Alessandro Esposito |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, in order to guarantee backups, desktop portability and integration f various advanced IT services, our institute established a SAN and virtual environments where all IT services are centralized. This works wonderfully but at the cost of downtimes caused by planned (which is not an issue) and unplanned downtime of all services. The obvious solution is to make research instruments independent from the main server with loss of some services and increase of maintenence cost on individual computers. Have anybody done a serious analysis of SAN requirements in order to maintain continuity of service? We are an institute serving between 150-200 employees and we are discussing on possible improvment to our infrastructure. I guess the most important question I am trying to answer is: 1) how much unplanned downtime is considered acceptable by other research institutes (in particular those with significant imaging facilities) and 2) if some key solutions have been identified for minimizing the impact on running microscopes during an IT outage. Again, although the basic solution is to isolate instruments, I hope we could discuss about a better compromise :) Cheers, Alessandro |
Chris Tully-2 |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Alessandro, This response in based not on personal experience but on helping a major hospital network setup a regional digital pathology network covering 5 hospitals separated by anywhere from 5 to 200 miles, while I was working for Aperio. For many reasons starting with the volume of data they expected to generate (multiple GB per day), they built not one but two SANs at two different sites. Then every connected instrument was configured to consider one of the sites (A) as it's primary, with automatic fail over to the secondary site (B). SAN site B was configured to live mirror site A so that in the event that A went down the most anyone would loose is the slide that was actively being scanned when the fail over happened. Further the scanning software was designed to recognize this failure and flag it for future correction. Obviously, mirrored SAN systems become expensive, but if minimal down time is required (desired) then that is the only solution I can come up with. Check into each instrument connected to your SAN. As with the Aperio Scanners that I was helping to implement, your instruments may have a third fail over mode either automatically setup of available to be configured for storing data locally when the SAN is unavailable. The challenge then becomes migrating data from local storage to the SAN once it is available again. You will want to push the manufacturers of each instrument to work with your IT department to automate that migration as much as possible (ideally, the SAN or the instrument will detect the restored connection and automatically push/pull the locally stored data). Then your only remaining challenge is to talk with your IT people and find out how long it would take them to recover from a worst case scenario and add enough local storage to each instrument to cover that event. How far you go down that path will obviously be governed by budget, backed up by need. The more failure modes you can plan for the less down time your users will experience. Obviously having data stored locally for some time will not be as convenient, but at least it is not lost. Chris Tully On 9/18/2013 3:37 AM, Alessandro Esposito wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > in order to guarantee backups, desktop portability and integration f various > advanced IT services, our institute established a SAN and virtual environments > where all IT services are centralized. > > This works wonderfully but at the cost of downtimes caused by planned > (which is not an issue) and unplanned downtime of all services. > > The obvious solution is to make research instruments independent from the > main server with loss of some services and increase of maintenence cost on > individual computers. > > Have anybody done a serious analysis of SAN requirements in order to maintain > continuity of service? We are an institute serving between 150-200 employees > and we are discussing on possible improvment to our infrastructure. > > I guess the most important question I am trying to answer is: 1) how much > unplanned downtime is considered acceptable by other research institutes (in > particular those with significant imaging facilities) and 2) if some key solutions > have been identified for minimizing the impact on running microscopes during > an IT outage. > > Again, although the basic solution is to isolate instruments, I hope we could > discuss about a better compromise :) > > Cheers, > > Alessandro -- *Chris Tully* Principal Consultant 240-475-9753 Image Incyte, LLC <http:%5C%5Cwww.ImageIncyte.com> [hidden email] <mailto:[hidden email]> |
Mark Cannell-2 |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** There are some simple tools to achieve folder mirroring e.g. http://www.techsoftpl.com/backup/index.php. For Unix systems this could be done by rsync or scripts called by CRON. HTH Cheers On 18/09/2013, at 2:08 PM, Chris Tully <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Alessandro, > > This response in based not on personal experience but on helping a major hospital network setup a regional digital pathology network covering 5 hospitals separated by anywhere from 5 to 200 miles, while I was working for Aperio. > > For many reasons starting with the volume of data they expected to generate (multiple GB per day), they built not one but two SANs at two different sites. Then every connected instrument was configured to consider one of the sites (A) as it's primary, with automatic fail over to the secondary site (B). SAN site B was configured to live mirror site A so that in the event that A went down the most anyone would loose is the slide that was actively being scanned when the fail over happened. Further the scanning software was designed to recognize this failure and flag it for future correction. > > Obviously, mirrored SAN systems become expensive, but if minimal down time is required (desired) then that is the only solution I can come up with. Check into each instrument connected to your SAN. As with the Aperio Scanners that I was helping to implement, your instruments may have a third fail over mode either automatically setup of available to be configured for storing data locally when the SAN is unavailable. The challenge then becomes migrating data from local storage to the SAN once it is available again. You will want to push the manufacturers of each instrument to work with your IT department to automate that migration as much as possible (ideally, the SAN or the instrument will detect the restored connection and automatically push/pull the locally stored data). Then your only remaining challenge is to talk with your IT people and find out how long it would take them to recover from a worst case scenario and add enough local storage to each instrument to cover that event. > > How far you go down that path will obviously be governed by budget, backed up by need. The more failure modes you can plan for the less down time your users will experience. > > Obviously having data stored locally for some time will not be as convenient, but at least it is not lost. > > Chris Tully > > On 9/18/2013 3:37 AM, Alessandro Esposito wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> in order to guarantee backups, desktop portability and integration f various >> advanced IT services, our institute established a SAN and virtual environments >> where all IT services are centralized. >> >> This works wonderfully but at the cost of downtimes caused by planned >> (which is not an issue) and unplanned downtime of all services. >> >> The obvious solution is to make research instruments independent from the >> main server with loss of some services and increase of maintenence cost on >> individual computers. >> >> Have anybody done a serious analysis of SAN requirements in order to maintain >> continuity of service? We are an institute serving between 150-200 employees >> and we are discussing on possible improvment to our infrastructure. >> >> I guess the most important question I am trying to answer is: 1) how much >> unplanned downtime is considered acceptable by other research institutes (in >> particular those with significant imaging facilities) and 2) if some key solutions >> have been identified for minimizing the impact on running microscopes during >> an IT outage. >> >> Again, although the basic solution is to isolate instruments, I hope we could >> discuss about a better compromise :) >> >> Cheers, >> >> Alessandro > > > -- > *Chris Tully* > Principal Consultant > 240-475-9753 > > > Image Incyte, LLC > > <http:%5C%5Cwww.ImageIncyte.com> [hidden email] <mailto:[hidden email]> Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
Alessandro Esposito |
In reply to this post by Alessandro Esposito
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Chris, thank you for your interesting example. I did not consider at all the possibility of the redirect to be handled by the microscope in itself and I should check if this is supported at all by a typical microscope (probably not). It would be smart that if in the absence of the SAN, the micropscope would switch saving to the local harddrive. I'll look into it. For other systems, such as screening platforms, that would be not possible and the solution you mention should be adopted, but manufacturers could implement saving locally with metadata the would be later recovered and transfered to the SQL server togehter with the data on the SAN. Thank you Alessandro On Wed, 18 Sep 2013 09:08:12 -0400, Chris Tully <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Alessandro, > >This response in based not on personal experience but on helping a major >hospital network setup a regional digital pathology network covering 5 >hospitals separated by anywhere from 5 to 200 miles, while I was working >for Aperio. > >For many reasons starting with the volume of data they expected to >generate (multiple GB per day), they built not one but two SANs at two >different sites. Then every connected instrument was configured to >consider one of the sites (A) as it's primary, with automatic fail over >to the secondary site (B). SAN site B was configured to live mirror site >A so that in the event that A went down the most anyone would loose is >the slide that was actively being scanned when the fail over happened. >Further the scanning software was designed to recognize this failure and >flag it for future correction. > >Obviously, mirrored SAN systems become expensive, but if minimal down >time is required (desired) then that is the only solution I can come up >with. Check into each instrument connected to your SAN. As with the >Aperio Scanners that I was helping to implement, your instruments may >have a third fail over mode either automatically setup of available to >be configured for storing data locally when the SAN is unavailable. The >challenge then becomes migrating data from local storage to the SAN once >it is available again. You will want to push the manufacturers of each >instrument to work with your IT department to automate that migration as >much as possible (ideally, the SAN or the instrument will detect the >restored connection and automatically push/pull the locally stored >data). Then your only remaining challenge is to talk with your IT people >and find out how long it would take them to recover from a worst case >scenario and add enough local storage to each instrument to cover that >event. > >How far you go down that path will obviously be governed by budget, >backed up by need. The more failure modes you can plan for the less >down time your users will experience. > >Obviously having data stored locally for some time will not be as >convenient, but at least it is not lost. > >Chris Tully > >On 9/18/2013 3:37 AM, Alessandro Esposito wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> in order to guarantee backups, desktop portability and >> advanced IT services, our institute established a SAN and virtual environments >> where all IT services are centralized. >> >> This works wonderfully but at the cost of downtimes caused by planned >> (which is not an issue) and unplanned downtime of all services. >> >> The obvious solution is to make research instruments independent from the >> main server with loss of some services and increase of maintenence cost on >> individual computers. >> >> Have anybody done a serious analysis of SAN requirements in order to maintain >> continuity of service? We are an institute serving between 150-200 employees >> and we are discussing on possible improvment to our infrastructure. >> >> I guess the most important question I am trying to answer is: 1) how much >> unplanned downtime is considered acceptable by other research institutes (in >> particular those with significant imaging facilities) and 2) if some key solutions >> have been identified for minimizing the impact on running microscopes during >> an IT outage. >> >> Again, although the basic solution is to isolate instruments, I hope we could >> discuss about a better compromise :) >> >> Cheers, >> >> Alessandro > > >-- >*Chris Tully* >Principal Consultant >240-475-9753 > > > Image Incyte, LLC > ><http:%5C%5Cwww.ImageIncyte.com> [hidden email] ><mailto:[hidden email]> |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We've thought about this and decided that regardless where data ultimately reside, all research instruments are stand alone when collecting data. Regardless of network status, they may collect data uninterrupted. We can tolerate a wait to recall data residing on a server that is temporarily down but we cannot tolerate interruption of ongoing experiments, especially those involving live material. Regards, Michael ________________________________________________________ Michael Cammer, Assistant Research Scientist Microscopy Core, NYU Langone Medical Center & Skirball Institute of Biomolecular Medicine Cell: (914) 309-3270 Microscopy Lab: (212) 263-7099 Dustin Lab: (212) 263-3208 |
Cameron, Lisa |
In reply to this post by Chris Tully-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello - Have people tested and found optimal 2-photon laser excitation wavelengths for some of the newer fluorophores like mTurquoise2 and Clover? I'd be interested to hear what people have tried / found. Our specific application is FLIM-FRET. We do the measurements with a Becker & Hickl time domain system on a Zeiss 710 with 2p laser. Thank you for any information - Lisa --------------------------------------- Lisa Cameron, Ph.D. Director of Confocal and Light Microscopy Core Dana Farber Cancer Institute 450 Brookline Ave.; JF 621 Boston, MA 02215 http://microscopy.dfci.harvard.edu The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
Tobias Rose |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Indeed - I'd like to second that question. Right now I'm using 850nm for mTurq2. This may be suboptimal, though. Well: I should probably just test it. Tobias -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cameron, Lisa Sent: Friday, September 20, 2013 15:40 To: [hidden email] Subject: 2-photon excitation wavelength recommendations for mTurq2 and clover ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello - Have people tested and found optimal 2-photon laser excitation wavelengths for some of the newer fluorophores like mTurquoise2 and Clover? I'd be interested to hear what people have tried / found. Our specific application is FLIM-FRET. We do the measurements with a Becker & Hickl time domain system on a Zeiss 710 with 2p laser. Thank you for any information - Lisa --------------------------------------- Lisa Cameron, Ph.D. Director of Confocal and Light Microscopy Core Dana Farber Cancer Institute 450 Brookline Ave.; JF 621 Boston, MA 02215 http://microscopy.dfci.harvard.edu The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
Free forum by Nabble | Edit this page |