Brenton Cavanagh |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, This might be a bit of a blast from the past, but I'm curious how often others change/changed their Sutter DG4 lamps (Ceremax PE175BF). The recommended end of life on the spec sheet is 1000 hr, which is define by the lamp reaching 50% of peak brightness. I inherited a system that had been running for 3000+ hours with no issue. Does the lamp ever actually cut out or are these lamps invincible? From the little I have found on the lifespan of these lamps. It appears to have a constant decline in intensity from the first ignition until it plateaus at 50% brightness for up to 7000+ hours. http://www.excelitas.com/Downloads/Cermax_Eng_Guide.pdf Very interested in the experience/advice that others have had and if they follow the recommended life time or another regime. Note: this lamp is used for long term live cell system where high intensity is not of high importance, more lamp life length and stability. Thanks Brenton |
Glenn Merrill-Skoloff |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, I took a standardized image every two weeks. For this I used a bead slides I had made fresh and took 3 fields of at least five beads each and signal averaged to get the total fluorescence of one bead. I would graph this and post the graph on the wall. The lamps have a burn-in time of 10 or so hours, then they would settle down and stay consistent for several months, then start to tail off. I changed lamps when the standard signal was down by 30%. I wish I could tell you how many hours this was but it has been awhile and I can't located the old graphs. - Glenn Glenn Merrill-Skoloff Laboratory Manager Director, Intravital Microscopy Core Beth Israel Deaconess Medical Center Division of Hemostasis and Thrombosis 3 Blackfan Circle - CLS 942 (617) 735-4040 [hidden email] From: Confocal Microscopy List <[hidden email]<mailto:[hidden email]>> on behalf of Brenton Cavanagh <[hidden email]<mailto:[hidden email]>> Reply-To: Confocal Microscopy List <[hidden email]<mailto:[hidden email]>> Date: Wednesday, June 13, 2018 at 12:39 PM To: "[hidden email]<mailto:[hidden email]>" <[hidden email]<mailto:[hidden email]>> Subject: [External] Sutter DG4 lamp life expectancy ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=WknmpdNpvrlj2B5K1aWVqL1SOiF30547pqSuOmtwXTQ&r=42NmABDnL7YbxR4oFPmDN-5XK3Kw7RY6XnrLkSD93J4&m=AFjJQHRICehRHnNhjNqA3qid6pPjibQngwWKmX_tRpg&s=bwDdSA_TKG5VT-KgQlf32l-hQDwhEo7ECDEKtqEtgMw&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=WknmpdNpvrlj2B5K1aWVqL1SOiF30547pqSuOmtwXTQ&r=42NmABDnL7YbxR4oFPmDN-5XK3Kw7RY6XnrLkSD93J4&m=AFjJQHRICehRHnNhjNqA3qid6pPjibQngwWKmX_tRpg&s=Hdxn9JWnHAFKTl7Z3DtZNqSI1-DxQ7_YVw-DwocFZl0&e= and include the link in your posting. ***** Hi All, This might be a bit of a blast from the past, but I'm curious how often others change/changed their Sutter DG4 lamps (Ceremax PE175BF). The recommended end of life on the spec sheet is 1000 hr, which is define by the lamp reaching 50% of peak brightness. I inherited a system that had been running for 3000+ hours with no issue. Does the lamp ever actually cut out or are these lamps invincible? From the little I have found on the lifespan of these lamps. It appears to have a constant decline in intensity from the first ignition until it plateaus at 50% brightness for up to 7000+ hours. https://urldefense.proofpoint.com/v2/url?u=http-3A__www.excelitas.com_Downloads_Cermax-5FEng-5FGuide.pdf&d=DwIFaQ&c=WknmpdNpvrlj2B5K1aWVqL1SOiF30547pqSuOmtwXTQ&r=42NmABDnL7YbxR4oFPmDN-5XK3Kw7RY6XnrLkSD93J4&m=AFjJQHRICehRHnNhjNqA3qid6pPjibQngwWKmX_tRpg&s=mUWS8XRP_G29K1w9VNoOJ8ve9I5jboNSFoXM6tKAauM&e= Very interested in the experience/advice that others have had and if they follow the recommended life time or another regime. Note: this lamp is used for long term live cell system where high intensity is not of high importance, more lamp life length and stability. Thanks Brenton ________________________________ This message is intended for the use of the person(s) to whom it may be addressed. It may contain information that is privileged, confidential, or otherwise protected from disclosure under applicable law. If you are not the intended recipient, any dissemination, distribution, copying, or use of this information is prohibited. If you have received this message in error, please permanently delete it and immediately notify the sender. Thank you. |
Free forum by Nabble | Edit this page |