Colocalization based upon "yellow" could be accurate, if and only if,  
the intensities are comparable and pixel (voxel) quantities in the  
suspected colocalized volumes are in roughly equal.  .  Otherwise,  
the yellow is masked by the predominate channel.  Something small,  
like lysosomes, would need to be sampled properly. Colocalization  
could be masked by blur unless deconvolved, even if images are  
collected with a confocal.
On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi all,
>
> It seems that in many papers from biologists or chemists, and i'm  
> talking
> high impact factors journals,  colocalisation of two elements is is  
> often
> assumed  by simple color superposition (ex: red and green fluoresce  
> yellow
> when colocalising), while microscopists (many physisists I suppose)  
> seem to
> need a more complex software-based confirmation.
> Is it ok, when using high end equipment and corrected objectives  
> (apochromat
> with high NA for ex.), to assume colocalisation by color  
> superposition,
> especially when fluorophore are confined to small volume entities,  
> like
> lysosomes ?
>
> Thanks
>
> Marc
M. van de corput M. van de corput
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

a good tutorial review  by Bolte & Cordelieres  J. Micr, 2006 is very
helpful understanding co-localization analysis.

Mariette

____________________________________________

Dr. M.P.C. Kemner-van de Corput,
____________________________________________

MGC - Dept. of Cell Biology & Genetics
Erasmus Medical Center
Dr. Molewaterplein 50, 3015 GE Rotterdam
POB 2040, 3000 CA Rotterdam, The Netherlands

Office: H-Ee751; tel: +31 10 704.3949
Lab: H-Ee710; tel lab: +31 10 704.3315
tel secr: +31 10 704.3169
____________________________________________

http://www2.eur.nl/fgg/ch1/cellbiology/
http://www.thesis.kemner.biz/
____________________________________________

Op Vr, 28 maart, 2008 10:54 pm, schreef Glen MacDonald:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Colocalization based upon "yellow" could be accurate, if and only if,
> the intensities are comparable and pixel (voxel) quantities in the
> suspected colocalized volumes are in roughly equal.  .  Otherwise,
> the yellow is masked by the predominate channel.  Something small,
> like lysosomes, would need to be sampled properly. Colocalization
> could be masked by blur unless deconvolved, even if images are
> collected with a confocal.
> On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi all,
>>
>> It seems that in many papers from biologists or chemists, and i'm
>> talking
>> high impact factors journals,  colocalisation of two elements is is
>> often
>> assumed  by simple color superposition (ex: red and green fluoresce
>> yellow
>> when colocalising), while microscopists (many physisists I suppose)
>> seem to
>> need a more complex software-based confirmation.
>> Is it ok, when using high end equipment and corrected objectives
>> (apochromat
>> with high NA for ex.), to assume colocalisation by color
>> superposition,
>> especially when fluorophore are confined to small volume entities,
>> like
>> lysosomes ?
>>
>> Thanks
>>
>> Marc
>
Jeremy Adler Jeremy Adler
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Glen MacDonald-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Problems with establishing colocalisation by overlaying images using different layers of an RGB image

1) A change to the detector gain and fiddling in Photoshop changes the apparent colocalisation.

2) The appearance of pixels, where the intensities are similar but are low, is highly dependent on the setup of the monitor and on limitations of printing.

3) Interpretation is therefore highly subjective and Journals should demand a higher level of evidence.

4) Software for quantitation is widely available. Most of which generates a very standard set of measurements.

It should be recognized that the quantitative and non quantitative methods are dependent upon having images of good quality and that are precisely aligned

   

Jeremy Adler
Cell Biology
The Wenner-Gren Inst.
Arrhenius Laboratories E5
Stockholm University
Stockholm 106 91
Sweden



-----Original Message-----
From: Confocal Microscopy List on behalf of Glen MacDonald
Sent: Fri 3/28/2008 22:54
To: [hidden email]
Subject: Re: colocalisation without software
 

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi all,
>
> It seems that in many papers from biologists or chemists, and i'm  
> talking
> high impact factors journals,  colocalisation of two elements is is  
> often
> assumed  by simple color superposition (ex: red and green fluoresce  
> yellow
> when colocalising), while microscopists (many physisists I suppose)  
> seem to
> need a more complex software-based confirmation.
> Is it ok, when using high end equipment and corrected objectives  
> (apochromat
> with high NA for ex.), to assume colocalisation by color  
> superposition,
> especially when fluorophore are confined to small volume entities,  
> like
> lysosomes ?
>
> Thanks
>
> Marc
Valeria Berno Valeria Berno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Glen MacDonald-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

This question just fit in perfectly on what I am trying to find out about
colocalization.

When and why do I need do deconvolve pictures collected with a confocal in
order to be sure about my colocalization (or not colocalization) results?

To be specific: I am working on pre and post-synaptic proteins.

Thanks

Valeria



> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Colocalization based upon "yellow" could be accurate, if and only if,
> the intensities are comparable and pixel (voxel) quantities in the
> suspected colocalized volumes are in roughly equal.  .  Otherwise,
> the yellow is masked by the predominate channel.  Something small,
> like lysosomes, would need to be sampled properly. Colocalization
> could be masked by blur unless deconvolved, even if images are
> collected with a confocal.
> On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi all,
>>
>> It seems that in many papers from biologists or chemists, and i'm
>> talking
>> high impact factors journals,  colocalisation of two elements is is
>> often
>> assumed  by simple color superposition (ex: red and green fluoresce
>> yellow
>> when colocalising), while microscopists (many physisists I suppose)
>> seem to
>> need a more complex software-based confirmation.
>> Is it ok, when using high end equipment and corrected objectives
>> (apochromat
>> with high NA for ex.), to assume colocalisation by color
>> superposition,
>> especially when fluorophore are confined to small volume entities,
>> like
>> lysosomes ?
>>
>> Thanks
>>
>> Marc
>
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal =
Hi Valeria, 

Here are a few thoughts:

Deconvolution improves your images in different ways: it removes out of focus blur (more so on widefield images), it reduces noise (widefield and confocal), improves signal to noise (widefield and confocal), and "tightens" the PSF (or the boundaries of the image of your objects). All these factors will improve your colocalization studies. For instance, with better signal to noise, you get sharper, cleaner intensity peaks for your objects... these will be easier to threshold or otherwise analyze for colocalization. Less noise in your images will also facilitate your colocalization studies. In addition to those practical considerations, a deconvolved image gives a more accurate 3-Dd representation of your objects, therefore leading to more accurate colocalization results. 

One important issue to consider is the PSF: the image of a point object is some sort of  blurred 3-D football (the American, elongated type). Deconvolution will shrink the boundaries of this football to some extent, and make them crisper (better defined). Even so, the image of a point object is a larger 3-D object (or PSF). The size of the PSF determines the degree of uncertainty in the actual size and boundaries of your object. The error is larger in the vertical axis, and lesser in the horizontal directions. For example, with a high NA 60x/1.4 oil lens, the dimensions of the PASF will be in the order of 0.2 microns in x/y and 0.5 microns in z.  From this you can see that if you have two molecules located 0.1 micron apart (let's say along the x axis, their deconvolved images will overlap to some extent, even though the molecules do not actually touch. If the two molecules are 0.2 or even 0.3 microns apart along the z axis, their images will also similarly appear to overlap.  However, by analyzing the images carefully (and better so in 3-D), and knowing the imaging properties of your objective (the PSF), you could probably determine that the images you are seeing are images of very small point objects, could determine the actual location of those objects (center of intensity of the images), and conclude there is no overlap. On non deconvolved images, this would be harder to do because there will be more blur, and more uncertainty in the actual size and location of the images (because of more noise in the images). 

Finally, you can see that since the image of an object is generally inflated by a well-defined quantity (namely, by the radius of the PSF). Therefore, the smaller and the closer the objects of interest are, the greater the potential for error: the difference between the size of a GFP molecule and its image is proportionally much greater than the difference between the size of a nucleus and the size of its image, because the dimensions of the PSF are relatively small compared to the size of the nucleus, but is quite large compared to the size of a molecule. Therefore, image quality, deconvolution, and knowing the properties of the microscope (especially the PSF of the particular objective used), will be much more important when you are looking at colocalization of small, faint objects, located close to each other, and will be less critical when looking at large objects...

Similarly, when you are looking at very small objects near the resolution limit, it is even more critical to use optimal sampling (have a pixel size that's about 2.5 times smaller than the (width at 50% max intensity of the) PSF). Again, a factor of two in pixel size will be proportionally huge when trying to analyze very small punctate objects or filaments of dimensions comparable to the PSF, while it will be less critical if you are looking at whole cells within a tissue, because in that case, the difference in pixel size is tiny compared to the size of the cell. 

Understanding the concepts of resolution, PSF, and sampling are critical to producing accurate quantitative colocalization (or other) data from microscopy images, especially of very small objects near the limit of resolution. I really recommend taking some time to read about these concepts, e.g. at this site:


(search for PSF, resolution, and such)

Hope this helps...


Julio.


--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024


==


On Mar 31, 2008, at 8:05 AM, Valeria Berno wrote:

Search the CONFOCAL archive at

Hi,

This question just fit in perfectly on what I am trying to find out about
colocalization.

When and why do I need do deconvolve pictures collected with a confocal in
order to be sure about my colocalization (or not colocalization) results?

To be specific: I am working on pre and post-synaptic proteins.

Thanks

Valeria



Search the CONFOCAL archive at

Colocalization based upon "yellow" could be accurate, if and only if,
the intensities are comparable and pixel (voxel) quantities in the
suspected colocalized volumes are in roughly equal.  .  Otherwise,
the yellow is masked by the predominate channel.  Something small,
like lysosomes, would need to be sampled properly. Colocalization
could be masked by blur unless deconvolved, even if images are
collected with a confocal.
On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:

Search the CONFOCAL archive at

Hi all,

It seems that in many papers from biologists or chemists, and i'm
talking
high impact factors journals,  colocalisation of two elements is is
often
assumed  by simple color superposition (ex: red and green fluoresce
yellow
when colocalising), while microscopists (many physisists I suppose)
seem to
need a more complex software-based confirmation.
Is it ok, when using high end equipment and corrected objectives
(apochromat
with high NA for ex.), to assume colocalisation by color
superposition,
especially when fluorophore are confined to small volume entities,
like
lysosomes ?

Thanks

Marc


Mayandi Sivaguru Mayandi Sivaguru
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Valeria Berno
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Valeria, my understanding is that you will be better off with deconvolving all your optical microscope data sets (widefield, confocal and etc) in general. With reference to colocalization analysis, you first sample the data following sequential scans (never simultaneous for the coloc analysis) Nyquist sampling in 3D (I would personally suggest a bit over sampling won't hurt, if you do not experience significant photobleaching), and then a deconvolution is a must with a plane by plane analysis.
Deconvolution will not change a "non-cocolalizing" data points in to "colocalizing" data points. But it can be otherwise, a colocaizing data points in raw data could become in fact not colocalizing anymore after deconvolution. But the parameters affecting your conlusion greatly is at much before you deconve the data i.e., the sample preparation, fixation, blocking, selection of antibodies, fluorophores, scan parameters and so on.
Shiv
   

At 10:05 AM 3/31/2008, you wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

This question just fit in perfectly on what I am trying to find out about
colocalization.

When and why do I need do deconvolve pictures collected with a confocal in
order to be sure about my colocalization (or not colocalization) results?

To be specific: I am working on pre and post-synaptic proteins.

Thanks

Valeria



> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Colocalization based upon "yellow" could be accurate, if and only if,
> the intensities are comparable and pixel (voxel) quantities in the
> suspected colocalized volumes are in roughly equal.  .  Otherwise,
> the yellow is masked by the predominate channel.  Something small,
> like lysosomes, would need to be sampled properly. Colocalization
> could be masked by blur unless deconvolved, even if images are
> collected with a confocal.
> On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi all,
>>
>> It seems that in many papers from biologists or chemists, and i'm
>> talking
>> high impact factors journals,  colocalisation of two elements is is
>> often
>> assumed  by simple color superposition (ex: red and green fluoresce
>> yellow
>> when colocalising), while microscopists (many physisists I suppose)
>> seem to
>> need a more complex software-based confirmation.
>> Is it ok, when using high end equipment and corrected objectives
>> (apochromat
>> with high NA for ex.), to assume colocalisation by color
>> superposition,
>> especially when fluorophore are confined to small volume entities,
>> like
>> lysosomes ?
>>
>> Thanks
>>
>> Marc
>

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu

Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Valeria Berno
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

First, when I sent that email on colocalization without software, I  
was actually cleaning out unfinished drafts in my email, with the  
intent of deleting it.  But, this is a topic that always initiates a  
discussion.

As recommended the paper by Boldt and Cordeliere is very good, as is  
Costes et al, 2004 for the most recent papers, both have refs to  
prior work. the former paper nicely summarizes different types of  
describing colocalization.

The confocal will not completely reject out of focus light due to  
residual aberrations in the optical pathway.  The sample is often  
overlooked as an element of the optical pathway and it will often  
contribute to spherical aberration and background.    Deconvolution  
of  confocal stacks will further reduce noise and effects of  
spherical aberration and scattering.  Of course, it is not a  
substitute for proper sample preparation, adequate sampling and a  
well maintained instrument.  Noise in your images will reduce the  
apparent colocalization by inserting random pixels above threshold,  
while factors contributing to background will raise the appearance of  
colocalization since the signals are spreading throughout the image.  
As mentioned in Jeremy's email, a number of other factors need to be  
considered.  Above all, and as well described in Costes etal,  
controls for your labeling are essential to avoid mistaking such  
things as autofluorescence, bleedthrough or non-specific labeling as  
intensities representing components for which you are analysing.  You  
are pushing the limits of optical resolution, so you will need to  
push the limits of good methodology.

Regards,
Glen



Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

************************************************************************
******
The box said "Requires WindowsXP or better", so I bought a Macintosh.
************************************************************************
******


On Mar 31, 2008, at 8:05 AM, Valeria Berno wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi,
>
> This question just fit in perfectly on what I am trying to find out  
> about
> colocalization.
>
> When and why do I need do deconvolve pictures collected with a  
> confocal in
> order to be sure about my colocalization (or not colocalization)  
> results?
>
> To be specific: I am working on pre and post-synaptic proteins.
>
> Thanks
>
> Valeria
>
>
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Colocalization based upon "yellow" could be accurate, if and only if,
>> the intensities are comparable and pixel (voxel) quantities in the
>> suspected colocalized volumes are in roughly equal.  .  Otherwise,
>> the yellow is masked by the predominate channel.  Something small,
>> like lysosomes, would need to be sampled properly. Colocalization
>> could be masked by blur unless deconvolved, even if images are
>> collected with a confocal.
>> On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:
>>
>>> Search the CONFOCAL archive at
>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>> Hi all,
>>>
>>> It seems that in many papers from biologists or chemists, and i'm
>>> talking
>>> high impact factors journals,  colocalisation of two elements is is
>>> often
>>> assumed  by simple color superposition (ex: red and green fluoresce
>>> yellow
>>> when colocalising), while microscopists (many physisists I suppose)
>>> seem to
>>> need a more complex software-based confirmation.
>>> Is it ok, when using high end equipment and corrected objectives
>>> (apochromat
>>> with high NA for ex.), to assume colocalisation by color
>>> superposition,
>>> especially when fluorophore are confined to small volume entities,
>>> like
>>> lysosomes ?
>>>
>>> Thanks
>>>
>>> Marc
>>
John Oreopoulos John Oreopoulos
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Mayandi Sivaguru
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Does anyone know of any freely available software that can deconvolve image data? I am only aware of one ImageJ deconvolution plugin that does a reasonable job under certain circumstances. I'd be interested to know if anyone has created any others.


John Oreopoulos, BSc,
PhD Candidate
University of Toronto
Institute For Biomaterials and Biomedical Engineering
Centre For Studies in Molecular Imaging

Tel: W:416-946-5022


On 31-Mar-08, at 3:02 PM, Mayandi Sivaguru wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Valeria, my understanding is that you will be better off with deconvolving all your optical microscope data sets (widefield, confocal and etc) in general. With reference to colocalization analysis, you first sample the data following sequential scans (never simultaneous for the coloc analysis) Nyquist sampling in 3D (I would personally suggest a bit over sampling won't hurt, if you do not experience significant photobleaching), and then a deconvolution is a must with a plane by plane analysis.
Deconvolution will not change a "non-cocolalizing" data points in to "colocalizing" data points. But it can be otherwise, a colocaizing data points in raw data could become in fact not colocalizing anymore after deconvolution. But the parameters affecting your conlusion greatly is at much before you deconve the data i.e., the sample preparation, fixation, blocking, selection of antibodies, fluorophores, scan parameters and so on.
Shiv
   

At 10:05 AM 3/31/2008, you wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

This question just fit in perfectly on what I am trying to find out about
colocalization.

When and why do I need do deconvolve pictures collected with a confocal in
order to be sure about my colocalization (or not colocalization) results?

To be specific: I am working on pre and post-synaptic proteins.

Thanks

Valeria



> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Colocalization based upon "yellow" could be accurate, if and only if,
> the intensities are comparable and pixel (voxel) quantities in the
> suspected colocalized volumes are in roughly equal.  .  Otherwise,
> the yellow is masked by the predominate channel.  Something small,
> like lysosomes, would need to be sampled properly. Colocalization
> could be masked by blur unless deconvolved, even if images are
> collected with a confocal.
> On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi all,
>>
>> It seems that in many papers from biologists or chemists, and i'm
>> talking
>> high impact factors journals,  colocalisation of two elements is is
>> often
>> assumed  by simple color superposition (ex: red and green fluoresce
>> yellow
>> when colocalising), while microscopists (many physisists I suppose)
>> seem to
>> need a more complex software-based confirmation.
>> Is it ok, when using high end equipment and corrected objectives
>> (apochromat
>> with high NA for ex.), to assume colocalisation by color
>> superposition,
>> especially when fluorophore are confined to small volume entities,
>> like
>> lysosomes ?
>>
>> Thanks
>>
>> Marc
>

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu



Nathan-64 Nathan-64
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Jon,

   If your university/department has a license, you might find Matlab's deconvblind, deconvwnr, deconvreg, and deconvlucy functions useful. They require basic Matlab skill and some knowledge about how to generate appropriate input PSFs, but I've used them successfully a few times.

Best,
Nate


Nathan O'Connor
Graduate Student
Physiology and Biophysics
Weill Medical College of Cornell University
NY, NY 10021

On Mon, Mar 31, 2008 at 3:40 PM, John Oreopoulos <[hidden email]> wrote:

Does anyone know of any freely available software that can deconvolve image data? I am only aware of one ImageJ deconvolution plugin that does a reasonable job under certain circumstances. I'd be interested to know if anyone has created any others.


John Oreopoulos, BSc,
PhD Candidate
University of Toronto
Institute For Biomaterials and Biomedical Engineering
Centre For Studies in Molecular Imaging

Tel: W:416-946-5022


On 31-Mar-08, at 3:02 PM, Mayandi Sivaguru wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Valeria, my understanding is that you will be better off with deconvolving all your optical microscope data sets (widefield, confocal and etc) in general. With reference to colocalization analysis, you first sample the data following sequential scans (never simultaneous for the coloc analysis) Nyquist sampling in 3D (I would personally suggest a bit over sampling won't hurt, if you do not experience significant photobleaching), and then a deconvolution is a must with a plane by plane analysis.
Deconvolution will not change a "non-cocolalizing" data points in to "colocalizing" data points. But it can be otherwise, a colocaizing data points in raw data could become in fact not colocalizing anymore after deconvolution. But the parameters affecting your conlusion greatly is at much before you deconve the data i.e., the sample preparation, fixation, blocking, selection of antibodies, fluorophores, scan parameters and so on.
Shiv
   

At 10:05 AM 3/31/2008, you wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

This question just fit in perfectly on what I am trying to find out about
colocalization.

When and why do I need do deconvolve pictures collected with a confocal in
order to be sure about my colocalization (or not colocalization) results?

To be specific: I am working on pre and post-synaptic proteins.

Thanks

Valeria



> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Colocalization based upon "yellow" could be accurate, if and only if,
> the intensities are comparable and pixel (voxel) quantities in the
> suspected colocalized volumes are in roughly equal.  .  Otherwise,
> the yellow is masked by the predominate channel.  Something small,
> like lysosomes, would need to be sampled properly. Colocalization
> could be masked by blur unless deconvolved, even if images are
> collected with a confocal.
> On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi all,
>>
>> It seems that in many papers from biologists or chemists, and i'm
>> talking
>> high impact factors journals,  colocalisation of two elements is is
>> often
>> assumed  by simple color superposition (ex: red and green fluoresce
>> yellow
>> when colocalising), while microscopists (many physisists I suppose)
>> seem to
>> need a more complex software-based confirmation.
>> Is it ok, when using high end equipment and corrected objectives
>> (apochromat
>> with high NA for ex.), to assume colocalisation by color
>> superposition,
>> especially when fluorophore are confined to small volume entities,
>> like
>> lysosomes ?
>>
>> Thanks
>>
>> Marc
>

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu





Farid Jalali Farid Jalali
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Valeria Berno
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Journal of Microscopy, Vol. 208, Pt 2 November 2002, pp. 134–147
Deconvolution improves colocalization analysis of multiple fluorochromes in 3D confocal data sets more than filtering
techniques
L. LANDMANN

Microscopy and Research Technique 64:103–112 (2004)
Colocalization Analysis Yields Superior Results After Image Restoration
LUKAS LANDMANN* AND PERMSIN MARBET

Two nice papers that specifically address the issue at hand. I am doing a great deal of co-localization analysis and the image intensity based correlative methods such as Pearson or Overlap greatly benefit from image restoration (I am using 3D Blind). The correlations generally end up being greater and using the method of Van Steensel (JaCOP plugin, pixel shift a re-calculation of Pearson) yields a steeper slope as one channel is shifted in registration relative to the other. Object based methods will as well benefit from this as noise is reduced, allowing for better image segmentation. The latter can yield centroids and for well sampled images, the distances between centroids can be meaningful. Centroids from the red and green channel with a distance of 0um between them is quite suggestive that the object defined by the two channels are close to each other. This can also be done quite easily in 3D as well with 3D object counter from Image J. As someone else had mentioned, this does operate at the limits of optical resolution, but I have found it very useful to develop an idea how close to targets are to each other and it can strengthen the image intensity based correlative methods.

JaCOP and the J.Microscopy article by Bolte and Cordelieres are excellent. Suzanne Bolte helped me with the centroid approach using Image J.



On Mon, Mar 31, 2008 at 11:05 AM, Valeria Berno <[hidden email]> wrote:
Hi,

This question just fit in perfectly on what I am trying to find out about
colocalization.

When and why do I need do deconvolve pictures collected with a confocal in
order to be sure about my colocalization (or not colocalization) results?

To be specific: I am working on pre and post-synaptic proteins.

Thanks

Valeria



> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Colocalization based upon "yellow" could be accurate, if and only if,
> the intensities are comparable and pixel (voxel) quantities in the
> suspected colocalized volumes are in roughly equal.  .  Otherwise,
> the yellow is masked by the predominate channel.  Something small,
> like lysosomes, would need to be sampled properly. Colocalization
> could be masked by blur unless deconvolved, even if images are
> collected with a confocal.
> On Feb 7, 2007, at 1:05 PM, Marc Thibault wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi all,
>>
>> It seems that in many papers from biologists or chemists, and i'm
>> talking
>> high impact factors journals,  colocalisation of two elements is is
>> often
>> assumed  by simple color superposition (ex: red and green fluoresce
>> yellow
>> when colocalising), while microscopists (many physisists I suppose)
>> seem to
>> need a more complex software-based confirmation.
>> Is it ok, when using high end equipment and corrected objectives
>> (apochromat
>> with high NA for ex.), to assume colocalisation by color
>> superposition,
>> especially when fluorophore are confined to small volume entities,
>> like
>> lysosomes ?
>>
>> Thanks
>>
>> Marc
>



--
Farid Jalali MSc
Senior Research Technician/ Lab Manager
Dr. Robert Bristow Lab
Applied Molecular Oncology
Princess Margaret Hospital
Toronto, Canada
416-946-4501 X4351 (Princess Margaret Hospital)
416-581-7754 STTARR at MaRS Building
416-581-7791 STTARR Microscopy Suite
Jeremy Adler Jeremy Adler
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Jeremy Adler
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: colocalisation without software

 
 
further thoughts on colocalisation in the context of deconvolution

1) A problem in measuring colocalisation is image quality -  an important assumption underlying the Pearson correlation coefficient is that the data, in our case pixel intensities, is accurate.  As a simple test acquire an image and immediately collect a second image, compare the same pixels in the 2 images - they are very unlikely to be identical, due to Poisson and background noise.
     Noise creates a difference between the measured Pearson correlation and the 'noise free' Pearson correlation. 
     However we can obtain a 'noise free' correlation by measuring and then factoring out image quality from the calculation.
 
Adler, Pagakis, Parmryd 2008
"Replicate based noise corrected correlation for accurate measurements of colocalization"
J. Microscopy   Vol. 230, issue 1   (in press shortly)


2) The Bolte and Cordelieres review has been cited (J Microscopy, 206, 213-232) favourably and indeed it is a useful review, but it also shows a fundamental misunderstanding of both data acquisition and of correlation measurement.
      This is apparent in the scattergrams in their Fig 5, which clearly show that some of the images have pixels that are either saturated or not even onscale, but were still included in the calculation of the correlation (correspondence with the authors). This may explain the strange correlation coefficients: Fig 5c is, by eye, uncorrelated, the data points are all over the place, but it has a reported correlation of 0.69, which is only slightly lower than the 0.8? associated with Fig5b, which clearly has an appreciable correlation. Equally strange is Fig 5c with a negative correlation of -0.3, despite that fact that there seem to be almost no pixels in which both fluorophores occur - this isn't a negative correlation or even uncorrelation (a Pearson correlation coeff of 0) but simply the absence of any relationship - the two fluorophores are never in the same place.
      The major problem in their measurements of correlation is that no attempt was made to exclude pixels that are devoid of fluorescence (background only) and to limit the analysis to pixels actually containing both fluorophores.
       I must add that my explanation is slightly speculative, since the authors have refused permission for the renanalysis of their images, because we had criticised  other aspects of the review (see ref). I would encourage everyone to examine Fig 5 and draw your own conclusions.

Adler & Parmryd, J Microscopy, 2007, 227, page 83

a reply by Bolte and Cordelieres appears in the same issue



Jeremy Adler
Cell Biology
The Wenner-Gren Inst.
Arrhenius Laboratories E5
Stockholm University
Stockholm 106 91
Sweden





-----Original Message-----
From: Confocal Microscopy List on behalf of Glen MacDonald
Sent: Fri 3/28/2008 22:54
To: [hidden email]
Subject: Re: colocalisation without software

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi all,
>
> It seems that in many papers from biologists or chemists, and i'm 
> talking
> high impact factors journals,  colocalisation of two elements is is 
> often
> assumed  by simple color superposition (ex: red and green fluoresce 
> yellow
> when colocalising), while microscopists (many physisists I suppose) 
> seem to
> need a more complex software-based confirmation.
> Is it ok, when using high end equipment and corrected objectives 
> (apochromat
> with high NA for ex.), to assume colocalisation by color 
> superposition,
> especially when fluorophore are confined to small volume entities, 
> like
> lysosomes ?
>
> Thanks
>
> Marc
Jeremy Adler Jeremy Adler
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software typo removed

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: colocalisation without software
 
my apologies for apparently resending this email, I noticed a significant typo
 
 
further thoughts on colocalisation in the context of deconvolution

1) A problem in measuring colocalisation is image quality -  an important assumption underlying the Pearson correlation coefficient is that the data, in our case pixel intensities, is accurate.  As a simple test acquire an image and immediately collect a second image, compare the same pixels in the 2 images - they are very unlikely to be identical, due to Poisson and background noise.
     Noise creates a difference between the measured Pearson correlation and the 'noise free' Pearson correlation. 
     However we can obtain a 'noise free' correlation by measuring and then factoring out image quality from the calculation.
 
Adler, Pagakis, Parmryd 2008
"Replicate based noise corrected correlation for accurate measurements of colocalization"
J. Microscopy   Vol. 230, issue 1   (in press shortly)


2) The Bolte and Cordelieres review has been cited (J Microscopy, 206, 213-232) favourably and indeed it is a useful review, but it also shows a fundamental misunderstanding of both data acquisition and of correlation measurement.
      This is apparent in the scattergrams in their Fig 5, which clearly show that some of the images have pixels that are either saturated or not even onscale, but were still included in the calculation of the correlation (correspondence with the authors). This may explain the strange correlation coefficients: Fig 5c is, by eye, uncorrelated, the data points are all over the place, but it has a reported correlation of 0.69, which is only slightly lower than the 0.8? associated with Fig5b, which clearly has an appreciable correlation. Equally strange is Fig 5d with a negative correlation of -0.3, despite that fact that there seem to be almost no pixels in which both fluorophores occur - this isn't a negative correlation or even uncorrelation (a Pearson correlation coeff of 0) but simply the absence of any relationship - the two fluorophores are never in the same place.
      The major problem in their measurements of correlation is that no attempt was made to exclude pixels that are devoid of fluorescence (background only) and to limit the analysis to pixels actually containing both fluorophores.
       I must add that my explanation is slightly speculative, since the authors have refused permission for the renanalysis of their images, because we had criticised  other aspects of the review (see ref). I would encourage everyone to examine Fig 5 and draw your own conclusions.

Adler & Parmryd, J Microscopy, 2007, 227, page 83

a reply by Bolte and Cordelieres appears in the same issue



Jeremy Adler
Cell Biology
The Wenner-Gren Inst.
Arrhenius Laboratories E5
Stockholm University
Stockholm 106 91
Sweden





-----Original Message-----
From: Confocal Microscopy List on behalf of Glen MacDonald
Sent: Fri 3/28/2008 22:54
To: [hidden email]
Subject: Re: colocalisation without software

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi all,
>
> It seems that in many papers from biologists or chemists, and i'm 
> talking
> high impact factors journals,  colocalisation of two elements is is 
> often
> assumed  by simple color superposition (ex: red and green fluoresce 
> yellow
> when colocalising), while microscopists (many physisists I suppose) 
> seem to
> need a more complex software-based confirmation.
> Is it ok, when using high end equipment and corrected objectives 
> (apochromat
> with high NA for ex.), to assume colocalisation by color 
> superposition,
> especially when fluorophore are confined to small volume entities, 
> like
> lysosomes ?
>
> Thanks
>
> Marc
M. van de corput M. van de corput
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: colocalisation without software

In reply to this post by Jeremy Adler
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Jeremy Adler wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  
> further thoughts on colocalisation in the context of deconvolution
>
> 1) A problem in measuring colocalisation is image quality -  an
> important assumption underlying the Pearson correlation coefficient is
> that the data, in our case pixel intensities, is accurate.  As a
> simple test acquire an image and immediately collect a second image,
> compare the same pixels in the 2 images - they are very unlikely to be
> identical, due to Poisson and background noise.
I am new in the field of colocalization studies  but am aware of the
importance of accurate image acquisition. I agree the image acquisition
is the first and very important step towards any analysis done on an
image. I use a 4 line scan and average the intensity of the 4 resulting
in an average intensity image.....isn't that a way to reduce noise and
determine the "left over noise/background"  more accurately? It surely
improves the image. Deconvolution then takes away a lot more noise as
well. by using measured SNR for each channel, and using a measure PSF
for each channel separately. After chromatic shift correction the image
is ready for colocalization analysis. There will always be some noise
left in the image as not all the noise is from equal intensity (due to
e.g. varying autofluorescence of biological structures).

I wonder if there is a list of do and do nots for doing colocalization
studies (besides the helpful list in cited articles). Or better: what is
allowed and what not on image processing.

Mariette van de Corput
Erasmus MC, Rotterdam, NL

>
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List on behalf of Glen MacDonald
> Sent: Fri 3/28/2008 22:54
> To: [hidden email]
> Subject: Re: colocalisation without software
>
> > Search the CONFOCAL archive at
> > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> >
> > Hi all,
> >
> > It seems that in many papers from biologists or chemists, and i'm
> > talking
> > high impact factors journals,  colocalisation of two elements is is
> > often
> > assumed  by simple color superposition (ex: red and green fluoresce
> > yellow
> > when colocalising), while microscopists (many physisists I suppose)
> > seem to
> > need a more complex software-based confirmation.
> > Is it ok, when using high end equipment and corrected objectives
> > (apochromat
> > with high NA for ex.), to assume colocalisation by color
> > superposition,
> > especially when fluorophore are confined to small volume entities,
> > like
> > lysosomes ?
> >
> > Thanks
> >
> > Marc