Michigan

USA

 

 

SUBSCRIPTION

 

Send an email to [hidden email] (and NOT

[hidden email])  and put the following into the body of the message:

 

*subscribe confocalmicroscopy*  (note that it *isn't* "subscribe confocal")

 

Alternatively, a subscriber can go to the website http://lists.umn.edu/ and follow the directions on the web page.

 

LEAVING

 

Send an email to [hidden email] (and NOT

[hidden email])  and put the following into the body of the message:

 

*signoff confocalmicroscopy*  (note that it *isn't* "signoff confocal")

 

Alternatively, a subscriber can go to the website http://lists.umn.edu/ and follow the directions on the web page.

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Eric Scarfone
Sent: Friday, July 31, 2009 11:32 AM
To: [hidden email]
Subject: Re: unsubscribe

 

> What's up with all the defectors? ;-)

looks like the Olympus crewmen are leaving the ship!!   

    : ^ )

 

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: Peter Pitrone <[hidden email]>
Date: Friday, July 31, 2009 11:27 am
Subject: Re: unsubscribe
To: [hidden email]

> What's up with all the defectors? ;-)
>
> On Jul 31, 2009, at 10:42 AM, Ian Dobbie wrote:
>
> Hi everyone,
>
> Mailing list control addresses are listed in the headers of every
> message.
>
> The relevant parts are:
> --------------------
> List-Help: <http://lists.umn.edu/cgi-bin/wa?LIST=CONFOCALMICROSCOPY>,
> <mailto:[hidden email]?body=INFO%
> 20CONFOCALMICROSCOPY>
> List-Unsubscribe: <mailto:CONFOCALMICROSCOPY-unsubscribe-
> [hidden email]>
> List-Subscribe: <mailto:CONFOCALMICROSCOPY-subscribe-
> [hidden email]>
> List-Owner: <mailto:[hidden email]>
> List-Archive: <http://lists.umn.edu/cgi-
> bin/wa?LIST=CONFOCALMICROSCOPY>--------------------
>
> Ian
>