Reenviar: Ablation with multiphoton laser

classic Classic list List threaded Threaded
♦
♦
2 messages Options
Katherina Garcia Garcia Katherina Garcia Garcia
Reply | Threaded
Open this post in threaded view
♦
♦
|

Reenviar: Ablation with multiphoton laser

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear all:


We are going to buy a Multiphoton Microscope and we would like to add a laser ablation module to this system. One company offers us to do the photoablation with the same laser we do multiphoton imaging. 


Do you have experience doing photoablation with a multiphoton laser?


Thanks in advance.
Best regards,
Katherina





------------------------------------------------------------------------------------------ 
Katherina García García
Técnico de Microscopía y Citometría
Tlf. 954977432
[hidden email]
Centro Andaluz de Biología del Desarrollo
Consejo Superior de Investigaciones Científicas
Universidad Pablo Olavide
Ctra. Utrera km1
41013 Sevilla
España
Jean-Yves Tinevez-3 Jean-Yves Tinevez-3
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Reenviar: Ablation with multiphoton laser

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

On May 4, 2011, at 9:41 AM, Katherina Garcia Garcia wrote:
>
> We are going to buy a Multiphoton Microscope and we would like to  
> add a laser ablation module to this system. One company offers us to  
> do the photoablation with the same laser we do multiphoton imaging.
>
>
> Do you have experience doing photoablation with a multiphoton laser?
>


Dear Katherine,

Two few years ago, we were doing laser ablation on the actin cortex of  
floating and adhering fibroblasts.
Jan Peychl set up an Olympus FlowView 1000, with the second scanner  
coupled to a 405nm PicoQuant pulsed laser (single photon ablation). We  
were happy with it: ablation was gentle enough for the cell to  
recover, and we had a large range of power for which the ablation  
effect could be tuned.

We also tried doing ablation on the samples on Nicola Maghelli's  
homemade 2 photons, using the imaging infrared pulsed laser.
Though our experience was very limited, we had a very hard time  
finding suitable parameters for the ablation to occur. Depending on  
the output power, we had either no effect (or bleaching), either the  
whole cell exploded. The parameters set for which the cortex was  
ablated was very sensitive, and once we have found it for one cell, it  
would not work for another cell.
So we did not pursue in this direction, and went back to single-photon  
ablation.

Cheers
jy


--