![]() ![]() |
Kilgore, Jason A. |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** ** vendor reply ** Hi, Sylvie, As Ulrike and others mentioned, we at Thermo Fisher Scientific do have several options to choose from in our FluoSpheres line of products. Table 1 in the product manual for FluoSpheres has a listing of the different colors and sizes offered in that range: http://tools.thermofisher.com/content/sfs/manuals/mp05000.pdf We also offer our PS-Speck beads, intended for point spread function determination, as catalog P7220, each microsphere has a diameter of 0.175 ± 0.005 µm (note the very small CV, for better reliability in sizing compared to the FluoSpheres products). https://assets.thermofisher.com/TFS-Assets/LSG/manuals/mp07220.pdf If you need more help in choosing, though, please feel free to email me separately from this thread and I'll be happy to help further. Cheers, Jason Jason A. Kilgore Technical Application Scientist Molecular Probes / EVOS Tech Support Life Sciences Solutions -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Boehm, Ulrike Sent: Tuesday, April 6, 2021 6:58 PM To: [hidden email] Subject: Re: Fluorescent beads CAUTION: This email originated from outside of Thermo Fisher Scientific. If you believe it to be suspicious, report using the Report Phish button in Outlook or send to [hidden email]. ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=L6Za7tOzk_HevvOyJ_FlKm-HlodLVxnHJkZpQzQpcJ4&s=_kKGIlSwhBNfpXTWqWpPouxBlZoC0x2uQSlm1UC4hl4&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=L6Za7tOzk_HevvOyJ_FlKm-HlodLVxnHJkZpQzQpcJ4&s=b6oj8I1qSzR7TkzCZMvloZXlx3RZcy5icALk8XTOefA&e= and include the link in your posting. ***** Dear Sylvie, I would contact ThermoFisher and ask if they have FluoSpheres in green and red in size 40 nm or 100 nm in storage (https://www.thermofisher.com/us/en/home/references/molecular-probes-the-handbook/tables/summary-of-molecular-probes-fluospheres-fluorescent-microspheres.html). Some of these beads are currently not listed on their website but believe it or not, you can still order them. I have performed some of my Ph.D. experiments with Crimson 100 nm beads that do not exist, according to their website. An entire custom-made batch costs approx. 12,000 Euro, but occasionally, labs order them custom-made, and ThermoFisher still might have some of these not listed beads in store. Best of luck. Best regards, Ulrike ____________________ Ulrike Boehm, Ph.D. Research Specialist | Advanced Imaging Center Howard Hughes Medical Institute | Janelia Research Campus 19700 Helix Drive, Ashburn, VA 20147 Email: [hidden email] Phone: 571-209-4000 x1356 -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Orestis Faklaris Sent: Tuesday, April 6, 2021 4:05 PM To: [hidden email] Subject: Re: Fluorescent beads External Email: Use Caution ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyP07XXoU$ Post images on https://urldefense.com/v3/__http://www.imgur.com__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNysq5N79E$ and include the link in your posting. ***** Hi, apart from the existing and widespread FluoSpheres, we just received and are going to test and compare the PSF of the yellow 50-90 nm beads from SpheroTech (https://urldefense.com/v3/__https://www.spherotech.com/up_to_5.htm__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNye1cA4tU$ , number FP-00552-2). We'll know more on their brightness the following days. Best Orestis __________________________________ Orestis Faklaris, PhD MRI, Montpellier Ressources Imagerie CRBM-CNRS 1919 route de Mende 34293 Montpellier FRANCE phone: +33.4.34.35.95.18 https://urldefense.com/v3/__http://www.mri.cnrs.fr__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyCwDWy4s$ -- Le 06/04/2021 à 15:27, Sylvie Le Guyader a écrit : > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confoca > lmicroscopy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKU > hhgFxghw1jWNyP07XXoU$ Post images on > https://urldefense.com/v3/__http://www.imgur.com__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNysq5N79E$ and include the link in your posting. > ***** > > Hello > > We are looking for very bright, monocolour (green or red), 50-100 nm fluorescent beads for some PSF test. > Does anyone know which beads are brightest? > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Facility Manager > Karolinska Institutet- Bionut Dpt > Blickagången 16, > Room 7362 (lab)/7840 (office) > 14157 Huddinge, Sweden > mobile: +46 (0) 73 733 5008 > LCI > website<https://urldefense.com/v3/__https://ki.se/en/bionut/welcome-to > -the-lci-facility__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3i > wmDKKUhhgFxghw1jWNyuluB-64$ > Follow our microscopy > blog!<https://urldefense.com/v3/__http://microscopykarolinska.se/__;!! > Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyYuc > 5jQI$ > > > > > > När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.com/v3/__https://ki.se/medarbetare/integritetsskyddspolicy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyoHM2xQo$ >. > > > Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://urldefense.com/v3/__https://ki.se/en/staff/data-protection-policy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyhiLgZEc$ >. |
![]() ![]() |
Cromey, Douglas W - (dcromey)-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Jason, Has Thermo considered adding an NIR dye to the four-color tetraspec beads? Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona Life Sciences North, Room 463 1333 N. Martin Ave, Tucson, AZ 85721 USA office: LSN 463 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 UA Microscopy Alliance - http://microscopy.arizona.edu/ A collaborative effort to bring information about shared microscopy facilities to the University of Arizona and the community. -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Kilgore, Jason A. Sent: Wednesday, April 7, 2021 10:31 AM To: [hidden email] Subject: [EXT]Re: Fluorescent beads **vendor reply** External Email ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** ** vendor reply ** Hi, Sylvie, As Ulrike and others mentioned, we at Thermo Fisher Scientific do have several options to choose from in our FluoSpheres line of products. Table 1 in the product manual for FluoSpheres has a listing of the different colors and sizes offered in that range: http://tools.thermofisher.com/content/sfs/manuals/mp05000.pdf We also offer our PS-Speck beads, intended for point spread function determination, as catalog P7220, each microsphere has a diameter of 0.175 ± 0.005 µm (note the very small CV, for better reliability in sizing compared to the FluoSpheres products). https://assets.thermofisher.com/TFS-Assets/LSG/manuals/mp07220.pdf If you need more help in choosing, though, please feel free to email me separately from this thread and I'll be happy to help further. Cheers, Jason Jason A. Kilgore Technical Application Scientist Molecular Probes / EVOS Tech Support Life Sciences Solutions -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Boehm, Ulrike Sent: Tuesday, April 6, 2021 6:58 PM To: [hidden email] Subject: Re: Fluorescent beads CAUTION: This email originated from outside of Thermo Fisher Scientific. If you believe it to be suspicious, report using the Report Phish button in Outlook or send to [hidden email]. ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=L6Za7tOzk_HevvOyJ_FlKm-HlodLVxnHJkZpQzQpcJ4&s=_kKGIlSwhBNfpXTWqWpPouxBlZoC0x2uQSlm1UC4hl4&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=L6Za7tOzk_HevvOyJ_FlKm-HlodLVxnHJkZpQzQpcJ4&s=b6oj8I1qSzR7TkzCZMvloZXlx3RZcy5icALk8XTOefA&e= and include the link in your posting. ***** Dear Sylvie, I would contact ThermoFisher and ask if they have FluoSpheres in green and red in size 40 nm or 100 nm in storage (https://www.thermofisher.com/us/en/home/references/molecular-probes-the-handbook/tables/summary-of-molecular-probes-fluospheres-fluorescent-microspheres.html). Some of these beads are currently not listed on their website but believe it or not, you can still order them. I have performed some of my Ph.D. experiments with Crimson 100 nm beads that do not exist, according to their website. An entire custom-made batch costs approx. 12,000 Euro, but occasionally, labs order them custom-made, and ThermoFisher still might have some of these not listed beads in store. Best of luck. Best regards, Ulrike ____________________ Ulrike Boehm, Ph.D. Research Specialist | Advanced Imaging Center Howard Hughes Medical Institute | Janelia Research Campus 19700 Helix Drive, Ashburn, VA 20147 Email: [hidden email] Phone: 571-209-4000 x1356 -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Orestis Faklaris Sent: Tuesday, April 6, 2021 4:05 PM To: [hidden email] Subject: Re: Fluorescent beads External Email: Use Caution ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyP07XXoU$ Post images on https://urldefense.com/v3/__http://www.imgur.com__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNysq5N79E$ and include the link in your posting. ***** Hi, apart from the existing and widespread FluoSpheres, we just received and are going to test and compare the PSF of the yellow 50-90 nm beads from SpheroTech (https://urldefense.com/v3/__https://www.spherotech.com/up_to_5.htm__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNye1cA4tU$ , number FP-00552-2). We'll know more on their brightness the following days. Best Orestis __________________________________ Orestis Faklaris, PhD MRI, Montpellier Ressources Imagerie CRBM-CNRS 1919 route de Mende 34293 Montpellier FRANCE phone: +33.4.34.35.95.18 https://urldefense.com/v3/__http://www.mri.cnrs.fr__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyCwDWy4s$ -- Le 06/04/2021 à 15:27, Sylvie Le Guyader a écrit : > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confoca > lmicroscopy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKU > hhgFxghw1jWNyP07XXoU$ Post images on > https://urldefense.com/v3/__http://www.imgur.com__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNysq5N79E$ and include the link in your posting. > ***** > > Hello > > We are looking for very bright, monocolour (green or red), 50-100 nm fluorescent beads for some PSF test. > Does anyone know which beads are brightest? > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Facility Manager > Karolinska Institutet- Bionut Dpt > Blickagången 16, > Room 7362 (lab)/7840 (office) > 14157 Huddinge, Sweden > mobile: +46 (0) 73 733 5008 > LCI > website<https://urldefense.com/v3/__https://ki.se/en/bionut/welcome-to > -the-lci-facility__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3i > wmDKKUhhgFxghw1jWNyuluB-64$ > Follow our microscopy > blog!<https://urldefense.com/v3/__http://microscopykarolinska.se/__;!! > Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyYuc > 5jQI$ > > > > > > När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.com/v3/__https://ki.se/medarbetare/integritetsskyddspolicy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyoHM2xQo$ >. > > > Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://urldefense.com/v3/__https://ki.se/en/staff/data-protection-policy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyhiLgZEc$ >. |
![]() ![]() |
Kilgore, Jason A. |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Not that I'm aware. But I've forwarded your suggestion to the product manager. Jason Jason A. Kilgore Technical Application Scientist Molecular Probes / EVOS Tech Support Life Sciences Solutions -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Cromey, Douglas W - (dcromey) Sent: Wednesday, April 7, 2021 11:43 AM To: [hidden email] Subject: Re: Fluorescent beads **vendor reply** CAUTION: This email originated from outside of Thermo Fisher Scientific. If you believe it to be suspicious, report using the Report Phish button in Outlook or send to [hidden email]. ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=DdeBMDW9NkVdXkGcSQiiVLII93ihVH3Y3TBEt732GZ4&s=8EVkN3XbewS4J8s-l3LWS8IdpTMxIO2LgO8cJjS3poA&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=DdeBMDW9NkVdXkGcSQiiVLII93ihVH3Y3TBEt732GZ4&s=_7zHsbJhGvjn0bv-Iwd_1E1jQK1jkzrC43Xrx5leCCw&e= and include the link in your posting. ***** Jason, Has Thermo considered adding an NIR dye to the four-color tetraspec beads? Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona Life Sciences North, Room 463 1333 N. Martin Ave, Tucson, AZ 85721 USA office: LSN 463 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 UA Microscopy Alliance - https://urldefense.proofpoint.com/v2/url?u=http-3A__microscopy.arizona.edu_&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=DdeBMDW9NkVdXkGcSQiiVLII93ihVH3Y3TBEt732GZ4&s=PQLm-mafVbl5sUKL8e9QC_1pkn77xFgxhK5UfpouxZs&e= A collaborative effort to bring information about shared microscopy facilities to the University of Arizona and the community. -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Kilgore, Jason A. Sent: Wednesday, April 7, 2021 10:31 AM To: [hidden email] Subject: [EXT]Re: Fluorescent beads **vendor reply** External Email ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=DdeBMDW9NkVdXkGcSQiiVLII93ihVH3Y3TBEt732GZ4&s=8EVkN3XbewS4J8s-l3LWS8IdpTMxIO2LgO8cJjS3poA&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=DdeBMDW9NkVdXkGcSQiiVLII93ihVH3Y3TBEt732GZ4&s=_7zHsbJhGvjn0bv-Iwd_1E1jQK1jkzrC43Xrx5leCCw&e= and include the link in your posting. ***** ** vendor reply ** Hi, Sylvie, As Ulrike and others mentioned, we at Thermo Fisher Scientific do have several options to choose from in our FluoSpheres line of products. Table 1 in the product manual for FluoSpheres has a listing of the different colors and sizes offered in that range: http://tools.thermofisher.com/content/sfs/manuals/mp05000.pdf We also offer our PS-Speck beads, intended for point spread function determination, as catalog P7220, each microsphere has a diameter of 0.175 ± 0.005 µm (note the very small CV, for better reliability in sizing compared to the FluoSpheres products). https://assets.thermofisher.com/TFS-Assets/LSG/manuals/mp07220.pdf If you need more help in choosing, though, please feel free to email me separately from this thread and I'll be happy to help further. Cheers, Jason Jason A. Kilgore Technical Application Scientist Molecular Probes / EVOS Tech Support Life Sciences Solutions -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Boehm, Ulrike Sent: Tuesday, April 6, 2021 6:58 PM To: [hidden email] Subject: Re: Fluorescent beads CAUTION: This email originated from outside of Thermo Fisher Scientific. If you believe it to be suspicious, report using the Report Phish button in Outlook or send to [hidden email]. ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=L6Za7tOzk_HevvOyJ_FlKm-HlodLVxnHJkZpQzQpcJ4&s=_kKGIlSwhBNfpXTWqWpPouxBlZoC0x2uQSlm1UC4hl4&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=L6Za7tOzk_HevvOyJ_FlKm-HlodLVxnHJkZpQzQpcJ4&s=b6oj8I1qSzR7TkzCZMvloZXlx3RZcy5icALk8XTOefA&e= and include the link in your posting. ***** Dear Sylvie, I would contact ThermoFisher and ask if they have FluoSpheres in green and red in size 40 nm or 100 nm in storage (https://www.thermofisher.com/us/en/home/references/molecular-probes-the-handbook/tables/summary-of-molecular-probes-fluospheres-fluorescent-microspheres.html). Some of these beads are currently not listed on their website but believe it or not, you can still order them. I have performed some of my Ph.D. experiments with Crimson 100 nm beads that do not exist, according to their website. An entire custom-made batch costs approx. 12,000 Euro, but occasionally, labs order them custom-made, and ThermoFisher still might have some of these not listed beads in store. Best of luck. Best regards, Ulrike ____________________ Ulrike Boehm, Ph.D. Research Specialist | Advanced Imaging Center Howard Hughes Medical Institute | Janelia Research Campus 19700 Helix Drive, Ashburn, VA 20147 Email: [hidden email] Phone: 571-209-4000 x1356 -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Orestis Faklaris Sent: Tuesday, April 6, 2021 4:05 PM To: [hidden email] Subject: Re: Fluorescent beads External Email: Use Caution ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyP07XXoU$ Post images on https://urldefense.com/v3/__http://www.imgur.com__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNysq5N79E$ and include the link in your posting. ***** Hi, apart from the existing and widespread FluoSpheres, we just received and are going to test and compare the PSF of the yellow 50-90 nm beads from SpheroTech (https://urldefense.com/v3/__https://www.spherotech.com/up_to_5.htm__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNye1cA4tU$ , number FP-00552-2). We'll know more on their brightness the following days. Best Orestis __________________________________ Orestis Faklaris, PhD MRI, Montpellier Ressources Imagerie CRBM-CNRS 1919 route de Mende 34293 Montpellier FRANCE phone: +33.4.34.35.95.18 https://urldefense.com/v3/__http://www.mri.cnrs.fr__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyCwDWy4s$ -- Le 06/04/2021 à 15:27, Sylvie Le Guyader a écrit : > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confoca > lmicroscopy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKU > hhgFxghw1jWNyP07XXoU$ Post images on > https://urldefense.com/v3/__http://www.imgur.com__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNysq5N79E$ and include the link in your posting. > ***** > > Hello > > We are looking for very bright, monocolour (green or red), 50-100 nm fluorescent beads for some PSF test. > Does anyone know which beads are brightest? > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Facility Manager > Karolinska Institutet- Bionut Dpt > Blickagången 16, > Room 7362 (lab)/7840 (office) > 14157 Huddinge, Sweden > mobile: +46 (0) 73 733 5008 > LCI > website<https://urldefense.com/v3/__https://ki.se/en/bionut/welcome-to > -the-lci-facility__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3i > wmDKKUhhgFxghw1jWNyuluB-64$ > Follow our microscopy > blog!<https://urldefense.com/v3/__http://microscopykarolinska.se/__;!! > Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyYuc > 5jQI$ > > > > > > När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.com/v3/__https://ki.se/medarbetare/integritetsskyddspolicy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyoHM2xQo$ >. > > > Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://urldefense.com/v3/__https://ki.se/en/staff/data-protection-policy__;!!Eh6p8Q!RacJT6MFQZREzM0VgS723rZx8jF4mfPN6Td8xcV3iwmDKKUhhgFxghw1jWNyhiLgZEc$ >. |
Free forum by Nabble | Edit this page |