On Oct 7, 2011, at 10:02 PM, Milton Charlton system wrote:

> Has anyone compared confocal images obtained with #1 (0.13-0.17mm) and
> #1.5(0.16-0.19mm) cover slips using 40X(1.25NA) and 63X(1.32 NA) oil
> objectives or similar?
>
> Mounting media:  Is there any consensus on the best mounting medium  
> that
> combines high refractive index (near RI= 1.5) with preservation of
> fluorescence?

Sorry to contribute to this conversation kind of late, but a few  
points –

Keep in mind that in practice there's *a lot* of variation in  
thickness between commercially available coverslips, even within the  
#1 and #1.5 range, respectively. A core director I know recently did a  
microcaliper survey of a number of brands of #1 and #1.5 cover slips  
and came up with descriptive statistics for them. I was shocked at how  
far off the 0.17 mm mark most fell, with #1.5's generally being  
notably thicker than the 0.17 average they are supposed to have.

The alternative is to buy coverslips specifically sold as 0.17 mm with  
very low variance (Zeiss, and I believe other manufacturers, carry  
these), which are expensive, or start picking through boxes of  
coverslips with microcalipers. But if you have correction-critical  
applications, such expense or effort may be called for.

This, of course, is assuming you're using an objective that is  
optimized for a 0.17 mm cover slip.

As for mounting media, keep in mind that even where mounting media  
matches glass, the further down below the coverslip the sample is, the  
more the mounting media contributes to optical aberration. In other  
words, optimal coverslip thickness is *really* optimal coverslip plus  
mounting media thickness. Again, for fully optimized optics, try to  
have the specimen actually in contact with the bottom of the coverslip.

Peter G. Werner
Program Assistant,
Merritt College Microscopy Program