Re: Gold Standard immersion oil

classic Classic list List threaded Threaded
♦
♦
8 messages Options
Claire Brown Claire Brown
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Gold Standard immersion oil

Just doing some catching up on the listserv so sorry for the late reply, but
I would also recommend the 37oC oil from Cargille. It has RI 1.52 at 37oC.

Claire
Beat Ludin Beat Ludin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Gold Standard immersion oil

What about autofluorscence? Last time I tested this oil (which is
quite a while ago) I found that it had very high autofluorescence. Of
course, this is much more critical for widefield than for confocal.

Beat


At 18:32 24-09-2008, you wrote:
>Just doing some catching up on the listserv so sorry for the late reply, but
>I would also recommend the 37oC oil from Cargille. It has RI 1.52 at 37oC.
>
>Claire
Neil Kad Neil Kad
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Gold Standard immersion oil

Hi all,

A quick update on the oil, I just tried out the DF from Cargille, again it has some fluorescence in the oil, its not too bad though but I will be trying the Olympus brand next - keep tuned!

Neil Kad

> Date: Wed, 24 Sep 2008 19:06:51 +0200
> From: [hidden email]
> Subject: Re: Gold Standard immersion oil
> To: [hidden email]
>
> What about autofluorscence? Last time I tested this oil (which is
> quite a while ago) I found that it had very high autofluorescence. Of
> course, this is much more critical for widefield than for confocal.
>
> Beat
>
>
> At 18:32 24-09-2008, you wrote:
> >Just doing some catching up on the listserv so sorry for the late reply, but
> >I would also recommend the 37oC oil from Cargille. It has RI 1.52 at 37oC.
> >
> >Claire


Win £3000 to spend on whatever you want at Uni! Click here to WIN!
Beat Ludin Beat Ludin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Gold Standard immersion oil

Thanks Neil. However, my question referred to the
type 37 oil mentioned by Claire, not the type DF.
In contrast to type DF (which is optimized for
fluorescence microscopy), type 37 (which is
optimized for 37°C) used to have very high
autofluorescence and I wondered whether that had been improved in the meantime.

At 10:35 25-09-2008, you wrote:

>Hi all,
>
>A quick update on the oil, I just tried out the
>DF from Cargille, again it has some fluorescence
>in the oil, its not too bad though but I will be
>trying the Olympus brand next - keep tuned!
>
>Neil Kad
>
> > Date: Wed, 24 Sep 2008 19:06:51 +0200
> > From: [hidden email]
> > Subject: Re: Gold Standard immersion oil
> > To: [hidden email]
> >
> > What about autofluorscence? Last time I tested this oil (which is
> > quite a while ago) I found that it had very high autofluorescence. Of
> > course, this is much more critical for widefield than for confocal.
> >
> > Beat
> >
> >
> > At 18:32 24-09-2008, you wrote:
> > >Just doing some catching up on the listserv
> so sorry for the late reply, but
> > >I would also recommend the 37oC oil from Cargille. It has RI 1.52 at 37oC.
> > >
> > >Claire
>
>
>----------
>Win £3000 to spend on whatever you want at Uni!
><http://clk.atdmt.com/UKM/go/111354032/direct/01/>Click here to WIN!
Jason Swedlow Jason Swedlow
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Gold Standard immersion oil

Sorry to beat a dead horse, but the autofluorescence issue is why we used LaserLiquids by Cargille (see previous posting).  A bit more money, but worth it, at least IMHO.

Cheers,

Jason

On Thu, Sep 25, 2008 at 9:55 AM, Beat Ludin <[hidden email]> wrote:
Thanks Neil. However, my question referred to the type 37 oil mentioned by Claire, not the type DF. In contrast to type DF (which is optimized for fluorescence microscopy), type 37 (which is optimized for 37°C) used to have very high autofluorescence and I wondered whether that had been improved in the meantime.


At 10:35 25-09-2008, you wrote:
Hi all,

A quick update on the oil, I just tried out the DF from Cargille, again it has some fluorescence in the oil, its not too bad though but I will be trying the Olympus brand next - keep tuned!

Neil Kad

> Date: Wed, 24 Sep 2008 19:06:51 +0200
> From: [hidden email]
> Subject: Re: Gold Standard immersion oil
> To: [hidden email]
>
> What about autofluorscence? Last time I tested this oil (which is
> quite a while ago) I found that it had very high autofluorescence. Of
> course, this is much more critical for widefield than for confocal.
>
> Beat
>
>
> At 18:32 24-09-2008, you wrote:
> >Just doing some catching up on the listserv so sorry for the late reply, but
> >I would also recommend the 37oC oil from Cargille. It has RI 1.52 at 37oC.
> >
> >Claire


----------
Win £3000 to spend on whatever you want at Uni! <http://clk.atdmt.com/UKM/go/111354032/direct/01/>Click here to WIN!


Hugo.Ostermann Hugo.Ostermann
Reply | Threaded
Open this post in threaded view
♦
♦
|

AW: Gold Standard immersion oil

In reply to this post by Beat Ludin
Hi Listers,
I can proof this. The 37 Degree oils ha a much higher background than the
Laser Fluid which has no measurable background.

With kind regards Hugo Ostermann

CHROMAPHOR Analysen-Technik GmbH
Mühlenkamp 37
D-59387 Ascheberg - Germany
Tel:   (++49) 0172- 28 28 373
mail:  [hidden email]
URL: http://www.chromaphor.de

-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List [mailto:[hidden email]] Im
Auftrag von Beat Ludin
Gesendet: Donnerstag, 25. September 2008 10:55
An: [hidden email]
Betreff: Re: Gold Standard immersion oil

Thanks Neil. However, my question referred to the
type 37 oil mentioned by Claire, not the type DF.
In contrast to type DF (which is optimized for
fluorescence microscopy), type 37 (which is
optimized for 37°C) used to have very high
autofluorescence and I wondered whether that had been improved in the
meantime.

At 10:35 25-09-2008, you wrote:

>Hi all,
>
>A quick update on the oil, I just tried out the
>DF from Cargille, again it has some fluorescence
>in the oil, its not too bad though but I will be
>trying the Olympus brand next - keep tuned!
>
>Neil Kad
>
> > Date: Wed, 24 Sep 2008 19:06:51 +0200
> > From: [hidden email]
> > Subject: Re: Gold Standard immersion oil
> > To: [hidden email]
> >
> > What about autofluorscence? Last time I tested this oil (which is
> > quite a while ago) I found that it had very high autofluorescence. Of
> > course, this is much more critical for widefield than for confocal.
> >
> > Beat
> >
> >
> > At 18:32 24-09-2008, you wrote:
> > >Just doing some catching up on the listserv
> so sorry for the late reply, but
> > >I would also recommend the 37oC oil from Cargille. It has RI 1.52 at
37oC.
> > >
> > >Claire
>
>
>----------
>Win £3000 to spend on whatever you want at Uni!
><http://clk.atdmt.com/UKM/go/111354032/direct/01/>Click here to WIN!
Lauran Oomen Lauran Oomen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Gold Standard immersion oil

Dear all,

Recently Cargille and we have developed and tested a new 37 degree oil for high-resolution live-cell fluorescence microscopy. This oil is based on the DF oil mentioned. The test results, including those on the level of autofluorescence, are described in a paper which will be published in a forthcoming issue of the J. Microscopy. If you want more information, please contact me off-list.

I would like to add that I have hesitated a while to respond to this thread, since it might look like self-advertisement and a commercial response. However, I do not have any financial interest in the product or the company.

Kind regards,
Lauran
--------------------------------------------------------------
Lauran Oomen
Manager Digital Microscopy Facility (H003)
NKI-AVL
Plesmanlaan 121
PO Box 90203
1006 BE Amsterdam
The Netherlands
 
phone +31 205126080
fax +31 205122909


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Hugo.Ostermann
Sent: donderdag 25 september 2008 21:32
To: [hidden email]
Subject: AW: Gold Standard immersion oil

Hi Listers,
I can proof this. The 37 Degree oils ha a much higher background than the
Laser Fluid which has no measurable background.

With kind regards Hugo Ostermann

CHROMAPHOR Analysen-Technik GmbH
Mühlenkamp 37
D-59387 Ascheberg - Germany
Tel:   (++49) 0172- 28 28 373
mail:  [hidden email]
URL: http://www.chromaphor.de

-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List [mailto:[hidden email]] Im
Auftrag von Beat Ludin
Gesendet: Donnerstag, 25. September 2008 10:55
An: [hidden email]
Betreff: Re: Gold Standard immersion oil

Thanks Neil. However, my question referred to the
type 37 oil mentioned by Claire, not the type DF.
In contrast to type DF (which is optimized for
fluorescence microscopy), type 37 (which is
optimized for 37°C) used to have very high
autofluorescence and I wondered whether that had been improved in the
meantime.

At 10:35 25-09-2008, you wrote:

>Hi all,
>
>A quick update on the oil, I just tried out the
>DF from Cargille, again it has some fluorescence
>in the oil, its not too bad though but I will be
>trying the Olympus brand next - keep tuned!
>
>Neil Kad
>
> > Date: Wed, 24 Sep 2008 19:06:51 +0200
> > From: [hidden email]
> > Subject: Re: Gold Standard immersion oil
> > To: [hidden email]
> >
> > What about autofluorscence? Last time I tested this oil (which is
> > quite a while ago) I found that it had very high autofluorescence. Of
> > course, this is much more critical for widefield than for confocal.
> >
> > Beat
> >
> >
> > At 18:32 24-09-2008, you wrote:
> > >Just doing some catching up on the listserv
> so sorry for the late reply, but
> > >I would also recommend the 37oC oil from Cargille. It has RI 1.52 at
37oC.
> > >
> > >Claire
>
>
>----------
>Win £3000 to spend on whatever you want at Uni!
><http://clk.atdmt.com/UKM/go/111354032/direct/01/>Click here to WIN!
Bruno Afonso Bruno Afonso
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Gold Standard immersion oil

Hi Lauren,

 
This is great news. When will this be available? I'm also looking to know if they will have different refraction indices :)

 
do you have a preprint you could send me?

 
best
b

On 9/26/08, [hidden email] <[hidden email]> wrote:
Dear all,

Recently Cargille and we have developed and tested a new 37 degree oil for high-resolution live-cell fluorescence microscopy. This oil is based on the DF oil mentioned. The test results, including those on the level of autofluorescence, are described in a paper which will be published in a forthcoming issue of the J. Microscopy. If you want more information, please contact me off-list.

I would like to add that I have hesitated a while to respond to this thread, since it might look like self-advertisement and a commercial response. However, I do not have any financial interest in the product or the company.

Kind regards,
Lauran
--------------------------------------------------------------
Lauran Oomen
Manager Digital Microscopy Facility (H003)
NKI-AVL
Plesmanlaan 121
PO Box 90203
1006 BE Amsterdam
The Netherlands

phone +31 205126080
fax +31 205122909



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Hugo.Ostermann
Sent: donderdag 25 september 2008 21:32
To: [hidden email]

Subject: AW: Gold Standard immersion oil

Hi Listers,
I can proof this. The 37 Degree oils ha a much higher background than the
Laser Fluid which has no measurable background.

With kind regards Hugo Ostermann

CHROMAPHOR Analysen-Technik GmbH
Mühlenkamp 37
D-59387 Ascheberg - Germany
Tel:   (++49) 0172- 28 28 373
mail:  [hidden email]
URL: http://www.chromaphor.de

-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List [mailto:[hidden email]] Im
Auftrag von Beat Ludin
Gesendet: Donnerstag, 25. September 2008 10:55
An: [hidden email]
Betreff: Re: Gold Standard immersion oil

Thanks Neil. However, my question referred to the
type 37 oil mentioned by Claire, not the type DF.
In contrast to type DF (which is optimized for
fluorescence microscopy), type 37 (which is
optimized for 37°C) used to have very high
autofluorescence and I wondered whether that had been improved in the
meantime.

At 10:35 25-09-2008, you wrote:
>Hi all,
>
>A quick update on the oil, I just tried out the
>DF from Cargille, again it has some fluorescence
>in the oil, its not too bad though but I will be
>trying the Olympus brand next - keep tuned!
>
>Neil Kad
>
> > Date: Wed, 24 Sep 2008 19:06:51 +0200
> > From: [hidden email]
> > Subject: Re: Gold Standard immersion oil
> > To: [hidden email]
> >
> > What about autofluorscence? Last time I tested this oil (which is
> > quite a while ago) I found that it had very high autofluorescence. Of
> > course, this is much more critical for widefield than for confocal.
> >
> > Beat
> >
> >
> > At 18:32 24-09-2008, you wrote:
> > >Just doing some catching up on the listserv
> so sorry for the late reply, but
> > >I would also recommend the 37oC oil from Cargille. It has RI 1.52 at
37oC.
> > >
> > >Claire
>
>
>----------
>Win £3000 to spend on whatever you want at Uni!
><http://clk.atdmt.com/UKM/go/111354032/direct/01/>Click here to WIN!



--
Bruno Afonso
http://brunoafonso.com (personal, mostly portuguese)
http://openwetware.org/wiki/User:BrunoAfonso (Professional, english)