To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear confocal list, I would have your advice about this ! Is-it possible to do spectral deconvolution on a dapi staining ? Regards Pascal |
|
Mark Cannell-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Yes, but YMMV. Regards On 29/06/2015, at 11:32 am, Pascal Weber <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear confocal list, > > I would have your advice about this ! > Is-it possible to do spectral deconvolution on a dapi staining ? > > Regards > > Pascal Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
|
|
In reply to this post by weber
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Sorry, What means YMMV ? |
|
Zdenek Svindrych-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Your mileage may vary... z. ---------- Původní zpráva ---------- Od: Pascal Weber <[hidden email]> Komu: [hidden email] Datum: 29. 6. 2015 7:51:24 Předmět: Re: Spectral deconvolution on DAPI "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Sorry, What means YMMV ?" |
|
Tim Feinstein |
|
|
In reply to this post by weber
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Yes and we have had good results doing it, but you have two potential problems. DAPI has an extremely broad emission spectrum, so it could get in the way of your 488 label and possibly the yellow-red as well if they are not very bright. Use a spectrum viewer like the ones from MolProbes/LifeTech/ThermoFisher, Chroma, Zeiss Light Lab etc to check how well your other emission peaks stand apart from DAPI. The other problem is kind of the inverse of the first one - 405 nm lasers only just touch the edge of DAPI's excitation spectrum (again check those spectra viewers), so imaging DAPI on a confocal can be quite inefficient compared with a widefield scope. As a result DAPI intensity can be hard to control between tissue types, preparations and even one experiment to the next. This is always the case but especially problematic when trying to balance your fluors for a spectral experiment. As a result we try to use far red stains like To-Pro-3 instead when staining DNA for confocal. In our experience To-Pro-3 is VERY bright in many tissues, so you have to titrate it down well below the recommended dosage to balance its intensity with other fluors. All the best, Tim Timothy Feinstein, Ph.D. Research Scientist University of Pittsburgh Department of Developmental Biology On 6/29/15, 6:32 AM, "Pascal Weber" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Dear confocal list, > >I would have your advice about this ! >Is-it possible to do spectral deconvolution on a dapi staining ? > >Regards > >Pascal |
|
|
In reply to this post by weber
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Infact i have lot of different bacteria and they will not have the same spectra. I want to discriminate them only this method. I have not different staining just this one but i'm not sure that is the wright way to proceed. My goal is to reach a good an simple method to verify if the sample is pure or not. Before to do DNA sequencing. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I think I missed a reply along the way but let me see if I understand what you want to do. You have several samples of dapi-stained bacteria and you want to detect differences in staining between the samples as a way to decide whether the bacteria are phylogenetically different? In other words, if the staining of A is different from B, then you want to sequence the two organisms, but if the staining between A and C is similar, you will consider them to be similar and not worth sequencing? On 6/29/15, 11:01 AM, "Pascal Weber" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Infact i have lot of different bacteria and they will not have the same >spectra. >I want to discriminate them only this method. I have not different >staining just >this one but i'm not sure that is the wright way to proceed. My goal is >to reach >a good an simple method to verify if the sample is pure or not. Before to >do >DNA sequencing. |
|
|
In reply to this post by weber
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Yes that's it ! |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Here is the opinion of a microbiologist. The signal you get from any stain other than a 16S rDNA FISH probe (or maybe a very specific antibody) is not valuable as a taxonomic indicator. I have not done this experiment, but I would expect that two completely different bacterial species could have similar DAPI signals, just as two bacteria with similar DAPI signals can be very different organisms. I think DAPI doesn¹t tell you anything about DNA other than where it is in a cell and how much of it is there (intensity). It may be theoretically possible to distinguish "high GC" organisms from others, but this parameter alone is not a very good indicator of what bacterium you have. I assume you have plenty of sample material, and that these bacteria can be grown again for nucleic acid extraction/analysis. If that is the case, and you do not have hundreds of samples to process, you are much better off simply sequencing the 16s ribosomal gene in each sample. Extraction is with a kit, you amplify the gene using a standard primer set to produce amplicons of around 500 bases. Sequence these, submit them to BLAST and you will get a reasonable idea of what you have. This procedure is being done by high-school students nowadays. Maybe I am missing something here. Happy to discuss in greater detail if you¹d like to take the discussion off the listserv. Robert J. Palmer Jr., PhD Bldg 30, Room 207 NIDCR/NIH Bethesda MD 20892 301-594-0025 On 6/29/15, 12:42 PM, "Pascal Weber" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Yes that's it ! |
|
Craig Brideau |
|
|
In reply to this post by weber
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If your spectral acquisition is very broad, you will catch most/all of the DAPI curve. The more of the curve shape you can catch, the better the unmix will be. As mentioned by Tim, 405nm is non-optimal for excitation, but it tends to be the most common laser for confocal of blue dyes. In my own experience it works fairly well, just watch out for scatter from the laser contaminating your spectra. Craig Brideau On Mon, Jun 29, 2015 at 10:42 AM, Pascal Weber <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Yes that's it ! > |
|
Tim Feinstein |
|
|
In reply to this post by weber
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Pascal, If these are photosynthetic bacteria then their autofluorescence spectra may help identify them. In a past life I tried that with dinoflagellates in a spectrofluorimeter. I do not think DAPI will be useful unless you have already ruled out every other possible way of classifying them. Of course it depends on your experimental design. If every dish can only be one out of four or five possible strains then you might get away with identifying a distinct spectrum for each strain, as long as you show that spectrum variabilities do not also come from life stage, nutrient conditions, stress level, ambient environment etc. If the list of possible strains is long or open-ended then you should try something else. Best, Tim Timothy Feinstein, Ph.D. Research Scientist University of Pittsburgh Department of Developmental Biology On 6/29/15, 11:01 AM, "Pascal Weber" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Infact i have lot of different bacteria and they will not have the same >spectra. >I want to discriminate them only this method. I have not different >staining just >this one but i'm not sure that is the wright way to proceed. My goal is >to reach >a good an simple method to verify if the sample is pure or not. Before to >do >DNA sequencing. |
|
Jurek Dobrucki |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Pascal, another issue to keep in mind when detecting DAPI, especially if the bacteria are also stained with some green dye or GFP, is photoconversion of this dye to a green emitting form (Zurek-Biesiada et al. Photoconversion of Hoechst 33258, DAPI, and Vybrant DyeCycle... Cytometry A. 2013 May;83(5):441-51). Hope this helps, Jurek Dobrucki On Mon, Jun 29, 2015 at 8:33 PM, Feinstein, Timothy N <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Pascal, > > If these are photosynthetic bacteria then their autofluorescence spectra > may help identify them. In a past life I tried that with dinoflagellates > in a spectrofluorimeter. I do not think DAPI will be useful unless you > have already ruled out every other possible way of classifying them. > > Of course it depends on your experimental design. If every dish can only > be one out of four or five possible strains then you might get away with > identifying a distinct spectrum for each strain, as long as you show that > spectrum variabilities do not also come from life stage, nutrient > conditions, stress level, ambient environment etc. If the list of > possible strains is long or open-ended then you should try something else. > > > Best, > > > Tim > > Timothy Feinstein, Ph.D. > Research Scientist > University of Pittsburgh Department of Developmental Biology > > > > > > On 6/29/15, 11:01 AM, "Pascal Weber" <[hidden email]> wrote: > >>***** >>To join, leave or search the confocal microscopy listserv, go to: >>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>Post images on http://www.imgur.com and include the link in your posting. >>***** >> >>Infact i have lot of different bacteria and they will not have the same >>spectra. >>I want to discriminate them only this method. I have not different >>staining just >>this one but i'm not sure that is the wright way to proceed. My goal is >>to reach >>a good an simple method to verify if the sample is pure or not. Before to >>do >>DNA sequencing. |