To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello everyone We have an encoded (motorized) stage that drifts around 0.5 um (500nm) per hour to mostly but not always one direction. Putting an enclosed box to surround the microscope seems did not help. Is this something unavoidable for a motorized stage? Is using fiducial marker the only solution if experiment requires high accuracy of XY position over long period of time? I would appreciate for any thought and advice. Jianbo Chen HIV Drug Resistance Program National Cancer Institute-Frederick Frederick MD 21702 [hidden email] 301-846-1841 https://ccrod.cancer.gov/confluence/display/CCRSSSCArchive/Home |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Jianbo, Although it is difficult to remotely solve your problem, but here are some ideas (based on our experiences with motorized stages). The first question is whether you really see a drift? Do you stay really at one location? Or do you acquire images at several locations and return to them? In this case what you see is just the reproducibility limitations of your system. Assuming that you see a "real drift": you stay at one location and the stage continuously drifts, there are a few things I could suggest to check. 1) Check the electrical connections of your stage and also whether all settings in your control software are correct. 2) If you have a control joystick - check it (exchange it) - as a defect (dying) joystick can easily cause such problems. 3) If you have linear encoders - check whether they are still clean. Dirty encoders can also lead to strange effects 4) a drift a 500 nm/hours is certainly not negligible but there are software tools to get rid of such movements (even without the use of any special markers). Cheers Gabor > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello everyone > We have an encoded (motorized) stage that drifts around 0.5 um (500nm) per hour to mostly but not always one direction. Putting an enclosed box to surround the microscope seems did not help. Is this something unavoidable for a motorized stage? Is using fiducial marker the only solution if experiment requires high accuracy of XY position over long period of time? I would appreciate for any thought and advice. > > Jianbo Chen > HIV Drug Resistance Program > National Cancer Institute-Frederick > Frederick MD 21702 > > [hidden email] > 301-846-1841 > > https://ccrod.cancer.gov/confluence/display/CCRSSSCArchive/Home -- Gabor Csucs Light Microscopy Centre, ETH Zurich Schafmattstrasse 18, HPM G41 CH-8093, Zurich, Switzerland Web: www.lmc.ethz.ch Phone: +41 44 633 6221 Mobile: +41 79 758 21 58 e-mail: [hidden email] |
![]() |
In reply to this post by Chen, Jianbo (NIH/NCI) [E]
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Good evening, I have experienced this problem earlier on an older motorized stage. The reason at that time had been a problem with a joystick. The joystick, which presented one out of two methods to move the table - the other being software and RS232 connection -, upon starting the system, was mechanically not entirely in its (0|0) position. It's position was hence some (x0|y0) at which neither x0 nor y0 were some small values, but not really 0. Touching the joystick after complete initialization would then bring it to the (0|0) position, which the joystick, then, interpreted as (-x0|-y0) and thus brought the table to a slight but steady movement. Possibly, this is your problem. Best wishes, Johannes > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello everyone > We have an encoded (motorized) stage that drifts around 0.5 um (500nm) per > hour to mostly but not always one direction. Putting an enclosed box to > surround the microscope seems did not help. Is this something unavoidable > for a motorized stage? Is using fiducial marker the only solution if > experiment requires high accuracy of XY position over long period of time? > I would appreciate for any thought and advice. > > Jianbo Chen > HIV Drug Resistance Program > National Cancer Institute-Frederick > Frederick MD 21702 > > [hidden email] > 301-846-1841 > > https://ccrod.cancer.gov/confluence/display/CCRSSSCArchive/Home > -- P. Johannes Helm Voice: (+47) 228 51159 (office) Fax: (+47) 228 51499 (office) -- P. Johannes Helm, M.Sc. PhD Seniorengineer CMBN University of Oslo Institute of Basic Medical Science Department of Anatomy Postboks 1105 - Blindern NO-0317 Oslo Voice: +47 228 51159 Fax: +47 228 51499 WWW: folk.uio.no/jhelm |
![]() |
In reply to this post by Chen, Jianbo (NIH/NCI) [E]
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Jianbo, You mention that you have enclosed your system with an incubator. If the drift you are experiencing is the sort of drift that Johannes is talking about, enclosing your system will not help at all. First, you need to determine whether or not your drift is "real" (drift caused by the automated stage moving itself because it is sticky) or not (possibly due to thermal fluctuations causing expansion/contraction of optical components). If it is "real" drift, then your automated stage should report the drift: just check the (x, y) position of your stage at t0 and at t1, before and after you notice a drift. if there is a change in the (x, y) position between t0 and t1, then the drift is probably due to a sticky joystick or poorly calibrated stage. If it is not "real", and enclosing your system with an incubator did not work, then you may be out of luck. You could try an objective warmer. You could also account for stage drift in your post-acquisition analysis or get/write a software program to re-align your stage during acquisition each time you go to take a new image (much like an autofocus program). Dan |
Free forum by Nabble | Edit this page |