Nico Stuurman Nico Stuurman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Does anyone know of a commercial (or other) source for a microscope slide with various diffraction grating patterns?  I want to use this for teaching purposes (demonstrate diffraction in the back focal plane), and ideally the slide would have multiple patterns (spacings from about 0.5 um to 5 um, various orientations, squares, etc..).  Basically, I am looking for a slide as was used in the Abbe diffraction kit.  All pointers will be appreciated!

Thanks for your help,

Nico
Zac Arrac Atelaz Zac Arrac Atelaz
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


There are several resources, this is only one of them
 
http://www.scientrific.com.au/product.php?p=774
 

> Date: Wed, 10 Aug 2011 21:29:25 -0700
> From: [hidden email]
> Subject: Source for diffraction grating slide
> To: [hidden email]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Does anyone know of a commercial (or other) source for a microscope slide with various diffraction grating patterns? I want to use this for teaching purposes (demonstrate diffraction in the back focal plane), and ideally the slide would have multiple patterns (spacings from about 0.5 um to 5 um, various orientations, squares, etc..). Basically, I am looking for a slide as was used in the Abbe diffraction kit. All pointers will be appreciated!
>
> Thanks for your help,
>
> Nico
     
Phillips, Thomas E. Phillips, Thomas E.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Does anyone know of a USA supplier for a similar diffraction grating slide?

Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
[hidden email]

http://www.biology.missouri.edu/faculty/phillips.html
http://www.biotech.missouri.edu/mcc/


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Zac Arrac Atelaz
Sent: Wednesday, August 10, 2011 11:38 PM
To: [hidden email]
Subject: Re: Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


There are several resources, this is only one of them
 
http://www.scientrific.com.au/product.php?p=774
 

> Date: Wed, 10 Aug 2011 21:29:25 -0700
> From: [hidden email]
> Subject: Source for diffraction grating slide
> To: [hidden email]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Does anyone know of a commercial (or other) source for a microscope slide with various diffraction grating patterns? I want to use this for teaching purposes (demonstrate diffraction in the back focal plane), and ideally the slide would have multiple patterns (spacings from about 0.5 um to 5 um, various orientations, squares, etc..). Basically, I am looking for a slide as was used in the Abbe diffraction kit. All pointers will be appreciated!
>
> Thanks for your help,
>
> Nico
     
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

EM grids will work, especially line grids. Maintaining alignment while mounting on a slide is a challenge.

Regards,
Glen


On Aug 11, 2011, at 5:48 AM, Phillips, Thomas E. wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Does anyone know of a USA supplier for a similar diffraction grating slide?
>
> Thomas E. Phillips, Ph.D
> Professor of Biological Sciences
> Director, Molecular Cytology Core
> 2 Tucker Hall
> University of Missouri
> Columbia, MO 65211-7400
> 573-882-4712 (office)
> 573-882-0123 (fax)
> [hidden email]
>
> http://www.biology.missouri.edu/faculty/phillips.html
> http://www.biotech.missouri.edu/mcc/
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Zac Arrac Atelaz
> Sent: Wednesday, August 10, 2011 11:38 PM
> To: [hidden email]
> Subject: Re: Source for diffraction grating slide
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
>
> There are several resources, this is only one of them
>
> http://www.scientrific.com.au/product.php?p=774
>
>
>> Date: Wed, 10 Aug 2011 21:29:25 -0700
>> From: [hidden email]
>> Subject: Source for diffraction grating slide
>> To: [hidden email]
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Does anyone know of a commercial (or other) source for a microscope slide with various diffraction grating patterns? I want to use this for teaching purposes (demonstrate diffraction in the back focal plane), and ideally the slide would have multiple patterns (spacings from about 0.5 um to 5 um, various orientations, squares, etc..). Basically, I am looking for a slide as was used in the Abbe diffraction kit. All pointers will be appreciated!
>>
>> Thanks for your help,
>>
>> Nico
>    
JOEL B. SHEFFIELD JOEL B. SHEFFIELD
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

We have used those toy glasses that create spectra when you look through
them.  In fact, they are grating replicas, --different patterns give, of
course, different results.  We've also used hemocytometers and stage
micrometer slides.  One more interesting source are the special gratings
used in "Chroma Depth" 3D system.  These last are gratings that shift the
spectrum either to the left or the right, depending on the eye.  As a
result, reds are moved forward, and blues to the rear.  This is great for
landscapes.  The bfp images from these gratings show spots on one side or
the other of the primary.

One more thought.  A standard EM standardization is a grating replica
mounted on an em grid.  You might see if someone like Pella can sell you
either fragments for you to mount, or even a larger sheet.

Joel


On Thu, Aug 11, 2011 at 11:24 AM, Glen MacDonald
<[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> EM grids will work, especially line grids. Maintaining alignment while
> mounting on a slide is a challenge.
>
> Regards,
> Glen
>
>
> On Aug 11, 2011, at 5:48 AM, Phillips, Thomas E. wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Does anyone know of a USA supplier for a similar diffraction grating
> slide?
> >
> > Thomas E. Phillips, Ph.D
> > Professor of Biological Sciences
> > Director, Molecular Cytology Core
> > 2 Tucker Hall
> > University of Missouri
> > Columbia, MO 65211-7400
> > 573-882-4712 (office)
> > 573-882-0123 (fax)
> > [hidden email]
> >
> > http://www.biology.missouri.edu/faculty/phillips.html
> > http://www.biotech.missouri.edu/mcc/
> >
> >
> > -----Original Message-----
> > From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Zac Arrac Atelaz
> > Sent: Wednesday, August 10, 2011 11:38 PM
> > To: [hidden email]
> > Subject: Re: Source for diffraction grating slide
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> >
> > There are several resources, this is only one of them
> >
> > http://www.scientrific.com.au/product.php?p=774
> >
> >
> >> Date: Wed, 10 Aug 2011 21:29:25 -0700
> >> From: [hidden email]
> >> Subject: Source for diffraction grating slide
> >> To: [hidden email]
> >>
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> *****
> >>
> >> Does anyone know of a commercial (or other) source for a microscope
> slide with various diffraction grating patterns? I want to use this for
> teaching purposes (demonstrate diffraction in the back focal plane), and
> ideally the slide would have multiple patterns (spacings from about 0.5 um
> to 5 um, various orientations, squares, etc..). Basically, I am looking for
> a slide as was used in the Abbe diffraction kit. All pointers will be
> appreciated!
> >>
> >> Thanks for your help,
> >>
> >> Nico
> >
>



--


Joel B. Sheffield, Ph.D
Department of Biology
Temple University
Philadelphia, PA 19122
Voice: 215 204 8839
e-mail: [hidden email]
URL:  http://astro.temple.edu/~jbs
Z.J. Zhang Z.J. Zhang
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Source for diffraction grating slide

In reply to this post by Phillips, Thomas E.
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

This might works for you - http://www.rainbowsymphonystore.com/scienanded1.html. They have different lines per inch, but many of them require a minimum order.



Zhaojie Zhang, Ph. D.
Director, Jenkins Microscopy Facility
University of Wyoming
Laramie, WY 82071
PHONE: 307-766-3038
FAX: 307-766-5625



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Phillips, Thomas E.
Sent: Thursday, August 11, 2011 6:49 AM
To: [hidden email]
Subject: Re: Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Does anyone know of a USA supplier for a similar diffraction grating slide?

Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
[hidden email]

http://www.biology.missouri.edu/faculty/phillips.html
http://www.biotech.missouri.edu/mcc/


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Zac Arrac Atelaz
Sent: Wednesday, August 10, 2011 11:38 PM
To: [hidden email]
Subject: Re: Source for diffraction grating slide

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


There are several resources, this is only one of them
 
http://www.scientrific.com.au/product.php?p=774
 

> Date: Wed, 10 Aug 2011 21:29:25 -0700
> From: [hidden email]
> Subject: Source for diffraction grating slide
> To: [hidden email]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Does anyone know of a commercial (or other) source for a microscope slide with various diffraction grating patterns? I want to use this for teaching purposes (demonstrate diffraction in the back focal plane), and ideally the slide would have multiple patterns (spacings from about 0.5 um to 5 um, various orientations, squares, etc..). Basically, I am looking for a slide as was used in the Abbe diffraction kit. All pointers will be appreciated!
>
> Thanks for your help,
>
> Nico