Dear list,
I'm looking for general information about the comparison between the
resonant scanners in the SP5 versus A1R. Apart other comments, are
possible future upgrades by software in the image format, panning and
quick switching between the high resolution scanner and the resonant one?

Thank you very much in advance.

--

Juan Luis Ribas
Servicio de Microscopía
Centro de Investigación, Tecnología e Innovación
Universidad de Sevilla
Av. Reina Mercedes 4b
41012 Sevilla
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
|

Re: Resonant scanner from A1R vs SP5 ...

Hi Juan,

On the SP5, when you start LAS AF the choice is resonant scanner or high resolution scanner. No way currently to quickly switch, but do you really need to? If you have a motorized stage, you can tile scan in resonant scanner mode faster than you could exit and restart.

George
p.s. I've noticed on both the Leica SP5, Zeiss LSM510 and Zeiss LSM710 that "faster is brighter" for many fluorophores. That is, a 0.4 us dwell time results in a brighter image than a 4 us which is brighter than a 40 us. This implies that <0.4 us might be even brighter - i.e. resonant scanner mode. A couple of possible explanations (not mutually exclusive):
a) photophysics - re TRex (PubMed 19337661), papers by Sanden/Spielmann/Widergren et al (PubMed 20375039, 20196585, 19007245, 17385841).
b) calibration method(s) by Zeiss and/or Leica that try to (but do not always) match output at all settings.
See also

high speed scanning has the potential to increase fluorescence yield and to reduce photobleaching. Borlinghaus RT. Microsc Res Tech. 2006 Sep;69(9):689-92.PMID: 16878313





At 12:37 PM 5/7/2010, you wrote:
Dear list,
I'm looking for general information about the comparison between the resonant scanners in the SP5 versus A1R. Apart other comments, are possible future upgrades by software in the image format, panning and quick switching between the high resolution scanner and the resonant one?

Thank you very much in advance.

--

Juan Luis Ribas
Servicio de Microscopía
Centro de Investigación, Tecnología e Innovación
Universidad de Sevilla
Av. Reina Mercedes 4b
41012 Sevilla







George McNamara, Ph.D.
Image Core Manager
Analytical Imaging Core Facility
University of Miami, Miller School of Medicine
Miami, FL 33136
[hidden email]
[hidden email]
305-243-8436 office
http://www.sylvester.org/AICF (Analytical Imaging Core Facility)
http://www.sylvester.org/AICF/pubspectra.zip (the entire 2000+ spectra .xlsx file is in the zip file)
http://home.earthlink.net/~geomcnamara

Maclyn McCarty Maclyn McCarty
Reply | Threaded
Open this post in threaded view
|

Re: Resonant scanner from A1R vs SP5 ...


What is the advantage of using resonant scanner?

Is there any technical comparison of the resonant scanner of A1 and SP5?

Mac.


On Sat, 8/5/10, George McNamara <[hidden email]> wrote:

From: George McNamara <[hidden email]>
Subject: Re: Resonant scanner from A1R vs SP5 ...
To: [hidden email]
Date: Saturday, 8 May, 2010, 2:45 PM

Hi Juan,

On the SP5, when you start LAS AF the choice is resonant scanner or high resolution scanner. No way currently to quickly switch, but do you really need to? If you have a motorized stage, you can tile scan in resonant scanner mode faster than you could exit and restart.

George
p.s. I've noticed on both the Leica SP5, Zeiss LSM510 and Zeiss LSM710 that "faster is brighter" for many fluorophores. That is, a 0.4 us dwell time results in a brighter image than a 4 us which is brighter than a 40 us. This implies that <0.4 us might be even brighter - i.e. resonant scanner mode. A couple of possible explanations (not mutually exclusive):
a) photophysics - re TRex (PubMed 19337661), papers by Sanden/Spielmann/Widergren et al (PubMed 20375039, 20196585, 19007245, 17385841).
b) calibration method(s) by Zeiss and/or Leica that try to (but do not always) match output at all settings.
See also

high speed scanning has the potential to increase fluorescence yield and to reduce photobleaching. Borlinghaus RT. Microsc Res Tech. 2006 Sep;69(9):689-92.PMID: 16878313





At 12:37 PM 5/7/2010, you wrote:
Dear list,
I'm looking for general information about the comparison between the resonant scanners in the SP5 versus A1R. Apart other comments, are possible future upgrades by software in the image format, panning and quick switching between the high resolution scanner and the resonant one?

Thank you very much in advance.

--

Juan Luis Ribas
Servicio de Microscopía
Centro de Investigación, Tecnología e Innovación
Universidad de Sevilla
Av. Reina Mercedes 4b
41012 Sevilla

Cameron Nowell Cameron Nowell
Reply | Threaded
Open this post in threaded view
|

eLabExperts Management System

Hi List,

Is anyone out there using the eLabExperts booking, charging etc system? see http://www.elabexperts.com/core/default.aspx

it looks really good and seems to be a more refined (and payed for) version of PPMS

If you have it, what has your experience been? Does it do everything it should?


Cheers


Cam



*Cameron J. Nowell
*Microscopy Manager
Centre for Advanced Microscopy
Ludwig Institute for Cancer Research
PO Box 2008
Royal Melbourne Hospital
Victoria, 3050
AUSTRALIA

*Office*: +61 3 9341 3155
*Mobile*: +61422882700
*Fax*: +61 3 9341 3104

Facility Website <http://www.ludwig.edu.au/confocal/>





-----Original Message-----
From: Confocal Microscopy List on behalf of Maclyn McCarty
Sent: Mon 24/05/2010 11:26 PM
To: [hidden email]
Subject: Re: Resonant scanner from A1R vs SP5 ...
 

What is the advantage of using resonant scanner?

Is there any technical comparison of the resonant scanner of A1 and SP5?

Mac.


On Sat, 8/5/10, George McNamara <[hidden email]> wrote:

From: George McNamara <[hidden email]>
Subject: Re: Resonant scanner from A1R vs SP5 ...
To: [hidden email]
Date: Saturday, 8 May, 2010, 2:45 PM


 
Hi Juan,


On the SP5, when you start LAS AF the choice is resonant scanner or high
resolution scanner. No way currently to quickly switch, but do you really
need to? If you have a motorized stage, you can tile scan in resonant
scanner mode faster than you could exit and restart.


George

p.s. I've noticed on both the Leica SP5, Zeiss LSM510 and Zeiss LSM710
that "faster is brighter" for many fluorophores. That is, a 0.4
us dwell time results in a brighter image than a 4 us which is brighter
than a 40 us. This implies that <0.4 us might be even brighter - i.e.
resonant scanner mode. A couple of possible explanations (not mutually
exclusive):

a) photophysics - re TRex (PubMed 19337661), papers by
Sanden/Spielmann/Widergren et al (PubMed 20375039, 20196585, 19007245,
17385841).
b) calibration method(s) by Zeiss and/or Leica that try to (but do
not always) match output at all settings.
See also


high speed
scanning has the potential to increase fluorescence yield and to reduce
photobleaching. Borlinghaus RT. Microsc Res Tech. 2006
Sep;69(9):689-92.PMID: 16878313








At 12:37 PM 5/7/2010, you wrote:

Dear list,

I'm looking for general information about the comparison between the
resonant scanners in the SP5 versus A1R. Apart other comments, are
possible future upgrades by software in the image format, panning and
quick switching between the high resolution scanner and the resonant
one?


Thank you very much in advance.


--


Juan Luis Ribas

Servicio de Microscopía

Centro de Investigación, Tecnología e Innovación

Universidad de Sevilla

Av. Reina Mercedes 4b

41012 Sevilla

 
yuansheng sun yuansheng sun
Reply | Threaded
Open this post in threaded view
|

Re: eLabExperts Management System

We just started using it.  I think it is a very good and useful tool
for managing a facility.  It meets all our needs for managing our
facility.  Cheers.  - sheng


Yuansheng Sun, Ph.D
Keck Center for Cellular Imaging
University of Virginia


On Tue, May 25, 2010 at 6:59 AM, Cameron Nowell
<[hidden email]> wrote:

> Hi List,
>
> Is anyone out there using the eLabExperts booking, charging etc system? see http://www.elabexperts.com/core/default.aspx
>
> it looks really good and seems to be a more refined (and payed for) version of PPMS
>
> If you have it, what has your experience been? Does it do everything it should?
>
>
> Cheers
>
>
> Cam
>
>
>
> *Cameron J. Nowell
> *Microscopy Manager
> Centre for Advanced Microscopy
> Ludwig Institute for Cancer Research
> PO Box 2008
> Royal Melbourne Hospital
> Victoria, 3050
> AUSTRALIA
>
> *Office*: +61 3 9341 3155
> *Mobile*: +61422882700
> *Fax*: +61 3 9341 3104
>
> Facility Website <http://www.ludwig.edu.au/confocal/>
>
>
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List on behalf of Maclyn McCarty
> Sent: Mon 24/05/2010 11:26 PM
> To: [hidden email]
> Subject: Re: Resonant scanner from A1R vs SP5 ...
>
>
> What is the advantage of using resonant scanner?
>
> Is there any technical comparison of the resonant scanner of A1 and SP5?
>
> Mac.
>
>
> On Sat, 8/5/10, George McNamara <[hidden email]> wrote:
>
> From: George McNamara <[hidden email]>
> Subject: Re: Resonant scanner from A1R vs SP5 ...
> To: [hidden email]
> Date: Saturday, 8 May, 2010, 2:45 PM
>
>
>
> Hi Juan,
>
>
> On the SP5, when you start LAS AF the choice is resonant scanner or high
> resolution scanner. No way currently to quickly switch, but do you really
> need to? If you have a motorized stage, you can tile scan in resonant
> scanner mode faster than you could exit and restart.
>
>
> George
>
> p.s. I've noticed on both the Leica SP5, Zeiss LSM510 and Zeiss LSM710
> that "faster is brighter" for many fluorophores. That is, a 0.4
> us dwell time results in a brighter image than a 4 us which is brighter
> than a 40 us. This implies that <0.4 us might be even brighter - i.e.
> resonant scanner mode. A couple of possible explanations (not mutually
> exclusive):
>
> a) photophysics - re TRex (PubMed 19337661), papers by
> Sanden/Spielmann/Widergren et al (PubMed 20375039, 20196585, 19007245,
> 17385841).
> b) calibration method(s) by Zeiss and/or Leica that try to (but do
> not always) match output at all settings.
> See also
>
>
> high speed
> scanning has the potential to increase fluorescence yield and to reduce
> photobleaching. Borlinghaus RT. Microsc Res Tech. 2006
> Sep;69(9):689-92.PMID: 16878313
>
>
>
>
>
>
>
>
> At 12:37 PM 5/7/2010, you wrote:
>
> Dear list,
>
> I'm looking for general information about the comparison between the
> resonant scanners in the SP5 versus A1R. Apart other comments, are
> possible future upgrades by software in the image format, panning and
> quick switching between the high resolution scanner and the resonant
> one?
>
>
> Thank you very much in advance.
>
>
> --
>
>
> Juan Luis Ribas
>
> Servicio de Microscopía
>
> Centro de Investigación, Tecnología e Innovación
>
> Universidad de Sevilla
>
> Av. Reina Mercedes 4b
>
> 41012 Sevilla
>
>
>
Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
|

Re: eLabExperts Management System

Hi!

Check our open source solution. It does everything this one does that is
really useful for a facility.

an example:

http://calendar.igc.gulbenkian.pt/weekview.php?resource=12

-yellow entries: pre-reservations
-green entries: with confirmation in situ at a certain time (entry time
+/- configurable interval ).
-blue entries: have waiting list with SMS or email warning in case of
entry deleted
-red entries: user did not show up

-4 different types of resource handling:
        a) direct booking
        b) pre-booking with confirmation in situ and during entry time +/-
configurable interval
        c) pre-booking with admin confirmation. Automatic warning for all
situations with .ics integration
        d) just monitoring, eg, -80, ultracentrifuges, etc

-Fully configurable reports with spread sheet export
-different levels of administration (resource, user, permissions, etc)
-harware monitoring integrated, tested on several confocals, computers,
-80c freezers, Facscan, Facscalibur (temperature and pressure included),
etc with sms or email warning in case of exceeding threshold

If you are not a fan of open source software, we have protocols with
companies to set servers and install the hardware monitoring.


All the best,
NM

PS1: I'm starting to make the documentation. Wish me good luck :(
PS2: Next week we should have a demo site. If interested please email me



On 5/25/10 2:38 PM, yuansheng sun wrote:

> We just started using it.  I think it is a very good and useful tool
> for managing a facility.  It meets all our needs for managing our
> facility.  Cheers.  - sheng
>
>
> Yuansheng Sun, Ph.D
> Keck Center for Cellular Imaging
> University of Virginia
>
>
> On Tue, May 25, 2010 at 6:59 AM, Cameron Nowell
> <[hidden email]>  wrote:
>> Hi List,
>>
>> Is anyone out there using the eLabExperts booking, charging etc system? see http://www.elabexperts.com/core/default.aspx
>>
>> it looks really good and seems to be a more refined (and payed for) version of PPMS
>>
>> If you have it, what has your experience been? Does it do everything it should?
>>
>>
>> Cheers
>>
>>
>> Cam
>>
>>
>>
>> *Cameron J. Nowell
>> *Microscopy Manager
>> Centre for Advanced Microscopy
>> Ludwig Institute for Cancer Research
>> PO Box 2008
>> Royal Melbourne Hospital
>> Victoria, 3050
>> AUSTRALIA
>>
>> *Office*: +61 3 9341 3155
>> *Mobile*: +61422882700
>> *Fax*: +61 3 9341 3104
>>
>> Facility Website<http://www.ludwig.edu.au/confocal/>
>>
>>
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List on behalf of Maclyn McCarty
>> Sent: Mon 24/05/2010 11:26 PM
>> To: [hidden email]
>> Subject: Re: Resonant scanner from A1R vs SP5 ...
>>
>>
>> What is the advantage of using resonant scanner?
>>
>> Is there any technical comparison of the resonant scanner of A1 and SP5?
>>
>> Mac.
>>
>>
>> On Sat, 8/5/10, George McNamara<[hidden email]>  wrote:
>>
>> From: George McNamara<[hidden email]>
>> Subject: Re: Resonant scanner from A1R vs SP5 ...
>> To: [hidden email]
>> Date: Saturday, 8 May, 2010, 2:45 PM
>>
>>
>>
>> Hi Juan,
>>
>>
>> On the SP5, when you start LAS AF the choice is resonant scanner or high
>> resolution scanner. No way currently to quickly switch, but do you really
>> need to? If you have a motorized stage, you can tile scan in resonant
>> scanner mode faster than you could exit and restart.
>>
>>
>> George
>>
>> p.s. I've noticed on both the Leica SP5, Zeiss LSM510 and Zeiss LSM710
>> that "faster is brighter" for many fluorophores. That is, a 0.4
>> us dwell time results in a brighter image than a 4 us which is brighter
>> than a 40 us. This implies that<0.4 us might be even brighter - i.e.
>> resonant scanner mode. A couple of possible explanations (not mutually
>> exclusive):
>>
>> a) photophysics - re TRex (PubMed 19337661), papers by
>> Sanden/Spielmann/Widergren et al (PubMed 20375039, 20196585, 19007245,
>> 17385841).
>> b) calibration method(s) by Zeiss and/or Leica that try to (but do
>> not always) match output at all settings.
>> See also
>>
>>
>> high speed
>> scanning has the potential to increase fluorescence yield and to reduce
>> photobleaching. Borlinghaus RT. Microsc Res Tech. 2006
>> Sep;69(9):689-92.PMID: 16878313
>>
>>
>>
>>
>>
>>
>>
>>
>> At 12:37 PM 5/7/2010, you wrote:
>>
>> Dear list,
>>
>> I'm looking for general information about the comparison between the
>> resonant scanners in the SP5 versus A1R. Apart other comments, are
>> possible future upgrades by software in the image format, panning and
>> quick switching between the high resolution scanner and the resonant
>> one?
>>
>>
>> Thank you very much in advance.
>>
>>
>> --
>>
>>
>> Juan Luis Ribas
>>
>> Servicio de Microscopía
>>
>> Centro de Investigación, Tecnología e Innovación
>>
>> Universidad de Sevilla
>>
>> Av. Reina Mercedes 4b
>>
>> 41012 Sevilla
>>
>>
>>
Jerry (Gerald) Sedgewick Jerry (Gerald) Sedgewick
Reply | Threaded
Open this post in threaded view
|

1st Annual Imaging in Research Workshop *commercial/academic announcement*

Hello All,

44 percent of cases investigated by the Office of Research Integrity in
2005-6 involved accusations of image fraud, compared with about 6 percent
a decade before that (1). These cases are rising, and most involve
graduate and post-doctoral students.  Students and staff who self-report
familiarity with imaging programs like Photoshop may be using it
improperly.

In response to this trend, the "1st Annual Imaging in Research Course:
Ethics, Acquisition, Post-Processing, Output and Segmenting" is being
held at the University of Minnesota Continuing Education Center in
Minneapolis/St. Paul, Minnesota, August 16 - 19, 2010, sponsored by the
Histochemical Society and the Adobe Corporation.  Attendees can choose to
attend the course for 3- or 4-days.  This workshop will educate those in
science, medicine and engineering about correct techniques when
acquiring, post-processing, and adjusting images for outputs; along with
techniques that work for segmenting complex, biological images (for
subsequent image analysis).

Other benefits of taking the course will likely result in:


Faster acceptance of submitted manuscripts

Authors better able to demonstrate outcomes to their target audience

Faster results from quantitation, with improved ability to segment
desired features

Better documentation of imaging procedures

Standardization of post-processing

Learning to adjust and modify images minimally and through the objective
use of numbers.

Jerry Sedgewick will present, along with invited speakers.  Jerry
directed a core light microscopy and imaging facility for 15 years at the
University of Minnesota, published 2 books on Photoshop and digital
imaging, and his quantitative work has led to FDA approval for start up
companies.

Please go to http://www.imagingandanalysis.com/seminars.html for more
information.  There is a limit of 30 seats.

If you are a core facility director, you may also wish to pass this along
to your users.  Information and techniques learned in this course will
aid in allaying concern about how your users acquire and post-process
images; and will lead to better grant and publication opportunities, as
well as improved images.

The cost is $590 for 3-days and $840 for 4-days, U.S. dollars. It
includes lunch, beverages and snacks.  Registrations for those who
received information about this course via their core facility will
receive discounts.

All the best,

Jerry Sedgewick

1. "Journals Find Many Images in Research Are Faked,"
Jeffrey R. Young, The Chronicle of Higher Education, September 9, 2009)
RICHARD BURRY RICHARD BURRY
Reply | Threaded
Open this post in threaded view
|

Re: 1st Annual Imaging in Research Workshop *commercial/academic announcement*

Bill
I see the HCS is sponsoring a meeting on image ethics this summer.  See bold below.
Dick

----- Original Message -----
From: "Jerry (Gerald) Sedgewick" <[hidden email]>
Date: Tuesday, May 25, 2010 12:34 pm
Subject: 1st Annual Imaging in Research Workshop *commercial/academic announcement*
To: [hidden email]

> Hello All,
>
> 44 percent of cases investigated by the Office of Research
> Integrity in
> 2005-6 involved accusations of image fraud, compared with about
> 6 percent
> a decade before that (1). These cases are rising, and most involve
> graduate and post-doctoral students.  Students and staff
> who self-report
> familiarity with imaging programs like Photoshop may be using it
> improperly.
>
> In response to this trend, the "1st Annual Imaging in Research Course:
> Ethics, Acquisition, Post-Processing, Output and Segmenting" is being
> held at the University of Minnesota Continuing Education Center in
> Minneapolis/St. Paul, Minnesota, August 16 - 19, 2010, sponsored
> by the
> Histochemical Society and the Adobe Corporation.  Attendees
> can choose to
> attend the course for 3- or 4-days.  This workshop will
> educate those in
> science, medicine and engineering about correct techniques when
> acquiring, post-processing, and adjusting images for outputs;
> along with
> techniques that work for segmenting complex, biological images (for
> subsequent image analysis).
>
> Other benefits of taking the course will likely result in:
>
>
> Faster acceptance of submitted manuscripts
>
> Authors better able to demonstrate outcomes to their target audience
>
> Faster results from quantitation, with improved ability to
> segment
> desired features
>
> Better documentation of imaging procedures
>
> Standardization of post-processing
>
> Learning to adjust and modify images minimally and through the
> objective
> use of numbers.
>
> Jerry Sedgewick will present, along with invited speakers.  Jerry
> directed a core light microscopy and imaging facility for 15
> years at the
> University of Minnesota, published 2 books on Photoshop and digital
> imaging, and his quantitative work has led to FDA approval for
> start up
> companies.
>
> Please go to http://www.imagingandanalysis.com/seminars.html for more
> information.  There is a limit of 30 seats.
>
> If you are a core facility director, you may also wish to pass
> this along
> to your users.  Information and techniques learned in this
> course will
> aid in allaying concern about how your users acquire and post-process
> images; and will lead to better grant and publication
> opportunities, as
> well as improved images.
>
> The cost is $590 for 3-days and $840 for 4-days, U.S. dollars. It
> includes lunch, beverages and snacks.  Registrations for
> those who
> received information about this course via their core facility will
> receive discounts.
>
> All the best,
>
> Jerry Sedgewick
>
> 1. "Journals Find Many Images in Research Are Faked,"
> Jeffrey R. Young, The Chronicle of Higher Education, September
> 9, 2009)
>
>
> --
> BEGIN-ANTISPAM-VOTING-LINKS
> ------------------------------------------------------
>
> Teach CanIt if this mail (ID 1043290605) is spam:
> Spam:       
> https://antispam.osu.edu/b.php?i=1043290605&m=855efe3218ec&c=sNot spam:    https://antispam.osu.edu/b.php?i=1043290605&m=855efe3218ec&c=n
> Forget vote:
> https://antispam.osu.edu/b.php?i=1043290605&m=855efe3218ec&c=f---
> ---------------------------------------------------
> END-ANTISPAM-VOTING-LINKS
>

Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849

Roshma Azeem Roshma Azeem
Reply | Threaded
Open this post in threaded view
|

Re: Resonant scanner from A1R vs SP5 ...

In reply to this post by Maclyn McCarty

Hi Mac,

Fast acquisition time is required to analyze rapid biological processes in a cell. Resonant scanners are about 10 times faster compared to the speed of conventional scanners that are able to acquire fast frame recording and provide real time live images. Due to the faster frame rate, they are useful in resolving the complicated dynamic changes in living cells.

Resonant scanners have some disadvantages like higher readout noise. In addition, their duty cycle is short as resonant scanners accelerate fast that may affect the scanning mirror .

Further technical information can be seen at the following sites:

http://www.microscopyu.com/tutorials/flash/resonantscanning/confocalresonantscanning/index.html

http://www.microscopyu.com/articles/confocal/resonantscanning.html


Nikon's A1R has a hybrid scanner and the SP5 has tandem scanner.


Roshma.
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
|

Re: Resonant scanner from A1R vs SP5 ...

In reply to this post by Maclyn McCarty
Hi Mac,

On the Leica SP5 standard scanner can be operated from 100 Hz to 1400 Hz, with a default of 400 Hz (lines per second). The default (or less) is needed is you want to operate at 1x zoom - higher scan speeds use smaller field of views (you can zoom way up at any speed). Resonant scanner operates at 8000 Hz (lines per second).

Upshot: resonant scanner operates at 5.7x faster than the fastest standard scanner speed.

George



George

At 09:26 AM 5/24/2010, you wrote:

What is the advantage of using resonant scanner?

Is there any technical comparison of the resonant scanner of A1 and SP5?

Mac.


On Sat, 8/5/10, George McNamara <[hidden email]> wrote:

From: George McNamara <[hidden email]>
Subject: Re: Resonant scanner from A1R vs SP5 ...
To: [hidden email]
Date: Saturday, 8 May, 2010, 2:45 PM

Hi Juan,

On the SP5, when you start LAS AF the choice is resonant scanner or high resolution scanner. No way currently to quickly switch, but do you really need to? If you have a motorized stage, you can tile scan in resonant scanner mode faster than you could exit and restart.

George
p.s. I've noticed on both the Leica SP5, Zeiss LSM510 and Zeiss LSM710 that "faster is brighter" for many fluorophores. That is, a 0.4 us dwell time results in a brighter image than a 4 us which is brighter than a 40 us. This implies that <0.4 us might be even brighter - i.e. resonant scanner mode. A couple of possible explanations (not mutually exclusive):
a) photophysics - re TRex (PubMed 19337661), papers by Sanden/Spielmann/Widergren et al (PubMed 20375039, 20196585, 19007245, 17385841).
b) calibration method(s) by Zeiss and/or Leica that try to (but do not always) match output at all settings.
See also

high speed scanning has the potential to increase fluorescence yield and to reduce photobleaching. Borlinghaus RT. Microsc Res Tech. 2006 Sep;69(9):689-92.PMID: 16878313





At 12:37 PM 5/7/2010, you wrote:
Dear list,
I'm looking for general information about the comparison between the resonant scanners in the SP5 versus A1R. Apart other comments, are possible future upgrades by software in the image format, panning and quick switching between the high resolution scanner and the resonant one?

Thank you very much in advance.

--

Juan Luis Ribas
Servicio de Microscopía
Centro de Investigación, Tecnología e Innovación
Universidad de Sevilla
Av. Reina Mercedes 4b
41012 Sevilla







George McNamara, Ph.D.
Image Core Manager
Analytical Imaging Core Facility
University of Miami, Miller School of Medicine
Miami, FL 33136
[hidden email]
[hidden email]
305-243-8436 office
http://www.sylvester.org/AICF (Analytical Imaging Core Facility)
http://www.sylvester.org/AICF/pubspectra.zip (the entire 2000+ spectra .xlsx file is in the zip file)
http://home.earthlink.net/~geomcnamara

David Strachan David Strachan
Reply | Threaded
Open this post in threaded view
|

Re: Resonant scanner from A1R vs SP5 ...

I am sure you can just type in the box to select other values than the nice round numbers given originally, even down to 1 hertz.

 

 

See below…..

On the Leica SP5 standard scanner can be operated from 100 Hz to 1400 Hz,

 

 

 

 

David Strachan

The Beatson Institute for Cancer Research

Glasgow  G61 1BD

( Direct Line +44 (0) 141 330 6872

( Fax                +44 (0) 141 942 6521

* E-mail          [hidden email]


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara
Sent: 31 May 2010 04:40
To: [hidden email]
Subject: Re: Resonant scanner from A1R vs SP5 ...

 

Hi Mac,

On the Leica SP5 standard scanner can be operated from 100 Hz to 1400 Hz, with a default of 400 Hz (lines per second). The default (or less) is needed is you want to operate at 1x zoom - higher scan speeds use smaller field of views (you can zoom way up at any speed). Resonant scanner operates at 8000 Hz (lines per second).

Upshot: resonant scanner operates at 5.7x faster than the fastest standard scanner speed.

George



George

At 09:26 AM 5/24/2010, you wrote:


What is the advantage of using resonant scanner?

Is there any technical comparison of the resonant scanner of A1 and SP5?

Mac.


On Sat, 8/5/10, George McNamara <[hidden email]> wrote:

From: George McNamara <[hidden email]>

Subject: Re: Resonant scanner from A1R vs SP5 ...

To: [hidden email]

Date: Saturday, 8 May, 2010, 2:45 PM

Hi Juan,

On the SP5, when you start LAS AF the choice is resonant scanner or high resolution scanner. No way currently to quickly switch, but do you really need to? If you have a motorized stage, you can tile scan in resonant scanner mode faster than you could exit and restart.

George

p.s. I've noticed on both the Leica SP5, Zeiss LSM510 and Zeiss LSM710 that "faster is brighter" for many fluorophores. That is, a 0.4 us dwell time results in a brighter image than a 4 us which is brighter than a 40 us. This implies that <0.4 us might be even brighter - i.e. resonant scanner mode. A couple of possible explanations (not mutually exclusive):

a) photophysics - re TRex (PubMed 19337661), papers by Sanden/Spielmann/Widergren et al (PubMed 20375039, 20196585, 19007245, 17385841).

b) calibration method(s) by Zeiss and/or Leica that try to (but do not always) match output at all settings.

See also

high speed scanning has the potential to increase fluorescence yield and to reduce photobleaching. Borlinghaus RT. Microsc Res Tech. 2006 Sep;69(9):689-92.PMID: 16878313




At 12:37 PM 5/7/2010, you wrote:

Dear list,

I'm looking for general information about the comparison between the resonant scanners in the SP5 versus A1R. Apart other comments, are possible future upgrades by software in the image format, panning and quick switching between the high resolution scanner and the resonant one?

Thank you very much in advance.

--

Juan Luis Ribas

Servicio de Microscopía

Centro de Investigación, Tecnología e Innovación

Universidad de Sevilla

Av. Reina Mercedes 4b

41012 Sevilla

 







George McNamara, Ph.D.
Image Core Manager
Analytical Imaging Core Facility
University of Miami, Miller School of Medicine
Miami, FL 33136
[hidden email]
[hidden email]
305-243-8436 office
http://www.sylvester.org/AICF (Analytical Imaging Core Facility)
http://www.sylvester.org/AICF/pubspectra.zip (the entire 2000+ spectra .xlsx file is in the zip file)