yuansheng sun |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Listers, We have two shutters and each is controlled by TTL pulses via a BNC cable. We want to buy a PC board that will allow us control the two shutters through our metamorph software. I would very much appreicate if anybody can recommend one. Thanks a lot in advance. Sheng Yuansheng Sun, Ph.D. Research Scientist, UVA |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I recommend NI's USB-6501; it's got 24 lines of digital I/O and is powered via USB. A related model, the 6008, has fewer digital I/O channels but has a bunch of analog in and a couple analog out if you are also looking for something more multipurpose. I use it to run some flip mirrors and do some external triggering on a couple of my rigs. Craig On Fri, Sep 28, 2012 at 7:49 AM, yuansheng sun <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Listers, > > We have two shutters and each is controlled by TTL pulses via a BNC > cable. We want to buy a PC board that will allow us control the two > shutters through our metamorph software. I would very much appreicate if > anybody can recommend one. Thanks a lot in advance. > > Sheng > Yuansheng Sun, Ph.D. > Research Scientist, UVA > |
|
Mark Cannell-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Another solution would be to use the PCs own parallel port -you control it by writing to it, look to the web for programming examples. It's fast and free. Might seem a bit old school but it is a simple solution…. CHeers On 1/10/2012, at 8:04 PM, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I recommend NI's USB-6501; it's got 24 lines of digital I/O and is powered > via USB. A related model, the 6008, has fewer digital I/O channels but has > a bunch of analog in and a couple analog out if you are also looking for > something more multipurpose. I use it to run some flip mirrors and do some > external triggering on a couple of my rigs. > > Craig > > > On Fri, Sep 28, 2012 at 7:49 AM, yuansheng sun <[hidden email]>wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear Listers, >> >> We have two shutters and each is controlled by TTL pulses via a BNC >> cable. We want to buy a PC board that will allow us control the two >> shutters through our metamorph software. I would very much appreicate if >> anybody can recommend one. Thanks a lot in advance. >> >> Sheng >> Yuansheng Sun, Ph.D. >> Research Scientist, UVA >> Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology& Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
| Free forum by Nabble | Edit this page |