http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, while performing z sections of mitochondria-stained living cells, I´m finding that the bottom part of the cell displays a brighter mitochondria fluoresecence, showing more contrast respect to cytosol than mitochondria in the top part of the cell. My system scans z from bottom to top. If it were simple bleaching due to laser light passing through the cell, I should observe that the bottom mitochondria are also less stained in a second series, but that´s not the case (the difference between top and bottom still there) Any idea to explain it? Thanks Pedro |
|
Ignatius, Mike |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal It is not uncommon to de-quench mitochondrial dyes with light. As the dye bleaches, it is no longer is available to quenches others around it. Eventually an optimum is reached, where the mitochondria are now brighter - or perfectly labeled. Same thing can happen with dyes on antibodies. One way to test is to reverse the stack order or to label with less dye. But sounds like a feature, not a flaw here? Mike Ignatius Molecular Probes/Invitrogen -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Pedro Camello Sent: Tuesday, May 06, 2008 3:44 PM To: [hidden email] Subject: Strange artifact in z series Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, while performing z sections of mitochondria-stained living cells, I´m finding that the bottom part of the cell displays a brighter mitochondria fluoresecence, showing more contrast respect to cytosol than mitochondria in the top part of the cell. My system scans z from bottom to top. If it were simple bleaching due to laser light passing through the cell, I should observe that the bottom mitochondria are also less stained in a second series, but that´s not the case (the difference between top and bottom still there) Any idea to explain it? Thanks Pedro |
|
|
In reply to this post by Pedro Camello
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal By bottom do you mean the part of the cell closest to the coverslip or furthest from the coverslip? It's not uncommon to see brighter fluorescence closer to the coverslip than farther away from the coverslip as both scattering and spherical aberration increase the further into a sample (i.e. from the coverslip) one images. Microscope objectives are corrected to give the best images immediately adjacent to the coverslip and their optical performance decreases the further into the sample you have to image. Careful matching of refractive indices of immersion media to the sample and correction collars can minimize these effects. Kurt Pedro Camello wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi all, > > while performing z sections of mitochondria-stained living cells, I´m > finding that the bottom part of the cell displays a brighter mitochondria > fluoresecence, showing more contrast respect to cytosol than mitochondria in > the top part of the cell. My system scans z from bottom to top. If it were > simple bleaching due to laser light passing through the cell, I should > observe that the bottom mitochondria are also less stained in a second > series, but that´s not the case (the difference between top and bottom still > there) > Any idea to explain it? > > Thanks > > Pedro > > -- Kurt Thorn, PhD Director, Nikon Imaging Center University of California San Francisco UCSF MC 2140 Genentech Hall Room S252 600 16th St. San Francisco, CA 94158-2517 http://nic.ucsf.edu phone 415.514.9709 fax 415.514.4300 |
|
|
In reply to this post by Pedro Camello
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Pedro: If you are using an inverted scope then the areas close to the bottom are less affected by spherical aberration and scattering. That's why they should be brighter. Mike ________________________________ From: Confocal Microscopy List on behalf of Pedro Camello Sent: Tue 5/6/2008 6:44 PM To: [hidden email] Subject: Strange artifact in z series Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, while performing z sections of mitochondria-stained living cells, I´m finding that the bottom part of the cell displays a brighter mitochondria fluoresecence, showing more contrast respect to cytosol than mitochondria in the top part of the cell. My system scans z from bottom to top. If it were simple bleaching due to laser light passing through the cell, I should observe that the bottom mitochondria are also less stained in a second series, but that´s not the case (the difference between top and bottom still there) Any idea to explain it? Thanks Pedro |
|
Stephen Cody |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal G'day Pedro, Are you using a water immersion lens? A water immersion lens with a coverslip correction collar will be your best chance at minimising the effects of spherical aberration etc. These artefacts will probably still affect your imaging at depth, even with a water immersion lens, but the problem will be greatly reduced. Cheers Stephen H. Cody Microscopy Manager Central Resource for Advanced Microscopy Ludwig Institute for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 Australia Tel: 61 3 9341 3155 Fax: 61 3 9341 3104 email: [hidden email] www.ludwig.edu.au/labs/confocal.html www.ludwig.edu.au/confocal Tip: Learn how to receive reminders about you microscope booking: www.ludwig.edu.au/confocal/Local/Booking_Hint.htm -----Original Message----- From: Confocal Microscopy List on behalf of Pedro Camello Sent: Tue 5/6/2008 6:44 PM To: [hidden email] Subject: Strange artifact in z series Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, while performing z sections of mitochondria-stained living cells, I´m finding that the bottom part of the cell displays a brighter mitochondria fluoresecence, showing more contrast respect to cytosol than mitochondria in the top part of the cell. My system scans z from bottom to top. If it were simple bleaching due to laser light passing through the cell, I should observe that the bottom mitochondria are also less stained in a second series, but that´s not the case (the difference between top and bottom still there) Any idea to explain it? Thanks Pedro This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research. |
|
|
In reply to this post by Pedro Camello
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Apologies for repeating the message. I hit the wrong button Pedro Pedro Camello escribió: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi all, > > while performing z sections of mitochondria-stained living cells, I´m > finding that the bottom part of the cell displays a brighter mitochondria > fluoresecence, showing more contrast respect to cytosol than mitochondria in > the top part of the cell. My system scans z from bottom to top. If it were > simple bleaching due to laser light passing through the cell, I should > observe that the bottom mitochondria are also less stained in a second > series, but that´s not the case (the difference between top and bottom still > there) > Any idea to explain it? > > Thanks > > Pedro > > |