Dear List
We have a Perkin Elmer UltraView ERS spinning disk confocal. Our 568 gas laser
has died. I would like to replace it with a diode laser. Can any one suggest a
company that will provide third party lasers and install them on these systems.
Thank you
Lloyd Williams
Director Bio-Imaging & Network Core Facilities
Department of Biological Sciences Rm. 826 HN
695 Park Avenue, New York, NY 10065
212 650 3872;  fax: 212 650-3656
Andrew Simmonds Andrew Simmonds
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Replacement Diode Laser Perkin Elmer Ultra View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Lloyd
I am going through the same process for my ERS system.
Aside from PE, I think the only other choice is...
http://www.spectral.ca/products/mergemodule/index.html
This product is even compatible with Volocity If you find additional
vendors - please let me know!
Andrew Simmonds
Associate Professor, Alberta Innovates-Health Solutions Senior Scholar
President – Canadian Society for Molecular Biosciences/ La Société
Canadienne pour les Biosciences Moléculaires Department of Cell Biology,
Faculty of Medicine and Dentistry University of Alberta


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Lloyd Williams
Sent: Wednesday, September 04, 2013 10:37 AM
To: [hidden email]
Subject: Replacement Diode Laser Perkin Elmer Ultra View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear List
We have a Perkin Elmer UltraView ERS spinning disk confocal. Our 568 gas
laser has died. I would like to replace it with a diode laser. Can any one
suggest a company that will provide third party lasers and install them on
these systems.
Thank you
Lloyd Williams
Director Bio-Imaging & Network Core Facilities Department of Biological
Sciences Rm. 826 HN
695 Park Avenue, New York, NY 10065
212 650 3872;  fax: 212 650-3656
Alison J. North Alison J. North
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Replacement Diode Laser Perkin Elmer Ultra View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Lloyd,
As Andrew said, Spectral Applied will install such lasers on
Perkin-Elmer-bought systems - they did so on mine a few years ago. We
run our system under MetaMorph but there are other options. You're most
welcome to swing by and take a look at them if you like, since we're
only just down the road from you....
Best,
Alison
(Rockefeller University)


On 9/4/2013 1:36 PM, Andrew Simmonds wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Lloyd
> I am going through the same process for my ERS system.
> Aside from PE, I think the only other choice is...
> http://www.spectral.ca/products/mergemodule/index.html
> This product is even compatible with Volocity If you find additional
> vendors - please let me know!
> Andrew Simmonds
> Associate Professor, Alberta Innovates-Health Solutions Senior Scholar
> President – Canadian Society for Molecular Biosciences/ La Société
> Canadienne pour les Biosciences Moléculaires Department of Cell Biology,
> Faculty of Medicine and Dentistry University of Alberta
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Lloyd Williams
> Sent: Wednesday, September 04, 2013 10:37 AM
> To: [hidden email]
> Subject: Replacement Diode Laser Perkin Elmer Ultra View
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear List
> We have a Perkin Elmer UltraView ERS spinning disk confocal. Our 568 gas
> laser has died. I would like to replace it with a diode laser. Can any one
> suggest a company that will provide third party lasers and install them on
> these systems.
> Thank you
> Lloyd Williams
> Director Bio-Imaging & Network Core Facilities Department of Biological
> Sciences Rm. 826 HN
> 695 Park Avenue, New York, NY 10065
> 212 650 3872;  fax: 212 650-3656

--
Alison J. North, Ph.D.,
Senior Director of the Bio-Imaging Resource Center and
Research Associate Professor,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab    ++ 212 327 7486
Fax:        ++ 212 327 7489
Cameron, Lisa Cameron, Lisa
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Replacement Diode Laser Perkin Elmer Ultra View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I think Spectral Applied Research distributes through BioVision in NYC (including lots of other places) - www.BioVis.com
Perhaps Alison can confirm or Spectral will let you know.
I have heard good things about both companies from multiple colleagues.

I have seen their products, but do not have equipment from them in my core (due to historical reasons).

No commercial interest.


---------------------------------------
Lisa Cameron, Ph.D.
Director of Confocal and Light Microscopy
Dana Farber Cancer Institute
Boston, MA


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Alison North
Sent: Wednesday, September 04, 2013 1:54 PM
To: [hidden email]
Subject: Re: [CONFOCALMICROSCOPY] Replacement Diode Laser Perkin Elmer Ultra View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Lloyd,
As Andrew said, Spectral Applied will install such lasers on
Perkin-Elmer-bought systems - they did so on mine a few years ago. We
run our system under MetaMorph but there are other options. You're most
welcome to swing by and take a look at them if you like, since we're
only just down the road from you....
Best,
Alison
(Rockefeller University)


On 9/4/2013 1:36 PM, Andrew Simmonds wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Lloyd
> I am going through the same process for my ERS system.
> Aside from PE, I think the only other choice is...
> http://www.spectral.ca/products/mergemodule/index.html
> This product is even compatible with Volocity If you find additional
> vendors - please let me know!
> Andrew Simmonds
> Associate Professor, Alberta Innovates-Health Solutions Senior Scholar
> President - Canadian Society for Molecular Biosciences/ La Société
> Canadienne pour les Biosciences Moléculaires Department of Cell Biology,
> Faculty of Medicine and Dentistry University of Alberta
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Lloyd Williams
> Sent: Wednesday, September 04, 2013 10:37 AM
> To: [hidden email]
> Subject: Replacement Diode Laser Perkin Elmer Ultra View
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear List
> We have a Perkin Elmer UltraView ERS spinning disk confocal. Our 568 gas
> laser has died. I would like to replace it with a diode laser. Can any one
> suggest a company that will provide third party lasers and install them on
> these systems.
> Thank you
> Lloyd Williams
> Director Bio-Imaging & Network Core Facilities Department of Biological
> Sciences Rm. 826 HN
> 695 Park Avenue, New York, NY 10065
> 212 650 3872;  fax: 212 650-3656

--
Alison J. North, Ph.D.,
Senior Director of the Bio-Imaging Resource Center and
Research Associate Professor,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab    ++ 212 327 7486
Fax:        ++ 212 327 7489


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.
Robin Battye-2 Robin Battye-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Replacement Diode Laser Perkin Elmer Ultra View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Commercial reply


Thank you for the kinds words and references.  Please feel free to contact
us off line if you would like to discuss some options.

Cheers Robin

Robin Battye, M.Sc., Ph. D.
Confocal Product Manager/
Technical Sales Specialist
Mobile: (416) 302-5934
 
Spectral Applied Research
9078 Leslie St., Unit 11
Richmond Hill, Ontario
L4B 3L8
Office: (905) 326-5040 ext. 406
Fax: (905) 326-5041
www.spectral.ca
 
We believe that performance comes before Compromise.
 


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Cameron, Lisa
Sent: Wednesday, September 04, 2013 4:02 PM
To: [hidden email]
Subject: Re: Replacement Diode Laser Perkin Elmer Ultra View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I think Spectral Applied Research distributes through BioVision in NYC
(including lots of other places) - www.BioVis.com Perhaps Alison can confirm
or Spectral will let you know.
I have heard good things about both companies from multiple colleagues.

I have seen their products, but do not have equipment from them in my core
(due to historical reasons).

No commercial interest.


---------------------------------------
Lisa Cameron, Ph.D.
Director of Confocal and Light Microscopy Dana Farber Cancer Institute
Boston, MA


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Alison North
Sent: Wednesday, September 04, 2013 1:54 PM
To: [hidden email]
Subject: Re: [CONFOCALMICROSCOPY] Replacement Diode Laser Perkin Elmer Ultra
View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Lloyd,
As Andrew said, Spectral Applied will install such lasers on
Perkin-Elmer-bought systems - they did so on mine a few years ago. We run
our system under MetaMorph but there are other options. You're most welcome
to swing by and take a look at them if you like, since we're only just down
the road from you....
Best,
Alison
(Rockefeller University)


On 9/4/2013 1:36 PM, Andrew Simmonds wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Lloyd
> I am going through the same process for my ERS system.
> Aside from PE, I think the only other choice is...
> http://www.spectral.ca/products/mergemodule/index.html
> This product is even compatible with Volocity If you find additional
> vendors - please let me know!
> Andrew Simmonds
> Associate Professor, Alberta Innovates-Health Solutions Senior Scholar
> President - Canadian Society for Molecular Biosciences/ La Société
> Canadienne pour les Biosciences Moléculaires Department of Cell
> Biology, Faculty of Medicine and Dentistry University of Alberta
>
>
> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]] On Behalf Of Lloyd Williams
> Sent: Wednesday, September 04, 2013 10:37 AM
> To: [hidden email]
> Subject: Replacement Diode Laser Perkin Elmer Ultra View
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear List
> We have a Perkin Elmer UltraView ERS spinning disk confocal. Our 568
> gas laser has died. I would like to replace it with a diode laser. Can
> any one suggest a company that will provide third party lasers and
> install them on these systems.
> Thank you
> Lloyd Williams
> Director Bio-Imaging & Network Core Facilities Department of
> Biological Sciences Rm. 826 HN
> 695 Park Avenue, New York, NY 10065
> 212 650 3872;  fax: 212 650-3656

--
Alison J. North, Ph.D.,
Senior Director of the Bio-Imaging Resource Center and Research Associate
Professor, The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab    ++ 212 327 7486
Fax:        ++ 212 327 7489


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the
e-mail
contains patient information, please contact the Partners Compliance
HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in
error
but does not contain patient information, please contact the sender and
properly
dispose of the e-mail.