http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, We have an iris on the 488 on a BioRad 1024 which is stuck, and doesn't open or close with the slider. The service quotation is to replace the PMT assembly. I'm wondering whether anyone has experience to share with a stuck iris, and whether particular parts, as opposed to the whole assembly, might be available. Thanks. Allan Kachelmeier Manager, confocal microscopy core Oregon Hearing Research Center OHSU |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi As far as I know the mrc1024 uses a futaba iris which is a brand for radio model cars. Try to have a look to the Part No. and ask them directly. http://www.futaba.com/ Regards, NM Quoting Allan Kachelmeier <[hidden email]>: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi, > > We have an iris on the 488 on a BioRad 1024 which is stuck, and > doesn't open or close with the slider. The service quotation is to > replace the PMT assembly. I'm wondering whether anyone has experience > to share with a stuck iris, and whether particular parts, as opposed > to the whole assembly, might be available. Thanks. > > Allan Kachelmeier > Manager, confocal microscopy core > Oregon Hearing Research Center > OHSU > |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Nuno,
What on earth does a radio-controlled model car need
an iris for? I'm intrigued.
Guy
Optical Imaging Techniques in Cell Biology
by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net From: Confocal Microscopy List on behalf of Nuno Moreno Sent: Sat 07/12/15 6:54 AM To: [hidden email] Subject: Re: stuck iris on a BioRad 1024 Search the CONFOCAL archive at |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Guy
I cannot confirm the use of the Futaba motor on the 1024 but I can comment on the attributes of that type of proportional control. The hobby industry uses a unique form of position control based on the phase change of a 50Hz signal where phase shifts between 5ms - 17.5ms within that 20ms period of the signal translate to proportional rotation of a lever. The interface circuitry is easy to build. When the lever moves it is reasonably repeatable, smooth and has plenty of torque to move an iris. It provides an inexpensive means of analog position control. It is not optimal for precision optical positioning but great for proof of concept. I was not aware that this type of control was used on a 1024 but I will investigate the next time I have an opportunity. I am too curious for my own good! Dan On Dec 14, 2007, at 5:33 PM, Guy Cox wrote: Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Nuno, What on earth does a radio-controlled model car need an iris for? I'm intrigued. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net From: Confocal Microscopy List on behalf of Nuno Moreno Sent: Sat 07/12/15 6:54 AM To: [hidden email] Subject: Re: stuck iris on a BioRad 1024 Search the CONFOCAL archive at |
|
|
In reply to this post by Guy Cox
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, They also make miniature machinery. I also made myself that question :) Regards, NM Guy Cox wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > Nuno, > > What on earth does a radio-controlled model car need > an iris for? I'm intrigued. > > > Guy > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) > Electron Microscope Unit, Madsen Building F09, > University of Sydney, NSW 2006 > ______________________________________________ > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > <https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=http://www.guycox.net> > > ------------------------------------------------------------------------ > *From:* Confocal Microscopy List on behalf of Nuno Moreno > *Sent:* Sat 07/12/15 6:54 AM > *To:* [hidden email] > *Subject:* Re: stuck iris on a BioRad 1024 > > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi > > As far as I know the mrc1024 uses a futaba iris which is a brand for > radio model > cars. Try to have a look to the Part No. and ask them directly. > http://www.futaba.com/ > > Regards, > NM > > Quoting Allan Kachelmeier <[hidden email]>: > > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > > Hi, > > > > We have an iris on the 488 on a BioRad 1024 which is stuck, and > > doesn't open or close with the slider. The service quotation is to > > replace the PMT assembly. I'm wondering whether anyone has experience > > to share with a stuck iris, and whether particular parts, as opposed > > to the whole assembly, might be available. Thanks. > > > > Allan Kachelmeier > > Manager, confocal microscopy core > > Oregon Hearing Research Center > > OHSU > > > -- Nuno Moreno Cell Imaging Unit Instituto Gulbenkian de Ciência http://uic.igc.gulbekian.pt http://www.igc.gulbekian.pt phone +351 214464606 fax +351 214407970 |
|
Allan Kachelmeier |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Thanks for the suggestions on how to get a stuck iris unstuck. It turned out that Simon's suggestion of moving the actuator arm back and forth a few times before turning on the machine worked. BTW the actuator mechanism, what opens and closes the iris, perhaps not the iris itself, is made by Futaba. Regards, Allan Allan Kachelmeier Manager, confocal microscopy core Oregon Hearing Research Center OHSU On 12/17/07, Nuno Moreno <[hidden email]> wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi, > > They also make miniature machinery. I also made myself that question :) > > Regards, > NM > > Guy Cox wrote: > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Nuno, > > > > What on earth does a radio-controlled model car need > > an iris for? I'm intrigued. > > > > > > Guy > > > > Optical Imaging Techniques in Cell Biology > > by Guy Cox CRC Press / Taylor & Francis > > http://www.guycox.com/optical.htm > > ______________________________________________ > > Associate Professor Guy Cox, MA, DPhil(Oxon) > > Electron Microscope Unit, Madsen Building F09, > > University of Sydney, NSW 2006 > > ______________________________________________ > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > > Mobile 0413 281 861 > > ______________________________________________ > > http://www.guycox.net > > <https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=http://www.guycox.net> > > > > ------------------------------------------------------------------------ > > *From:* Confocal Microscopy List on behalf of Nuno Moreno > > *Sent:* Sat 07/12/15 6:54 AM > > *To:* [hidden email] > > *Subject:* Re: stuck iris on a BioRad 1024 > > > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > > Hi > > > > As far as I know the mrc1024 uses a futaba iris which is a brand for > > radio model > > cars. Try to have a look to the Part No. and ask them directly. > > http://www.futaba.com/ > > > > Regards, > > NM > > > > Quoting Allan Kachelmeier <[hidden email]>: > > > > > Search the CONFOCAL archive at > > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > > > > Hi, > > > > > > We have an iris on the 488 on a BioRad 1024 which is stuck, and > > > doesn't open or close with the slider. The service quotation is to > > > replace the PMT assembly. I'm wondering whether anyone has experience > > > to share with a stuck iris, and whether particular parts, as opposed > > > to the whole assembly, might be available. Thanks. > > > > > > Allan Kachelmeier > > > Manager, confocal microscopy core > > > Oregon Hearing Research Center > > > OHSU > > > > > > > -- > Nuno Moreno > Cell Imaging Unit > Instituto Gulbenkian de Ciência > http://uic.igc.gulbekian.pt > http://www.igc.gulbekian.pt > phone +351 214464606 > fax +351 214407970 > |