Giang, William |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear list, Has anyone compared the emitted light gathering capabilities of a Yokogawa CSU-X1 vs a CrestOptics X-Light V3? Neither have a microlens array on the emission path, and both units would have 50um diameter pinholes with 250um spacing. Techs from a large microscopy company said the CSU-X1 is superior for fast live-cell imaging in cell culture, but it's not clear to me why. Is Yokogawa able to get a higher density of pinholes on their disk? Am I misunderstanding how pinhole spacing works? x-posted to Microforum at https://forum.microlist.org/t/yokogawa-csu-x1-vs-crest-x-light-v3-light-output/1240 Thanks in advance, Will William Giang | Imaging Specialist Whitehead Building, Room 465 615 Michael Street, Atlanta, GA 30322 [hidden email]<mailto:[hidden email]> ici.emory.edu<http://www.cores.emory.edu/ici/> (304) 853-DATA Remember to acknowledge ICI in your publications.<https://www.cores.emory.edu/ici/Acknowledgements/index.html> ________________________________ This e-mail message (including any attachments) is for the sole use of the intended recipient(s) and may contain confidential and privileged information. If the reader of this message is not the intended recipient, you are hereby notified that any dissemination, distribution or copying of this message (including any attachments) is strictly prohibited. If you have received this message in error, please contact the sender by reply e-mail message and destroy all copies of the original message (including attachments). |
Dr. K N Ganesh |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi 1. CSU X1 being dual disc spinning system has micro lenses on excitation side that enhances the excitation efficiency 2.CSUX1 is laser based has an ability to synchronised disc rotation with respect exposure time of camera . (1800 rpm or 5000 rpm ) Whereas crest offers simple single Nipkow Disc without micro lenses and more importantly it uses bright LED light if I am not wrong. Naturally CSUX1 should perform better. Regards Ganesh Sent from my iPhone > On 19-Dec-2020, at 05:34, Giang, William <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear list, > > Has anyone compared the emitted light gathering capabilities of a Yokogawa CSU-X1 vs a CrestOptics X-Light V3? > Neither have a microlens array on the emission path, and both units would have 50um diameter pinholes with 250um spacing. > > Techs from a large microscopy company said the CSU-X1 is superior for fast live-cell imaging in cell culture, but it's not clear to me why. Is Yokogawa able to get a higher density of pinholes on their disk? Am I misunderstanding how pinhole spacing works? > > x-posted to Microforum at https://forum.microlist.org/t/yokogawa-csu-x1-vs-crest-x-light-v3-light-output/1240 > > Thanks in advance, > Will > > > William Giang | Imaging Specialist > > Whitehead Building, Room 465 > 615 Michael Street, Atlanta, GA 30322 > [hidden email]<mailto:[hidden email]> > ici.emory.edu<http://www.cores.emory.edu/ici/> > (304) 853-DATA > > > Remember to acknowledge ICI in your publications.<https://www.cores.emory.edu/ici/Acknowledgements/index.html> > > > > ________________________________ > > This e-mail message (including any attachments) is for the sole use of > the intended recipient(s) and may contain confidential and privileged > information. If the reader of this message is not the intended > recipient, you are hereby notified that any dissemination, distribution > or copying of this message (including any attachments) is strictly > prohibited. > > If you have received this message in error, please contact > the sender by reply e-mail message and destroy all copies of the > original message (including attachments). |
Alessandra Scarpellini |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** **COMMERCIAL REPLY** Dear Confocal List, I'd like to point out a few key points on the CrestOptics solutions, especially the X-light V3: - Yes, there are no micro-lenses to focus excitation light, however this is compensated by the use of more powerful laser sources (cost-effective multi-mode sources) to achieve comparable excitation power on the sample - On The X-light V3 micro-lenses are employed to achieve homogeneous illumination, so that quantitative imaging can be done on the full field of view (FOV). This FOV measures 25mm, meaning a much faster data collection (more cells in a single FOV) - 2 cameras with FOV 25mm can be used simultaneously for even faster data collection - CrestOptics disk spins at 15K RPM, meaning acquisition speed >1kHz without artefacts - Disk pattern can be customized for the best ratio of confocality and throughput depending on the sample These and many more info can be found on our website: www.crestoptics.com Best regards, Alessandra Alessandra Scarpellini, PhD | Head of Sales and Marketing Via di Torre Rossa, 66 00165 ROME - Italy M: +39 3357151933 P: +39 0661660508 CONFIDENTIALITY & PRIVACY NOTICE The information in this email is confidential and protected by civil and criminal law. Unauthorized storing thereof is unlawful. Please notify to the sender of any misdelivered email and therefore destroy it. -----Messaggio originale----- Da: Confocal Microscopy List <[hidden email]> Per conto di Ganesh Kadasoor Inviato: sabato 19 dicembre 2020 11:29 A: [hidden email] Oggetto: Re: Spinning disk comparison - CSU-X1 vs CrestOptics X-Light V3 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi 1. CSU X1 being dual disc spinning system has micro lenses on excitation side that enhances the excitation efficiency 2.CSUX1 is laser based has an ability to synchronised disc rotation with respect exposure time of camera . (1800 rpm or 5000 rpm ) Whereas crest offers simple single Nipkow Disc without micro lenses and more importantly it uses bright LED light if I am not wrong. Naturally CSUX1 should perform better. Regards Ganesh Sent from my iPhone > On 19-Dec-2020, at 05:34, Giang, William <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear list, > > Has anyone compared the emitted light gathering capabilities of a Yokogawa CSU-X1 vs a CrestOptics X-Light V3? > Neither have a microlens array on the emission path, and both units would have 50um diameter pinholes with 250um spacing. > > Techs from a large microscopy company said the CSU-X1 is superior for fast live-cell imaging in cell culture, but it's not clear to me why. Is Yokogawa able to get a higher density of pinholes on their disk? Am I misunderstanding how pinhole spacing works? > > x-posted to Microforum at > https://forum.microlist.org/t/yokogawa-csu-x1-vs-crest-x-light-v3-ligh > t-output/1240 > > Thanks in advance, > Will > > > William Giang | Imaging Specialist > > Whitehead Building, Room 465 > 615 Michael Street, Atlanta, GA 30322 > [hidden email]<mailto:[hidden email]> > ici.emory.edu<http://www.cores.emory.edu/ici/> > (304) 853-DATA > > > Remember to acknowledge ICI in your > publications.<https://www.cores.emory.edu/ici/Acknowledgements/index.h > tml> > > > > ________________________________ > > This e-mail message (including any attachments) is for the sole use of > the intended recipient(s) and may contain confidential and privileged > information. If the reader of this message is not the intended > recipient, you are hereby notified that any dissemination, > distribution or copying of this message (including any attachments) is > strictly prohibited. > > If you have received this message in error, please contact the sender > by reply e-mail message and destroy all copies of the original message > (including attachments). |
Knecht, David |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I have never understood the rationale for micro-lenses in a spinning disk system. These systems are aimed at live cell imaging. I have never used anywhere near full laser power on either our Yokogawa system or our Crest system. Cells do not tolerate high excitation light in general and so we always work with the minimum light level possible and that is never near the limit of laser power. So I don’t see the rationale for the added cost of a micro-lens system, unless you have a really low power laser source. I would be curioius to know if anyone has a different opinion. As to the question that was asked, I have not directly compared the two, but not sure how I would since they are on different microscopes, objectives, cameras, laser launches, etc. I can say that both work to give nice images of live cells. I like the ability of the Crest to function in confocal or non-confocal mode which our Yokogawa system cannot do. Dave On Dec 19, 2020, at 8:16 AM, Alessandra Scarpellini <[hidden email]<mailto:[hidden email]>> wrote: *Message sent from a system outside of UConn.* ***** To join, leave or search the confocal microscopy listserv, go to: https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353507262%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=N%2F%2FWVlqzR9RuPU%2FwxmrzknJ%2BYcBNfQO4u5yRteBJ%2BXs%3D&reserved=0 Post images on https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=x0G1MzUn7idEPdcCf7Xfo1fr8fTjUHjQrMp9YuZdxdY%3D&reserved=0 and include the link in your posting. ***** **COMMERCIAL REPLY** Dear Confocal List, I'd like to point out a few key points on the CrestOptics solutions, especially the X-light V3: - Yes, there are no micro-lenses to focus excitation light, however this is compensated by the use of more powerful laser sources (cost-effective multi-mode sources) to achieve comparable excitation power on the sample - On The X-light V3 micro-lenses are employed to achieve homogeneous illumination, so that quantitative imaging can be done on the full field of view (FOV). This FOV measures 25mm, meaning a much faster data collection (more cells in a single FOV) - 2 cameras with FOV 25mm can be used simultaneously for even faster data collection - CrestOptics disk spins at 15K RPM, meaning acquisition speed >1kHz without artefacts - Disk pattern can be customized for the best ratio of confocality and throughput depending on the sample These and many more info can be found on our website: https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.crestoptics.com%2F&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=0D05rdwBnXFYFyNOFfiJEvODpD4QF1vJksr%2B%2FDPGuhU%3D&reserved=0 Best regards, Alessandra Alessandra Scarpellini, PhD | Head of Sales and Marketing Via di Torre Rossa, 66 00165 ROME - Italy M: +39 3357151933 P: +39 0661660508 CONFIDENTIALITY & PRIVACY NOTICE The information in this email is confidential and protected by civil and criminal law. Unauthorized storing thereof is unlawful. Please notify to the sender of any misdelivered email and therefore destroy it. -----Messaggio originale----- Da: Confocal Microscopy List <[hidden email]> Per conto di Ganesh Kadasoor Inviato: sabato 19 dicembre 2020 11:29 A: [hidden email] Oggetto: Re: Spinning disk comparison - CSU-X1 vs CrestOptics X-Light V3 ***** To join, leave or search the confocal microscopy listserv, go to: https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=ZRnA7nbJYpXVVobEEPU3AvsXJWk4XLtMmsslRfD%2Fbtk%3D&reserved=0 Post images on https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=x0G1MzUn7idEPdcCf7Xfo1fr8fTjUHjQrMp9YuZdxdY%3D&reserved=0 and include the link in your posting. ***** Hi 1. CSU X1 being dual disc spinning system has micro lenses on excitation side that enhances the excitation efficiency 2.CSUX1 is laser based has an ability to synchronised disc rotation with respect exposure time of camera . (1800 rpm or 5000 rpm ) Whereas crest offers simple single Nipkow Disc without micro lenses and more importantly it uses bright LED light if I am not wrong. Naturally CSUX1 should perform better. Regards Ganesh Sent from my iPhone On 19-Dec-2020, at 05:34, Giang, William <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=ZRnA7nbJYpXVVobEEPU3AvsXJWk4XLtMmsslRfD%2Fbtk%3D&reserved=0 Post images on https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=x0G1MzUn7idEPdcCf7Xfo1fr8fTjUHjQrMp9YuZdxdY%3D&reserved=0 and include the link in your posting. ***** Dear list, Has anyone compared the emitted light gathering capabilities of a Yokogawa CSU-X1 vs a CrestOptics X-Light V3? Neither have a microlens array on the emission path, and both units would have 50um diameter pinholes with 250um spacing. Techs from a large microscopy company said the CSU-X1 is superior for fast live-cell imaging in cell culture, but it's not clear to me why. Is Yokogawa able to get a higher density of pinholes on their disk? Am I misunderstanding how pinhole spacing works? x-posted to Microforum at https://nam10.safelinks.protection.outlook.com/?url=https%3A%2F%2Fforum.microlist.org%2Ft%2Fyokogawa-csu-x1-vs-crest-x-light-v3-ligh&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=aeFi3tg5EB3Je7jnM7rA1BRZGE3ytQyQPqQFwJ9WnIs%3D&reserved=0 t-output/1240 Thanks in advance, Will William Giang | Imaging Specialist Whitehead Building, Room 465 615 Michael Street, Atlanta, GA 30322 [hidden email]<mailto:[hidden email]> ici.emory.edu<https://nam10.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.cores.emory.edu%2Fici%2F&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=TnrIfWzf0sUe4qkgoX%2BbkMpwpYBekD5ddgF1NUo8LaE%3D&reserved=0> (304) 853-DATA Remember to acknowledge ICI in your publications.<https://nam10.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.cores.emory.edu%2Fici%2FAcknowledgements%2Findex.h&data=04%7C01%7Cdavid.knecht%40UCONN.EDU%7Cf930fd5e77584f2e2ce408d8a421ca48%7C17f1a87e2a254eaab9df9d439034b080%7C0%7C0%7C637439812353517251%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000&sdata=J8p%2BGHL%2Bnkf4h8LQ4f%2BPYKABZ4lOaH7cEFFeYNLYUyY%3D&reserved=0 tml> ________________________________ This e-mail message (including any attachments) is for the sole use of the intended recipient(s) and may contain confidential and privileged information. If the reader of this message is not the intended recipient, you are hereby notified that any dissemination, distribution or copying of this message (including any attachments) is strictly prohibited. If you have received this message in error, please contact the sender by reply e-mail message and destroy all copies of the original message (including attachments). Dr. David Knecht Professor of Molecular and Cell Biology University of Connecticut 91 N. Eagleville Rd. Storrs, CT 06269-3125 |
Zdenek Svindrych-2 |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi David, the rationale is simple. The purpose of the spinning disk is to attenuate out-of-focus light. If the CSU-X1 attenuates this unwanted emission by a factor of 30 (estimate, I didn't do the math), it will also attenuate the laser excitation by the same factor. With a strong laser that shouldn't be a problem, right? Well, 1:30 pinhole crosstalk may not be enough. If you bleach a (small, let's say 1 um by 1 um) cone of light in a thick fluorescent layer you won't be able to see it with the X1, but you can see it clearly with a point scanner. That's why CSU-W1 or Dragonfly have sparser pinholes and higher attenuation. Without the microlenses you would end up with very little excitation even with powerful lasers. And as a matter of fact, we use quite strong excitation (30% with a 150 mW laser) to capture a z-stack quickly and do longer pauses between z-stacks - this helps with motion artifacts and allows for deconvolution... Back to the original question - with identical pinhole pattern and overall optical configuration (minus the excitation pinholes), the detection efficiency should be the same. The ultimate limiting factor will be how much light can the disk handle, how you deal with the reflected laser light from the disk (essentially 100%), and the autofluorescence of any element that is common to the strong excitation (before the disk) and emission paths... Best, zdenek |
Craig Brideau |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** As Zdenek says, you can need quite substantial laser powers under some scenarios without the microlenses concentrating the energy. The cost for high-power versions of certain lasers can also be quite high, if the required energy density is available at all, so on occasion a weaker laser is your only option. A lower power laser also requires less heat sinking and electrical current, which simplifies the overall design and in some cases can lead to longer laser life if thermal design is still taken seriously by the manufacturer. Craig On Sun, Dec 20, 2020 at 11:28 AM Zdenek Svindrych <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi David, > > the rationale is simple. The purpose of the spinning disk is to attenuate > out-of-focus light. If the CSU-X1 attenuates this unwanted emission by a > factor of 30 (estimate, I didn't do the math), it will also attenuate the > laser excitation by the same factor. With a strong laser that shouldn't be > a problem, right? > Well, 1:30 pinhole crosstalk may not be enough. If you bleach a (small, > let's say 1 um by 1 um) cone of light in a thick fluorescent layer you > won't be able to see it with the X1, but you can see it clearly with a > point scanner. That's why CSU-W1 or Dragonfly have sparser pinholes and > higher attenuation. Without the microlenses you would end up with very > little excitation even with powerful lasers. > And as a matter of fact, we use quite strong excitation (30% with a 150 mW > laser) to capture a z-stack quickly and do longer pauses between z-stacks - > this helps with motion artifacts and allows for deconvolution... > > Back to the original question - with identical pinhole pattern and > overall optical configuration (minus the excitation pinholes), the > detection efficiency should be the same. The ultimate limiting factor will > be how much light can the disk handle, how you deal with the reflected > laser light from the disk (essentially 100%), and the autofluorescence of > any element that is common to the strong excitation (before the disk) and > emission paths... > > Best, zdenek > |
Giang, William |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Ganesh - #2 - I've encountered no issues with hardware triggering on a Nikon scope with the X-Light V2 LFOV, and the light engine was the Lumencor CELESTA with 1W laser lines. Thanks Alessandra for covering #1. Dave - I concur. I've never wanted/needed more power in the context of live cell imaging, and I've been pleased with the V2 LFOV. If you had a power meter that could ensure the same power density at the sample plane, I'd love to see what the difference between your systems would be. But it does sound like there are too many things that'd have to be estimated to make the comparison conclusive. Zdenek - Yes, I've heard many people suggesting at least 100mW lasers for the W1. Some have also suggested (for when you're fine with diffraction limited resolution) using the SoRa microlensed emission disk with 1x mag to collect roughly 3x the light vs the standard 50um W1 disk. But I'd need more than a couple $600 relief checks for the SoRa disk. For the X-Light V2 LFOV, I believe the emission filters have high (6+) OD to block (reflected) laser light from making it to the detector. I never really considered if this reflected laser light would excite out-of-focus regions of the sample--and either increase background or induce unnecessary photodamage--but maybe there's a little bit of that going on? I suppose the V3's software controlled square iris also partially sidesteps the issue of cranking up the laser power. Craig - As Alessandra said, they use multimode fiber coupled light engines (Lumencor CELESTA or 89 North LDI) which are cheaper than the Yokogawa SDs which require single mode optical fiber input. I'm still shocked that Andrew Seeber got "an x-light V3 with the LDI from 89North, two Hamamatsu Fusion cameras, all the optics I need, a VisiView license, a second hand Zeiss Observer 7 body and a table for ~$200k. (source at https://forum.microlist.org/t/crest-x-light-v3-spinning-disk-confocal-reviews/565/15). Really not sure how he swung that. I promise this conversation was started in good faith and that I'm not a shill! Thanks all for your responses. -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Craig Brideau Sent: Sunday, December 20, 2020 5:49 PM To: [hidden email] Subject: [External] Re: Spinning disk comparison - CSU-X1 vs CrestOptics X-Light V3 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** As Zdenek says, you can need quite substantial laser powers under some scenarios without the microlenses concentrating the energy. The cost for high-power versions of certain lasers can also be quite high, if the required energy density is available at all, so on occasion a weaker laser is your only option. A lower power laser also requires less heat sinking and electrical current, which simplifies the overall design and in some cases can lead to longer laser life if thermal design is still taken seriously by the manufacturer. Craig On Sun, Dec 20, 2020 at 11:28 AM Zdenek Svindrych <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi David, > > the rationale is simple. The purpose of the spinning disk is to > attenuate out-of-focus light. If the CSU-X1 attenuates this unwanted > emission by a factor of 30 (estimate, I didn't do the math), it will > also attenuate the laser excitation by the same factor. With a strong > laser that shouldn't be a problem, right? > Well, 1:30 pinhole crosstalk may not be enough. If you bleach a > (small, let's say 1 um by 1 um) cone of light in a thick fluorescent > layer you won't be able to see it with the X1, but you can see it > clearly with a point scanner. That's why CSU-W1 or Dragonfly have > sparser pinholes and higher attenuation. Without the microlenses you > would end up with very little excitation even with powerful lasers. > And as a matter of fact, we use quite strong excitation (30% with a > 150 mW > laser) to capture a z-stack quickly and do longer pauses between > z-stacks - this helps with motion artifacts and allows for deconvolution... > > Back to the original question - with identical pinhole pattern and > overall optical configuration (minus the excitation pinholes), the > detection efficiency should be the same. The ultimate limiting factor > will be how much light can the disk handle, how you deal with the > reflected laser light from the disk (essentially 100%), and the > autofluorescence of any element that is common to the strong > excitation (before the disk) and emission paths... > > Best, zdenek > ________________________________ This e-mail message (including any attachments) is for the sole use of the intended recipient(s) and may contain confidential and privileged information. If the reader of this message is not the intended recipient, you are hereby notified that any dissemination, distribution or copying of this message (including any attachments) is strictly prohibited. If you have received this message in error, please contact the sender by reply e-mail message and destroy all copies of the original message (including attachments). |
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, One limitation we have run into with the CSU (we have a W1) is the availability of suitable short-pass dichroic beamsplitters with sufficient flatness (I guess the option to use standard long-pass dichroics would be a benefit of the Crest system). We are working on a custom configuration of the CSU where we are trying to find short-pass dichroics with edge cut-offs at 500, 550, or 600 nm. Does anyone know of suppliers who might be able to meet this need? Thanks! Best regards, Silas On 12/20/2020 4:49 PM, Craig Brideau wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > As Zdenek says, you can need quite substantial laser powers under some > scenarios without the microlenses concentrating the energy. The cost for > high-power versions of certain lasers can also be quite high, if the > required energy density is available at all, so on occasion a weaker laser > is your only option. A lower power laser also requires less heat sinking > and electrical current, which simplifies the overall design and in some > cases can lead to longer laser life if thermal design is still taken > seriously by the manufacturer. > > Craig > > On Sun, Dec 20, 2020 at 11:28 AM Zdenek Svindrych <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi David, >> >> the rationale is simple. The purpose of the spinning disk is to attenuate >> out-of-focus light. If the CSU-X1 attenuates this unwanted emission by a >> factor of 30 (estimate, I didn't do the math), it will also attenuate the >> laser excitation by the same factor. With a strong laser that shouldn't be >> a problem, right? >> Well, 1:30 pinhole crosstalk may not be enough. If you bleach a (small, >> let's say 1 um by 1 um) cone of light in a thick fluorescent layer you >> won't be able to see it with the X1, but you can see it clearly with a >> point scanner. That's why CSU-W1 or Dragonfly have sparser pinholes and >> higher attenuation. Without the microlenses you would end up with very >> little excitation even with powerful lasers. >> And as a matter of fact, we use quite strong excitation (30% with a 150 mW >> laser) to capture a z-stack quickly and do longer pauses between z-stacks - >> this helps with motion artifacts and allows for deconvolution... >> >> Back to the original question - with identical pinhole pattern and >> overall optical configuration (minus the excitation pinholes), the >> detection efficiency should be the same. The ultimate limiting factor will >> be how much light can the disk handle, how you deal with the reflected >> laser light from the disk (essentially 100%), and the autofluorescence of >> any element that is common to the strong excitation (before the disk) and >> emission paths... >> >> Best, zdenek >> Silas J. Leavesley, Ph.D. Professor Department of Chemical and Biomolecular Engineering Department of Pharmacology Center for Lung Biology University of South Alabama 150 Student Services Drive, SH4129 Mobile, AL 36688 ph: (251)-460-6160 fax: (251)-461-1485 web: http://www.southalabama.edu/centers/bioimaging <http://www.southalabama.edu/centers/bioimaging> <http://www.usahealthsystem.com/clb> |
Mark Cannell-2 |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Silas Chroma can make such filter with very high performance on flat substrates but for one offs it' an expensive route... Can you use a suitable wide band pass filter such as their https://www.chroma.com/products/parts/et510-80m which is a short pass at 475-500 ? Note that the filter will tune to a shorter wavelength if used at 45 degrees and Chroma support will be able to advise the best choice from their existing filters (No commercial interest -but tell them I sent you __). HTH Mark Mark B. Cannell. Ph.D. FRSNZ FISHR Department of Physiology, Pharmacology & Neuroscience School of Medical Sciences University Walk Bristol BS8 1TD On 7/01/21, 10:31 PM, "Confocal Microscopy List on behalf of Silas Leavesley" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, One limitation we have run into with the CSU (we have a W1) is the availability of suitable short-pass dichroic beamsplitters with sufficient flatness (I guess the option to use standard long-pass dichroics would be a benefit of the Crest system). We are working on a custom configuration of the CSU where we are trying to find short-pass dichroics with edge cut-offs at 500, 550, or 600 nm. Does anyone know of suppliers who might be able to meet this need? Thanks! Best regards, Silas On 12/20/2020 4:49 PM, Craig Brideau wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > As Zdenek says, you can need quite substantial laser powers under some > scenarios without the microlenses concentrating the energy. The cost for > high-power versions of certain lasers can also be quite high, if the > required energy density is available at all, so on occasion a weaker laser > is your only option. A lower power laser also requires less heat sinking > and electrical current, which simplifies the overall design and in some > cases can lead to longer laser life if thermal design is still taken > seriously by the manufacturer. > > Craig > > On Sun, Dec 20, 2020 at 11:28 AM Zdenek Svindrych <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi David, >> >> the rationale is simple. The purpose of the spinning disk is to attenuate >> out-of-focus light. If the CSU-X1 attenuates this unwanted emission by a >> factor of 30 (estimate, I didn't do the math), it will also attenuate the >> laser excitation by the same factor. With a strong laser that shouldn't be >> a problem, right? >> Well, 1:30 pinhole crosstalk may not be enough. If you bleach a (small, >> let's say 1 um by 1 um) cone of light in a thick fluorescent layer you >> won't be able to see it with the X1, but you can see it clearly with a >> point scanner. That's why CSU-W1 or Dragonfly have sparser pinholes and >> higher attenuation. Without the microlenses you would end up with very >> little excitation even with powerful lasers. >> And as a matter of fact, we use quite strong excitation (30% with a 150 mW >> laser) to capture a z-stack quickly and do longer pauses between z-stacks - >> this helps with motion artifacts and allows for deconvolution... >> >> Back to the original question - with identical pinhole pattern and >> overall optical configuration (minus the excitation pinholes), the >> detection efficiency should be the same. The ultimate limiting factor will >> be how much light can the disk handle, how you deal with the reflected >> laser light from the disk (essentially 100%), and the autofluorescence of >> any element that is common to the strong excitation (before the disk) and >> emission paths... >> >> Best, zdenek >> -- Silas J. Leavesley, Ph.D. Professor Department of Chemical and Biomolecular Engineering Department of Pharmacology Center for Lung Biology University of South Alabama 150 Student Services Drive, SH4129 Mobile, AL 36688 ph: (251)-460-6160 fax: (251)-461-1485 web: http://www.southalabama.edu/centers/bioimaging <http://www.southalabama.edu/centers/bioimaging> <http://www.usahealthsystem.com/clb> |
Free forum by Nabble | Edit this page |