Jean-Yves Tinevez-3 |
Dear friends of imaging,
I have been running in temperature problems, trying to image C.elegans at 20-22ºC, in a very warm microscopy room. At best, the objective is at 25ºC, and that seriously compromises my imaging condition. I have been looking for a temperature controller that would allow to *cool* down a slide to the 20ºC range, and was wondering if you had any experience with such a thing. I typically have a Zeiss stage, and use classical microscope glass slides. I was directed to the Harvard Apparatus microincubation chamber: http://www.harvardapparatus.com/webapp/wcs/stores/servlet/product_11051_10001_39704_-1_HAI_ProductDetail___ It works on slides, but this is a bit overkill, and does not work on upright microscopes (only inverted). I was wondering if you had a product in mind that allow the close temperature control of slides, and that would fit on an upright microscope. Any help would be greatly appreciated. Best jyt |
Beat Ludin |
[Commercial interest]
Dear Jean-Yves Our iceCUBE&BOX system allows you to work at temperatures down to 10°C below ambient and can be supplied for any microscope. Because the entire microscope is cooled down, you can use it with any stage and sample format and there's no need for a cumbersome objective cooler to avoid a heat bridge in the area you are imaging. If you are interested, please contact me off-list for further information. Best regards, Beat At 16:15 08-03-10, you wrote: >Dear friends of imaging, > >I have been running in temperature problems, trying to image C.elegans at >20-22ºC, in a very warm microscopy room. At best, the objective is at 25ºC, >and that seriously compromises my imaging condition. I have been looking for >a temperature controller that would allow to *cool* down a slide to the 20ºC >range, and was wondering if you had any experience with such a thing. > >I typically have a Zeiss stage, and use classical microscope glass slides. I >was directed to the Harvard Apparatus microincubation chamber: > >http://www.harvardapparatus.com/webapp/wcs/stores/servlet/product_11051_10001_39704_-1_HAI_ProductDetail___ > >It works on slides, but this is a bit overkill, and does not work on upright >microscopes (only inverted). > >I was wondering if you had a product in mind that allow the close >temperature control of slides, and that would fit on an upright microscope. > >Any help would be greatly appreciated. >Best >jyt |
Dale Callaham |
In reply to this post by Jean-Yves Tinevez-3
There is a very good section on the Molecular Expressions website for
live cell imaging and culture chambers. http://micro.magnet.fsu.edu/primer/techniques/livecellimaging/index.html The link for culture chambers takes you to the Nikon MicroscopyU website and they discuss many types of chambers. http://www.microscopyu.com/articles/livecellimaging/culturechambers.html Nothing is going to be simple, but a closed chamber with circulated cooled culture medium is an option. A blower with a Peltier cooled air stream will cool the whole stage environment including the objective. You are very close to ambient, but you still may need to dehumidify the room to keep from getting condensation on things. Dale Jean-Yves Tinevez wrote: > Dear friends of imaging, > > I have been running in temperature problems, trying to image C.elegans at > 20-22ºC, in a very warm microscopy room. At best, the objective is at 25ºC, > and that seriously compromises my imaging condition. I have been looking for > a temperature controller that would allow to *cool* down a slide to the 20ºC > range, and was wondering if you had any experience with such a thing. > > I typically have a Zeiss stage, and use classical microscope glass slides. I > was directed to the Harvard Apparatus microincubation chamber: > > http://www.harvardapparatus.com/webapp/wcs/stores/servlet/product_11051_10001_39704_-1_HAI_ProductDetail___ > > It works on slides, but this is a bit overkill, and does not work on upright > microscopes (only inverted). > > I was wondering if you had a product in mind that allow the close > temperature control of slides, and that would fit on an upright microscope. > > Any help would be greatly appreciated. > Best > jyt |
In reply to this post by Jean-Yves Tinevez-3
On Mon, 2010-03-08 at 09:15 -0600, Jean-Yves Tinevez wrote:
> Dear friends of imaging, > > I have been running in temperature problems, trying to image C.elegans at > 20-22ºC, in a very warm microscopy room. At best, the objective is at 25ºC, > and that seriously compromises my imaging condition. I have been looking for > a temperature controller that would allow to *cool* down a slide to the 20ºC > range, and was wondering if you had any experience with such a thing. > > I typically have a Zeiss stage, and use classical microscope glass slides. I > was directed to the Harvard Apparatus microincubation chamber: > > http://www.harvardapparatus.com/webapp/wcs/stores/servlet/product_11051_10001_39704_-1_HAI_ProductDetail___ > > It works on slides, but this is a bit overkill, and does not work on upright > microscopes (only inverted). > > I was wondering if you had a product in mind that allow the close > temperature control of slides, and that would fit on an upright microscope. > > Any help would be greatly appreciated. > Best > jyt we also image c.elegans and had some temperature problems but those got resolved once we kicked out our halogen lamps and designed our own LED system (one LED driven by a konrad microcontroller). can't speak much about temperature control otherwise (we just monitor it, with the same konrad device) but if you're handy then you can build some rather fancy stuff yourself very cheaply. a sensor that you put near the slide, peltier elements for cooling, some plexiglass, the controller, and you are done. in your case, maybe you even can attach the sensor and peltier on the slide with CPU paste /Johan |
In reply to this post by Jean-Yves Tinevez-3
I have several Linkam microscope stages that I use with upright microscopes. I use this Peltier PE120 stage one for cooling from ambient down to ~-10oC:
http://www.linkam.co.uk/index.php?option=com_virtuemart&page=shop.browse&category_id=12&Itemid=70 I just use normal microscope slides resting on the Peltier-cooled surface. Condensation could be a problem and usually requires piped cooled air/nitrogen gas. However, using a small fan close to the objective might be OK, since you are only cooling from 25 to 20C. Regards Mark Dr Mark Auty Manager, National Food Imaging Centre Food Processing & Functionality Department Moorepark Food Research Centre Teagasc Moorepark Fermoy, Co Cork Ireland tel: +353 25 42442 fax: +353 25 42340 [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jean-Yves Tinevez Sent: 08 March 2010 15:15 To: [hidden email] Subject: Temperature controler for upright microscopes Dear friends of imaging, I have been running in temperature problems, trying to image C.elegans at 20-22ºC, in a very warm microscopy room. At best, the objective is at 25ºC, and that seriously compromises my imaging condition. I have been looking for a temperature controller that would allow to *cool* down a slide to the 20ºC range, and was wondering if you had any experience with such a thing. I typically have a Zeiss stage, and use classical microscope glass slides. I was directed to the Harvard Apparatus microincubation chamber: http://www.harvardapparatus.com/webapp/wcs/stores/servlet/product_11051_10001_39704_-1_HAI_ProductDetail___ It works on slides, but this is a bit overkill, and does not work on upright microscopes (only inverted). I was wondering if you had a product in mind that allow the close temperature control of slides, and that would fit on an upright microscope. Any help would be greatly appreciated. Best jyt ------------------------------------------------------------------------------ Attention: This e-mail is privileged and confidential. If you are not the intended recipient please delete the message and notify the sender. Any views or opinions presented are solely those of the author. This email was scanned by Teagasc and has been certified virus free with the pattern file currently in use. This however cannot guarantee that it does not contain malicious content. Tabhair aire: Tá an r-phost seo faoi phribhléid agus faoi rún. Mura tusa an duine a bhí beartaithe leis an teachtaireacht seo a fháil, scrios é le do thoil agus cuir an seoltóir ar an eolas. Is leis an údar amháin aon dearcaí nó tuairimí a léirítear. Scanadh an r-phost seo le Teagasc agus deimhníodh go raibh sé saor ó víoras leis an bpatrúnchomhad atá in úsáid faoi láthair. Ní féidir a ráthú leis seo áfach nach bhfuil ábhar mailíseach ann. |
Louis Kerr |
In reply to this post by Jean-Yves Tinevez-3
We have a stage system on our Zeiss LSM 510 inverted microscope that can cool and heat based on a Pecon stage insert http://www.pecon.biz/?page_id=314 and a Thermo NESLAB RTE refrigerated bath/circulator with water and ethylene glycol. We have operated it between 2-40C. I would think you could find a plate for an upright or I have seen home built models of aluminum plate with tubes attached or embedded. Because you are so close to ambient temperatures you would be best off with a device that cools and heats.
Louie ----- Original Message ----- From: "Dale Callaham" <dac@RESEARCH.UMASS.EDU> To: CONFOCALMICROSCOPY@LISTS.UMN.EDU Sent: Monday, March 8, 2010 10:40:02 AM Subject: Re: Temperature controler for upright microscopes There is a very good section on the Molecular Expressions website for live cell imaging and culture chambers. http://micro.magnet.fsu.edu/primer/techniques/livecellimaging/index.html The link for culture chambers takes you to the Nikon MicroscopyU website and they discuss many types of chambers. http://www.microscopyu.com/articles/livecellimaging/culturechambers.html Nothing is going to be simple, but a closed chamber with circulated cooled culture medium is an option. A blower with a Peltier cooled air stream will cool the whole stage environment including the objective. You are very close to ambient, but you still may need to dehumidify the room to keep from getting condensation on things. Dale Jean-Yves Tinevez wrote: > Dear friends of imaging, > > I have been running in temperature problems, trying to image C.elegans at > 20-22ºC, in a very warm microscopy room. At best, the objective is at 25ºC, > and that seriously compromises my imaging condition. I have been looking for > a temperature controller that would allow to *cool* down a slide to the 20ºC > range, and was wondering if you had any experience with such a thing. > > I typically have a Zeiss stage, and use classical microscope glass slides. I > was directed to the Harvard Apparatus microincubation chamber: > > http://www.harvardapparatus.com/webapp/wcs/stores/servlet/product_11051_10001_39704_-1_HAI_ProductDetail___ > > It works on slides, but this is a bit overkill, and does not work on upright > microscopes (only inverted). > > I was wondering if you had a product in mind that allow the close > temperature control of slides, and that would fit on an upright microscope. > > Any help would be greatly appreciated. > Best > jyt -- Louis Kerr lkerr@mbl.edu Research and Education Support Coordinator Marine Biological Laboratory 7 MBL Street Woods Hole, MA 02543 508-289-7273 508-540-6902 (FAX) 508-292-0289 (Cell phone) VISIT OUR WEB SITES: www.mbl.edu www.courses.mbl.edu |
In reply to this post by Jean-Yves Tinevez-3
Jean-Yves
Bioptechs develops and manufacturers a unique line of micro-environmental control systems for time lapse imaging. We pay very close attention to your exact needs. There are many approaches to micro-environmental control. To make the selection process easier and make sure we cover all the factors to ensure a successfully working system, please fill out our User Profile Questionnaire available from home page. www.bioptechs.com Temperature control on an upright scope, close to ambient is a little more tricky but we can do it and eliminate condensation at the same time. Dan On Mar 8, 2010, at 10:15 AM, Jean-Yves Tinevez wrote: Dear friends of imaging, I have been running in temperature problems, trying to image C.elegans at 20-22ºC, in a very warm microscopy room. At best, the objective is at 25ºC, and that seriously compromises my imaging condition. I have been looking for a temperature controller that would allow to *cool* down a slide to the 20ºC range, and was wondering if you had any experience with such a thing. I typically have a Zeiss stage, and use classical microscope glass slides. I was directed to the Harvard Apparatus microincubation chamber: http://www.harvardapparatus.com/webapp/wcs/stores/servlet/product_11051_10001_39704_-1_HAI_ProductDetail___ It works on slides, but this is a bit overkill, and does not work on upright microscopes (only inverted). I was wondering if you had a product in mind that allow the close temperature control of slides, and that would fit on an upright microscope. Any help would be greatly appreciated. Best jyt Dan Focht Bioptechs Inc P: (724)282-7145 F: (724)282-0745 [hidden email] |
Andreas Bruckbauer |
Hello,
we recently had some trouble caused by the curtain in our microscope room, it builds up electric charge under dry weather conditions leading to small electric shocks when users touch the optical table. Has anyone encountered such a problem and found a solution? Are there any antistatic black curtains which would do better? Andreas |
EricMarino |
We’ve purchased curtains from a local hospital supply
distributer, since they deal with static sensitive medical devices they’re
antistatic, however, you will not get them in black so we tested white versus
black curtains and saw no difference in ambient light leaking into the scopes. Eric Marino From: Confocal Microscopy
List [mailto:[hidden email]] On Behalf Of Andreas
Bruckbauer Hello, No virus
found in this incoming message.
Eric Marino
Senior Imaging Specialist Program in Cellular and Molecular Medicine Boston Children's Hospital |
In reply to this post by Andreas Bruckbauer
Hello:
We have a static dissipating floor. This was done when the lab was built (I assume it is grounded). The drawback is that most shoes soles are good insulators. Therefore the use of shoe grounding strap is necessary. Sophie ____________________________________________________ Sophie M. K. Brunet, Ph. D. Research Officer Optical Spectroscopy, Laser Systems and Applications [hidden email] 306-966-1719 (office) 306-966-1702 (fax) ____________________________________________________ Saskatchewan Structural Sciences Centre University of Saskatchewan Thorvaldson Bldg. 110 Science Place Saskatoon, Sk S7N 5C9 ____________________________________________________ Quoting Andreas Bruckbauer <[hidden email]>: > > Hello, > we recently had some trouble caused by the curtain in our microscope room, it > builds up electric charge under dry weather conditions leading to small > electric shocks when users touch the optical table. Has anyone encountered > such a problem and found a solution? Are there any antistatic black curtains > which would do better? > > Andreas > > > > > |
Jean-Yves Tinevez-3 |
In reply to this post by Jean-Yves Tinevez-3
Dear friends of imaging,
Thank you all for the fantastic input! I will carefully look into all of it. Thanks again Best jy -- Jean-Yves Tinevez PFID - Imagopole Institut Pasteur 25-28, rue du Docteur Roux 75724 Paris cedex 15 France tel: +33 1 40 61 31 77 |
Free forum by Nabble | Edit this page |