Silicone oil objective correction collar stuck

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello all,

We have a 30x Olympus silicone immersion objective (UPLSAPO30XS, 1.05 NA) with a stuck correction collar.  Does anyone have any experience / suggestions on how to get it unstuck? Or would you recommend to get it professionally repaired?

Any advice would be great.

Thanks, Kristina

_____________________
DR KRISTINA JAHN
Imaging Support Specialist
Centenary Institute
 
T: 61 2 9565 6180
E: [hidden email]
Mail Address: Locked Bag No. 6, Newtown 2042, NSW Australia
Delivery Address: Building 93 RPA, Missenden Rd Camperdown 2050, NSW Australia
 


--
weber weber
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Silicone oil objective correction collar stuck

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Maybe you can heat it with a gun heat. I did it on a Leica objectiv and turn the collar.
kspencer007 kspencer007
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Silicone oil objective correction collar stuck

In reply to this post by Kristina Jahn
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Kristina;
        I've used Marvel oil to unstick correction collars. It's a penetrating oil. Minimal amount, let it sit, try to work it loose, add some more oil (small amounts), repeat.
Kathy

The Scripps Research Institute
Dept of Molecular and Cellular Neuroscience
10550 N. Torrey Pines Road
DNC 216
La Jolla, Ca 92037



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kristina Jahn
Sent: Tuesday, June 27, 2017 11:18 PM
To: [hidden email]
Subject: Silicone oil objective correction collar stuck

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello all,

We have a 30x Olympus silicone immersion objective (UPLSAPO30XS, 1.05 NA) with a stuck correction collar.  Does anyone have any experience / suggestions on how to get it unstuck? Or would you recommend to get it professionally repaired?

Any advice would be great.

Thanks, Kristina

_____________________
DR KRISTINA JAHN
Imaging Support Specialist
Centenary Institute
 
T: 61 2 9565 6180
E: [hidden email]
Mail Address: Locked Bag No. 6, Newtown 2042, NSW Australia Delivery Address: Building 93 RPA, Missenden Rd Camperdown 2050, NSW Australia
 


--