Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***
Commercial Response:
 
precisExcite LED light source in USA
 
Following earlier discussions on the Confocal Listserver, CoolLED is now pleased to advise that a licensing agreement is in effect.  As such, our precisExcite LED source is now available in the USA.
 
The US research community can now enjoy the benefits of precisExcite LED excitation which include stability, long-lifetime, ease of control, and integration with leading imaging software packages.
 
We will be pleased to answer further specific questions sent to us directly.  You can contact precisExcite using the details below or via our website at www.precisExcite.com.
 
 
Best Regards
 
JIM Beacher / precisExcite

CIL House, Charlton Road
Andover, Hants. SP10 3JL
England, U.K.
 
USA (Toll Free)          1.800.877.0128
UK/Worldwide   +44 (0) 1264 320989
 
precisExcite - Simply Better Control
 
This email and its attachments may be confidential and are intended solely for the use of the individual(s) to whom it is addressed. Please contact the sender if you believe that you have received this email in error.
 
 
-----Original Message-----
From: "James Beacher" <[hidden email]>
Sent: 08/05/2008 10:28
To: "Confocal Microscopy List" <[hidden email]>
Subject: RE: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***


 
Commercial Response
 
We would like to clarify the situation regarding of sales of our precisExcite LED light source in the USA.  We are currently not in a position to do so due to a patent issue.  We are in discussion with the owners of the patent and are seeking a resolution.  We will make an announcement as soon as we are in a position to do so.
 
We have a number of distribution agreements with microscope manufacturers and imaging software companies for sales worldwide.  At this time these agreements are similarly limited to exclude the USA.
 
We very much hope that this will be resolved soon.
 
Separately, we can advise that the flexibility of our LED system is now established with 11 discrete LED wavelengths available.  We will continue to add new wavelengths in the future.  LEDs will certainly address the problems discussed in this recent thread.
 
 
JIM Beacher
CoolLED Ltd.
 
-----Original Message-----
From: "Confocal Microscopy List" <[hidden email]>
Sent: 07/05/2008 16:44
To: "[hidden email]" <[hidden email]>
Subject: Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

Is this true?  Thanks for the response Gary but we had an e-mail from
Jim Beacher recently informing us that CoolLED could not currently
supply it to the USA.  Jim, could we have clarification on this issue
please?  Is it that only particular models are available here?

Thanks,
Alison

P.S.  Since an Andor person has taken the trouble to help me out,
I feel I should be allowed to comment on an Andor product at this point
- we finally got round to demo-ing the iXon 512 EMCCD on our spinning
disk yesterday and now I see what all the fuss is about - fabulous
camera guys!!!





Gary Laevsky wrote:


> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Alison and all,
>
> CoolLed has a 3 channel version available in the states for $8,500 list.
>
> Best,
>
> 
>
> Gary Laevsky, Ph.D.
>
> Imaging Application Specialist
>
> 
>
> Andor Technology
>
> discover new ways of seeing
>
> 
>
> [hidden email]
>
> Cell          (774) 291 - 9992
> Office       (860) 290 - 9211 x219
> Fax          (860) 290 - 9566
> Web:       www.andor.com
>
> -----Original Message-----
> From: Confocal Microscopy List [[hidden email]] On
> Behalf Of Alison North
> Sent: Wednesday, May 07, 2008 10:51 AM
> To: [hidden email]
> Subject: Re: HBO/Mercury lamp lifetimes
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Jim,
>
> Oh you cruel man, you are mercilessly TAUNTING those of us based in the
> USA who would love to buy your LEDs but are prevented from doing so by
> exceptionally irritating patent restrictions (hello to the company
> concerned - your LED system is prohibitively expensive!). Or am I
> missing something and this situation has recently changed?  Do keep us
> posted please?  And yes, I am intentionally sending this to the
> listserver, so that other people based in this country don't get all
> excited by your LEDs and have their hopes dashed, like us.  Plus, if I
> am incorrect about the patent situation and we can in fact obtain an
> LED-based lamp here at a reasonable price (without having to assemble
> everything ourselves), we would like to know.
>
> Thanks and best wishes,
> Alison
>
>
> James Beacher wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>> Commercial Interest
>> 
>> 
>> Having followed this discussion, we can contain ourselves no
>> longer.........*LEDs!*
>> 
>> 
>> 
>> JIM Beacher
>> CoolLED Ltd
>> 
>> 
>>
>>     -----Original Message-----
>>     *From:* "Confocal Microscopy List" <[hidden email]>
>>     *Sent:* 07/05/2008 14:50
>>     *To:* "[hidden email]"
> <[hidden email]>
>>     *Subject:* Re: HBO/Mercury lamp lifetimes
>>
>>
>>     Search the CONFOCAL archive at
>>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>     Hi all,
>>
>>     I would like to go back to an earlier post, and reiterate that
> there
>>     are several different types of power supplies. The
> constant-current
>>     ones increase the applied voltage as the "conductivity" of the
> bulb
>>     decreases. This means that the amount of power deposited in the
> arc
>>     slowly INCREASES as the electrodes age and eventually, this will
>>     cause overheating, and softening of the fused-quartz envelope.
>>     Constant-power supplies will not cause this problem, but the
>>     intensity of the brightest part of the arc will slowly deteriorate
>>     with time.
>>
>>     Another factor is dirt. Either finger grease or perhaps some
>>     accidental event such as a small, fried insect or a very small
> piece
>>     of flaked-off black paint. In short, anything that absorbs light
> and
>>     gets hot while attached in some way to the quartz envelope. Any
> "dark
>>     spot" will cause local overheating and asymmetrical envelope
> failure.
>>     A second possibility has to to with the connections between the
> bulb
>>     and the housing. If the mating surfaces of these metal parts
> become
>>     corroded, or oxidized, the contact resistance will increase as
> will
>>     the working temperature. This also can lead to premature envelope
>>     failure.
>>
>>     Even the alignment of the spherical rear mirror can be important.
> If
>>     the reflected image of the arc is formed not to the side of the
> arc
>>     itself but concentrated onto a small section of the envelope, this
>>     can exacerbate any asymmetrical heating cause by dirt or
>>     contamination.
>>
>>     The effect of these "operator" factors will vary somewhat
> depending
>>     on the type of power supply used.
>>
>>     Therefore, it seems to me that the list of important variables
> should
>>     include not just be the bulb manufacturer, the bulb type  and the
>>     design of the power supply but the care with which the new bulb
> has
>>     been installed and aligned.
>>
>>     And then there is the whole matter of variable-power Hg systems.
> We
>>     usually use our Atto-arc 50W at about 20% power and the bulbs
> often
>>     seem to last for over a thousand hours. More than enough
> brightness
>>     to find  the part of your sample best suited to confocal analysis.
>>
>>     Finally, there is the matter of xenon arcs. These have all the
> same
>>     problems mentioned above, but in addition, because they are
> reputed
>>     to start off having a much higher gas pressure, any resulting
>>     explosions are said to be much more impressive, in fact even
>>     dangerous (albeit, without the additional nuisance of Hg
>>     contamination). Has anyone had any experience here?
>>
>>     Cheers,
>>
>>     Jim Pawley
>>
>>
>>      >Search the CONFOCAL archive at
>>      >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>      >
>>      >Hello everyone
>>      >
>>      >I think we are barking under the wrong tree.  Please correct me
> if i
>>      >am wrong, but many explosions I read reported in this thread
> (and
>>      >only one -in 25 years- which happened to me @50hrs) happened at
> way
>>      >less than the rated hours.
>>      >
>>      >Shouldn't  we revise the myth that changing the lamp early
> protects
>>      >from explosions?
>>      >
>>      >It seems to me that it is either due to material failure (in
> which
>>      >case I think the lamp manufacturers should pay the damage bill)
> or
>>      >too many ignitions, in which case the manufacturers should
> include
>>      >this count in the specs.
>>      >
>>      >Regards,
>>      >
>>      >*********************************
>>      >Stamatis Pagakis Ph.D.
>>      >Biological Imaging Unit
>>      >Biomedical Research Foundation, Academy of Athens,  Greece
>>      >
>>      >>On 4 May 2008, at 00:54, Russell Spear wrote:
>>      >>Search the CONFOCAL archive at
>>      >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>      >>
>>      >>So it isn't worth the risk to use them past rated hours.
>>      >>
>>      >>Russ
>>      >>
>>      >>
>>      >>
>>      >>Russell N. Spear
>>      >>Sr. Research Specialist
>>      >>Dept. of Plant Pathology
>>      >>Univ. of Wisconsin
>>      >>1630 Linden Dr.
>>      >>Madison WI 53706
>>      >>
>>      >>voice 608.263.2093
>>      >>fax     608.263.2626
>>
>>
>>     --
>>                    **********************************************
>>     Prof. James B. Pawley,                                      Ph.
>>     608-263-3147
>>     Room 223, Zoology Research Building,             
>>     FAX  608-265-5315
>>     1117 Johnson Ave., Madison, WI, 53706
>>     [hidden email]
>>     3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>>     Vancouver Canada
>>     Info: http://www.3dcourse.ubc.ca/            Applications due by
>>     March 15, 2008
>>                    "If it ain't diffraction, it must be statistics."
> Anon.
>

--