Dear Renato,

Ironically, the big issue here is the NA of the objective.  The larger the NA, the shorter the working distance.  The shorter working distance wil limit how far you can go into the sample without running into it.  It was a lesson I learned the hard way, in my early days as technical marketing manager and field support for Sarastro, one of the pioneers in confocal microscopy.  We were running high resolution scans which took a great deal of time.  After about the first 10 sections, the image just began to go out of focus. It didn't occur to me for over 20 minutes that we were working with a high NA objective, with a VERY short working distance, and that the rest of the scans were just pushing the objective into the sample!  How embarrassing!  But it was one trial learning. 

Please remember that working with a lower NA is a trade-off. While you will be able to image deeply into the tissue, your individual sections will not be as thin as usual.  If you are interested in the general structure of the tissue, this will not be an issue.  If you are trying to image very fine particles within the tissue, that might be another story. 

George's comments are well founded.  While at Sarastro, I  imaged 300microns into a proto-type of a J&J wet bandage, showing the gel layer and then the non-woven fabric underneath, as well as 200+ microns into poly-urethane foams and, if memory serves, about 200microns into central nervous system tissue, all with lower mag/lower NA objectives. 

As with any microscopy technique, now that you know the fundamentals, give it a try and find out what works best.

Good hunting!
Barbara Foster, President and Sr. Consultant

Microscopy/Microscopy Education
7101 Royal Glen Trail, Suite A
McKinney TX 75070
P: (972)924-5310  Skype: fostermme
W: www.MicroscopyEducation.com

NEWS! Visit the NEW and IMPROVED www.MicroscopyEducation.com! And don't forget:  MME is now scheduling customized, on-site courses for the balance of the year.  Call me for a free assessment and quote.

At 05:24 PM 10/5/2009, you wrote:
Commercial interest:

We have set up several CSU-10s and CSUX systems for similar purposes. One paper describing this system was published last year.

Disease Models Mech.
1, 155 (2008)  cited as an editors Choice in Science Magazine.

80 to 100 ums into live tissue was obtainable in 4 colors using air lens.
Water or glycerin immersion lens should improve this.

Best regards

George A. Peeters MD, MS

President,  Solamere Technology Group Inc

1427 Perry Ave

Salt Lake City, UT 84103

www.solameretech.com

801 322-2645 office          801 322-2645 fax

801 232-6911 cell

On Oct 5, 2009, at 2:50 PM, Kathryn Spencer wrote:

Hi Renato;
            We routinely image 200 micron mouse brain sections with our Yokogawa CSU-10 spinning disk. Effectively, we can see about 100 microns well. We are using a 20x LUCPlan FluorN 0.45 dry objective. This is mostly due to constraints of using Millipore chambers for culturing the mouse brains; we need the extra working distance. I realize the mismatch between the pinholes and the objective NA. Our four day, time-lapse images are really very nice with this objective, although our expression levels need to be rather high for the cell processes to be seen.
            Best,
            Kathy
 
 
From: Confocal Microscopy List [[hidden email][hidden email]] On Behalf Of Renato Mortara
Sent: Monday, October 05, 2009 1:04 PM
To: [hidden email]
Subject: How thick can samples be to be imaged on a Spinning Disk Confocal ?
 
Hello,
 
I am in the process of deciding the best possible configuration to assemble a spinning disk confocal with the Yokogawa CSU-X1 scanning head.
 
It is common knowledge that imaging 'thick' samples can be tricky or simply not feasible with spinning disk confocals.
 
Does anyone there have practical experience for instance, with mouse brain sections ?
 
Many thanks for the input,
 
Best
 
Renato
 