Alberto Diaspro |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear following a discussion with the MRT staff and Wiley, they informed me that the paper is now available for free for one month. Hope this can help. Please, have a check and let me know if it works at http://onlinelibrary.wiley.com/doi/10.1002/jemt.22834/full Best Alby |
|
James Pawley |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, Those attending the sadly-gone UBC 3D-course may remember that our first slogan was, "If it isn’t diffraction, it’s statistics.” Colin’s wonderful, inclusive paper does indeed wrap up the diffraction part superbly. But how many photons/nm*3 would you have to collect in order to obtain data that would allow you to fill in those beautifully narrow lines? Millions? (if the line width is say 1% of its “value” surely you need 10,000 detected photons to determine it And how big must a pixel be to trace out these lovely curves without them looking blocky? Given that biologists habitually work with specimens capable of producing at most tens of detectable photons/pixel, surely it is the statistics that places the most severe spatial limits on the data we finally obtain. Where is the young mathematician (statistician?) who can give us an equally comprehensive analysis of the number of photons per cubic nm that we must elicit and detectin order to “resolve” two closely spaced point objects? And why do we always assume that these two objects have the same intensity when this is so seldom the case? What visibility criteria would be appropriate to model the problem of seeing the Earth from Alpha Centauri? Please someone take up this quest. Jim Pawley James and Christine Pawley, 5446 Burley Place, Box 2348, Sechelt BC, Canada, V0N3A0 [hidden email]<mailto:[hidden email]>, Phone 1-604-885-0840, cell 1-604-989-6146 On Feb 13, 2017, at 7:11 AM, Alberto Diaspro <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear following a discussion with the MRT staff and Wiley, they informed me that the paper is now available for free for one month. Hope this can help. Please, have a check and let me know if it works at http://onlinelibrary.wiley.com/doi/10.1002/jemt.22834/full Best Alby |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Doesn't an infinitely sharp edge require infinite frequency content? Therefore infinite data? Craig On Mon, Feb 13, 2017 at 6:57 PM, JAMES B PAWLEY <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > Those attending the sadly-gone UBC 3D-course may remember that our first > slogan was, "If it isn’t diffraction, it’s statistics.” > > Colin’s wonderful, inclusive paper does indeed wrap up the diffraction > part superbly. > > But how many photons/nm*3 would you have to collect in order to obtain > data that would allow you to fill in those beautifully narrow lines? > Millions? (if the line width is say 1% of its “value” surely you need > 10,000 detected photons to determine it And how big must a pixel be to > trace out these lovely curves without them looking blocky? Given that > biologists habitually work with specimens capable of producing at most tens > of detectable photons/pixel, surely it is the statistics that places the > most severe spatial limits on the data we finally obtain. > > Where is the young mathematician (statistician?) who can give us an > equally comprehensive analysis of the number of photons per cubic nm that > we must elicit and detectin order to “resolve” two closely spaced point > objects? And why do we always assume that these two objects have the same > intensity when this is so seldom the case? What visibility criteria would > be appropriate to model the problem of seeing the Earth from Alpha Centauri? > > Please someone take up this quest. > > Jim Pawley > > James and Christine Pawley, 5446 Burley Place, Box 2348, Sechelt BC, > Canada, V0N3A0 [hidden email]<mailto:[hidden email]>, Phone > 1-604-885-0840, cell 1-604-989-6146 > > > > On Feb 13, 2017, at 7:11 AM, Alberto Diaspro <[hidden email]<mailto: > [hidden email]>> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear > following a discussion with the MRT staff and Wiley, they informed me that > the paper is now available for free for one month. Hope this can help. > Please, have a check and let me know if it works at > http://onlinelibrary.wiley.com/doi/10.1002/jemt.22834/full > > Best > Alby > > |
|
Sripad Ram-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jim, Great question!! Well about 10 years ago we attempted to answer the very same question that you ask, i.e. how many photon do you need to resolve two closely spaced point sources. Stated differently, how does photon statistics impact 2D/3D resolution of optical microscopes? Here are some references to that work: 1. Beyond Rayleigh’s criterion: A resolution measure with application to single-molecule microscopy, PNAS, 2006, 103, 4457-4462 2. A comparative study of high resolution microscopy imaging modalities using a three-dimensional resolution measure, Optics Express, 2009,17:24377-24402 3. A resolution measure for three-dimensional microscopy, Optics Communications, 2009, 282:1751-1761. All of these papers should be on pubmed central as they were part of an NIH funded study and the final accepted manuscripts were uploaded sometime in 2012. Hope this helps. Sripad On Mon, Feb 13, 2017 at 6:01 PM, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Doesn't an infinitely sharp edge require infinite frequency content? > Therefore infinite data? > > Craig > > On Mon, Feb 13, 2017 at 6:57 PM, JAMES B PAWLEY <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi all, > > > > Those attending the sadly-gone UBC 3D-course may remember that our first > > slogan was, "If it isn’t diffraction, it’s statistics.” > > > > Colin’s wonderful, inclusive paper does indeed wrap up the diffraction > > part superbly. > > > > But how many photons/nm*3 would you have to collect in order to obtain > > data that would allow you to fill in those beautifully narrow lines? > > Millions? (if the line width is say 1% of its “value” surely you need > > 10,000 detected photons to determine it And how big must a pixel be to > > trace out these lovely curves without them looking blocky? Given that > > biologists habitually work with specimens capable of producing at most > tens > > of detectable photons/pixel, surely it is the statistics that places the > > most severe spatial limits on the data we finally obtain. > > > > Where is the young mathematician (statistician?) who can give us an > > equally comprehensive analysis of the number of photons per cubic nm > that > > we must elicit and detectin order to “resolve” two closely spaced point > > objects? And why do we always assume that these two objects have the same > > intensity when this is so seldom the case? What visibility criteria would > > be appropriate to model the problem of seeing the Earth from Alpha > Centauri? > > > > Please someone take up this quest. > > > > Jim Pawley > > > > James and Christine Pawley, 5446 Burley Place, Box 2348, Sechelt BC, > > Canada, V0N3A0 [hidden email]<mailto:[hidden email]>, Phone > > 1-604-885-0840, cell 1-604-989-6146 > > > > > > > > On Feb 13, 2017, at 7:11 AM, Alberto Diaspro <[hidden email]<mailto: > > [hidden email]>> wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear > > following a discussion with the MRT staff and Wiley, they informed me > that > > the paper is now available for free for one month. Hope this can help. > > Please, have a check and let me know if it works at > > http://onlinelibrary.wiley.com/doi/10.1002/jemt.22834/full > > > > Best > > Alby > > > > > |
|
Zdenek Svindrych-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, there is also a concept of "Stochastic Transfer Function" that may be able to answer that question. See e.g. here: https://doi.org/10.1364/JOSAA.26.001630 It's from 2009, but it seems nobody followed that idea further... It would be nice if someone compared various SIM methods ('Gustaffson' SIM, re-scan, OPRA, image-scanning microscopy (airyscan), etc...) in terms of photon- limited performance. zdenek ---------- Původní zpráva ---------- Od: S Ram <[hidden email]> Komu: [hidden email] Datum: 13. 2. 2017 23:42:08 Předmět: Re: Sheppard's paper on Resolution - wall removed "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jim, Great question!! Well about 10 years ago we attempted to answer the very same question that you ask, i.e. how many photon do you need to resolve two closely spaced point sources. Stated differently, how does photon statistics impact 2D/3D resolution of optical microscopes? Here are some references to that work: 1. Beyond Rayleigh’s criterion: A resolution measure with application to single-molecule microscopy, PNAS, 2006, 103, 4457-4462 2. A comparative study of high resolution microscopy imaging modalities using a three-dimensional resolution measure, Optics Express, 2009,17:24377-24402 3. A resolution measure for three-dimensional microscopy, Optics Communications, 2009, 282:1751-1761. All of these papers should be on pubmed central as they were part of an NIH funded study and the final accepted manuscripts were uploaded sometime in 2012. Hope this helps. Sripad On Mon, Feb 13, 2017 at 6:01 PM, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Doesn't an infinitely sharp edge require infinite frequency content? > Therefore infinite data? > > Craig > > On Mon, Feb 13, 2017 at 6:57 PM, JAMES B PAWLEY <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi all, > > > > Those attending the sadly-gone UBC 3D-course may remember that our first > > slogan was, "If it isn’t diffraction, it’s statistics.” > > > > Colin’s wonderful, inclusive paper does indeed wrap up the diffraction > > part superbly. > > > > But how many photons/nm*3 would you have to collect in order to obtain > > data that would allow you to fill in those beautifully narrow lines? > > Millions? (if the line width is say 1% of its “value” surely you need > > 10,000 detected photons to determine it And how big must a pixel be to > > trace out these lovely curves without them looking blocky? Given that > > biologists habitually work with specimens capable of producing at most > tens > > of detectable photons/pixel, surely it is the statistics that places the > > most severe spatial limits on the data we finally obtain. > > > > Where is the young mathematician (statistician?) who can give us an > > equally comprehensive analysis of the number of photons per cubic nm > that > > we must elicit and detectin order to “resolve” two closely spaced point > > objects? And why do we always assume that these two objects have the same > > intensity when this is so seldom the case? What visibility criteria would > > be appropriate to model the problem of seeing the Earth from Alpha > Centauri? > > > > Please someone take up this quest. > > > > Jim Pawley > > > > James and Christine Pawley, 5446 Burley Place, Box 2348, Sechelt BC, > > Canada, V0N3A0 [hidden email]<mailto:[hidden email]>, Phone > > 1-604-885-0840, cell 1-604-989-6146 > > > > > > > > On Feb 13, 2017, at 7:11 AM, Alberto Diaspro <[hidden email]<mailto: > > [hidden email]>> wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear > > following a discussion with the MRT staff and Wiley, they informed me > that > > the paper is now available for free for one month. Hope this can help. > > Please, have a check and let me know if it works at > > http://onlinelibrary.wiley.com/doi/10.1002/jemt.22834/full > > > > Best > > Alby > > > > > |