I have to stain antibodies with an antibody staining kit designed for 100µg of protein from molecular probes, my problem is that I can not spend 100µg of my precious antibody for this staining... Do you think that staining of only 50 µg of antibody with this kit would be OK? Do you have any tips to improve the staining of low concentration/mass of antibodies with this type of staining kit? Thank you benoit -- Benoît Labarthe, PharmD, MSc Etudiant PhD - PhD trainee Institut de Cardiologie de Montréal - Montreal Heart Institute Centre de Recherche, Research Center Laboratoire du Dr.Théroux 5000 Bélanger Est Montréal Québec Canada H1T 1C8 Tel (514)-376 3330 ext 3017 Fax (514)- 376 1076 |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Benoit, If I am understanding you correctly it sounds like you might be better off using a secondary antibody to "tag" your antibody instead of directly tagging your primary. In this manner you will use a lot less. Hope this helps. Scott J. Howell, Ph.D. Manager, Imaging Module Visual Sciences Research Center Case Western Reserve University 2085 Adelbert Rd. Institute of Pathology Room 106 Cleveland, Ohio 44106 216-368-2300 http://www.case.edu/med/vsrc/ ----- Original Message ----- From: Benoit Labarthe <[hidden email]> Date: Tuesday, May 6, 2008 2:55 pm Subject: Staining of antibodies... To: [hidden email] > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Dear All, > > I have to stain antibodies with an antibody staining kit designed > for 100µg > of protein from molecular probes, my problem is that I can not > spend 100µg > of my precious antibody for this staining... Do you think that > staining of > only 50 µg of antibody with this kit would be OK? Do you have any > tips to > improve the staining of low concentration/mass of antibodies with > this type > of staining kit? > > Thank you > > benoit > > > > -- > Benoît Labarthe, PharmD, MSc > Etudiant PhD - PhD trainee > Institut de Cardiologie de Montréal - Montreal Heart Institute > Centre de Recherche, Research Center > Laboratoire du Dr.Théroux > 5000 Bélanger Est > Montréal Québec Canada H1T 1C8 > Tel (514)-376 3330 ext 3017 > Fax (514)- 376 1076 > |
In reply to this post by Benoit Labarthe
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
We designed those microscale labeling kits for 20 to
100ugs. 50 should be fine.
Mike Ignatius From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Benoit Labarthe Sent: Tuesday, May 06, 2008 11:55 AM To: [hidden email] Subject: Staining of antibodies... I have to stain antibodies with an antibody staining kit designed for 100µg of protein from molecular probes, my problem is that I can not spend 100µg of my precious antibody for this staining... Do you think that staining of only 50 µg of antibody with this kit would be OK? Do you have any tips to improve the staining of low concentration/mass of antibodies with this type of staining kit? Thank you benoit -- Benoît Labarthe, PharmD, MSc Etudiant PhD - PhD trainee Institut de Cardiologie de Montréal - Montreal Heart Institute Centre de Recherche, Research Center Laboratoire du Dr.Théroux 5000 Bélanger Est Montréal Québec Canada H1T 1C8 Tel (514)-376 3330 ext 3017 Fax (514)- 376 1076 |
Benoit Labarthe |
In reply to this post by Scott Howell-3
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Thank you Scott
Sadly, we do not have this choice, our experiment does not allow us to use secondary stainings.... benoit On Tue, May 6, 2008 at 3:02 PM, Scott Howell <[hidden email]> wrote: Search the CONFOCAL archive at -- Benoît Labarthe, PharmD, MSc Etudiant PhD - PhD trainee Institut de Cardiologie de Montréal - Montreal Heart Institute Centre de Recherche, Research Center Laboratoire du Dr.Théroux 5000 Bélanger Est Montréal Québec Canada H1T 1C8 Tel (514)-376 3330 ext 3017 Fax (514)- 376 1076 |
Benoit Labarthe |
In reply to this post by Ignatius, Mike
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Good news! what are the critical points of the protocol, would an increase of the incubation time to 2h help for the staining?
benoit On Tue, May 6, 2008 at 3:12 PM, Ignatius, Mike <[hidden email]> wrote:
-- Benoît Labarthe, PharmD, MSc Etudiant PhD - PhD trainee Institut de Cardiologie de Montréal - Montreal Heart Institute Centre de Recherche, Research Center Laboratoire du Dr.Théroux 5000 Bélanger Est Montréal Québec Canada H1T 1C8 Tel (514)-376 3330 ext 3017 Fax (514)- 376 1076 |