I have a user who wants to use our objective based TIRF to
measure particle absorption to the coverslip using scattering. They are
currently using a 633 nm laser and looking at scatter with a 10x lens and a PMT
on a prism based TIRF. My system has a 633 laser and we have typical filters to
look at fluorescence but I’m not sure if there is something special I would
need to do to look at the scattering. I’m not sure if the bandpass filter
we have for fluorescence will pick it up or not. Any suggestions would be appreciated. Sincerely, Claire Claire
M. Brown, PhD Life
Sciences Complex Imaging Facility Director |
|
|
Hi Claire,
Hope all is well up in Montreal. The simplest thing to do to image the scattered light of TIR illumination would be to just remove the emission filter from your filter cube and perhaps replace the dichroic with a half-silvered mirror like you find in reflected brightfield filter cubes, but this second part isn't totally necessary. Couple of things to keep in mind: 1. Be very careful looking at the sample through the eyepieces (if any are present) since you are now viewing laser light directly. You'll want to have a way to adjust and turn down the laser power that is incident on the sample either with a gradient neutral density filter wheel or electronically if your laser has this ability. The same applies for your detector or imaging device. Don't expose these to too much light or they could be damaged. 2. Scattered TIR images don't look great - usually there is a lot of background and laser interference fringes in the image. Nevertheless, the intensity of the scattering object that is being imaged can be used to do some very interesting quantitative work. For example, see these publications: TOTAL INTERNAL-REFLECTION MICROSCOPY - A SURFACE INSPECTION TECHNIQUE Author(s): TEMPLE PA Source: APPLIED OPTICS Volume: 20 Issue: 15 Pages: 2656-2664 Published: 1981 Measurement of colloidal forces with TIRM. Author(s): Prieve DC Source: ADVANCES IN COLLOID AND INTERFACE SCIENCE Volume: 82 Issue: 1-3 Pages: 93-125 Published: OCT 10 1999 Measuring colloidal forces using evanescent wave scattering Author(s): Bike SG Source: CURRENT OPINION IN COLLOID & INTERFACE SCIENCE Volume: 5 Issue: 1-2 Pages: 144-150 Published: MAR 2000 A new approach for analyzing particle motion near an interface using total internal reflection microscopy Author(s): Oetama RJ, Walz JY Source: JOURNAL OF COLLOID AND INTERFACE SCIENCE Volume: 284 Issue: 1 Pages: 323-331 Published: APR 1 2005 Surface Plasmon Resonance Imaging Using a High Numerical Aperture Microscope Objective Author(s): Huang B (Huang, Bo), Yu F (Yu, Fang), Zare RN (Zare, Richard N.) Source: ANALYTICAL CHEMISTRY Volume: 79 Issue: 7 Pages: 2979-2983 Published: APR 1 2007 These publications also contain instrumental diagrams which are (with the exception of the last paper) essentially prism-based TIRF microscopes without an emission filter to block scattered laser light. It is interesting to note that "TIR" microscopy ("TIRF" microscopy with the F-fluorescence dropped from the name) seems to have developed in the materials science and analytical chemistry fields independent of the the advances made in the biological sciences with TIRF microscopy. 3. The first and the last papers also show that additional contrast and information about the orientation of scattering particles can be gained by varying the polarization of the incident illumination beam, but you probably won't have the ability to easily change that on a commercial system. Try rotating the optical fiber connector that delivers the laser light into the microscope system. If it's a "polarization maintaining" fiber, you may see some variation in the scatting particle's intensity. John Oreopoulos, BSc, PhD Candidate University of Toronto Institute For Biomaterials and Biomedical Engineering Centre For Studies in Molecular Imaging On 11-Jan-10, at 12:22 PM, Claire Brown, Dr. wrote: > I have a user who wants to use our objective based TIRF to measure > particle absorption to the coverslip using scattering. They are > currently using a 633 nm laser and looking at scatter with a 10x > lens and a PMT on a prism based TIRF. > > My system has a 633 laser and we have typical filters to look at > fluorescence but I’m not sure if there is something special I would > need to do to look at the scattering. I’m not sure if the bandpass > filter we have for fluorescence will pick it up or not. > > Any suggestions would be appreciated. > > Sincerely, > > Claire > > Claire M. Brown, PhD > Life Sciences Complex Imaging Facility Director > |
| Free forum by Nabble | Edit this page |