Hello all,
We sometimes see vertical line artifacts in images acquired with our slit-scanning confocal. The pattern is highly reproducible and can be divided out with a "blank" image. However, we sometimes have to filter out the artifacts in Fourier space by masking with a thin horz. rectangle through the middle of the FFT map, save the very center. Can someone please point out references where line artifacts have been removed with a filter (not by dividing out), whether in Fourier space or not?
Those interested can see sample images and FFT mask at http://www.mediafire.com/?sharekey=7b16b66ca9129ce991b20cc0d07ba4d2a83703a20834b252
Thanks,
Esteban
-- G. Esteban Fernandez, Ph.D. Associate Director Molecular Cytology Core Facility University of Missouri 120 Bond Life Sciences Center Columbia, MO 65211 http://www.biotech.missouri.edu/mcc/ 573-882-4895 573-884-9395 fax |
|
Tim Holmes-2 |
|
|
Esteban: Check out Fig’s. 53 – 56, and afterwards in: The
Image Processing Handbook by John C. Russ, CRC Press, 1992. If you want, I can
scan the pages and email a pdf to you. Regards Tim From: Confocal Microscopy
List [mailto:[hidden email]] On Behalf Of G. Esteban
Fernandez Hello all, We sometimes see vertical line artifacts in images acquired
with our slit-scanning confocal. The pattern is highly
reproducible and can be divided out with a "blank"
image. However, we sometimes have to filter out the artifacts
in Fourier space by masking with a thin horz. rectangle through the middle
of the FFT map, save the very center. Can someone please point out references where line artifacts have
been removed with a filter (not by dividing out), whether in Fourier space
or not? Those
interested can see sample images and FFT mask at http://www.mediafire.com/?sharekey=7b16b66ca9129ce991b20cc0d07ba4d2a83703a20834b252 Thanks, Esteban
No virus
found in this incoming message. |
|
Tim Holmes-2 |
|
|
In reply to this post by G. Esteban Fernandez
Forgot to mention: the examples are on pg. 204 of that book. From: Tim Holmes Esteban: Check out Fig’s. 53 – 56, and afterwards
in: The Image Processing Handbook by John C. Russ, CRC Press, 1992.
If you want, I can scan the pages and email a pdf to you. Regards Tim From: Confocal Microscopy
List [mailto:[hidden email]] On Behalf Of G. Esteban
Fernandez Hello all, We sometimes see vertical line artifacts in images acquired
with our slit-scanning confocal. The pattern is highly
reproducible and can be divided out with a "blank"
image. However, we sometimes have to filter out the artifacts
in Fourier space by masking with a thin horz. rectangle through the middle
of the FFT map, save the very center. Can someone please point out references where line artifacts have
been removed with a filter (not by dividing out), whether in Fourier space
or not? Those
interested can see sample images and FFT mask at http://www.mediafire.com/?sharekey=7b16b66ca9129ce991b20cc0d07ba4d2a83703a20834b252 Thanks, Esteban
No virus
found in this incoming message. |
|
|
In reply to this post by G. Esteban Fernandez
Hi Esteban,
I can't give you a reference, unfortunately, but I can confirm that we had the same problem years ago with a Zeiss LSM5 live; I wrote an automated pre-processing tool in IDL that removed the critical part from the FFT and did the inverse FFT for all images. The problem was eventually solved by Zeiss.
Hope this helps, Zoltan On Tue, Dec 9, 2008 at 7:25 PM, G. Esteban Fernandez <[hidden email]> wrote:
-- Zoltan Cseresnyes Facility manager, Imaging Suite |
|
JOEL B. SHEFFIELD |
|
|
Hi,
ImageJ has an FFT/Bandpass filter built in that can selectively remove either vertical or horizontal lines without a need to construct an FFT mask. Let me know if you need more information. Joel > > Hi Esteban, I can't give you a reference, unfortunately, but I can > confirm that we had the same problem years ago with a Zeiss LSM5 > live; I wrote an automated pre-processing tool in IDL that removed > the critical part from the FFT and did the inverse FFT for all > images. The problem was eventually solved by Zeiss. > Hope this helps, > Zoltan > > > On Tue, Dec 9, 2008 at 7:25 PM, G. Esteban Fernandez > <[hidden email]> wrote: > Hello all, > > We sometimes see vertical line artifacts in images acquired with our > slit-scanning confocal. The pattern is highly reproducibleand can > be divided out with a "blank" image.However, we sometimes have > tofilter out the artifacts in Fourier spaceby masking with a thin > horz. rectangle through the middle of the FFT map, save the very > center. Can someone please point out references where line artifacts > have been removedwith a filter (not by dividing out), whether in > Fourier space or not? > > Those interested can see sample images and FFT mask at > http://www.mediafire.com/?sharekey=7b16b66ca9129ce991b20cc0d07ba4d2a83 > 703a20834b252 > > > Thanks, > Esteban > > -- > G. Esteban Fernandez, Ph.D. > Associate Director > Molecular Cytology Core Facility > University of Missouri > 120 Bond Life Sciences Center > Columbia, MO 65211 > > http://www.biotech.missouri.edu/mcc/ > > 573-882-4895 > 573-884-9395 fax > > > > -- > > Zoltan Cseresnyes > Facility manager, Imaging Suite -- Joel B. Sheffield, Ph.D. Biology Department, Temple University 1900 North 12th Street Philadelphia, PA 19122 [hidden email] (215) 204 8839, fax (215) 204 0486 http://astro.temple.edu/~jbs |
|
Roman Veith |
|
|
In reply to this post by G. Esteban Fernandez
Hi!
We had the same problem with a Zeiss LSM Live microscope. The problem was first thought to be related to dust at the CCD line (and therefore producing a vertical line, because one or more pixels are "covered"). The ZEISS service tried to solve the problem, it resulted at least in some hardware exchange at the LIVE unit, so maybe the dust was not the only porblem. Now it's better, but in certain cases one can still see some lines. If the lines recently occured maybe it's really dust. Regards, Roman G. Esteban Fernandez schrieb: > Hello all, > > We sometimes see vertical line artifacts in images acquired with our > slit-scanning confocal. The pattern is highly reproducible and can be > divided out with a "blank" image. However, we sometimes have > to filter out the artifacts in Fourier space by masking with a thin > horz. rectangle through the middle of the FFT map, save the very > center. Can someone please point out references where line artifacts > have been removed with a filter (not by dividing out), whether in > Fourier space or not? > > Those interested can see sample images and FFT mask at > http://www.mediafire.com/?sharekey=7b16b66ca9129ce991b20cc0d07ba4d2a83703a20834b252 > <https://webmail.um.umsystem.edu/exchweb/bin/redir.asp?URL=http://www.mediafire.com/?sharekey=7b16b66ca9129ce991b20cc0d07ba4d2a83703a20834b252> > > Thanks, > Esteban > > -- > G. Esteban Fernandez, Ph.D. > Associate Director > Molecular Cytology Core Facility > University of Missouri > 120 Bond Life Sciences Center > Columbia, MO 65211 > > http://www.biotech.missouri.edu/mcc/ > > 573-882-4895 > 573-884-9395 fax -- -- Roman Veith Diplom-Biologe Institut für physikalische und theoretische Chemie AG Biophysikalische Chemie Universität Bonn Wegelerstraße 12 53115 Bonn Tel-Nr.: 0228-733089 Fax-Nr.: 0228-739424 http://www.thch.uni-bonn.de/pc/kubitscheck/ |
|
Arwed Weigel |
|
|
In reply to this post by G. Esteban Fernandez
A similar problem exists in structured illumination microscopy that uses
a linear grid. the following references might help: http://dx.doi.org/10.1046/j.1365-2818.2001.00894.x <http://dx.doi.org/10.1046/j.1365-2818.2001.00894.x> http://dx.doi.org/10.1111/j.0022-2720.2004.01411.x <http://dx.doi.org/10.1111/j.0022-2720.2004.01411.x> -- arwed weigel abteilung fuer neurophysiology und zellulaere biophysik universitaet goettingen humboldtallee 23 37073 goettingen phon +49 (0)551 395918 fax +49 (0)551 398399 |
|
Stephen Cody-2 |
|
|
G'day,
Interestingly the old Bio-Rad DVC-250 slit scanner was also prone to this problem; it was caused by dust accumutating in the slit. Perhaps the 5Live may be prone to dust settling on the "AchroGate" beam splitter, confocal slit or as already suggested at the detector. Cheers Stephen Cody |