Re: new facility planning … flow rate air co nditioning

classic Classic list List threaded Threaded
♦
♦
1 message Options
Christian Kukat Christian Kukat
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: new facility planning … flow rate air co nditioning

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Silke,
 
Very often heat is produced at a certain point (e.g. Argon laser ventilator, multiphoton (MP) laser chiller). In our institute, we could reduce the heat load by discharging heat into an exhaust system directly at the point of heat generation. This can reduce the air exchange rate, but we still have 8 volume changes per hour.  A tight connection to the exhaust system should be avoided, because this can cause overheating if there is a failure in the exhaust system. If the heat load is too high, consider insulating the hot air exhaust tubes to prevent thermal radiation into the room. Additionally, we have cooling convector units in the rooms which can cool down the room by another 2 °C.
 
Best wishes,
 
Christian
_______________________________________________________________________

Dr. Christian Kukat
Leitung Serviceeinheit FACS & Mikroskopie / Head of FACS & Imaging Core Facility
Max-Planck-Institut für Biologie des Alterns / Max Planck Institute for Biology of Ageing

Joseph-Stelzmann-Str. 9b, D-50931 Köln / Cologne, Germany
Tel.: +49 221 37970 830
Fax: +49 221 37970 88830
E-mail: [hidden email]
www.age.mpg.de
_______________________________________________________________________