Re: destroy alexa fluor fluorescence to allow restaining **Vendor Reply**

classic Classic list List threaded Threaded
♦
♦
1 message Options
Kilgore, Jason A. Kilgore, Jason A.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: destroy alexa fluor fluorescence to allow restaining **Vendor Reply**

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****


** Vendor Reply **

Antibody stripping, as another commenter had mentioned, may be an option.  A low-pH glycine can be used.  Given that these are whole mounts, though, it may be more difficult to effectively strip.  

Other than antibody stripping, the only other way I'm familiar with is photobleaching.  The AF dyes are very photostable, but as long as you aren't using an antifade mounting medium, then you could blast the cells with intense visible wavelength light, such as with a mercury lamp, and it should photobleach the dyes most or all the way.

Jason

Jason A. Kilgore
Technical Application Scientist
Molecular Probes / EVOS Tech Support
Life Sciences Solutions

Thermo Fisher Scientific
29851 Willow Creek Rd.
Eugene, OR  97402-9132
1-800-955-6288 then option 4, then option 6, then option 2.
Or dial direct at +1 541 335 0353
[hidden email]
www.lifetechnologies.com

This communication is intended solely for the individual/entity to whom it is addressed. It may contain confidential or legally privileged information.  Any unauthorized disclosure or copying is prohibited and may be unlawful. If you have received this communication in error, please notify the sender immediately and delete it from your system.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dale Moulding
Sent: Thursday, February 16, 2017 8:13 AM
To: [hidden email]
Subject: destroy alexa fluor fluorescence to allow restaining

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,
does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes?
We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin.
Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding?
Cheers
Dale